  209   Fri Dec 14 10:07:14 2012 EvgeniDRS-4 trigger

 

Evgeni wrote:

How to configure DRS oscilloscope for the oscillations with an amplitude greater than the value of the exposed
in the trigger (internal). 

 

 

  210   Fri Dec 14 10:07:54 2012 EvgeniDRS-4 trigger

 

Stefan Ritt wrote:

 

Evgeni wrote:

 

Stefan Ritt wrote:

 

Evgeni wrote:

How to configure DRS oscilloscope for the oscillations with an amplitude greater than the value of the exposed
in the trigger (internal). 

 

Sorry, I don't understand that question. The DRS4 Evaluation board input signal range is 1V. If you have larger signals, you have to attenuante them externally.

/Stefan 

 

 

Can I adjust the internal trigger DRS oscilloscope for signal extraction 0.5 volts (for example) from any of the four channels. In our case, there is a lot of noise with low amplitude, which must be removed during the registration. We need to record the individual pulses of higher amplitude than the noise. So we want to use the internal trigger DRS oscilloscope to cut this noise by setting up its threshold amplitude noise.

 

 

 Sure, you can set the trigger level with the vertical slider (see attached figure). The trigger level works form -0.5V to +0.5V. Just like with a normal oscilloscope. I thought this would be obvious...

 

    That's it? Thank you for your comprehensive answer.

  211   Fri Dec 14 21:49:29 2012 Stefan RittEVM rev4 board trigger change and drs_example

Martin Petriska wrote:

 I switched from rev 3 to rev 4 board, but have some problems with triggering, board is now waiting for trigger (rev.3 is working). How to do in drs_exam.cpp for example triggering on Ch0 && CH1 ?

Software 4.0.0, windows version.

Here is old trigger initialisation: 

b->EnableTrigger(0,1);

b->SetTriggerSource(0);

b->SetTriggerLevel(0.25, false);

b->SetTriggerDelayNs(0);

 

Btw. Is it possible to set up different trigger Levels for each channel ?

 

(If there is some interest here is my code in Qt, still aplha) http://sourceforge.net/p/qtpals/code

Sorry the late reply.

In V4, triggering has changed. You can trigger now on an OR or AND of channels. Therefore you have to supply a bitmask, where the 1st bit = CH1, 2nd bit = CH2 and so on. Have a look at the most recent drs_exam. It contains code:

 

   /* use following lines to enable hardware trigger on CH1 at 50 mV positive edge */
   if (b->GetBoardType() == 8) {     // Evaluaiton Board V4 
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<0);        // set CH1 as source
   } else {                          // Evaluation Board V3
      b->EnableTrigger(0, 1);           // lemo off, analog trigger on
      b->SetTriggerSource(0);           // use CH1 as source
   } 

So if you want CH1 && CH2, you look at the source code of SetTriggerSource. It contains

 

      // Set trigger configuration
   // OR  0=CH1, 1=CH2, 2=CH3, 3=CH4, 4=EXT
   // AND 8=CH1, 9=CH2, 10=CH3, 11=CH4, 12=EXT

 

So an AND between CH1 and CH2 needs a

    b->SetTriggerSource(1<<8 | 1<<9);

Your code looks interesting. Do you have a screenshot or can you explain what it does? 

  212   Thu Dec 27 00:12:12 2012 Jinhong Wangvariation of sampling capacitors

Hi Stefan,

A quick question, what is the typical variation of the sampling capacitors in DRS4?  Will this variation be significant to affect your sampling result?

 Best,

Jinhong

  213   Thu Dec 27 09:49:17 2012 Stefan Rittvariation of sampling capacitors

Jinhong Wang wrote:

Hi Stefan,

A quick question, what is the typical variation of the sampling capacitors in DRS4?  Will this variation be significant to affect your sampling result?

 Best,

Jinhong 

The capacitors sample the input voltage, not the charge, so the actual size of the capacitors does not matter on first order (the variations might be in the order of 5%). A bigger effect is the variation of the analog switches in the front of the capacitors, which is about 15%. So the actual bandwidth each cell sees varies by maybe 20% (given by the R and the C), but this comes only into play when sampling steep edges.

