Message ID: 940     Entry time: Mon Jun 15 04:39:45 2026     In reply to: 939     Reply to this: 941
Author: Mingxin Liu 
Subject: Inquiry about DC-coupled DRS4 hardware design scheme 

Dear Dr. Stefan Ritt,

Thank you so much for your detailed and helpful reply. It’s very clear to me now why the standard DRS4 design uses AC coupling, and I appreciate your practical suggestions on biasing and protection.

I have a follow-up question about the THS4508 amplifier I plan to use at the input stage, if you don’t mind. If I want to make the signal path DC-coupled (both input and output), would it be possible to power the THS4508 with dual supplies, e.g., +3.5 V and -1.5 V, so that the input can accept signals in the range of -1 V to +1 V? In this configuration, could the amplifier still achieve a bandwidth of 700–800 MHz, or is that high bandwidth only guaranteed when using a single +5 V supply?

I would really appreciate your insight on this point.

Thanks again for your help and advice!

Best regards,

Mingxin Liu

Stefan Ritt wrote:

We do not have any DC-coupled design for the DRS4. The reason for that is that for many applications we use SiPMs, which are directly powered through the DRS4 board. Since the bias voltage of ~50V would kill the DRS4, we always go with AC-coupling. 

The reason the DRS4 evaluation board cannot fully capture signals below -0.5V is the fact that the board only has a single +5V power supply and no +-5V power rails. Having an additional -5V power rail would make the evaluation board too bulky and it could not be powered over USB.

In principle you could bias the point after the AC coupling capacitor to +5V, then you can capture up to -5V signals. Add a divider by two to go up to -10V. But then the input protection diodes won't work as it, you would need some fast 5V zener diodes which open for signals below -5V to protect the following circuitry.

Stefan

 