Fri Jan 16 14:12:19 2015, Stefan Ritt, Mac OSX Yosemite 10.10
|
> Hello,
>
> I can compile version 5.0.3 of DRS4sc on Mac OSX 10.0 without errors but when I want to execute the program I get the following error:
>
> [home]$ ./DRSOsc
> DRSOsc(48068,0x7fff7ac5e300) malloc: *** error for object 0x7f88d9434a80: incorrect checksum for freed object - object was probably
> modified after being freed.
> *** set a breakpoint in malloc_error_break to debug
>
> Is this a known error? Will it be fixed in the next release?
When I compile on OSX 10.10.1 with XCode 6.1.1 I get no error. I'm using wxWidgets 3.0.0. Same when I compile with the command line
tools (which you first have to install from Apple) via "make", "make app", "open ./DRSOsc.app". You cannot start a graphical program
directly from the command line like under Linux, you have to make an app and the do "open <app>".
Best regards,
Stefan |
Mon Nov 28 22:28:34 2016, Randall Gladen, Long timing between two channels
|
I don't believe I fully understand how the timing works between multiple channels on DRS4 board, even after reading the manual, but I am hoping to measure a time difference between two channels longer than 1024/sampling rate. So far, I have written a program in Matlab to extract timing and voltage information from the binary file to find the time difference between two channels that are set with the AND trigger logic and appear within approximately 80 ns of each other at a sampling rate of 1 GSPS. This works as intended, but I would now like to try to measure time differences of anywhere between 50 ns and several ms within a single spectrum. Since this is out of the range of only 1024 channels above 1GSPS, is it possible for the board to keep track of the time between two trigger pulses that occur at time differences longer than 1024/sampling rate?
Thank you very much for your help, and if I am severely misunderstanding how the board works, please forgive my ignorance and feel free to correct me,
~Randall
Edit: I forgot to mention that I am collecting the data using the provided DRS4 Oscilloscope software. |
Wed Nov 30 10:45:29 2016, Stefan Ritt, Long timing between two channels
|
You cannot measure times longer than 1024/sampling rate.
Stefan
Randall Gladen wrote: |
I don't believe I fully understand how the timing works between multiple channels on DRS4 board, even after reading the manual, but I am hoping to measure a time difference between two channels longer than 1024/sampling rate. So far, I have written a program in Matlab to extract timing and voltage information from the binary file to find the time difference between two channels that are set with the AND trigger logic and appear within approximately 80 ns of each other at a sampling rate of 1 GSPS. This works as intended, but I would now like to try to measure time differences of anywhere between 50 ns and several ms within a single spectrum. Since this is out of the range of only 1024 channels above 1GSPS, is it possible for the board to keep track of the time between two trigger pulses that occur at time differences longer than 1024/sampling rate?
Thank you very much for your help, and if I am severely misunderstanding how the board works, please forgive my ignorance and feel free to correct me,
~Randall
Edit: I forgot to mention that I am collecting the data using the provided DRS4 Oscilloscope software.
|
|
Tue Nov 3 22:37:56 2015, Will Flanagan, Latest macro for DRS4 V5
|
Hi DRS4 Experts,
I have an extremely naive question: Is there any official macro to unpack the DRS4 binary files? All I am looking to do is to plot a few of my waveforms and manipulate them in root. I am using OSX 10.10 and ROOT 5.34.
Thanks in advance,
Will |
Tue Nov 3 23:15:38 2015, Will Flanagan, Latest macro for DRS4 V5
|
I should of course mention that I looked through the DRS4 website and didn't see anything obvious: https://www.psi.ch/drs/evaluation-board
Thanks,
Will
Will Flanagan wrote: |
Hi DRS4 Experts,
I have an extremely naive question: Is there any official macro to unpack the DRS4 binary files? All I am looking to do is to plot a few of my waveforms and manipulate them in root. I am using OSX 10.10 and ROOT 5.34.
Thanks in advance,
Will
|
|
Wed Nov 4 15:40:10 2015, Stefan Ritt, Latest macro for DRS4 V5
|
Have a look here: elog:361
Will Flanagan wrote: |
Hi DRS4 Experts,
I have an extremely naive question: Is there any official macro to unpack the DRS4 binary files? All I am looking to do is to plot a few of my waveforms and manipulate them in root. I am using OSX 10.10 and ROOT 5.34.
Thanks in advance,
Will
|
|
Thu Nov 5 00:18:42 2015, Will Flanagan, Latest macro for DRS4 V5
|
Hi Stefan,
This is absolutely perfect.
Thanks,
Will
Stefan Ritt wrote: |
Have a look here: elog:361
Will Flanagan wrote: |
Hi DRS4 Experts,
I have an extremely naive question: Is there any official macro to unpack the DRS4 binary files? All I am looking to do is to plot a few of my waveforms and manipulate them in root. I am using OSX 10.10 and ROOT 5.34.