Stefan

  214   Thu Dec 27 18:15:14 2012 Jinhong Wangvariation of sampling capacitors

Stefan Ritt wrote:

Jinhong Wang wrote:

Hi Stefan,

A quick question, what is the typical variation of the sampling capacitors in DRS4?  Will this variation be significant to affect your sampling result?

 Best,

Jinhong 

The capacitors sample the input voltage, not the charge, so the actual size of the capacitors does not matter on first order (the variations might be in the order of 5%). A bigger effect is the variation of the analog switches in the front of the capacitors, which is about 15%. So the actual bandwidth each cell sees varies by maybe 20% (given by the R and the C), but this comes only into play when sampling steep edges.

Stefan

 Great to know this! Thanks~

Jinhong

  215   Fri Feb 1 17:43:48 2013 Jinhong Wangvariation of sampling capacitors

Jinhong Wang wrote:

Stefan Ritt wrote:

Jinhong Wang wrote:

Hi Stefan,

A quick question, what is the typical variation of the sampling capacitors in DRS4?  Will this variation be significant to affect your sampling result?

 Best,

Jinhong 

The capacitors sample the input voltage, not the charge, so the actual size of the capacitors does not matter on first order (the variations might be in the order of 5%). A bigger effect is the variation of the analog switches in the front of the capacitors, which is about 15%. So the actual bandwidth each cell sees varies by maybe 20% (given by the R and the C), but this comes only into play when sampling steep edges.

Stefan

 Great to know this! Thanks~

Jinhong

 Hi Dr. Stefan,

So the sampling capacitors store the input voltage instead of the charge. What about the readout circuits? I saw there is a buffer followed each sampling capacitor. Do you buffer the charge (like a charge sensitive amplifier)  or the voltage? From Fig.12, 14 in datasheet, it seems most probably the readout is a charging or discharging of a capacitor. Could you please add some comments on this? 

Cheers,

Jinhong

  216   Tue Feb 5 14:38:35 2013 Stefan Rittvariation of sampling capacitors

Jinhong Wang wrote:

Hi Dr. Stefan,

So the sampling capacitors store the input voltage instead of the charge. What about the readout circuits? I saw there is a buffer followed each sampling capacitor. Do you buffer the charge (like a charge sensitive amplifier)  or the voltage? From Fig.12, 14 in datasheet, it seems most probably the readout is a charging or discharging of a capacitor. Could you please add some comments on this? 

Cheers,

Jinhong

The buffer buffers the voltage, not the charge. The curves in Fig. 12, 14 indicate that the voltage followers take some time until they settle. 

Stefan

  217   Wed Feb 13 16:58:40 2013 Martin PetriskaNonuniform sampling

 Are there any plans to include reconstruction of nonuniform sampling  in DRS4 to get uniformly sampled data?

Im now reading article IEEE Trans on Circ. ans Systems I, Vol.55 No.8 sept. 2008 Reconstruction of Nonuniformly Sampled Bandlimited Signals Usinga Differentiator–Multiplier Cascade by Stefan Tertinek and Christian Vogel

and plan to implement it, but may be somebody has it done before me.

 

  218   Wed Feb 13 17:03:53 2013 Stefan RittNonuniform sampling

Martin Petriska wrote:

 Are there any plans to include reconstruction of nonuniform sampling  in DRS4 to get uniformly sampled data?

Im now reading article IEEE Trans on Circ. ans Systems I, Vol.55 No.8 sept. 2008 Reconstruction of Nonuniformly Sampled Bandlimited Signals Usinga Differentiator–Multiplier Cascade by Stefan Tertinek and Christian Vogel

and plan to implement it, but may be somebody has it done before me.

 

Interesting paper. I was not aware of this method. Sounds interesting. AFAIK, nobody has implemented it so far.

My (old) plan was to linearly interpolate samples (something you could do in an FPGA as well), but this will introduce (small) errors. The next best thing would be to do some spline interpolation, but this is time consuming and not suited for an FPGA.

If you get good results with the method above, please let others know about it.

/Stefan 

  219   Fri Feb 22 11:46:17 2013 Yury GolodDRS4 trigger, different polarity

I need to synchronize two signals. These signals have a different polarity.

I can set triggers on different levels. But I can't set different polarity of triggers.

Now I can set (T1 and T2), I need to set (T1 and (not T2))

Is it possible?

