    Reply  Mon Oct 14 11:45:06 2019, Danyang, how to acquire the stop position with channel cascading Capture.PNG
I tried the logic in my designed board.  The results are shown in the picture: Srout keeps low when A3-A0  is
set to 1101 and srclk is set as you mentioned. And the drs4 chip does not output sine wave in such configuration.

Srout signal only reacts after the rsrload signal is pulled high and A3-A0 is not 1101.  
    Reply  Mon Oct 14 12:56:13 2019, Stefan Ritt, how to acquire the stop position with channel cascading 
Note that you have to read out the Write Shift Register only if you do channel cascading, e.g. configuring the chip with 4x2048 bins by setting the Write
Shift Register to 01010101b. Then the Write Shift Register tells you in which 1024-bin segment the Domino Wave has been stopped. If you use the normal
8x1024 bin mode, you don't have to read out the Write Shift Register since it continas only 1's. 
    Reply  Mon Oct 14 13:44:26 2019, Danyang, how to acquire the stop position with channel cascading 
Yes, firstly I configured the chip with 4x2048 bins by setting the Write Shift Register to 01010101b, A3-A0 keeps 1101----> secondly
I enabled the domino wave, wait  some time for stable,  A3-A0 keeps 1111  ---->thirdly stops the domino wave when the trigger comes,
A3-A0 keeps 1101 (or 1010, 0000)----> forthly send the clock pulse to the srclk pin, A3-A0 keeps 1101,  srout
    Reply  Mon Oct 14 15:27:09 2019, Stefan Ritt, how to acquire the stop position with channel cascading 
If you configure the Write Shift Register with 01010101b, then all you have to do after a trigger is to set A3-A0 to 1101. The WSROUT pin shows you then
either ther state 01010101b or 10101010b, you the pin should be 1 or 0, and that's all you need. The Write Shift Register is NOT routed to the SROUT
pin, you only see it at the WSROUT pin.
    Reply  Tue Oct 15 08:14:17 2019, Danyang, how to acquire the stop position with channel cascading 
Thanks a lot. The problem is solved when A3-A0 is set 1101 and srclk keeps low.

Best Regards,
Danyang
Entry  Sat Feb 19 17:25:29 2011, S S Upadhya, how to synchronize Sampling frequency of two evaluation boards 
 Dear sir,
We have two evaluation boards of DRS4. We would like to use 8 inputs to be recorded on a trigger and we would like to find the relative time difference
of inputs. So is it possible to synchronize the sampling frequency of the two evaluation boards. 
    Reply  Sat Feb 19 22:46:35 2011, Stefan Ritt, how to synchronize Sampling frequency of two evaluation boards 


    
        
            S S Upadhya wrote:
        
        
            
             Dear sir,
    Reply  Mon Feb 21 08:10:31 2011, Stefan Ritt, how to synchronize Sampling frequency of two evaluation boards 


    
        
            Stefan Ritt wrote:
        
        
            
            
            
    Reply  Mon Feb 21 12:42:33 2011, S S Upadhya, how to synchronize Sampling frequency of two evaluation boards 


    
        
            Stefan Ritt wrote:
        
        
            
            
            
Entry  Mon Aug 29 09:36:34 2016, benjamin legeyt, increment write config register on the fly? 
Hello,

I have a question about using the write config register to enable/disable sampling on the fly.  I am looking to instrument an experiment
at EPFL where multiple short events need to be captured during a 20us period followed by an 80us quiet period during which we could read out the chip.
    Reply  Mon Aug 29 10:57:33 2016, Stefan Ritt, increment write config register on the fly? 
The issue with "stopping at cell 767" would also affect this mode of operation. Furthermore, the DRS4 chip has only 10 bit register which
records in which cell the event has occured, and where the readout must be started. If you record 8 separate events, you don't know where to start
the readout.
    Reply  Mon Aug 29 12:18:49 2016, benjamin legeyt, increment write config register on the fly? 
If I may trouble you for a little more information, the critical point then is that there should not be any zeroes in the write config register
while the sampling is active?  In case it was unclear I would only be reading out once sampling was stopped (dwrite = 0).  

As for the readout, I know that I would have to read out all 1024 samples each time, and keep track of where each channel stopped in the FPGA.
    Reply  Mon Aug 29 12:51:48 2016, Stefan Ritt, increment write config register on the fly? 
The problem is when you change the write config register from 11111111 to 01111111, or from 00001111 to 00000111, then the last 256 sampels of the previous
channel (in the first case #0, in the scond #4) would be overwritten as soon as dwrite =1 again. So you loose 1/4 ef each channel.

Concerning the readout, indeed you can keep track in the FPGA, but only with a certainty of a few cells. This gives some timing inacccuracy of
Entry  Fri Dec 13 10:37:18 2013, Dmitry Hits, input protection in DRS4 evaluation board 
 Dear Stefan
Last month I was using a DRS4 evaluation board to digitise the signal from the charged particles in piM1 beam line at PSI. The beam in piM1 is
a mixture of pions with a small percentage of protons. Pions are close to minimum ionising and were producing the signals on the order of 250 mV (Landau
    Reply  Fri Dec 13 11:37:58 2013, Stefan Ritt, input protection in DRS4 evaluation board 


    
        
            Dmitry Hits wrote:
        
        
            
            Last month I was using a DRS4 evaluation board to digitise the signal from the charged particles in
Entry  Thu Oct 14 15:19:00 2021, Keita Mizukoshi, livetime (or deadtime) of DRS4 evaluation board 
Dear experts,

 

I would like to use the DRS4 evaluation board for actual physics experiment.
    Reply  Thu Oct 14 15:25:07 2021, Stefan Ritt, livetime (or deadtime) of DRS4 evaluation board 
The one thing you can do easily is to look at the scaler values. If one channel counts all physical events, and you have all read out events, then the
ratio give you the live/deadtime. The hardware scalers also keep running during the DRS readout.

Stefan
    Reply  Thu Oct 14 18:03:52 2021, Keita Mizukoshi, livetime (or deadtime) of DRS4 evaluation board 
Thank you very much for your response.
Excuse me for my very stupid confirmation.
If I take N events finally and the hardware scaler value is M, the livetime is realtime*(N/M). Is this correct?
    Reply  Thu Oct 14 18:42:31 2021, Stefan Ritt, livetime (or deadtime) of DRS4 evaluation board 
I would say not exactly, but it's a good approximation.




Keita
Mizukoshi wrote:



Thank you very much for your response.
    Reply  Fri Oct 15 06:15:53 2021, Keita Mizukoshi, livetime (or deadtime) of DRS4 evaluation board 
Thank you very much.




Stefan
Ritt wrote:



I would say not exactly, but it's a good approximation.