Rebecca Schmitz | AND Trigger problems with 2-3 channels | Hello,
It seems that a coincidence with two fixed channels suddenly works. I don't know why.
Screenshot 1 shows the trigger settings for the coincidence with two channels.
Screenshot 2 shows the oscilloscope surface.
Screenshot 3 shows the configuration. In the background there is a coincidence with three channels.
In contrast to the coincidence with two channels this doesn't look good.
Rebecca
Stefan Ritt wrote: |
Can you post a screenshot where I can see the channel waveforms, the configuration and the trigger settings?
Stefan
Rebecca Schmitz wrote: |
Hello,
I work with the DRS4 Evaluation Board V5 and I have a problem with the software.
I have a problem with the AND trigger setting.
For this I have chosen the AND trigger function. However, when ONE channel senses an impulse, it triggers. I'm not able to see signals in the other channels, which I selected for the coincidence.
What is the cause of this? Maybe the time window?
I would like to measure a coincidence with two or three random channels. Is it possible?
Thanks for the help.
Rebecca
|
|
|
617
|
Thu Jun 8 15:52:20 2017 |
Stefan Ritt | AND Trigger problems with 2-3 channels | Can you post a screenshot where I can see the channel waveforms, the configuration and the trigger settings?
Stefan
Rebecca Schmitz wrote: |
Hello,
I work with the DRS4 Evaluation Board V5 and I have a problem with the software.
I have a problem with the AND trigger setting.
For this I have chosen the AND trigger function. However, when ONE channel senses an impulse, it triggers. I'm not able to see signals in the other channels, which I selected for the coincidence.
What is the cause of this? Maybe the time window?
I would like to measure a coincidence with two or three random channels. Is it possible?
Thanks for the help.
Rebecca
|
|
616
|
Thu Jun 8 14:26:23 2017 |
Rebecca Schmitz | AND Trigger problems with 2-3 channels | Hello,
I work with the DRS4 Evaluation Board V5 and I have a problem with the software.
I have a problem with the AND trigger setting.
For this I have chosen the AND trigger function. However, when ONE channel senses an impulse, it triggers. I'm not able to see signals in the other channels, which I selected for the coincidence.
What is the cause of this? Maybe the time window?
I would like to measure a coincidence with two or three random channels. Is it possible?
Thanks for the help.
Rebecca |
615
|
Tue May 30 21:22:10 2017 |
Esperienza Giove | Setting input range | Thank you
Stefan Ritt wrote: |
See elog:531
Esperienza Giove wrote: |
Hello,
is it possible to set a completely negative input range like -1 to 0 or -0.95 to 0.05 ?
|
|
|
614
|
Tue May 30 21:00:26 2017 |
Stefan Ritt | Setting input range | See elog:531
Esperienza Giove wrote: |
Hello,
is it possible to set a completely negative input range like -1 to 0 or -0.95 to 0.05 ?
|
|
613
|
Tue May 30 20:45:30 2017 |
Esperienza Giove | Setting input range | Hello,
is it possible to set a completely negative input range like -1 to 0 or -0.95 to 0.05 ? |
612
|
Fri May 26 08:48:25 2017 |
Stefan Ritt | Invalid magic number 0000 | There is no other way to reset the board. As I said, people running this under Windows or MacOS are fine, so maybe this calls for a change of OS.
Esperienza Giove wrote: |
Hello, thanks for your answer. Unluckily if i try to reset in this way it keeps hanging
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
Invalid magic number: 0000
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
I also tried with sudo tee /sys/bus/usb/drivers/usb/unbind and binding again; same thing happens. It seems the board needs to be reset when this happens. Is there a way to do that - to reset the board instead of usb?
