DRS4 Forum
  DRS4 Discussion Forum, Page 1 of 16  Not logged in ELOG logo
New entries since:Thu Jan 1 01:00:00 1970
Entry  Mon Aug 18 06:52:51 2025, Jonathan Bradshaw, Unexpected behaviour following RSRLOAD 
Hello

I'm working to bring up a new capture board using a DRS4 and I'm having a minor problem and a major problem.

Minor problem: if I send a reset signal into the DRS4, the PLL doesn't work right.  If I leave NRSESET pin with a wek pullup (and never
    Reply  Tue Aug 19 02:40:58 2025, Jonathan Bradshaw, Unexpected behaviour following RSRLOAD 6x
Some images

Notes:


top of the puicture shows the logic channels
Red: SRCLK
Blue: SRIN
Green: SROUT
Orange: normally
       Reply  Tue Aug 19 23:10:30 2025, Jonathan Bradshaw, Unexpected behaviour following RSRLOAD 
Turns out it was a damaged DRS4 IC.

I ported the drs4_eval5_app code onto our board and observed much the same misbehaviour.  So I bit the bullet and replaced the DRS4 IC,
and things are going better.
Entry  Sat Jul 5 04:36:13 2025, Sandeep Godiyal, Wrong Firmware Version: board has 13279, required is 15147. Board may not work properly drseval1.png
I am using DRS4 Evaluation Board Rev.: 2.0.. Command Line interface: "drscl" detecting the board with message: Found DRS4 board
0 on USB, serial #2034, firmware revision 13279, also info command also working.

In "drsosc" interface: Getting this message: Wrong Firmware Version: board has 13279, required is 15147. Board may
    Reply  Mon Jul 7 16:53:26 2025, Stefan Ritt, Wrong Firmware Version: board has 13279, required is 15147. Board may not work properly 
You have to use the software belonging to that board. You cannot use the newest software with an old board. Look here:

  https://www.psi.ch/en/ltp-muon-physics/software-download

So you need teh software version 2.0.
Entry  Thu May 8 23:41:03 2025, Jonathan Bradshaw, Handling of Write Shift Register and Write Config Register 
Hi all

We're building a product which will use two different operating modes; firstly a long capcture using channel daisy chaining (2048 samples)
and secondly a segmented capture (2 separate captures of 1024 samples each). 
    Reply  Fri May 9 08:17:50 2025, Stefan Ritt, Handling of Write Shift Register and Write Config Register 
This is correct. Setting A0-A3 to 0b1101 multiplexes the Shift Write Register to SROUT, so you will either a "0" or a "1" depending
on which of the two channels was written last.

Your segmented capture does unfortunately not work. Due to a bug in the silicon, the first (e.g. even) written channel gets half overwritten
       Reply  Tue May 13 04:10:30 2025, Jonathan Bradshaw, Handling of Write Shift Register and Write Config Register 
Hi Stefan

Just so I'm 100% clear; is there no reliable way to perform 2 segmented captures with a single DRS4 IC?

While not a showstopper, this is a bit disappointing.
          Reply  Tue May 13 08:51:34 2025, Stefan Ritt, Handling of Write Shift Register and Write Config Register 
Yes this is correct. Anyhow, even if it would be working, you would not be happy with it. After having designed ~10 boards with the DRS4 chip, I learned
the hard way that any digital activity on the board during the sampling phase is strictly forbidden. You see crosstalk up to 100's of mV in some cases
(with a preamplifier on the board, 10-20mV without preamp). So rule #1 is to keep the board as "quite" as possible when sampling the input. If
             Reply  Thu May 15 00:01:20 2025, Jonathan Bradshaw, Handling of Write Shift Register and Write Config Register 
All right, thank you for the clarification.




Stefan
Ritt wrote:



Yes this is correct. Anyhow, even if it would be working, you would
Entry  Thu May 8 23:23:19 2025, Jonathan Bradshaw, Clarification of full channel readout Screenshot_2025-05-09_093332.png
Hi all

We're working on a new product using the DRS4 IC, and want to do a full readout from cell 0 (not just Region of Interest).  I have a
couple of questions I hope you can help me with:
    Reply  Fri May 9 08:26:17 2025, Stefan Ritt, Clarification of full channel readout 
The full readout mode is not really recommended since you have to pull out the stop position separately. Just do the ROI readout using the RSRLOAD signal,
and then do 1024 samples, which also gives you the full waveform, but also the stop position in a single readout cyclce. The "full readout mode"
is more there for "historical reasons", but nobody really uses it any more.
Entry  Tue Mar 25 16:31:41 2025, Matías Tobar, drs_exam.cpp not compile DRS.hdrs_exam.cpp
Hi! i'm trying to compile drs_exam.cpp but it yields the following error lines:

I do have the "DRS.h" file on the same folder, which is clearly the file that's causing troubles. I also tried to run "DRS.cpp"
but it yields a similar error lines. ('sin definir' means 'not defined').
    Reply  Wed Mar 26 08:42:08 2025, Stefan Ritt, drs_exam.cpp not compile 
You have to link against the DRS.cpp library, plus usblib, plus ... Note there is both a Makefile and a CMakeLists.txt for it. Google how to use "make"
or "cmake".