 

           

d->SetTriggerLevel(-0.4,0.4,0.0,0.0,false);

d->EnableTrigger(1, 0);  // Enable trigger

d->SetTriggerSource(1<<8 | 1<<9);  // T1 and T2

 

file DRS.cpp:

int DRSBoard::SetTriggerLevel(double voltage1,double voltage2, double voltage3,double voltage4,bool negative)

{

      SetDAC(fDAC_TLEVEL1, voltage1/2 + 0.8);

      SetDAC(fDAC_TLEVEL2, voltage2/2 + 0.8);

 

  220   Fri Feb 22 11:56:57 2013 Stefan RittDRS4 trigger, different polarity

Yury Golod wrote:

I need to synchronize two signals. These signals have a different polarity.

I can set triggers on different levels. But I can't set different polarity of triggers.

Now I can set (T1 and T2), I need to set (T1 and (not T2))

Is it possible?

 

           

 

d->SetTriggerLevel(-0.4,0.4,0.0,0.0,false);

d->EnableTrigger(1, 0);  // Enable trigger

d->SetTriggerSource(1<<8 | 1<<9);  // T1 and T2

 

file DRS.cpp:

int DRSBoard::SetTriggerLevel(double voltage1,double voltage2, double voltage3,double voltage4,bool negative)

{

      SetDAC(fDAC_TLEVEL1, voltage1/2 + 0.8);

      SetDAC(fDAC_TLEVEL2, voltage2/2 + 0.8);

 

There is no way to select different polarities, it is not implemented in the firmware. Like your digital oscilloscope has also only one polarity switch.

The only way to do it is to use a (passive) inverter, so that you have two signals of the same polarity. Something like this:

http://www.phillipsscientific.com/pdf/460ds.pdf

 

  222   Wed Feb 27 13:47:32 2013 Georg WinnerChip Test - Cell Error

When starting Chip Test in DRS Command Line Interface, I receive the following message:

Cell error on channel 1, cell 5: -154.4 mV instead 0 mV

Chip Error!

What does this mean? The maximal peak-to-peak Amplitude given to channel was for a short time 10V.

The graphical interface shows no artefacts when using channel 1.

 

  223   Thu Feb 28 10:47:14 2013 Dmitry Hitsclock and trigger outs
Hi,
I am considering using the DRS4 evaluation board as an ADC card for the wire chamber in the physics lab (VP) experiment at ETH. However, the wire 
chamber has 8 outputs, so I would need to have two of such boards. Is it possible to synchronise them, online or offline? From the website, it looks 
like yes, but the documentation says that these features (trigger and clock out) may not have been implemented in firmware yet. Could you tell me 
the status?

Thank you very much,

Dmitry.
  224   Thu Feb 28 12:58:44 2013 Stefan Rittclock and trigger outs
> Hi,
> I am considering using the DRS4 evaluation board as an ADC card for the wire chamber in the physics lab (VP) experiment at ETH. However, the wire 
> chamber has 8 outputs, so I would need to have two of such boards. Is it possible to synchronise them, online or offline? From the website, it looks 
> like yes, but the documentation says that these features (trigger and clock out) may not have been implemented in firmware yet. Could you tell me 
> the status?
> 
> Thank you very much,
> 
> Dmitry.

I'm right now working on it. If you only need 2-3 ns accuracy  between the two boards then you can do this already now without firmware upgrade. The software for this is in principle ready, but I have to 
finish the documentation. Since I'm on a business travel right now this might take me some time (weeks?). 

If you want better timing (O(100ps)) between the boards, then you will need a firmware update. Or you wait until we ship boards with the new firmware. I will announce this through this forum.

Stefan
  225   Wed Mar 6 12:35:38 2013 Osip LishilinDRS4- analog pulse counting

Hello, Stefan. Have you implemented pulse counting yet?

Best, Osip.

  226   Wed Mar 6 12:37:14 2013 Stefan RittDRS4- analog pulse counting

Osip Lishilin wrote:

Hello, Stefan. Have you implemented pulse counting yet?

Best, Osip.

Nope. 

  227   Wed Mar 6 13:08:03 2013 Stefan RittChip Test - Cell Error

Georg Winner wrote:

When starting Chip Test in DRS Command Line Interface, I receive the following message:

Cell error on channel 1, cell 5: -154.4 mV instead 0 mV

Chip Error!

