  604   Thu Apr 13 17:10:58 2017 Christian FarinaStand-alone Time Calibration for PSI Board

Thank you for your help Stefan. I will try to get the TC part isolated.

Stefan Ritt wrote:

Than you can try to isolate the code. Note that different SCAs might work differently. Like the DRS4 has a channel-to-channel jitter which others might not. But you will see.

Stefan

Christian Farina wrote:

Hi Stefan,

Thank you for your reply. I have read the paper already. I looked through the code and I understand that the LTC and GTC are performed by the AnalyzeSlope and AnalyzePeriod functions, respectively, correct? It seems to me to be a complicated business to re-write that part from scratch, at least for an inexperienced programmer like me. It made more sense to try to isolate that part from the original DRS.cpp. Ideally, I would like to have a stand-alone program that would work on any SCA without references to the drs hardware specifics.

Stefan Ritt wrote:

Hard to say. Timing calibration is quite delicate. If you start from scratch, better read this paper: https://arxiv.org/abs/1405.4975

If you try to extract the code from DRS.cpp, better read the paper, too. Probably it will not be possible to develop or extract the code without knowing how it works. 

Best,
Stefan

Christian Farina wrote:

Hello everybody,

I was trying to create a stand-alone program that would perform a time calibration on the board. My goal would be the following.

- acquire  about 10k sinus waveforms

- write them to disk (also for later reanalysis)

- run the time calibration on the recorded data

- store the clibration results in a file / database

Being not an expert, my question here is the following. Would it be easier to try to isolate the time calibration part from the DRS.cpp source code or re-write entirely the code from scratch?

Thanks.

 

 

 

 

  409   Wed May 13 00:52:51 2015 Cosmin DeaconuGetting Trigger Source

I'd like to be able to know which channel (0,1,2,3 or external) was responsible for the trigger.  DRSBoard::GetTriggerSource() seems to always return 1.  Is there a way to get this information?  Using the DRS4 evaluation board and software version 5.0.3.

Thanks,

Cosmin

 

  410   Wed May 13 01:07:36 2015 Cosmin DeaconuDRS4 Evaluation Board + Powered USB Hub

I am trying to use 4 evaluation boards with a powered USB hub (since eventually, I will have to do this on a laptop).  It seems like destroying the DRS object is insufficent to properly close the boards when on the hub (i.e. I get usb read errors next time I run my program). When all the boards are plugged into the computer, all is fine.  This is on Linux using libusb1. My guess is something about resetting the port doesn't work properly (but maybe that's this particular hub's fault?).  Has anyone else experienced a similar issue. If not, can someone recommend a hub that is known to work?

  485   Mon Mar 21 10:38:27 2016 Daniel DribinDRS Oscilloscope freezing after a long run

Dear Stefan Ritt,

I am using a DRS4 v5 to do timing measurements of Positron lifetime. I use the DRS Oscilloscope with triggering on 2 channels when I have a coincidence. Attached is a picture with all the setting that I use. When I use the DRS4 for a long measurements of 5 million events for a couple of hours, the DRS Oscilloscope stops showing any signal .After the first restart of the program I get a strange signal which is at the bottom of the scope range of voltage picture below(in the picture I changed the vertical positions of the channels for better viewing). Only after a couple of DRS Oscilloscope restarts and USB reconnections do I get the results again.

I currently am using another DRS4 v5 and the same situation occurs again although with lower frequency.

What can I do to solve this problem?

thank you very much,

Daniel

 

Attachment 1: drs_settings.png
drs_settings.png
Attachment 2: empty_drs.png
empty_drs.png
Attachment 3: drs_ofset.png
drs_ofset.png
  501   Mon Apr 4 11:41:26 2016 Daniel DribinDRS Oscilloscope freezing after a long run

Dear Stefan Ritt,

Yes I use Windows 7, If the DRS Oscilloscope program stays on for a couple of hours without saving the data, the problem will occur. It seems it happens more often when there is data writing and when the rate of events is slow, about 100 events per second, at high rates it almost doesn't happen. Can it be temperature related?