Stefan
       Reply  Tue Apr 1 16:24:33 2025, Matías Tobar, drs_exam.cpp not compile 
Thanks! I solved it by running ./drs_exam after execute Makefile (there was an inconsistency in the files directory).

Matías




Stefan
Entry  Thu Mar 27 15:53:10 2025, Justin Tabbett, Noisy counts with adapted drs_exam.cpp Channel_1_2.pngdrs_exam.cpp
Greetings, 

I have adapted the drs_exam.cpp to allow for a user input number of channels and trigger levels.

The program mostly works well, however there are counts which form a noise peak, imposed on the regular channel response.
Entry  Fri Dec 20 20:35:31 2024, Matias Henriquez, Problem with C++ script to use DRS4 evaluation board. Not taking data. 1.pngdrs_test.cpp
Hello,

I need to write a script in C++ to take data using the DRS4 evaluation board v4. For that, I used the drs_exam.cpp example as a reference. This
is my code (see attachement 2), which is very similar to the provided example, however the difference is that I need to trigger on CH1 OR CH2. In the next
    Reply  Fri Dec 27 22:04:48 2024, Matias Henriquez, Problem with C++ script to use DRS4 evaluation board. Not taking data. 
Hello, some updates:

     4. I was able to capture correct waveforms using c++ code. I needed to use the function SetTriggerDelayNs() to properly capture
my waveforms.
       Reply  Mon Jan 6 12:52:23 2025, Stefan Ritt, Problem with C++ script to use DRS4 evaluation board. Not taking data. 
1. Transparent mode is not needed for the hardware trigger, no idea why the code is there. You can probably remove it.

2. EnableTCal is only for the sake of having some waveforms at the input. Indeed you have to disable it to sample real signals.

3. EnableTrigger() is there to enable the hardware trigger. The flag2 is ther for historical reasons (used in older versions of the board).
Entry  Sun Sep 23 02:22:46 2018, Gerard Arino-Estrada, Trigger OUT pulse width variable from 100 us up to 100 ms 
Hello Stefan,

I am using the DRS4 board connected to a Raspberry PI and through the drsosc application. I am interested on using the "Trigger OUT"
signal to do some extra data processing with NIM modules. According to the manual, for each hardware trigger a TTL pulse of 150 ns width should be send
    Reply  Wed Sep 26 14:44:14 2018, Stefan Ritt, Trigger OUT pulse width variable from 100 us up to 100 ms 
The "Trigger OUT" has changed recently. It goes high on a new trigger, but then STAYS high until the board has been read out by the PC and
re-started. This allows better synchronization with some external trigger, which can be re-armed with the falling edge of the trigger out signal. The signal
can be quite long, since readout of an event via USB typically takes 2 ms, but can be more if the PC is busy. If you need back your 150 ns pulse,
       Reply  Wed Sep 26 18:28:20 2018, Gerard Arino-Estrada, Trigger OUT pulse width variable from 100 us up to 100 ms 
Thank you very much for the answer, I really appreciate your help.

Thanks!

Gerard
          Reply  Wed Sep 26 19:21:03 2018, Stefan Ritt, Trigger OUT pulse width variable from 100 us up to 100 ms 
In meantime I even updated the manual.

Stefan




Gerard
       Reply  Mon Dec 23 19:31:31 2024, Matias Henriquez, Trigger OUT pulse width variable from 100 us up to 100 ms 
Given this new scenario, what is the maximum rate of events that can be processed then (a rough estimation would be great, 1/2ms?)? is it mainly limited
by the USB data transmission and the PC? How does the logic of the trigger and DRS4 data sampling works inside the FPGA in general terms? e.g: trigger
activated -> dwrite ON -> ADC acquisition -> busy until data has been shipped off to the PC -> free to process new events. 
Entry  Tue May 21 18:13:08 2024, Rebecca Hicks, Error when running drsosc 
Hi, I'm a student trying to figure out the DRS4 board. I cloned the github repo, but when I run drsosc, I get an error: Gtk-Message: 10:06:38.376:
Failed to load module "canberra-gtk-module". I'm not sure what that means. The oscilloscope window does open up for me though. Thanks for
any help!
    Reply  Fri Jun 28 23:33:51 2024, Patricia Lecomti, Error when running drsosc 
Salut !

