DRS4 Forum
  DRS4 Discussion Forum, Page 29 of 46  Not logged in ELOG logo
ID Date Authordown Subject
  225   Wed Mar 6 12:35:38 2013 Osip LishilinDRS4- analog pulse counting

Hello, Stefan. Have you implemented pulse counting yet?

Best, Osip.

  245   Mon May 20 08:42:16 2013 Osip LishilinDRS4- analog pulse counting

Stefan Ritt wrote:

Osip Lishilin wrote:

Hello, Stefan. Have you implemented pulse counting yet?

Best, Osip.

Nope. 

 Will it be done in the foreseeable future?

  351   Mon Jun 9 12:03:26 2014 Osip LishilinAnnouncement of new Evaluation Board V5

Stefan Ritt wrote:

Hardware scalers for all four channels and the trigger working up to 200 MHz. With the trigger scaler one can measure for example coincidence rates between two channels.

 Does it give the ability to measure triggering rate? I'm talking again about possibility of use DRS4 as pulse counter for PMT's. If yes, do I need new v5 board or it is possible to use v4 board?

  356   Mon Jun 16 15:35:59 2014 Osip LishilinAnnouncement of new Evaluation Board V5

Stefan Ritt wrote:

Osip Lishilin wrote:

Stefan Ritt wrote:

Hardware scalers for all four channels and the trigger working up to 200 MHz. With the trigger scaler one can measure for example coincidence rates between two channels.

 Does it give the ability to measure triggering rate? I'm talking again about possibility of use DRS4 as pulse counter for PMT's. If yes, do I need new v5 board or it is possible to use v4 board?

Yes it is possible to measure the raw trigger rate, with a resolution of 10 Hz. You need a new V5 board for that. 

I'm not sure if I understand you correctly. The trigger rate could be up to 200 MHz, and it's possible to measure it  with 10 Hz resolution. Is it right?

Does it possible to measure independent trigger rate for each channel?

  183   Fri Oct 12 14:06:04 2012 Moritz von WitzlebenDRS abbreviation

Hello,

what is the abbreviation of DRS?

Thanks and kind Regards,

Moritz

  393   Mon Nov 17 16:36:18 2014 Mickey ChiuRaspberry Pi drsosc does not exit properly

When running drsosc on a raspberry pi, it seems the exit doesn't seem to work at all.  This is true for the "exit" button on the window, or the file menu exit, or the "x" on the window.  I end up having to kill drsosc manually from the command line.  This wouldn't be such a bad thing except that it doesn't seem to store any settings when killed in this way.  I'm wondering if anyone else sees the same thing, or if there is a fix out there, before I go and delve into why.

  138   Fri Dec 9 17:45:48 2011 Michael BükerFixes to DOScreen.cpp for recent built on linux
> I was just building version 3.1.0 and ran into some problems in DOScreen.cpp.  Basically the conversions from
> char* to wxString were generating "ambiguous overload" errors (in gcc 4.4.3, wx-2.8)
> 
> The simple fix is given in  the following diff output.

Today, I ran into the same problem and was happy to find your fix. I've incorporated it into a unified diff file,
that can easily be applied with the patch program by saving it into a file ('drsosc-3.1.0-wxfix.patch', say), and
in the drs-3.1.0 directory running:

patch -1 < drsosc-3.1.0-wxfix.patch

This is the file:

--- src/DOScreen.cpp.orig	2011-12-09 15:49:48.682201902 +0100
+++ src/DOScreen.cpp		2011-12-09 15:51:45.666000111 +0100
@@ -234,7 +234,7 @@ void DOScreen::DrawWaveform(wxDC& dc, wx
 