Daniel

Stefan Ritt wrote:

Dear Daniel,

sorry my late reply, I'm pretty busy these days. The behavior you report has not been seen before, but I guess no one tried to take such long runs of data yet. Can you confirm that the problem also occurs without writing data to disk, or is it disk-related? I guess you use it under Windows 7, right?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

I am using a DRS4 v5 to do timing measurements of Positron lifetime. I use the DRS Oscilloscope with triggering on 2 channels when I have a coincidence. Attached is a picture with all the setting that I use. When I use the DRS4 for a long measurements of 5 million events for a couple of hours, the DRS Oscilloscope stops showing any signal .After the first restart of the program I get a strange signal which is at the bottom of the scope range of voltage picture below(in the picture I changed the vertical positions of the channels for better viewing). Only after a couple of DRS Oscilloscope restarts and USB reconnections do I get the results again.

I currently am using another DRS4 v5 and the same situation occurs again although with lower frequency.

What can I do to solve this problem?

thank you very much,

Daniel

 

 

 

  506   Wed Apr 6 09:43:52 2016 Daniel DribinDRS Oscilloscope freezing after a long run

At hight rates I worked with files of up to 20 GB so I don't think this is the problem.

I will try to run it under Ubuntu and see if i can recreate the problem.

Thank you very much for the quick responses and help.

Stefan Ritt wrote:

Even with writing for one night no problem (see below). Have you checked how big your data file is? I guess there is a limit under Windows of 2 GB. If that's the case, you have to write shorter files.

 

 

  507   Wed Apr 6 09:46:10 2016 Daniel DribinDRS Oscilloscope freezing after a long run
Martin Petriska wrote:

 

Stefan Ritt wrote:

I tried this night to run the board at a 10 Hz rate with an external pulser, without writing, and it did not freeze after ~14 hours of running on Mac OSX. This night I will try again with writing.

Stefan

Stefan Ritt wrote:

Then it seems that there is some USB communication problem. I heard this also from other people, that the USB data transfer under Windows has sometimes problems. I develop and run the board under Mac OSX, and there the same software runs for days without problem. So I guess it's related to the underlying libusb lib which is used by the DRS oscilloscope, on which I have no influence. So the only advice I can give is to take shorter series of data. Anyhow the board is not considered a full DAQ system, just an "evaluation board" which means one can try the DRS4 chip and play with it. For serious business one should build own electronics with the chip. Anyhow we are currently developping an Ethernet board which allows much faster acquisition rates, so USB will be obsolete some day. Nevertheless I will try to reproduce your problem and see if I can do anything. At what trigger rate does it show up most prominently?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

Yes I use Windows 7, If the DRS Oscilloscope program stays on for a couple of hours without saving the data, the problem will occur. It seems it happens more often when there is data writing and when the rate of events is slow, about 100 events per second, at high rates it almost doesn't happen. Can it be temperature related?

Daniel

Stefan Ritt wrote:

Dear Daniel,

sorry my late reply, I'm pretty busy these days. The behavior you report has not been seen before, but I guess no one tried to take such long runs of data yet. Can you confirm that the problem also occurs without writing data to disk, or is it disk-related? I guess you use it under Windows 7, right?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

I am using a DRS4 v5 to do timing measurements of Positron lifetime. I use the DRS Oscilloscope with triggering on 2 channels when I have a coincidence. Attached is a picture with all the setting that I use. When I use the DRS4 for a long measurements of 5 million events for a couple of hours, the DRS Oscilloscope stops showing any signal .After the first restart of the program I get a strange signal which is at the bottom of the scope range of voltage picture below(in the picture I changed the vertical positions of the channels for better viewing). Only after a couple of DRS Oscilloscope restarts and USB reconnections do I get the results again.

I currently am using another DRS4 v5 and the same situation occurs again although with lower frequency.

