DRS4 Forum
  DRS4 Discussion Forum  Not logged in ELOG logo
Entry  Sat Feb 12 13:06:56 2022, Matias Senger, Cannot trigger on pulses, have to trigger on undershoot 
    Reply  Tue Feb 15 12:02:29 2022, Stefan Ritt, Cannot trigger on pulses, have to trigger on undershoot 
Message ID: 864     Entry time: Tue Feb 15 12:02:29 2022     In reply to: 862
Author: Stefan Ritt 
Subject: Cannot trigger on pulses, have to trigger on undershoot 

The trigger comparator is a ADCMP601 unit which requires a minimum pulse width of 3-4 ns. I see that your pulses are only 1-2 ns wide. You have to make your pulses wider in order to trigger on them.

Stefan

ELOG V3.1.5-fe60aaf