Thank you
Stefan Ritt wrote:
|
Under linux, many people observed that the USB connection is unstable to the evaluation board. This must be related to the linux USB stack, since my code runs fine under MacOSX and Windows, where I use the same USB library (libusb-1.0). So I can't do anything from my side. Baybe the linux system has some tools to reset an USB endpoint. I googled it and found some proposals here:
https://askubuntu.com/questions/645/how-do-you-reset-a-usb-device-from-the-command-line
Regards,
Stefan
Esperienza Giove wrote: |
Hello everybody!
After some times i init my board, or if i stop the program during the acquisition, i get the error message "Invalid magic 0000". The only way i can solve this problem is to physically disconnect and plug in again the USB cable.
What's the point? Is there a bash command / code i could use to reset it?
Thank you very much
|
|
|
|
611
|
Thu May 25 20:20:57 2017 |
Esperienza Giove | Invalid magic number 0000 | Hello, thanks for your answer. Unluckily if i try to reset in this way it keeps hanging
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
Invalid magic number: 0000
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
I also tried with sudo tee /sys/bus/usb/drivers/usb/unbind and binding again; same thing happens. It seems the board needs to be reset when this happens. Is there a way to do that - to reset the board instead of usb?
Thank you
Stefan Ritt wrote:
|
Under linux, many people observed that the USB connection is unstable to the evaluation board. This must be related to the linux USB stack, since my code runs fine under MacOSX and Windows, where I use the same USB library (libusb-1.0). So I can't do anything from my side. Baybe the linux system has some tools to reset an USB endpoint. I googled it and found some proposals here:
https://askubuntu.com/questions/645/how-do-you-reset-a-usb-device-from-the-command-line
Regards,
Stefan
Esperienza Giove wrote: |
Hello everybody!
After some times i init my board, or if i stop the program during the acquisition, i get the error message "Invalid magic 0000". The only way i can solve this problem is to physically disconnect and plug in again the USB cable.
What's the point? Is there a bash command / code i could use to reset it?
Thank you very much
|
|
|
610
|
Thu May 25 20:17:41 2017 |
Esperienza Giove | Invalid magic number 0000 | Hello, thanks for your answer. Unluckily if i try to reset in this way it keeps hanging
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
Invalid magic number: 0000
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
I also tried with sudo tee /sys/bus/usb/drivers/usb/unbind and binding again; same thing happens. It seems the board needs to be reset when this happens. Is there a way to do that - to reset the board instead of usb?
Thank you
Stefan Ritt wrote:
|
Under linux, many people observed that the USB connection is unstable to the evaluation board. This must be related to the linux USB stack, since my code runs fine under MacOSX and Windows, where I use the same USB library (libusb-1.0). So I can't do anything from my side. Baybe the linux system has some tools to reset an USB endpoint. I googled it and found some proposals here:
https://askubuntu.com/questions/645/how-do-you-reset-a-usb-device-from-the-command-line
Regards,
Stefan
Esperienza Giove wrote: |
Hello everybody!
After some times i init my board, or if i stop the program during the acquisition, i get the error message "Invalid magic 0000". The only way i can solve this problem is to physically disconnect and plug in again the USB cable.
What's the point? Is there a bash command / code i could use to reset it?
Thank you very much
|
|
|
609
|
Tue May 23 10:24:47 2017 |
Stefan Ritt | Invalid magic number 0000 | Under linux, many people observed that the USB connection is unstable to the evaluation board. This must be related to the linux USB stack, since my code runs fine under MacOSX and Windows, where I use the same USB library (libusb-1.0). So I can't do anything from my side. Baybe the linux system has some tools to reset an USB endpoint. I googled it and found some proposals here:
https://askubuntu.com/questions/645/how-do-you-reset-a-usb-device-from-the-command-line
Regards,
Stefan
Esperienza Giove wrote: |
Hello everybody!
After some times i init my board, or if i stop the program during the acquisition, i get the error message "Invalid magic 0000". The only way i can solve this problem is to physically disconnect and plug in again the USB cable.
What's the point? Is there a bash command / code i could use to reset it?
Thank you very much
|
|
608
|
Mon May 22 18:27:56 2017 |
Esperienza Giove | Invalid magic number 0000 | Hello everybody!