What can I do to solve this problem?

thank you very much,

Daniel

 

 

 

 

 

Hi I have also positron annihilation system based on DRS4v4 cards. Its running several weeks, sometimes months, without freezing in windows7 64bit system (Pentium Core Quad 2Ghz, 4GRam).  Problem was when widows was trying to install updates and restarted PC. In beginning I had some problem with memory leak in my application, but it was simple seen in task manager that application memory was rising and was need to find memory leak in application code. Now I remember card was sometimes freezing when room air conditioning with 2kW was starting and high electricity pulses were reason of USB problems, it helped to put air conditioner and PC in different power line input. Hope it help to solve zour problems.

Martin

I too have air conditioning in the room, not sure to which power line it is connected, but I'll try to check this aswell.

Thank you very much.

  400   Thu Mar 19 07:37:52 2015 Daniel Stricker-ShaverRunning 2 instances of a DRS DAQ program

I don't know if it helps, but we measured the time resolution between two independendly running v3 boards using a single PC (latest software) in Linux. (http://arxiv.org/abs/1405.4975)

You start the DRS DAQ program with only one USB board connected, first. Afterwards connect the second board and start another session. If you externally trigger (global) both boards with less than 1 Hz, you can garantee that both programs save the same events independently (from two individual DRS boards).

Daniel

Stefan Ritt wrote:

I never had in mind running two systems in parallel, that's why the code claims all interfaces when started. You have to dig into the usb code which is located in musbstd.c at function musb_open(). There you will find a line libusb_claim_interface() which requests exclusive access to the usb subsystem. The code is there because I copied it from some standard example for the libusb library. You have to read the documentation for libusb (http://libusb.sourceforge.net/api-1.0/) and see if you can get rid of that. Probaby you have to claim/release the interface on each access, but I never tried that.

Stefan

Hermann-Josef Mathes wrote:

Hi,

we want to run two instances of our little DRS DAQ program but obviously the first instance started always claims all DRS boards for itself and the other one exits with an error. The 2 boards used in the example below have the serial number # 2413 and #2414 and are v5 boards.

The first one:

mathes@ikauger5:~/src/DRS4/Cpp> ./drsdaq -b 2413
DRSController: found board with serial number #2413
DRSController: found board with serial number #2414
DRSController: using board with serial number #2413
CalibratedFrequency= 1.00721
====================================
DRS type:            DRS4
Board type:          9
Serial number:       2413
Firmware revision:   21260

...
And the second one:

mathes@ikauger5:~/src/DRS4/Cpp> ./drsdaq -b 2414
musb_open: usb_set_configuration() error -6
musb_open: Found USB device 0x04b4:0x1175 instance 0, but cannot initialize it: please check permissions on "/proc/bus/usb/1/7" and "/dev/bus/usb/1/7"
USB successfully scanned, but no boards found
...


How can our goal be achieved?

Thanks

Hermann-Josef

 

 

  407   Tue Apr 21 13:03:38 2015 Daniel Stricker-ShaverDRS4 Evaluation Board Baseline/Voltage Calibration

I also use Ubuntu 14.04 LTS and for my V3 borad I have to use drsosc 4.x or ealier to perform the calibration.

Stefan Ritt wrote:

1) I tried to cablirate a V5 board with drsosc 5.0.3 and it just worked fine for me. No idea what went wrong in your case.

2) The "found 4096 stuck pixels on this board" can be safely ignored. It comes from the fact that the standard evaluation board has four cannels unconnected (the DRS4 chip has 8 channels, four are connected to in the evaluation board and four are unconnected). So the software sees wrong values on four channels because they are unconnected and thinks something is wrong. Unfortunately the software cannot determine if the channels are connected or not. So just ignore it.

3) I heard several people having to reset their boards under Linux in a similar way than you. This is probalby due to some instability in the USB part of the linux kernel, since the problem does not occur on other systems (Windows, Mac OSX). So I cannot do anything from the software side.

