  619   Fri Jun 16 17:34:20 2017 Laura GonellaDriver installation on Windows 10

Hello,

I am trying to get a DRS4 board to run on Windows 10. I am having problems with the driver installation. I am getting the follwoing message

"There is no driver selected for the device information set or element"

I had specified the path to look for the driver as C:\ProgramFilesx86\DRS\driver\. I also tried the option to look online for the driver. None works. Can anyone help?

Thanks,

Laura

  467   Wed Jan 6 15:51:58 2016 Larry ByarsUse of Channel Cascading in drs_exam.cpp

Hello Stefan,

Here in Rockford, TN we just got a new DRS4 evaluation board (Serail # 2612, Board Type 9, Firmware 21305) which is labeled as combined 2048.

It looks like the drs_exam.cpp only works with 1024 samples per channel. We'd like to be able to get 2048 samples from each of the four channels but I am uncertain what code modifications are necessary support this.

Could you offer a suggestion? I've searched the forum for cascade and read several threads but they are pretty old. One even says it isn't supported in the evaluation board, but I think that is no longer the case.

 

Thanks for your help,

 

Larry Byars

 

  469   Tue Jan 12 15:42:31 2016 Larry ByarsUse of Channel Cascading in drs_exam.cpp

An update. I have been successful in making modifications to drs_exam.cpp so that I can get 2048 samples per channel.. The main changes were to the size of the time_array and wave_array and adding a call to Set ChannelConfig(0,8,4). It was also necessary to change the parameters to GetWave so that the Trigger Cell and WSR values were passed to get the channel combinations correct (2048 channel.ppt).

I've moved on to try to increase the speed of acquisition (I get only about 500 events/sec) and trying to understand the corrections.Working through the source code slowly...

Regards,

Larry Byars

Larry Byars wrote:

Hello Stefan,

Here in Rockford, TN we just got a new DRS4 evaluation board (Serail # 2612, Board Type 9, Firmware 21305) which is labeled as combined 2048.

It looks like the drs_exam.cpp only works with 1024 samples per channel. We'd like to be able to get 2048 samples from each of the four channels but I am uncertain what code modifications are necessary support this.

Could you offer a suggestion? I've searched the forum for cascade and read several threads but they are pretty old. One even says it isn't supported in the evaluation board, but I think that is no longer the case.

 

Thanks for your help,

 

Larry Byars

 

 

  509   Thu Apr 21 22:16:43 2016 Kyle WeinfurtherNegative fCellDT values from GetTimeCalibration()

Hello Stefan,

I am using four DRS4 v5 eval boards to digitize 16 channels of data. I have recently changed from saving the timing information of the waveform using GetTime() to GetTimeCalibration(). When changing over, I noticed that some values for fCellDT for cell 498 are negative. Over the 16 channels used, 4 of them have negative time bin widths for cell 498 while the other 12 channels are very close to 0 (in the ~10 ps range). One of the eval boards has no negative fCellDT whereas the other three boards have one or two channels with negative values.

Upon further inspection, I checked the time between samples of GetTime() and found the same results in cell 498. After finding this, I did a timing calibration again with CalibrateTiming() even though in a different post on the discussion forum you said it was valid for a wide range of temperatures and a long time (years). This still allowed the negative fCellDT values to persist.

Is this a common occurance? If so, is there a method to fix this issue? Is there a reason for cell 498 to have a small value for fCellDT? I searched the discussion forum and did not find anything relating to this issue.

Attached are a couple waveform traces using GetTime() zoomed in on cell 498.

Thanks,

Kyle Weinfurther

Attachment 1: ch5.png
ch5.png
Attachment 2: ch7.png
ch7.png
Attachment 3: ch9.png
ch9.png
  563   Fri Nov 18 05:52:45 2016 Kurtis NishimuraChannel offsets in GetTime()

Hello,

I have a question about the GetTime() method in DRS.cpp.  I understand how the DT values are applied for all channels, and I also understand from the evaluation board manual that the timing of each channel is synchronized at sample 0, so samples should really be aligned from channel-to-channel relative to sample 0.