Je vois que tu rencontres un petit problème avec ton installation. Le message "Gtk-Message: Failed to load module 'canberra-gtk-module'"
indique que ton système essaie de charger un module GTK spécifique qui n'est pas installé. Heureusement, ce n'est pas un problème
Entry  Thu Feb 22 01:21:11 2024, Rod McInnis, Simulation of FPGA 
Hello:

A bit of background:  I am working on a project that is utilizing the DRS4 Evaluation board as a prototype platform for a dedicated, special
use capture. We will only be utilizing one channel of the ADC capture, and the 1024 samples is more than enough. 
    Reply  Thu Feb 22 10:37:03 2024, Stefan Ritt, Simulation of FPGA 
The Cypress has its own firmware, contained in the distribution under firmware/CY7C68013A/drs_eval.c. There you can see how the data is fetched. I kind
of forgot how exactly it worked, since I wrote that code back in 2011. But most if the Cypress code is just the configuration of the USB, the communication
with the FPGA is kind of straight forward in the Cypress implementation. But you have to read the manual of that chip to understand it.
Entry  Wed Oct 25 19:44:25 2023, John Westmoreland, WaveDREAM Design 
Hello All,

Are there any design resources available for the WaveDREAM PCBA's?

Thanks In Advance,
    Reply  Wed Oct 25 19:47:23 2023, Stefan Ritt, WaveDREAM Design 
No. This is a proprietary design.

Best,
Stefan
       Reply  Wed Oct 25 19:52:33 2023, John Westmoreland, WaveDREAM Design 
Stefan,

Oh, didn't realize that.

Thanks!
Entry  Wed Jun 10 12:46:43 2009, Stefan Ritt, Input range switch added in Version 2.1.3 Capture.png
 A new software verison for the DRS4 Evaluation Board has been has been released. Version 2.1.3 adds a switch for the input range of the DRS4 board.
Once can choose between -0.5V...0.5V and 0V...1V:
    Reply  Tue Sep 5 03:28:52 2023, Matias Henriquez, Input range switch added in Version 2.1.3 
Hello,

It is not quite clear to me yet how the input range is only determined by the front end and not the DRS4 chip. According to the datasheet, the
selection of ROFS determines whether the input differential range is -0.5V to 0.5V (ROFS=1.55V) or 0V to 1V (ROFS=1.05V) or -0.05V to 0.95V (ROFS=1.1V).
       Reply  Wed Sep 13 13:18:45 2023, Stefan Ritt, Input range switch added in Version 2.1.3 
To achieve an input range of -1V to 0V, you need an external buffer which can shift this range into the DRS4 range of -0.5V to +0.5V. This external buffer
has then to operate with bipolar power supplies, like -2.5V to +2.5V, which are not present on the evaluation board.

Best regards,
Entry  Fri Jun 9 04:11:40 2023, Javier Caravaca, Different sampling rates in multi-board configuration 
Hello,

Is it possible to have different sampling rates in multi-board configuration? I tried using the scope application but I am unable to change
the sampling rate independently.
    Reply  Mon Jun 12 14:22:04 2023, Stefan Ritt, Different sampling rates in multi-board configuration 
No, that's unfortunately not possible.

Stefan




Javier
Entry  Mon Oct 17 16:29:37 2022, Sebastian Infante, DRS4 installation via tar in ubuntu not working 
Hello i cant install any the last versions that i downloaded from the dropbox, i can untar the file called drs-5.0.6 and when i type "make"
while inside the extracted folder that starts working properly till a point and i get an error, its worth mention that i installed wxWidgets and could
make a simple hello world that worked properly in wxWidgets.
    Reply  Mon Feb 6 13:28:28 2023, Stefan Ritt, DRS4 installation via tar in ubuntu not working 
I fixed the described error. Can you try the new version from https://bitbucket.org/ritt/drs4eb/commits/80b3af753ed32eb365725f0f3244a4109347c01b




Sebastian
Infante wrote:



Hello i cant install any the last versions that i downloaded from
Entry  Sat Oct 22 13:24:20 2022, Phan Van Chuan, Channel Cascading Option in the 2048-bin DRS4V51.png
Dear Stefan,
We are using DRS4 evaluation board version 5.1 and firmware version 30000 (as the picture attached). Now, I am in need one channel with length 2048
bin. However, I can't find the resistors R99, ... ,R106 on the hardware of evaluation board; it seems my DRS4 evaluation board doesn't use 2048
    Reply  Mon Oct 24 12:50:24 2022, Stefan Ritt, Channel Cascading Option in the 2048-bin 
The board is delivered in one or the other mode and not meant to be changed by the user, since this requires very delicate soldering which is not easy.
If you try anyhow, you loose the quarantee. You can send the board back to the manufacturer for the modification, but this costs quite some moeny.