/Stefan

Julien Wulf wrote:

Hi,

I`m trying to calibrate my DRS4 evoluation board to an input range of 0-1V but it doesn`t work.

1) First I tried to calibrate it with the drsosc (version 5.0.3) Software. The -0.5V - 0.5V calibration works, but during the 0 - 1V calibration the Software crashes.

2) I also tried to calibrate the input range with a C++ DAQ Package (based on drs_exam). Here the code of the calibration:

    ....

    b->SetInputRange(0.) (Center at 0 V )

    b->CalibrateVolt(NULL);

   ....

   Calibration Works

   ....

   b->SetInputRange(0.5) (Center at 0.5 V )

   b->CalibrateVolt(NULL);

  ....

   Results in: Found 4096 stuck pixels on this board.

Did I do a mistake or is this a normal behaviour of the board? Also the board often crashes and I get a magic number 0000 after restarting the DAQ. Then the board needs to be restarted via pulling the plug. ( I ensured that I terminate the USB connection before I close the program with "delete drs"). Is there a possibility to avoid this error?

My OS: Ubuntu 14.04 LTS.

Ciao,

Julien

 

 

  511   Sat Apr 23 12:33:17 2016 Daniel Stricker-ShaverNegative fCellDT values from GetTimeCalibration()

Hi Kyle,

If I remember right the negative sampling width happens only for 498 and at high sampling speeds. It is described in a paper from Stefan:

http://arxiv.org/pdf/1405.4975.pdf

or

“Novel Calibration Method for Switched Capacitor Arrays Enables Time Measurements With Sub-Picosecond Resolution”( IEEE Transactions on Nuclear Science 61 (2014),Nr. 6, 3607–3617)

Kyle Weinfurther wrote:

Hello Stefan,

I am using four DRS4 v5 eval boards to digitize 16 channels of data. I have recently changed from saving the timing information of the waveform using GetTime() to GetTimeCalibration(). When changing over, I noticed that some values for fCellDT for cell 498 are negative. Over the 16 channels used, 4 of them have negative time bin widths for cell 498 while the other 12 channels are very close to 0 (in the ~10 ps range). One of the eval boards has no negative fCellDT whereas the other three boards have one or two channels with negative values.

Upon further inspection, I checked the time between samples of GetTime() and found the same results in cell 498. After finding this, I did a timing calibration again with CalibrateTiming() even though in a different post on the discussion forum you said it was valid for a wide range of temperatures and a long time (years). This still allowed the negative fCellDT values to persist.

Is this a common occurance? If so, is there a method to fix this issue? Is there a reason for cell 498 to have a small value for fCellDT? I searched the discussion forum and did not find anything relating to this issue.

Attached are a couple waveform traces using GetTime() zoomed in on cell 498.

Thanks,

Kyle Weinfurther

 

  542   Sun Oct 9 10:43:35 2016 Danny Petschketime difference between 2 channels only ~30-35ps @ 5GSmples/s

(Board Type:9, DRS4)

Hello,

I´m trying to reach the timig resolution of about 2.5ps as written in the manual. 

My settings are:

5GSamples/s

+/-0.5V

I followed the instructions of the manual. The chip was warm and ran about 10h. Then, Timing- followed by Voltage-Calibration.

The test-signal is a splittet sine-wave of 20MHz (function-generator) brought on A0 and A1 (A1 signal is delayed by 1ns-cable).

I´ve been testing different trigger-logic: (Chn1 AND Chn2), (Chn1 OR Chn2) and only Chn1 or Chn2. 

Trigger-levels were changed too.

All setups show the same result of 1.009ns +/- 30-35ns (results from the DRS-Oscilloscope).

What is wrong from my side? 

Thanks a lot for your help

  544   Mon Oct 10 11:30:37 2016 Danny Petschketime difference between 2 channels only ~30-35ps @ 5GSmples/s

Hello Stefan,

Chn2 & Chn3 were used for delay-determination as you can see on the second picture.