What does this mean? The maximal peak-to-peak Amplitude given to channel was for a short time 10V.

The graphical interface shows no artefacts when using channel 1.

 

The "Chip Test" command is made for a special test board we use for chip testing. This command will not work with the evaluation board, since only four of the 8 DRS channels are connected there. So just ignore it and verify the board functionality by looking at the graphical interface.

/Stefan 

  228   Mon Mar 25 11:12:53 2013 Georg WinnerDifferences in Source Code

I have noticed some differences in the source code between Windows (4.0.0) and Linux (4.0.1) Version.

drs_exam.cpp: In the windows version when setting the trigger there is no part "if (b->GetBoardType() == 8) {...} else {...}" like in Linux version. So under Windows drs_exam does not start readout of DRS 4 Evalutation Board V4, because it does not get the trigger, under linux the board can be read out succesfull. I have found out, that adding the missing part solves the problem for the windows version.

drs.cpp (Windows Version), line 2101, function "int DRSBoard::SetTriggerDelayNs(int delay)":

There is no operation which  calculates the variable "fTriggerDelayNs" out of variable "ticks" like in function "int DRSBoard::SetTriggerDelayPercent(int delay)" (Line 2073). So "fTriggerDelayNs" can get diverse values when using one of the Trigger Setting Functions. Was this intended?

 

  229   Tue Mar 26 01:17:59 2013 Jill Russekcascading -- DRS4 Osci.cpp & DRS.cpp

 

All I'm trying to do is cascade one input signal, though all available channels, so that I end up with 8*1024 bins per event.

Here is the read out on my board/chip:

Mezz. Board index:    0
DRS type:             DRS4
Board type:           8
Serial number:        2249
Firmware revision:    17662
Temperature:          35.2 C
Input range:          -0.5V...0.5V
Calibrated range:     -0.5V...0.5V
Calibrated frequency: 5.120 GHz
Status reg.:          0000001A
Control reg.:         00000010
  DMODE circular
Trigger bus:          00000000
Frequency:            5.120 GHz

 

What I've tried thus far:

In Osci.cpp, in the method/function  SelectSource(int board, int firstChannel, int chnSection), I added a line.. (in bold)

//----------------------------------------------------------------------------------------------------------------------------------------------

 if (b->GetBoardType() == 5 || b->GetBoardType() == 7 || b->GetBoardType() == 8) {
         
        
             if (chnSection == 2)
                b->SetChannelConfig(0, 8, 4);
             //added
             else if(chnSection == 1)
                 b->SetChannelConfig(0, 8, 2);
             //added

             else
                b->SetChannelConfig(0, 8, 8);

//----------------------------------------------------------------------------------------------------------------------------------------------

I've also tried doing settings such as SetChannelConfig(0, 8, 1);   , SetChannelConfig(0, 8, 2); , SetChannelConfig(0, 1, 2); , etc..

Which always ends up making the run fail.. and sometimes I get index errors..

As far as I understanding the program now, this is what I know:

fChannelCascading determines getChannelCascading,
this determines the  if (casc == 2) line in configDialogue.cpp, which sets:
 b->SetChannelConfig(config, 8, 4);

fChannelCascading is being set by:
 switch (nConfigChannels) {
      case 1:
         fChannelConfig = 0x01;
         fChannelCascading = 8;
         break;
      case 2:
         fChannelConfig = 0x11;
         fChannelCascading = 4;
         break;
      case 4:
         fChannelConfig = 0x55;
         fChannelCascading = 2;
         break;
      case 8:
         fChannelConfig = 0xFF;
         fChannelCascading = 1;
         break;
      default:
         printf("Invalid channel configuration\n");
         return 0;
      }

which is being set by nConfigChannels in DRS.cpp, in the method:
SetChannelConfig(int firstChannel, int lastChannel, int nConfigChannels)

SetChannelConfig is being called in the ConfigDialogue.cpp,  but the default Osci program is such that you can't do a configuration for a cascade of one signal using all the channels. At least, not that I am aware of. 

So what buttons do I need to enable, or what do I need to call, or write, so that I can cascade a signal to end up with 8*1024 bins per event?

This has had me going in circles for weeks, so thank you for your help!!!!

 

 