  703   Tue Jun 19 06:42:23 2018 Phan Van ChuanThe data acquisition speedDear Stefan,

We are using an DRS4 board V5.1
for building a metering system for the scintillator
 wavech0.png 
  702   Wed Jun 13 16:34:28 2018 Julian KempMaximum analog input voltageThank you! That solves my problem.




  
  701   Wed Jun 13 13:42:47 2018 Stefan RittMaximum analog input voltageIn principle the numbers in the manual
are correct. But they relate to pulses of
a certain length, because the input protection
  
  700   Wed Jun 13 13:23:17 2018 Julian KempMaximum analog input voltageDear all,

I have been wondering what the
maximum analog input voltage for the DRS4
  
  699   Fri Jun 8 08:11:05 2018 Stefan Ritt Several people reported this problem, but
we cannot reproduce it at our lab. Both the
oscilloscope and the command line interface
  
  698   Thu Jun 7 16:27:21 2018 Phan Van Chuan Dear Stefan,

I am using an DRS4 board to test
the signal from an scintillator detector;
 figure1.pngfigure2.png 
  697   Thu May 17 13:29:34 2018 Stefan Ritt"Symmetric spikes" fixedGood news for all DRS4 users. After many
years, I finally understand where the "symmetric
spikes" come from and how to fix them.
 with.pngScreen_Shot_2018-05-17_at_13.30.23_.pngwithout.png 
  696   Mon May 14 09:21:29 2018 Alessio BertiWIndows Connection problem with drs507 SOLVEDHi,

I have a machine with Windows 10
and the solution provided by Steven works
  
  695   Wed May 9 14:07:10 2018 Alec ShacklefordWIndows Connection problem with drs507 SOLVEDThank you for this fantastic solution.
I had almost reinstalled windows 7 to see
if that would solve the issue!
  
  694   Wed May 9 09:03:52 2018 Stefan RittManual Rev5.1 Figure 1, optional componentsI updated the picture in the manual with
a current picture of a Rev5.1 board, and
also added a picture of the bottom side.
  
  693   Tue May 8 23:58:35 2018 Sean QuinnManual Rev5.1 Figure 1, optional componentsDear All,

 

I'm troubleshooting a board
  
  692   Tue May 8 14:43:03 2018 Stefan RittPeak at 0 mV in tracesThe DRS chip is read out with a 12 bit
ADC, thus the phyical resolution is roughly
1V/4096 = 0.24 mV. I say roughly since the
  
  691   Tue May 8 12:15:54 2018 Alessio BertiPeak at 0 mV in tracesHi Stefan,

following your example, we tried
to perform the same measurement, using drs_exam
 20180508_drs4_drs_exam_1000_events_81_bins_linear.png20180508_drs4_drs_exam_1000_events_81_bins_log.png20180508_drs4_drs_exam_1000_events_23_bins_linear.png20180508_drs4_drs_exam_1000_events_23_bins_log.png 
  690   Sun May 6 11:45:09 2018 Stefan Rittconfusion about the description in drs.cppThe locbus_addr is indeed 32 bits wide,
since the firmware was originally derived
from some firmware running in a VME crate,
  
  689   Sun May 6 08:13:37 2018 chen wenjunconfusion about the description in drs.cppHi Stefan:

  I'm still confused that
althought the 8 bits buffer is enough,the
  
  688   Fri May 4 12:11:57 2018 Stefan RittRunning drs_example.cppAnd here is the second part of your answer:
When you change the input range, you have
to redo the voltage calibration. Best is
  
  687   Fri May 4 11:56:08 2018 Stefan RittVoltage and Timing Calibration in drs_exam.cppHave you set the sampling frequency 

b->SetFrequency(5, true);

before the calibration?
  
  686   Fri May 4 11:35:20 2018 Stefan RittPeak at 0 mV in tracesI tried the following:

- trigger on a 10 MHz sine wave
on CH0, CH1 was open
 Screen_Shot_2018-05-04_at_11.36.24_.png 
  685   Wed May 2 12:23:16 2018 Alessio BertiPeak at 0 mV in tracesHi,

thank you for the quick reply.
All the bins in the previous histograms have
 zero_peak_after_spike_removal_ch2_1000_bins.png 
  684   Wed May 2 12:12:42 2018 Stefan RittPeak at 0 mV in tracesI note that your peak at zero is exactly
twice as high as the bins left and right,
so this looks to me like a binning problem
  