 

The second picture shows all 4 Channels without any voltage input.

On Channel 4 streaks (red circle) occur often and Channel 1 has totally different Offset (Picture 1). 

Thanks

 

Stefan Ritt wrote:

Can you post a screenshot of your measurement?

Stefan

Danny Petschke wrote:

(Board Type:9, DRS4)

Hello,

I´m trying to reach the timig resolution of about 2.5ps as written in the manual. 

My settings are:

5GSamples/s

+/-0.5V

I followed the instructions of the manual. The chip was warm and ran about 10h. Then, Timing- followed by Voltage-Calibration.

The test-signal is a splittet sine-wave of 20MHz (function-generator) brought on A0 and A1 (A1 signal is delayed by 1ns-cable).

I´ve been testing different trigger-logic: (Chn1 AND Chn2), (Chn1 OR Chn2) and only Chn1 or Chn2. 

Trigger-levels were changed too.

All setups show the same result of 1.009ns +/- 30-35ns (results from the DRS-Oscilloscope).

What is wrong from my side? 

Thanks a lot for your help

 

 

Attachment 1: allChannels_zero_scaled.png
allChannels_zero_scaled.png
Attachment 2: Chn2_Chn3_1ns_delay_scaled.png
Chn2_Chn3_1ns_delay_scaled.png
  546   Tue Oct 11 09:04:33 2016 Danny Petschketime difference between 2 channels only ~30-35ps @ 5GSmples/s

Hello Stefan,

thanks for the paper. That makes sense. I thought about sth. like that but wasn`t sure. Couldn´t check higher frequencies (limit of my function generator).

What do think about the other picture I attached yesterday where Chn1 shows a totally different offset than Chn2-4. Moreover Chn4 shows some streaks (red circle) ?

Best regards

Danny

Stefan Ritt wrote:

Ok, I got it. The timing resolution is affected by the signal-to-noise ratio over the rise-time of your signal. You find the full formula herer:

https://arxiv.org/abs/1405.4975

Your sine wave input signal has a slow rise time, and therefore limits the time resolution. I reproduced your measurement with a 20 MHz sine wave and got the same result:

 

If I increase the frequency to 100 MHz and increase the amplitude, I get a better resolution:

 

This is 5 ps which is better than 37 ps, but still not 2.5 ps. This can only be reached by sending single pulses to the evaluation board which have a rise time of > 300 mV / ns, which can be seen here:

 

It is important to understand the relation timing - resolution vs. rise time / noise as explained in the quoted paper. If you have tiny pulses from your detector, you never will be able to measure excellent timing. This is physics, and not related to the specific electronics you are using.

Best regards,
Stefan

 

 

 

 

 

 

  584   Sat Jan 28 14:11:58 2017 Danny PetschkeAND trigger problems

 Dear Stefan,

I have 2 identical pulses as a splittet signal with an amplitude of 300mV. Range is -0.5-0.5V, 5.12GSamp using the Evaluation-Board. Both signals are triggered in AND logic. One of the signals is delayed by a fixed value of 1-50ns for testing. On increasing the trigger Level from 10% to 50% of amplitude (pulse rise time is 2.5ns) pulses cannot anymore triggered above 4-5ns delay. It means there is a proportionality between the trigger level and the available range where 2 signals can be triggered in AND logic (Time-difference between 2 pulses). Do I anything misunderstand or is the time the comparator needs by higher trigger Levels for comparation longer than the 200ns at 5.12GSamp?

Board was timing and voltage calibrated before.

Thx

Danny

  774   Mon Oct 14 09:32:33 2019 Danyanghow to acquire the stop position with channel cascading

Hi Steffan,

       In DSR4 DATASHEET Rev.0.9 Page13,  there is a paragraph "If the DRS4 is configured for channel cascading or daisy chaining, it is necessary to know which the current channel is where the sampling has been stopped. This can be
determined by addressing the Write Shift Register withA3-A0 = 1101b and by applying clock pulses to the SRCLK input ...".

