Entry  Sat Jul 5 04:36:13 2025, Sandeep Godiyal, Wrong Firmware Version: board has 13279, required is 15147. Board may not work properly drseval1.png

I am using DRS4 Evaluation Board Rev.: 2.0.. Command Line interface: "drscl" detecting the board with message: Found DRS4 board 0 on USB, serial #2034, firmware revision 13279, also info command also working.

In "drsosc" interface: Getting this message: Wrong Firmware Version: board has 13279, required is 15147. Board may not work properly. In "config" I checked it is detecting the board.

Software version, I am using is: "drs505.exe"

How can I resolve these issues, please guide!

    Reply  Mon Jul 7 16:53:26 2025, Stefan Ritt, Wrong Firmware Version: board has 13279, required is 15147. Board may not work properly 

You have to use the software belonging to that board. You cannot use the newest software with an old board. Look here:

  https://www.psi.ch/en/ltp-muon-physics/software-download

So you need teh software version 2.0.

/Stefan

 