After some times i init my board, or if i stop the program during the acquisition, i get the error message "Invalid magic 0000". The only way i can solve this problem is to physically disconnect and plug in again the USB cable.
What's the point? Is there a bash command / code i could use to reset it?
Thank you very much |
607
|
Thu Apr 20 06:30:13 2017 |
Strahinja Lukic | Wave rotation during transfer from the board? | Thanks.
Strahinja
Stefan Ritt wrote: |
This is correct. Actually the amplitude array is rotated already inside the DRS4 chip. So the readout starts with the stop cell plus one. If you do not do anything, the waveform is already "rotated". If you want the waveform to start with physical cell #0, you have to "unrotate" it.
Stefan
Strahinja Lukic wrote: |
Is this correct, and does this mean that the amplitude array is "rotated" already during the transfer from the board?
|
|
|
606
|
Wed Apr 19 12:17:25 2017 |
Stefan Ritt | Wave rotation during transfer from the board? | This is correct. Actually the amplitude array is rotated already inside the DRS4 chip. So the readout starts with the stop cell plus one. If you do not do anything, the waveform is already "rotated". If you want the waveform to start with physical cell #0, you have to "unrotate" it.
Stefan
Strahinja Lukic wrote: |
Is this correct, and does this mean that the amplitude array is "rotated" already during the transfer from the board?
|
|
605
|
Sat Apr 15 03:48:31 2017 |
Strahinja Lukic | Wave rotation during transfer from the board? | I don't know if this question is already documented elsewhere.
I am developing a DAQ code for the DRS evaluation board, v4 for a test beam experiment. I link parts of the existing DRS code as a library.
To understand the effect of various flags used in calls to the functions DRSBoard::GetTime() and DRSBoard::GetWave(), I performed several tests with the 100 MHz signal connected to the inputs of the chip (DRSBoard::EnableTcal()), and several tests with signals from scintillation counters.
My question is about the flag "adjustToClock" in the call to DRSBoard::GetWave(). From looking at the code, I expected it to cause the waveforms to be "rotated" to start from the trigger cell, in a similar way that the flag "rotated" in the call to DRSBoard::GetTime() does for the time array. However, "adjustToClock" seems to shift the waveforms wrongly. I.e., if I want both the time and the amplitude arrays "rotated" to start from the trigger cell, I should set rotated=true for time and adjustToClock=false for the amplitude. This is also how these functions are called in e.g., Osci::ReadWaveforms().
Is this correct, and does this mean that the amplitude array is "rotated" already during the transfer from the board?
I am using DRS evaluation board serial #2733, firmware revision 30000.
Many thanks,
Strahinja
|
604
|
Thu Apr 13 17:10:58 2017 |
Christian Farina | Stand-alone Time Calibration for PSI Board | Thank you for your help Stefan. I will try to get the TC part isolated.
Stefan Ritt wrote: |
Than you can try to isolate the code. Note that different SCAs might work differently. Like the DRS4 has a channel-to-channel jitter which others might not. But you will see.
Stefan
Christian Farina wrote: |
Hi Stefan,
Thank you for your reply. I have read the paper already. I looked through the code and I understand that the LTC and GTC are performed by the AnalyzeSlope and AnalyzePeriod functions, respectively, correct? It seems to me to be a complicated business to re-write that part from scratch, at least for an inexperienced programmer like me. It made more sense to try to isolate that part from the original DRS.cpp. Ideally, I would like to have a stand-alone program that would work on any SCA without references to the drs hardware specifics.
Stefan Ritt wrote: |
Hard to say. Timing calibration is quite delicate. If you start from scratch, better read this paper: https://arxiv.org/abs/1405.4975
If you try to extract the code from DRS.cpp, better read the paper, too. Probably it will not be possible to develop or extract the code without knowing how it works.