       My question is the timing details about srclk, srout, A3-A0 in the above control and its timing relation with stop shift register (Figure 15).  And can this configuration be used in the full readout mode with output MUXOUT?               

Best Regards,
Danyang (sun2222@mail.ustc.edu.cn)

Attachment 1: Capture.PNG
Capture.PNG
  776   Mon Oct 14 11:45:06 2019 Danyanghow to acquire the stop position with channel cascading

I tried the logic in my designed board.  The results are shown in the picture: Srout keeps low when A3-A0  is set to 1101 and srclk is set as you mentioned. And the drs4 chip does not output sine wave in such configuration.

Srout signal only reacts after the rsrload signal is pulled high and A3-A0 is not 1101.  

The number of srclk is not enough?  Is there any recommended time to configure the command?

 

Best Regards,
Danyang

 

Stefan Ritt wrote:

You first set A3-A0, on the next clock cycle you issue pulses on srclk, and about 10ns after each clock pulse the output shows up at srout. Best is to verity this with an oscilloscope.

The radout of the shift register is independent of the readout mode, so you can use with with MUXOUT as well.

Stefan

Danyang wrote:

Hi Steffan,

       In DSR4 DATASHEET Rev.0.9 Page13,  there is a paragraph "If the DRS4 is configured for channel cascading or daisy chaining, it is necessary to know which the current channel is where the sampling has been stopped. This can be
determined by addressing the Write Shift Register withA3-A0 = 1101b and by applying clock pulses to the SRCLK input ...".

       My question is the timing details about srclk, srout, A3-A0 in the above control and its timing relation with stop shift register (Figure 15).  And can this configuration be used in the full readout mode with output MUXOUT?               

Best Regards,
Danyang (sun2222@mail.ustc.edu.cn)

 

 

Attachment 1: Capture.PNG
Capture.PNG
  778   Mon Oct 14 13:44:26 2019 Danyanghow to acquire the stop position with channel cascading

Yes, firstly I configured the chip with 4x2048 bins by setting the Write Shift Register to 01010101b, A3-A0 keeps 1101----> secondly I enabled the domino wave, wait  some time for stable,  A3-A0 keeps 1111  ---->thirdly stops the domino wave when the trigger comes, A3-A0 keeps 1101 (or 1010, 0000)----> forthly send the clock pulse to the srclk pin, A3-A0 keeps 1101,  srout pin keeps low----> fifthly enable rsrload, A3-A0 (0000-1000),  srout pin reacts nomally.   I think the cascading is worked when I checked the waveform on the oscilloscope. Is there any step I missed?

Best Regards,
Danyang

Stefan Ritt wrote:

Note that you have to read out the Write Shift Register only if you do channel cascading, e.g. configuring the chip with 4x2048 bins by setting the Write Shift Register to 01010101b. Then the Write Shift Register tells you in which 1024-bin segment the Domino Wave has been stopped. If you use the normal 8x1024 bin mode, you don't have to read out the Write Shift Register since it continas only 1's. 

Stefan

Danyang wrote:

I tried the logic in my designed board.  The results are shown in the picture: Srout keeps low when A3-A0  is set to 1101 and srclk is set as you mentioned. And the drs4 chip does not output sine wave in such configuration.

Srout signal only reacts after the rsrload signal is pulled high and A3-A0 is not 1101.  

The number of srclk is not enough?  Is there any recommended time to configure the command?

 

Best Regards,
Danyang

 

Stefan Ritt wrote:

You first set A3-A0, on the next clock cycle you issue pulses on srclk, and about 10ns after each clock pulse the output shows up at srout. Best is to verity this with an oscilloscope.

The radout of the shift register is independent of the readout mode, so you can use with with MUXOUT as well.