    // display optional debug messages
    if (*m_frame->GetOsci()->GetDebugMsg()) {
-      wxst = m_frame->GetOsci()->GetDebugMsg();
+      wxst = wxString(m_frame->GetOsci()->GetDebugMsg(),wxConvUTF8);
       dc.SetPen(wxPen(*wxLIGHT_GREY, 1, wxSOLID));
       dc.SetBrush(*wxGREEN);
       dc.SetTextForeground(*wxBLACK);
@@ -243,7 +243,7 @@ void DOScreen::DrawWaveform(wxDC& dc, wx
       dc.DrawText(wxst, m_x1+4, m_y1+2);
    }
    if (m_debugMsg[0]) {
-      wxst = m_debugMsg;
+      wxst = wxString(m_debugMsg,wxConvUTF8);
       dc.SetPen(wxPen(*wxLIGHT_GREY, 1, wxSOLID));
       dc.SetBrush(*wxGREEN);
       dc.SetTextForeground(*wxBLACK);
@@ -474,9 +474,9 @@ void DOScreen::DrawWaveform(wxDC& dc, wx
    if (m_osci->GetNumberOfBoards() && m_osci->IsIdle()) {
       dc.SetTextForeground(*wxGREEN);
       if (m_osci->GetTriggerMode() == TM_AUTO)
-         wxst = "AUTO";
+         wxst = wxString("AUTO",wxConvUTF8);
       else
-         wxst = "TRIG?";
+         wxst = wxString("TRIG?",wxConvUTF8);
       dc.GetTextExtent(wxst, &w, &h);
       dc.DrawText(wxst, m_x2 - w - 2, m_y1 + 1);
    }
  431   Tue Jun 16 20:45:54 2015 Michael BuadelkDRS4 Evaluation Board Osc Application

Hi, I have a DRS4 v5 evaluation board and I have a novice question about the oscilliscop application. When I connect it to a photo-detector (silicon photo-multiplier to be exact), the signal appears only on one half of the screen, and I cannot change it to be full screen, and pulse to be centered. I tried changing delay time and played around with the settings of the applicaton but no success. I'd apprecite if someone help me on this, probably very simple, problem.

  528   Sun Jun 12 08:45:52 2016 Michaelproblems of DRS4

Hi

I want to use DRS4 to digitize 16 channels of signals. The width of signal is about 20 ns, with frequency of 50Hz. The time differences between these 16 signals are not constant, arranging from 3us to 0. I am confused about this in some aspects.

  1. Can I use SIMULTANEOUS WRITINT AND READING to realize this? I saw the VHDL program, and if I understand it correctly, it did not work at this state.
  2. Or sampling at 1GSPS, using CASCADING OF CHANNELS, I can sample signal at most 4us or 8us, then digitizing all signals of one chip. Have you tested 4 or more channels cascading before?

Besides, any advice will be helpful!

Thank you.

  Draft   Sun Jun 12 08:49:54 2016 Michaelproblems of DRS4

Hi

I want to use DRS4 to digitize 16 channels of signals. The width of signal is about 20 ns, with frequency of 50Hz. The time differences between these 16 signals are not constant, arranging from 3us to 0. I am confused about this in some aspects.

  1. Can I use SIMULTANEOUS WRITINT AND READING to realize this? I saw the VHDL program, and if I understand it correctly, it did not work at this state.
  2. Or sampling at 1GSPS, using CASCADING OF CHANNELS, I can sample signal at most 4us or 8us, then digitizing all signals of one chip. Have you tested 4 or more channels cascading before?

Besides, any advice will be helpful!

Thank you.

  829   Wed Jul 14 14:55:09 2021 Mehrpad MonajemC code to read the 4 channel with external trigger

Hi there,

Recently I bought a 5GSPS evaluation board with 2048 sampling points.
I want to read 4 inputs of the evaluation bord ar 5 GSPS or 2.5GSP and use an external trigger.
I've checked your website and download drs-5.0.5 which contains the source code in C. It seems that the file drs_exam.cpp can do what I am looking for.
So far I could make and compile the project in Linux Ubuntu, but I couldn't compile it in Windows 10.  I've used Cygwin64 to compile the project in windows 10.

I have the following questions:

1- Since I only need to compile the drs_exam.cpp file, could please help me with how can I compile it directly(without making the entire project). Or tell me which version of Wxwidget and libusb I have to install.

2- If you have any sample code that can read 4 inputs with an external trigger, please tell me where can I find it.

In the end, I want to write a wrapper on this C file(which returns digitized data) and run it from my python program. Thank you in advance.
Best regards,
Mehrpad

 

  831   Tue Aug 10 13:57:09 2021 Mehrpad MonajemC code to read the 4 channel with external trigger

Thank you for the reply.

In the version that I have, I cannot find drs_exam_2048.cpp file. Could you please send me the link to download the software folder, which contain this file.

Best,

Mehrpad

Stefan Ritt wrote:

Sorry the late reply, I was on vacation. 

Here are some answers:

1. I'm sorry I can't help much here, since I currently don't have a Windows 10 computer here to compile any code. I moved now completely to MacOSX, being very similar to Linux. I'm not allowed to run a Windows 7 computer any more for security reasons. Last time this worked for me was with Wxwidget version 3.0 and libusb 1.0, but I guess libusb is not critical so you can use a newer version. If you just compile drs_exam.cpp, you don't need any Wxwidget library. That one is only used for the oscilloscope program.