However, DRS.cpp has the following snippet in DrsBoard::GetTime():

   if (channelIndex > 0) {
      // correct all channels to channel 0 (Daniel's method)
      iend = tc >= 700 ? 700+1024 : 700;
      for (i=tc,gt0=0 ; i<iend ; i++)
         gt0 += fCellDT[chipIndex][0][i % 1024];
      
      for (i=tc,gt=0 ; i<iend ; i++)
         gt += fCellDT[chipIndex][channelIndex][i % 1024];
      
      for (i=0 ; i<fChannelDepth ; i++)
         time[i] += (float)(gt0 - gt);
   }

I can see what this is calculating and applying such an offset, but I don't understand why things seem to be referenced to sample 700.  Is there a particular reason why sample 700 is chosen here?  This does not seem like a straightforward application of the attached instructions from the evaluation board user's manual.

Any insight would be much appreciated!

Thanks so much,

-Kurtis

Attachment 1: offsetInstructions.png
offsetInstructions.png
  890   Tue Sep 27 10:17:58 2022 Kunal ShindeRequired Firmware for DRS4 Evaluation Board Version 2.0

Hi, I am working on an old DRS4 board Version "2.0" with firmware revision "13191", I was unable to find this specific firmware source files ("VHDL source code"), please help me where could I find this or send me the required.

Regards,

Kunal

  892   Tue Sep 27 10:52:41 2022 Kunal ShindeRequired Firmware for DRS4 Evaluation Board Version 2.0

I checked the link you provided but it seems that the link doesnt exist please send me valid one.

Regards,

Kunal

Stefan Ritt wrote:

You find each software version at the usual download location at

https://www.dropbox.com/home/drs/drs4/distribution/Download/Linux

The one you need is probably drs-2.1.3.tar.gz which was the last version for the 2.0 board which is now more than 10 years old.

Best,
Stefan

 

Kunal Shinde wrote:

Hi, I am working on an old DRS4 board Version "2.0" with firmware revision "13191", I was unable to find this specific firmware source files ("VHDL source code"), please help me where could I find this or send me the required.

Regards,

Kunal

 

 

  508   Fri Apr 15 12:58:46 2016 Konstantin GusevDRS4 purchase information

    Hi,

 I can't contact with Anita Van Loon about DSR4 chip's price and delivery.

Did you still sell it? Can you provide me this information?

  784   Thu May 21 07:18:48 2020 Keita MizukoshiDRS4 Evaluation board control tool 'drscl' with macro file

Dear experts,

 

I would like to use DRS4 evaluation board as DAQ system for small, table-top experiment.

I need waveforms capture as binary file on some trigger based on command line without GUI.

I found `drscl` tool in official software, but it require interactive command. I'd rather use static macro or so on to control DAQ as same behaviour in each time.

I guess, experts are thinking users should develop DAQ code by themselves for their experiment specifically, but my request is very common so someone has already developed these tool.

 

Best regards,

Keita

  785   Thu May 21 07:38:05 2020 Keita MizukoshiType check at DOFrame.h in official software

Hi,

 

I've failured to compile official software. The cause is the following line.

DOFrame.h L.111    bool GetRefclk()        { return m_refClk > 0; }

 

m_refClk is pointer to bool. I guess these line is for null-check of the pointer.

Can I replace the following line as 

bool GetRefclk()        { return m_refClk != nullptr; }

?

The latest compilers may not accept C-style check.

 

My compiler version is

Apple clang version 11.0.3 (clang-1103.0.32.59)
Target: x86_64-apple-darwin19.4.0
Thread model: posix
InstalledDir: /Applications/Xcode.app/Contents/Developer/Toolchains/XcodeDefault.xctoolchain/usr/bin

Best regards,

Keita

  789   Mon May 25 03:36:12 2020 Keita MizukoshiDRS4 Evaluation board control tool 'drscl' with macro file

Thank you very much. That is what I wanted.