Stefan

Danyang wrote:

Hi Steffan,

       In DSR4 DATASHEET Rev.0.9 Page13,  there is a paragraph "If the DRS4 is configured for channel cascading or daisy chaining, it is necessary to know which the current channel is where the sampling has been stopped. This can be
determined by addressing the Write Shift Register withA3-A0 = 1101b and by applying clock pulses to the SRCLK input ...".

       My question is the timing details about srclk, srout, A3-A0 in the above control and its timing relation with stop shift register (Figure 15).  And can this configuration be used in the full readout mode with output MUXOUT?               

Best Regards,
Danyang (sun2222@mail.ustc.edu.cn)

 

 

 

 

  780   Tue Oct 15 08:14:17 2019 Danyanghow to acquire the stop position with channel cascading

Thanks a lot. The problem is solved when A3-A0 is set 1101 and srclk keeps low.

Best Regards,
Danyang

Stefan Ritt wrote:

If you configure the Write Shift Register with 01010101b, then all you have to do after a trigger is to set A3-A0 to 1101. The WSROUT pin shows you then either ther state 01010101b or 10101010b, you the pin should be 1 or 0, and that's all you need. The Write Shift Register is NOT routed to the SROUT pin, you only see it at the WSROUT pin.

Stefan

Danyang wrote:

Yes, firstly I configured the chip with 4x2048 bins by setting the Write Shift Register to 01010101b, A3-A0 keeps 1101----> secondly I enabled the domino wave, wait  some time for stable,  A3-A0 keeps 1111  ---->thirdly stops the domino wave when the trigger comes, A3-A0 keeps 1101 (or 1010, 0000)----> forthly send the clock pulse to the srclk pin, A3-A0 keeps 1101,  srout pin keeps low----> fifthly enable rsrload, A3-A0 (0000-1000),  srout pin reacts nomally.   I think the cascading is worked when I checked the waveform on the oscilloscope. Is there any step I missed?

Best Regards,
Danyang

Stefan Ritt wrote:

Note that you have to read out the Write Shift Register only if you do channel cascading, e.g. configuring the chip with 4x2048 bins by setting the Write Shift Register to 01010101b. Then the Write Shift Register tells you in which 1024-bin segment the Domino Wave has been stopped. If you use the normal 8x1024 bin mode, you don't have to read out the Write Shift Register since it continas only 1's. 

Stefan

Danyang wrote:

I tried the logic in my designed board.  The results are shown in the picture: Srout keeps low when A3-A0  is set to 1101 and srclk is set as you mentioned. And the drs4 chip does not output sine wave in such configuration.

Srout signal only reacts after the rsrload signal is pulled high and A3-A0 is not 1101.  

The number of srclk is not enough?  Is there any recommended time to configure the command?

 

Best Regards,
Danyang

 

Stefan Ritt wrote:

You first set A3-A0, on the next clock cycle you issue pulses on srclk, and about 10ns after each clock pulse the output shows up at srout. Best is to verity this with an oscilloscope.

The radout of the shift register is independent of the readout mode, so you can use with with MUXOUT as well.

Stefan

Danyang wrote:

Hi Steffan,

       In DSR4 DATASHEET Rev.0.9 Page13,  there is a paragraph "If the DRS4 is configured for channel cascading or daisy chaining, it is necessary to know which the current channel is where the sampling has been stopped. This can be
determined by addressing the Write Shift Register withA3-A0 = 1101b and by applying clock pulses to the SRCLK input ...".

       My question is the timing details about srclk, srout, A3-A0 in the above control and its timing relation with stop shift register (Figure 15).  And can this configuration be used in the full readout mode with output MUXOUT?               