2. The program drs_exam_2048.cpp is meant to read channels in 2048-bin mode.

3. To adjust the delay between the trigger and the readout, use the function b->SetTriggerDelayNs(xxx)

Best,
Stefan

Mehrpad Monajem wrote:

Hi there,

Recently I bought a 5GSPS evaluation board with 2048 sampling points.
I want to read 4 inputs of the evaluation bord ar 5 GSPS or 2.5GSP and use an external trigger.
I've checked your website and download drs-5.0.5 which contains the source code in C. It seems that the file drs_exam.cpp can do what I am looking for.
So far I could make and compile the project in Linux Ubuntu, but I couldn't compile it in Windows 10.  I've used Cygwin64 to compile the project in windows 10.

I have the following questions:

1- Since I only need to compile the drs_exam.cpp file, could please help me with how can I compile it directly(without making the entire project). Or tell me which version of Wxwidget and libusb I have to install.

2- If you have any sample code that can read 4 inputs with an external trigger, please tell me where can I find it.

In the end, I want to write a wrapper on this C file(which returns digitized data) and run it from my python program. Thank you in advance.
Best regards,
Mehrpad

 

 

 

  843   Tue Oct 26 10:41:46 2021 Mehrpad MonajemExternal trigger and drs_exam

Hi Stefan,


I have two problems regarding using the drs_exam file with external trigger:


1- I connected a 200Khz signal with 20ns rising edge, 50 ohm load, and 27% duty cycle as an external trigger. The output of the drs_exam file starts from 0 to 200ns. Since I use an external trigger, I think it should be starting from 0 to 5ns and then again starting from 0. Could you please tell me where the problem is?

2- How is it possible to change from 1024 to 2048 bins in the drs_exam example?

 

You can find my code in the attachment.

Best regards,
Mehrpad

  846   Tue Oct 26 15:05:18 2021 Mehrpad MonajemExternal trigger and drs_exam

Thanks for your reply.

1- I want to have a window size of 25.6ns instead of 200ns at 5GSPS. I have a 200khz high voltage pulser, which applies a pulse to my sample. I want to digitize the detector signal for each pulse (each pulse has a 25.6ns period). The pulser and digitizer use same 200khz trigger signal from each channel of the signal generator.

2- My DRS board has a 2048 combined stick on it. But the software distribution that I have doesn't contain the drs_exam_2048.cpp program. Could you please send the link that I can download this program? I can't find it under the link below.

link: https://www.dropbox.com/sh/clqo7ekr0ysbrip/AACoWJzrQAbf3WiBJHG89bGGa?dl=0

Best regards,

Mehrpad

Stefan Ritt wrote:

1. Why should your waveform start from 0 to 5ns? I don't get your point. Whenever you trigger a readout, you get a 200ns wide time window, and by definition it starts at zero.

2. In the software distribution you have a drs_exam_2048.cpp program. Note that your board needs to be physically modified before delivery to switch to 2048 bins.

Best,
Stefan

Mehrpad Monajem wrote:

Hi Stefan,


I have two problems regarding using the drs_exam file with external trigger:


1- I connected a 200Khz signal with 20ns rising edge, 50 ohm load, and 27% duty cycle as an external trigger. The output of the drs_exam file starts from 0 to 200ns. Since I use an external trigger, I think it should be starting from 0 to 5ns and then again starting from 0. Could you please tell me where the problem is?

2- How is it possible to change from 1024 to 2048 bins in the drs_exam example?

 

You can find my code in the attachment.

Best regards,
Mehrpad

 

 

  173   Wed Aug 1 17:42:32 2012 Mayank S. RajguruCalculation of loop filter parameters (R,C1and C1) for 1 GHz

 Hi,

we are planning to use the DRS4 in our board for 1 GHz sampling and digitization.

I have seen in the data sheet that "For the PLL to work, an external loop filter is required. This filter ensures quick locking and stable operation at the desired sampling frequency".

What formula do you use to calculate the values of R, C1 and C2?

Can we use the same given value for different frequencies?

Thanks,

Mayak

  914   Tue Mar 25 16:31:41 2025 Matías Tobardrs_exam.cpp not compile

Hi! i'm trying to compile drs_exam.cpp but it yields the following error lines:

I do have the "DRS.h" file on the same folder, which is clearly the file that's causing troubles. I also tried to run "DRS.cpp" but it yields a similar error lines. ('sin definir' means 'not defined').

