Entry  Wed Aug 7 15:05:59 2013, Hermann-Josef Mathes, Repeated time calibration 
    Reply  Wed Aug 7 15:10:57 2013, Stefan Ritt, Repeated time calibration 
       Reply  Wed Aug 7 15:20:33 2013, Hermann-Josef Mathes, Repeated time calibration 
          Reply  Wed Feb 5 13:41:42 2014, Stefan Ritt, Repeated time calibration 
Message ID: 330     Entry time: Wed Feb 5 13:41:42 2014     In reply to: 284
Author: Stefan Ritt 
Subject: Repeated time calibration 

Hermann-Josef Mathes wrote:

Hi Stefan,

thanks for the quick reply, I know that this solution works with drscl but not within my code.

I tried to track it down, but gave up very soon. Seems as if AnalyzeWF() which is called by CalibrateTiming() finds to much zero-crossings when it is called the second time.

Regards

  -- Hermann-Josef

Time calibration has been changed completely in meantime. With the new V5 boards, we have a new oscillator on the board where on can calibrate each channel individually. This is necessary to obtain a good timing down to a few ps. With the current code the above problem has vanished. We also learned that the time calibration is very stable (less than a ps) over several months, so no need to repeat the calibration over and over again.

/Stefan 