Best Regards,
Danyang (sun2222@mail.ustc.edu.cn)

 

 

 

 

 

 

  724   Thu Nov 8 11:44:35 2018 Davide DepaoliTiming Issue
Hi,

We are using the DRS4 Evaluation Board as a digitizer in our laboratory.
We found a strange behavior in the saved files, more specifically the time difference between two consecutive points is not constant, also after the Timing Calibration.
As an example, I paste a piece of a xml file saved using the drsosc program, acquiring CH1 (open):
	
	---------------------------
	---[ START XML EXAMPLE ]---
	---------------------------
	
	<?xml version="1.0" encoding="ISO-8859-1"?>
	<!-- created by MXML on Thu Nov  8 11:13:27 2018 -->
	<DRSOSC>
	<Event>
	<Serial>1</Serial>
	<Time>2018/11/08 11:13:27.163</Time>
	<HUnit>ns</HUnit>
	<VUnit>mV</VUnit>
	<Board_2796>
	<Trigger_Cell>216</Trigger_Cell>
	<Scaler0>0</Scaler0>
	<CHN1>
	<Data>0.000,-1.0</Data>
	<Data>1.083,-1.0</Data>
	<Data>2.143,-1.0</Data>
	<Data>2.926,-1.0</Data>
	<Data>4.249,-0.1</Data>
	<Data>4.929,-0.6</Data>
	<Data>6.075,-0.4</Data>
	<Data>7.042,0.0</Data>
	<Data>8.299,0.2</Data>
	
	[...]
	
	-------------------------
	---[ END XML EXAMPLE ]---
	-------------------------

We found the same behavior saving events in the binary format, and then reading them with the read_binary.cpp

Is there a way to fix our issue?

Thanks a lot

Davide and Alessio
  726   Thu Nov 8 12:02:34 2018 Davide DepaoliTiming Issue
Thanks a lot for the quick response.
We will do as you suggest.

Best regards

Davide and Alessio

> That's not a bug, but a feature of the DRS4 chip. The time bins have different values by the properties of the chip. They are generated by a chain of inverters, which all have different 
propagation times. This delay is measured by the time calibration and then applied. If you want equidistant bins, 
> you have to interpolate your data points (linearly or by splines) and resample the signal. You can find more details in the DRS4 data sheet.
> 
> Best,
> Stefan
> 
> 
> > Hi,
> > 
> > We are using the DRS4 Evaluation Board as a digitizer in our laboratory.
> > We found a strange behavior in the saved files, more specifically the time difference between two consecutive points is not constant, also after the Timing Calibration.
> > As an example, I paste a piece of a xml file saved using the drsosc program, acquiring CH1 (open):
> > 	
> > 	---------------------------
> > 	---[ START XML EXAMPLE ]---
> > 	---------------------------
> > 	
> > 	<?xml version="1.0" encoding="ISO-8859-1"?>
> > 	<!-- created by MXML on Thu Nov  8 11:13:27 2018 -->
> > 	<DRSOSC>
> > 	<Event>
> > 	<Serial>1</Serial>
> > 	<Time>2018/11/08 11:13:27.163</Time>
> > 	<HUnit>ns</HUnit>
> > 	<VUnit>mV</VUnit>
> > 	<Board_2796>
> > 	<Trigger_Cell>216</Trigger_Cell>
> > 	<Scaler0>0</Scaler0>
> > 	<CHN1>
> > 	<Data>0.000,-1.0</Data>
> > 	<Data>1.083,-1.0</Data>
> > 	<Data>2.143,-1.0</Data>
> > 	<Data>2.926,-1.0</Data>
> > 	<Data>4.249,-0.1</Data>
> > 	<Data>4.929,-0.6</Data>
> > 	<Data>6.075,-0.4</Data>
> > 	<Data>7.042,0.0</Data>
> > 	<Data>8.299,0.2</Data>
> > 	
> > 	[...]
> > 	
> > 	-------------------------
> > 	---[ END XML EXAMPLE ]---
> > 	-------------------------
> > 
> > We found the same behavior saving events in the binary format, and then reading them with the read_binary.cpp
> > 
> > Is there a way to fix our issue?
> > 
> > Thanks a lot
> > 
> > Davide and Alessio