Thanks in advance,
Will
|
|
|
Mon Aug 13 19:44:59 2018, Martin Petriska, Latch delay support
|
Hi,
https://forge.physik.rwth-aachen.de/projects/drs4-rwth
Not sure about their licensing, but is it possible to add latch delay support to official firmware ?
Best regards
Martin |
Tue Aug 14 06:10:49 2018, Stefan Ritt, Latch delay support
|
I put that on the wish list, but I won't have time for that in the next months.
Stefan
Martin Petriska wrote: |
Hi,
https://forge.physik.rwth-aachen.de/projects/drs4-rwth
Not sure about their licensing, but is it possible to add latch delay support to official firmware ?
Best regards
Martin
|
|
Wed Mar 9 09:57:20 2016, Christian D, LabView
|
Hi,
I would like to use the DRS4 board with LabView for fast readout.
Do you know anyone who has written a VI for that?
Thanks,
Christian |
Fri Nov 18 16:38:42 2016, Gerard Montarou, LabView
|
Hello,
Did you start to write some VI to interface DRS4board with labview ?
i also have in mind to do that.I am surprised that nobody alraedy did it since there is no answer toyour question
gerard
Christian D wrote: |
Hi,
I would like to use the DRS4 board with LabView for fast readout.
Do you know anyone who has written a VI for that?
Thanks,
Christian
|
|
Tue Jan 31 01:37:35 2017, VO HONG HAI, LLD and ULD discriminations,
|
Dear Stefan,
Is there any way to develop LLD and ULD discrimination in DSR-4 evaluation board?
Best regards,
V.H.Hai |
Tue Jan 31 08:40:04 2017, Stefan Ritt, LLD and ULD discriminations,
|
Not inside the board. Each channel has a single discriminator. You can select to trigger on a rising or falling edge, but you don't have two levels. What you can do however is to make an external trigger, like using old NIM logic. You can make discrimaiton with different levels and use a coincidence unit to combine them. Then feed the trigger into the external trigger input of the evaluation board (5V TTL level, not NIM level!).
Stefan
VO HONG HAI wrote: |
Dear Stefan, Is there any way to develop LLD and ULD discrimination in DSR-4 evaluation board? Best regards, V.H.Hai |
|
Sat May 23 11:03:20 2015, Felix Bachmair, Issue with Trigger rates below ~100Hz   
|
Hi
We are working with the DRS 4 V5 version and we investigated an issue with the trigger at rates below ~120 Hz.
As long as we have a trigger rate of more than 125 Hz. everything seems to work fine and we are recording more or less all events.
As soon as we go lower in input trigger rate to 100Hz, we see a drop in trigger rates to approx 15 - 20 Hz.
When we use the new firmware we can see that the busy signal is 0 for much longer times than usual up to .5 seconds.
We made a plot of input trigger rate vs trigger rate of drs: https://plot.ly/~simon.corrodi/316
In the oscilloscope plots one can see the the trigger in in yellow and the trig out from drs board in blue.
Do you have any idea what could be the reason?
We also |
Thu Jul 2 08:53:17 2015, Felix Bachmair, Issue with Trigger rates below ~100Hz
|
Hi,
We did a further investigation of this problem:
We figured out that this issue seems to be related to the kernel.
We tested it now on two machines with Ubuntu 14.04.2 LTS (kernel 3.16.0-41), one with RHEL 6.6 (kernel 2.6.32) , one with Fedora 20 (kernel 3.18.7) and one with Mac OSX. We see this issue with the Ubuntu and the fedroa machines.Both have a kernel above 3.0 while RHEL has a kernel of 2.6
We can repoduce the problem on all input channels as a trigger.
I will try to find out what could be the cause of it.
Cheers
Felix
Felix Bachmair wrote: |
Hi
We are working with the DRS 4 V5 version and we investigated an issue with the trigger at rates below ~120 Hz.
As long as we have a trigger rate of more than 125 Hz. everything seems to work fine and we are recording more or less all events.
As soon as we go lower in input trigger rate to 100Hz, we see a drop in trigger rates to approx 15 - 20 Hz.
When we use the new firmware we can see that the busy signal is 0 for much longer times than usual up to .5 seconds.
We made a plot of input trigger rate vs trigger rate of drs: https://plot.ly/~simon.corrodi/316
In the oscilloscope plots one can see the the trigger in in yellow and the trig out from drs board in blue.
Do you have any idea what could be the reason?
We also
|
|
Mon May 22 18:27:56 2017, Esperienza Giove, Invalid magic number 0000
|
Hello everybody!
After some times i init my board, or if i stop the program during the acquisition, i get the error message "Invalid magic 0000". The only way i can solve this problem is to physically disconnect and plug in again the USB cable.