Best regards,
Entry  Tue Sep 27 10:17:58 2022, Kunal Shinde, Required Firmware for DRS4 Evaluation Board Version 2.0 
Hi, I am working on an old DRS4 board Version "2.0" with firmware revision "13191", I was unable to find this specific firmware source
files ("VHDL source code"), please help me where could I find this or send me the required.

Regards,
    Reply  Tue Sep 27 10:37:11 2022, Stefan Ritt, Required Firmware for DRS4 Evaluation Board Version 2.0 
You find each software version at the usual download location at

https://www.dropbox.com/home/drs/drs4/distribution/Download/Linux

The one you need is probably drs-2.1.3.tar.gz which was the last version for the 2.0 board which is now more than 10 years old.
       Reply  Tue Sep 27 10:52:41 2022, Kunal Shinde, Required Firmware for DRS4 Evaluation Board Version 2.0 
I checked the link you provided but it seems that the link doesnt exist please send me valid one.

Regards,

Kunal
          Reply  Tue Sep 27 15:20:55 2022, Stefan Ritt, Required Firmware for DRS4 Evaluation Board Version 2.0 
Sorry, got the wrong link. Here the right one: https://www.dropbox.com/sh/clqo7ekr0ysbrip/AACoWJzrQAbf3WiBJHG89bGGa?dl=0

If you untar the archive, you will find a "firmware" subdirectory with all VHDL code.

Stefan
Entry  Wed Sep 7 10:13:41 2022, Prajjalak Chattopadhyay, Register status after reset 
What are the default register statuses after DRS4 gets reset?
Entry  Fri Apr 9 20:29:45 2021, Sean Quinn, Spikes/noise sensitive to clock settings? spikes_16MHz.pngspike_period.pngbetter_spikes_15MHz.pngspike_period_15MHz.png
Dear DRS4 team,

I'm trying to troubleshoot some odd spike behavior. If I run the ADC and SR CLK at 16 MHz (behavior also seen at 33 MHz) we get very noisy
data (post-calibration) with periodic spikes.
    Reply  Fri Apr 9 21:38:59 2021, Stefan Ritt, Spikes/noise sensitive to clock settings? 
elog:824




Sean
Quinn wrote:



Dear DRS4 team,
       Reply  Fri Jun 24 09:57:36 2022, LynseyShun, Spikes/noise sensitive to clock settings? 
Hello, I now have periodic spikes in CH0 and CH1 output. How can I eliminate these spikes? I'm sorry I didn't understand your elimination
method. Please explain the method in detail. Thank you very much




Stefan
          Reply  Fri Jul 29 17:23:43 2022, Stefan Ritt, Spikes/noise sensitive to clock settings? 
Look at the DRS4 data sheet, Figure 12. You see there the rising SRCLK pulse which outputs the next analog value. You also see tSAMP which describes
the sampling piont (strobe or clock sent to your ADC). The value of tSAMP must be such that the values is sampled at the point where it flattens out, just
2-3 ns BEFORE the next analog sample is clocked out, as written in the text. So you have to phase shift your clock going to SRCLK and the one going to
Entry  Tue Jul 19 02:35:04 2022, Jingyu Zhang, Increase event rate, use ROI mode, and install sw from source in Mac 
Dear experts, 

We are trying to increase the event rate of the DRS4. We looked into the ROI
but couldn’t figure out how to run in ROI mode. We are wondering if there is pre-existing firmware for this? We also tried to download and build
    Reply  Fri Jul 29 14:09:35 2022, Stefan Ritt, Increase event rate, use ROI mode, and install sw from source in Mac 
The firmware from the website always reads 1024 bins. You have to modify it to stop before that, like reading only 128 samples or so. For compiling under
MacOSX, this should work, since I do it myself. 

Regards,
Entry  Tue Apr 12 10:40:36 2022, LynseyShun,  
Hello, I am Lynsey. now I set A3-A0 to 1001 in ROI mode, but only OUT0 has output, and the other seven channels(OUT1-OUT7) do not output corresponding
waveforms.

In ROI mode, can OUT0-OUT7 output sampled waveforms at the same time?
    Reply  Tue Apr 12 10:49:27 2022, Stefan Ritt,  
A3-A0 = 1001 should be all you need to activate OUT0-OUT7. It works in our designs. Maybe double check the address lines with an oscilloscope.

Stefan




LynseyShun
       Reply  Thu Jun 16 05:31:25 2022, LynseyShun,  
Thank you very much for your help!




Stefan
Ritt wrote:



A3-A0 = 1001 should be all you need to activate OUT0-OUT7. It works
ELOG V3.1.5-3fb85fa6