Stefan Ritt wrote:

There is an example program in the distribution under software/drscl/drs_exam.cpp which is a stand-alone program to do what you need. It uses the C library coming with the distribution. It configureres the board, defines a trigger, and then writes a few waveforms into a file. You can use it as a starting point for your development. If you need any other language, you have to develop bindings to the C library.

Stefan

Keita Mizukoshi wrote:

Dear experts,

 

I would like to use DRS4 evaluation board as DAQ system for small, table-top experiment.

I need waveforms capture as binary file on some trigger based on command line without GUI.

I found `drscl` tool in official software, but it require interactive command. I'd rather use static macro or so on to control DAQ as same behaviour in each time.

I guess, experts are thinking users should develop DAQ code by themselves for their experiment specifically, but my request is very common so someone has already developed these tool.

 

Best regards,

Keita

 

 

  837   Thu Oct 14 15:19:00 2021 Keita Mizukoshilivetime (or deadtime) of DRS4 evaluation board

Dear experts,

 

I would like to use the DRS4 evaluation board for actual physics experiment.

I made a CUI script based on the drs_exam, https://github.com/mzks/drs4_tools/blob/main/build/source/drscmd.cpp.

In this framework, how can we obtain DAQ livetime (or deadtime)?

Has some function already provided to evaluate them from firmware?

 

Best regards,

Keita

  839   Thu Oct 14 18:03:52 2021 Keita Mizukoshilivetime (or deadtime) of DRS4 evaluation board

Thank you very much for your response.
Excuse me for my very stupid confirmation.
If I take N events finally and the hardware scaler value is M, the livetime is realtime*(N/M). Is this correct?

Stefan Ritt wrote:

The one thing you can do easily is to look at the scaler values. If one channel counts all physical events, and you have all read out events, then the ratio give you the live/deadtime. The hardware scalers also keep running during the DRS readout.

Stefan

Keita Mizukoshi wrote:

Dear experts,

 

I would like to use the DRS4 evaluation board for actual physics experiment.

I made a CUI script based on the drs_exam, https://github.com/mzks/drs4_tools/blob/main/build/source/drscmd.cpp.

In this framework, how can we obtain DAQ livetime (or deadtime)?

Has some function already provided to evaluate them from firmware?

 

Best regards,

Keita

 

 

  841   Fri Oct 15 06:15:53 2021 Keita Mizukoshilivetime (or deadtime) of DRS4 evaluation board

Thank you very much.

Stefan Ritt wrote:

I would say not exactly, but it's a good approximation.

Keita Mizukoshi wrote:

Thank you very much for your response.
Excuse me for my very stupid confirmation.
If I take N events finally and the hardware scaler value is M, the livetime is realtime*(N/M). Is this correct

 

  866   Tue Mar 1 19:03:37 2022 Keita MizukoshiScaler issue to evaluate live time

Hi. I'm trying to evaluate livetime of the evaluation board with the hardware scaler. I'm facing a strange issue.

I took the rate with the function, DRS->GetScaler(int channel).
I guess that channels 0--3 mean the rate for the channel, and channel 4 means the counter of the trigger.
I took the 1,000 pulses generated by a pulse generator with 50 Hz.
The scaler values are ~ 39.83, not 50.
The timestamp difference between the initial event and the final event is 19.98 seconds.
1000/19.98 ~ 50, thus, the evaluation board took the pulses with enough livetime.

Can we believe the scaler value for the livetime evaluation?

  870   Fri Mar 4 03:55:33 2022 Keita MizukoshiScaler issue to evaluate live time

Thank you very much for your explanation.

 

I would like to show you a pulse example ('black line is the threshold).

Still, pulse generator rate and DRS4 rate are a bit different more than 10 Hz.

 

Stefan Ritt wrote:

The scalers are read out 10x per seconds, so they have an accuracy of 10 Hz. I tried a 50 Hz pulser, and measured 40 Hz, I tried 52 Hz and measured 50 Hz. This is about what you can expect.