Thanks a lot for the help!

$ g++ drs_exam.cpp  -o drs_exam
/usr/bin/ld: /tmp/ccroK16H.o: en la función `main':
drs_exam.cpp:(.text+0x48): referencia a `DRS::DRS()' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x135): referencia a `DRSBoard::Init()' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x155): referencia a `DRSBoard::SetFrequency(double, bool)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x169): referencia a `DRSBoard::SetTranspMode(int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x184): referencia a `DRSBoard::SetInputRange(double)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x1b6): referencia a `DRSBoard::EnableTrigger(int, int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x1ca): referencia a `DRSBoard::SetTriggerConfig(int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x1fe): referencia a `DRSBoard::EnableTrigger(int, int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x212): referencia a `DRSBoard::SetTriggerConfig(int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x22d): referencia a `DRSBoard::SetTriggerLevel(double)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x241): referencia a `DRSBoard::SetTriggerPolarity(bool)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x255): referencia a `DRSBoard::SetTriggerDelayNs(int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x2b6): referencia a `DRSBoard::StartDomino()' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x2e9): referencia a `DRSBoard::IsBusy()' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x30b): referencia a `DRSBoard::TransferWaves(int, int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x326): referencia a `DRSBoard::GetTriggerCell(unsigned int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x350): referencia a `DRSBoard::GetTime(unsigned int, int, int, float*, bool, bool)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x377): referencia a `DRSBoard::GetWave(unsigned int, unsigned char, float*)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x392): referencia a `DRSBoard::GetTriggerCell(unsigned int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x3c3): referencia a `DRSBoard::GetTime(unsigned int, int, int, float*, bool, bool)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x3f1): referencia a `DRSBoard::GetWave(unsigned int, unsigned char, float*)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x52e): referencia a `DRS::~DRS()' sin definir
collect2: error: ld returned 1 exit status
 

 

  917   Tue Apr 1 16:24:33 2025 Matías Tobardrs_exam.cpp not compile

Thanks! I solved it by running ./drs_exam after execute Makefile (there was an inconsistency in the files directory).

Matías

Stefan Ritt wrote:

You have to link against the DRS.cpp library, plus usblib, plus ... Note there is both a Makefile and a CMakeLists.txt for it. Google how to use "make" or "cmake".

Stefan

 

  809   Thu Feb 25 17:56:39 2021 Matthias PlumDRS spike removal for multiple waveforms

Hi,

Is there a way that someone can help me and my student to enable RemoveSymmetricSpikes function in the drs_exam.cpp? We are not 100% sure how to call the function if you want to read out four waveforms.

Cheers,

Matthias

  862   Sat Feb 12 13:06:56 2022 Matias SengerCannot trigger on pulses, have to trigger on undershoot

I am using the DRS4 board trying to measure pulses produced by an LGAD. I have no prior experience with this board, have just installed the `drsosc` application and am exploring. I am experiencing some strange trigger behavior. Consider the following screenshot:

 

Here nothing is strange, the board is triggering on the undershoot and it is working fine, I can trigger on rising/falling edge, different levels, etc.

Now, the strange thing is that if I pull the trigger up to trigger on the pulse itself it stops triggering:

I have tried many different setups for the trigger (rising, falling edge, different levels, etc) and nothing works. In the undershoot, everything works.

I have tried with the internal test signal and it works fine:

What could be the problem?

I have run the voltage and time calibrations as suggested in the manual.

  867   Wed Mar 2 17:25:10 2022 Matias SengerHow to convert samples to volt?

I am using the `drscl` app. My prior experience is practically zero, sorry if this is a very naive question. When I read using `read 0 1` (channel 0, with calibration) I get this:

```
Calibration not valid for board #2946
  10    3    7    4   10    8   14    5    5    9    3    4    9    8    9    4
   3    3   12    5    5   13    3    8    1    5    0    4    8    6    6    3
...etc...
```

Why does it says that the calibration is not valid? How am I supposed to go from integers to volts?

If I run the `info` command I get this:

```
==============================
Mezz. Board index:    0
DRS type:             DRS4
Board type:           9
Serial number:        2946
Firmware revision:    30000
Temperature:          43.4 C
Input range:          -0.5V...0.5V
Calibrated range:     -0.5V...0.5V
Calibrated frequency: 0.000 GHz
Status reg.:          0000009A
Control reg.:         00000000
  DMODE circular
Trigger bus:          00000000
Frequency:            1.007 GHz
```

ELOG V3.1.5-3fb85fa6