What's the point? Is there a bash command / code i could use to reset it?
Thank you very much |
Tue May 23 10:24:47 2017, Stefan Ritt, Invalid magic number 0000
|
Under linux, many people observed that the USB connection is unstable to the evaluation board. This must be related to the linux USB stack, since my code runs fine under MacOSX and Windows, where I use the same USB library (libusb-1.0). So I can't do anything from my side. Baybe the linux system has some tools to reset an USB endpoint. I googled it and found some proposals here:
https://askubuntu.com/questions/645/how-do-you-reset-a-usb-device-from-the-command-line
Regards,
Stefan
Esperienza Giove wrote: |
Hello everybody!
After some times i init my board, or if i stop the program during the acquisition, i get the error message "Invalid magic 0000". The only way i can solve this problem is to physically disconnect and plug in again the USB cable.
What's the point? Is there a bash command / code i could use to reset it?
Thank you very much
|
|
Thu May 25 20:17:41 2017, Esperienza Giove, Invalid magic number 0000
|
Hello, thanks for your answer. Unluckily if i try to reset in this way it keeps hanging
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
Invalid magic number: 0000
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
I also tried with sudo tee /sys/bus/usb/drivers/usb/unbind and binding again; same thing happens. It seems the board needs to be reset when this happens. Is there a way to do that - to reset the board instead of usb?
Thank you
Stefan Ritt wrote:
|
Under linux, many people observed that the USB connection is unstable to the evaluation board. This must be related to the linux USB stack, since my code runs fine under MacOSX and Windows, where I use the same USB library (libusb-1.0). So I can't do anything from my side. Baybe the linux system has some tools to reset an USB endpoint. I googled it and found some proposals here:
https://askubuntu.com/questions/645/how-do-you-reset-a-usb-device-from-the-command-line
Regards,
Stefan
Esperienza Giove wrote: |
Hello everybody!
After some times i init my board, or if i stop the program during the acquisition, i get the error message "Invalid magic 0000". The only way i can solve this problem is to physically disconnect and plug in again the USB cable.
What's the point? Is there a bash command / code i could use to reset it?
Thank you very much
|
|
|
Thu May 25 20:20:57 2017, Esperienza Giove, Invalid magic number 0000
|
Hello, thanks for your answer. Unluckily if i try to reset in this way it keeps hanging
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
Invalid magic number: 0000
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
I also tried with sudo tee /sys/bus/usb/drivers/usb/unbind and binding again; same thing happens. It seems the board needs to be reset when this happens. Is there a way to do that - to reset the board instead of usb?
Thank you
Stefan Ritt wrote:
|
Under linux, many people observed that the USB connection is unstable to the evaluation board. This must be related to the linux USB stack, since my code runs fine under MacOSX and Windows, where I use the same USB library (libusb-1.0). So I can't do anything from my side. Baybe the linux system has some tools to reset an USB endpoint. I googled it and found some proposals here:
https://askubuntu.com/questions/645/how-do-you-reset-a-usb-device-from-the-command-line
Regards,
Stefan
Esperienza Giove wrote: |
Hello everybody!
After some times i init my board, or if i stop the program during the acquisition, i get the error message "Invalid magic 0000". The only way i can solve this problem is to physically disconnect and plug in again the USB cable.
What's the point? Is there a bash command / code i could use to reset it?
Thank you very much
|
|
|
Fri May 26 08:48:25 2017, Stefan Ritt, Invalid magic number 0000
|
There is no other way to reset the board. As I said, people running this under Windows or MacOS are fine, so maybe this calls for a change of OS.
Esperienza Giove wrote: |
Hello, thanks for your answer. Unluckily if i try to reset in this way it keeps hanging
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
Invalid magic number: 0000
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
musb_write: requested 10, wrote 0, errno -7 (Unknown error 18446744073709551609)
musb_read error 0
I also tried with sudo tee /sys/bus/usb/drivers/usb/unbind and binding again; same thing happens. It seems the board needs to be reset when this happens. Is there a way to do that - to reset the board instead of usb?
Thank you
Stefan Ritt wrote:
|
Under linux, many people observed that the USB connection is unstable to the evaluation board. This must be related to the linux USB stack, since my code runs fine under MacOSX and Windows, where I use the same USB library (libusb-1.0). So I can't do anything from my side. Baybe the linux system has some tools to reset an USB endpoint. I googled it and found some proposals here:
https://askubuntu.com/questions/645/how-do-you-reset-a-usb-device-from-the-command-line
Regards,
Stefan
Esperienza Giove wrote: |
Hello everybody!
After some times i init my board, or if i stop the program during the acquisition, i get the error message "Invalid magic 0000". The only way i can solve this problem is to physically disconnect and plug in again the USB cable.
What's the point? Is there a bash command / code i could use to reset it?
Thank you very much
|
|
|
|
|