Best,
Stefan
Christian Farina wrote: |
Hello everybody,
I was trying to create a stand-alone program that would perform a time calibration on the board. My goal would be the following.
- acquire about 10k sinus waveforms
- write them to disk (also for later reanalysis)
- run the time calibration on the recorded data
- store the clibration results in a file / database
Being not an expert, my question here is the following. Would it be easier to try to isolate the time calibration part from the DRS.cpp source code or re-write entirely the code from scratch?
Thanks.
|
|
|
|
|
603
|
Thu Apr 13 17:02:01 2017 |
Stefan Ritt | Stand-alone Time Calibration for PSI Board | Than you can try to isolate the code. Note that different SCAs might work differently. Like the DRS4 has a channel-to-channel jitter which others might not. But you will see.
Stefan
Christian Farina wrote: |
Hi Stefan,
Thank you for your reply. I have read the paper already. I looked through the code and I understand that the LTC and GTC are performed by the AnalyzeSlope and AnalyzePeriod functions, respectively, correct? It seems to me to be a complicated business to re-write that part from scratch, at least for an inexperienced programmer like me. It made more sense to try to isolate that part from the original DRS.cpp. Ideally, I would like to have a stand-alone program that would work on any SCA without references to the drs hardware specifics.
Stefan Ritt wrote: |
Hard to say. Timing calibration is quite delicate. If you start from scratch, better read this paper: https://arxiv.org/abs/1405.4975
If you try to extract the code from DRS.cpp, better read the paper, too. Probably it will not be possible to develop or extract the code without knowing how it works.
Best,
Stefan
Christian Farina wrote: |
Hello everybody,
I was trying to create a stand-alone program that would perform a time calibration on the board. My goal would be the following.
- acquire about 10k sinus waveforms
- write them to disk (also for later reanalysis)
- run the time calibration on the recorded data
- store the clibration results in a file / database
Being not an expert, my question here is the following. Would it be easier to try to isolate the time calibration part from the DRS.cpp source code or re-write entirely the code from scratch?
Thanks.
|
|
|
|
602
|
Thu Apr 13 16:54:32 2017 |
Christian Farina | Stand-alone Time Calibration for PSI Board | Hi Stefan,
Thank you for your reply. I have read the paper already. I looked through the code and I understand that the LTC and GTC are performed by the AnalyzeSlope and AnalyzePeriod functions, respectively, correct? It seems to me to be a complicated business to re-write that part from scratch, at least for an inexperienced programmer like me. It made more sense to try to isolate that part from the original DRS.cpp. Ideally, I would like to have a stand-alone program that would work on any SCA without references to the drs hardware specifics.
Stefan Ritt wrote: |
Hard to say. Timing calibration is quite delicate. If you start from scratch, better read this paper: https://arxiv.org/abs/1405.4975
If you try to extract the code from DRS.cpp, better read the paper, too. Probably it will not be possible to develop or extract the code without knowing how it works.
Best,
Stefan
Christian Farina wrote: |
Hello everybody,
I was trying to create a stand-alone program that would perform a time calibration on the board. My goal would be the following.
- acquire about 10k sinus waveforms
- write them to disk (also for later reanalysis)
- run the time calibration on the recorded data
- store the clibration results in a file / database
Being not an expert, my question here is the following. Would it be easier to try to isolate the time calibration part from the DRS.cpp source code or re-write entirely the code from scratch?
Thanks.
|
|
|
601
|
Thu Apr 13 16:50:18 2017 |
Stefan Ritt | Stand-alone Time Calibration for PSI Board | Hard to say. Timing calibration is quite delicate. If you start from scratch, better read this paper: https://arxiv.org/abs/1405.4975
If you try to extract the code from DRS.cpp, better read the paper, too. Probably it will not be possible to develop or extract the code without knowing how it works.
Best,
Stefan
Christian Farina wrote: |
Hello everybody,
I was trying to create a stand-alone program that would perform a time calibration on the board. My goal would be the following.
