DRS4 Forum
  DRS4 Discussion Forum  Not logged in ELOG logo
Entry  Wed Aug 28 13:07:51 2013, Andrey Kuznetsov, Some bug fixes and questions 
    Reply  Thu Sep 5 10:01:00 2013, Andrey Kuznetsov, Some bug fixes and questions 
       Reply  Mon Sep 9 06:49:36 2013, Andrey Kuznetsov, Some bug fixes and questions 
          Reply  Wed Jan 15 17:11:14 2014, Stefan Ritt, Some bug fixes and questions 
       Reply  Wed Jan 15 17:02:58 2014, Stefan Ritt, Some bug fixes and questions 
    Reply  Wed Jan 15 16:15:00 2014, Stefan Ritt, Some bug fixes and questions 
Message ID: 325     Entry time: Wed Jan 15 16:15:00 2014     In reply to: 289
Author: Stefan Ritt 
Subject: Some bug fixes and questions 

Andrey Kuznetsov wrote:

  For http://www.psi.ch/drs/DocumentationEN/manual_rev20.pdf:

0 0x02 15..8 board_type 5 for DRS4 USB Evaluation Board 1.1 ---> should instead say Evaluation Board 2.0 ?

Type 5 is both for board V1.1 and V2. I consider V1.1 obsolete (nobody uses it). I added code 8 for V4 and 9 for V5 

Andrey Kuznetsov wrote:

 1) if(i==100) should be if(i==1000) in function int DRSBoard::SetFrequency(double demand, bool wait)

Otherwise if PLL did not lock, i = 1000, and the if statement is evaluating it against 100, not 1000 so it never gets triggered and the error goes unnoticed.

      if (wait) {
         StartDomino();
         for (i=0 ; i<1000 ; i++)
         if (GetStatusReg() & BIT_PLL_LOCKED0)
            break;
         SoftTrigger();
         if (i==100) {
            printf("PLL did not lock for frequency %lf\n", demand);
            return 0;
         }
      }

 Absolutely correct, I changed it in the current software.

 

Andrey Kuznetsov wrote:

 2) int DRSBoard::RegulateFrequency(double demand) does not seem to work at all, the frequency does not lock for either 2 or 5 GHz, tested using DRS4 v2.0 eval board with DRS v4.0.1 and v2.0.1 software's drscl tool.

 This function is for the old DRS2 chip back in 2003 or so. That chip did not have an on-chip PLL for frequency regulation, so this was done in the FPGA. Just don't us it (who told you to???). 

 

Andrey Kuznetsov wrote:

3) In int DRSBoard::SetTriggerDelayPercent(int delay) and int DRSBoard::SetTriggerDelayNs(int delay), what is the purpose of Read and setting of "reg" if it's not being used or exported anywhere else outside of that function? Seems like Read and reg are called for nothing.

      Read(T_CTRL, &reg, REG_TRG_DELAY, 2);
      reg = (reg & 0xFF00) | ticks;
      Write(T_CTRL, REG_TRG_DELAY, &ticks, 2);

Also, I don't understand why in int DRSBoard::SetSyncDelay(int ticks), the code changes to 
      Read(T_CTRL, &reg, REG_TRG_DELAY, 2);
      reg = (reg & 0xFF) | (ticks << 8);
      Write(T_CTRL, REG_TRG_DELAY, &reg, 2);

In particular, reg = (reg & 0xFF00) | ticks; and reg = (reg & 0xFF) | (ticks << 8);  I'm not really sure but doesn't Read() with size 2 return a value that has a maximum value of 0xFF, or 8bits?   But ticks << 8, since ticks == 255 max, makes 255 << 8 => 65280, which is now a 16bit value and size 4. No? I might be wrong here, and if I am then I don't understand what's going on.   Can you please explain? In v2.0.1 the ticks were a maximum of 511 or 9bits, why did it change to 8bits?

The register REG_TRG_DELAY contains two 8-bit values, one for the trigger delay, one for the sync delay (which is currently not used). So to set one half of the register containing 16 bits, the register is read (16 bits are read with size 2, since these are two bytes), and the upper or lower 8 bits are replaced with the new value. Then the 16 bits are written back. The firmware in the FPGA does not allow to access only 8 bits of a 16-bit register.

 

Andrey Kuznetsov wrote:

 

4) A function is being called incorrectly in GetWave() in DRS.cpp

int DRSBoard::GetWave(unsigned int chipIndex, unsigned char channel, float *waveform)
{
   return GetWave(chipIndex, channel, waveform, true, fStopCell[chipIndex], false, 0, true);
}

The return is calling the following overloaded function:
int DRSBoard::GetWave(unsigned int chipIndex, unsigned char channel, float *waveform, bool responseCalib, int triggerCell, int wsr, bool adjustToClock, float threshold, bool offsetCalib)
the problem is that int wsr is not passed to the function and it thus causes implicit conversions where false is being cast into int, 0 into bool, and true into float.

That's correct. Fortunately the wrong values for the wsr does not have any influence for "normal" users. It is only used for channel cascading which is not used in the evaluation board. I will fix it in the code anyhow.

ELOG V3.1.5-2eba886