DRS4 Forum
  DRS4 Discussion Forum  Not logged in ELOG logo
Entry  Tue May 19 14:14:45 2015, Ilja Bekman, DRS4 firmware UCF constraints  
    Reply  Fri May 22 14:25:45 2015, Stefan Ritt, DRS4 firmware UCF constraints  firmware.zip
       Reply  Tue May 26 11:27:27 2015, Felix Bachmair, DRS4 firmware UCF constraints  
          Reply  Fri Jun 5 12:07:38 2015, Stefan Ritt, DRS4 firmware UCF constraints  
             Reply  Fri Jun 5 13:15:35 2015, Felix Bachmair, DRS4 firmware UCF constraints  
                Reply  Fri Jun 5 13:29:55 2015, Stefan Ritt, DRS4 firmware UCF constraints  
                   Reply  Fri Jun 5 13:32:03 2015, Stefan Ritt, DRS4 firmware UCF constraints  
Message ID: 429     Entry time: Fri Jun 5 13:29:55 2015     In reply to: 428     Reply to this: 430
Author: Stefan Ritt 
Subject: DRS4 firmware UCF constraints  
Do the following: 

Use the TRG OUT of the evaluation board as a "busy". Only if this signal goes low (meaning that the readout of the board is complete and the board has been restarted), then re-enable triggers in your trigger logic.

/Stefan

> Hi Stefan,
> No we only use one evaluation board. We use the evaluation board as a part of our beam test setup. It includes a telescope based on the current PSI46V2.1 CMS Pixel chip and a trigger logic board for triggering the telescope and the evaluation board. This includes a 
> handshake between every device and the tlu
> e.g. the tlu expects an answer for each trigger.
> If the trigger comes within this first 200 mus it seems that not every trigger is accepted.
> In this moment our readout would 'die' since the tlu is waiting for the handshake.
ELOG V3.1.5-fe60aaf