Thu Dec 13 12:03:29 2012, Evgeni, DRS-4 trigger
|
Thu Dec 13 12:14:35 2012, Stefan Ritt, DRS-4 trigger
|
Thu Dec 13 19:49:47 2012, Evgeni, DRS-4 trigger
|
Fri Dec 14 08:42:53 2012, Stefan Ritt, DRS-4 trigger
|
Fri Dec 14 10:07:54 2012, Evgeni, DRS-4 trigger
|
Fri Dec 14 10:07:14 2012, Evgeni, DRS-4 trigger
|
|
Message ID: 210
Entry time: Fri Dec 14 10:07:54 2012
In reply to: 208
|
Author: |
Evgeni |
Subject: |
DRS-4 trigger |
|
|
Stefan Ritt wrote: |
Evgeni wrote: |
Stefan Ritt wrote: |
Evgeni wrote: |
How to configure DRS oscilloscope for the oscillations with an amplitude greater than the value of the exposed
in the trigger (internal).
|
Sorry, I don't understand that question. The DRS4 Evaluation board input signal range is 1V. If you have larger signals, you have to attenuante them externally.
/Stefan
|
Can I adjust the internal trigger DRS oscilloscope for signal extraction 0.5 volts (for example) from any of the four channels. In our case, there is a lot of noise with low amplitude, which must be removed during the registration. We need to record the individual pulses of higher amplitude than the noise. So we want to use the internal trigger DRS oscilloscope to cut this noise by setting up its threshold amplitude noise.
|
Sure, you can set the trigger level with the vertical slider (see attached figure). The trigger level works form -0.5V to +0.5V. Just like with a normal oscilloscope. I thought this would be obvious...
|
That's it? Thank you for your comprehensive answer. |