- acquire about 10k sinus waveforms
- write them to disk (also for later reanalysis)
- run the time calibration on the recorded data
- store the clibration results in a file / database
Being not an expert, my question here is the following. Would it be easier to try to isolate the time calibration part from the DRS.cpp source code or re-write entirely the code from scratch?
Thanks.
|
|
600
|
Thu Apr 13 16:42:21 2017 |
Christian Farina | Stand-alone Time Calibration for PSI Board | Hello everybody,
I was trying to create a stand-alone program that would perform a time calibration on the board. My goal would be the following.
- acquire about 10k sinus waveforms
- write them to disk (also for later reanalysis)
- run the time calibration on the recorded data
- store the clibration results in a file / database
Being not an expert, my question here is the following. Would it be easier to try to isolate the time calibration part from the DRS.cpp source code or re-write entirely the code from scratch?
Thanks. |
599
|
Tue Apr 11 09:41:44 2017 |
Stefan Ritt | drs4 registers behaviour | What I do is the following: Have the RESET input unconnected. When you power up, this makes an internal reset during the power up, and that's all you need. Then configure your registers using the sequences described in the manual. Then do not touch the RESET any more.
Stefan
Giovanni Bruni wrote: |
Thank you Stefan for replying!
I have still the RESET issue in mind: how would you suggest to reset properly the DRS? Is there a particular procedure to follow instead of just sending a negative pulse to the RESET pin? Is it preferable to turn the DRS off and then restart?
Thanks!
Giovanni
Stefan Ritt wrote: |
1. WRITE SHIFT register and CONFIG registers are initialized to "1" on power up, but if you want to change that, use A0-A3 etc. as you indicated.
2. If you address the READ SHIFT register by applyin "1011" to A0-A3, the input of the register is connected to SRIN. So in fig. 11, you apply 1023x"0" plus 1x"1", which effectively clears the register and keeps one "1" at the last position, so on the next rising clock this gets shifted into position #0. If you do the readout, and NOT addresing the READ SHIFT register, then the input of that register is connected to it's output internally. Therefore the single "1" keep rotating on every 1024 clock cycles.
Giovanni Bruni wrote: |
Hej Stefan! Thank you for your answer!
Just to be sure to have understood properly:
1. Using the RESET line should be avoided. And in any case, the CONFIG register and the WRITE SHIFT register need to be initialized "by hand" using the A0-A3, SRCLK and SRIN pins. Is it correct?
2. Doing the procedure shown in Figure 11 will always inject a "1" in cell #0 of the READ SHIFT register, regardless if (before starting the procedure) there was a "1" in any other cell, right?
Thank you!
Giovanni
Stefan Ritt wrote: |
Using the RESET line to reset registers is not a good idea since it can have some bad side-effects. The READ SHIFT register is NOT affected by RESET, so you have to inititialize these registers differently. To set a "1"-value at a defined position, you have to follow figure 11 in the data sheet. Once you executed that, your "1" is always at the same posiiton (namely cell #0), so after 1024 clock cycles you arrive at the same state, and do not have to re-do fig. 11 again.
Stefan
Giovanni Bruni wrote: |
Hej everyone!
I have some questions regarding what happens to some DRS registers in some scenarios:
1. How are the registers affected by a RESET? According to the data sheet all the CONFIG REGISTER bits are initilialized to 1. But what about the WRITE SHIFT and the READ SHIFT registers? Are they affected somehow after a RESET has been applied?
2. Suppose the DRS is happily running and I have done some readouts in ROI mode, so that the only "1"-value bit in the READ SHIFT register is in a random position. If now I want to execute a FULL READOUT, should I use the procedure explained in the data sheet (figure 11) for the FULL READOUT mode? or is this procedure useless since my "1"-value bit is already set somewhere in the READ SHIFT register and therefore a ROI readout of 1024 cells would be the solution (and getting the initial position from the SROUT pin)?
Thanks a lot!
Have a nice day!