The scaler rate is measured after the discriminator of the trigger, so the trigger level also affects the scaler reading. If you have a 100 mV pulse and your threshold is 200 mV, your scaler rate drops to zero. That can be seen best with the DRSOsc and sliding the trigger value. If you have a 50 Hz pulse with narrow (< us) pulses, things are fine. But if you use a 50 Hz square wave, then you get distorted signals due to the AC coupling which can also be confusing. See for example here: https://www.daqarta.com/dw_gg0o.htm

Keita Mizukoshi wrote:

Hi. I'm trying to evaluate livetime of the evaluation board with the hardware scaler. I'm facing a strange issue.

I took the rate with the function, DRS->GetScaler(int channel).
I guess that channels 0--3 mean the rate for the channel, and channel 4 means the counter of the trigger.
I took the 1,000 pulses generated by a pulse generator with 50 Hz.
The scaler values are ~ 39.83, not 50.
The timestamp difference between the initial event and the final event is 19.98 seconds.
1000/19.98 ~ 50, thus, the evaluation board took the pulses with enough livetime.

Can we believe the scaler value for the livetime evaluation?

 

 

Attachment 1: pulse_example.png
pulse_example.png
Attachment 2: rate.png
rate.png
  916   Thu Mar 27 15:53:10 2025 Justin TabbettNoisy counts with adapted drs_exam.cpp

Greetings, 

I have adapted the drs_exam.cpp to allow for a user input number of channels and trigger levels.

The program mostly works well, however there are counts which form a noise peak, imposed on the regular channel response.

To illustrate, I acquired 10,000 counts (measuring peak to peak) with the drsosc, and with my adapted script, with two channels and OR trigger logic.

 

Is there something missing in my code that could explain the cause of this noise peak? I have attached the .cpp file.

Many thanks,

Justin

Attachment 1: Channel_1_2.png
Channel_1_2.png
Attachment 2: drs_exam.cpp
/********************************************************************\

  Name:         drs_exam.cpp
  Created by:   Stefan Ritt

  Contents:     Simple example application to read out a DRS4
                evaluation board

  $Id: drs_exam.cpp 21308 2014-04-11 14:50:16Z ritt $

\********************************************************************/

#include <math.h>

#ifdef _MSC_VER

#include <windows.h>

#elif defined(OS_LINUX)

#define O_BINARY 0

#include <unistd.h>
#include <ctype.h>
#include <sys/ioctl.h>
#include <errno.h>

#define DIR_SEPARATOR '/'

#endif

#include <stdio.h>
#include <string.h>
#include <stdlib.h>

#include "strlcpy.h"
#include "DRS.h"

#include <math.h>

#include <cstdio>

#include <fcntl.h>
#include <time.h>
#include <sys/stat.h>
#include <assert.h>
#include <iostream>
#include <chrono>
#include <vector>
#include <algorithm>
#include <iostream>
#include <conio.h>  // For _kbhit() and _getch() functions (Windows only)

/*------------------------------------------------------------------*/

int main()
{
    int i, j, k, l, m, n, p, numC, nBoards, i_start, i_end, csel;
    char str[256], param[10][100];
    DRS* drs;
    DRSBoard* b;
    float time_array[8][1024];
    float wave_array[8][1024];
    FILE* f;
    FILE* fph;

    /* do initial scan */
    drs = new DRS();

    /* show any found board(s) */
    for (i = 0; i < drs->GetNumberOfBoards(); i++) {
        b = drs->GetBoard(i);
        printf("Found DRS4 evaluation board, serial #%d, firmware revision %d\n",
            b->GetBoardSerialNumber(), b->GetFirmwareVersion());
    }

    /* exit if no board found */
    nBoards = drs->GetNumberOfBoards();
    if (nBoards == 0) {
        printf("No DRS4 evaluation board found\n");
        return 0;
    }

    /* continue working with first board only */
    b = drs->GetBoard(0);

    /* initialize board */
    b->Init();

    /* set sampling frequency */
    b->SetFrequency(5, true);

    /* enable transparent mode needed for analog trigger */
    b->SetTranspMode(1);

    /* set input range to -0.5V ... +0.5V */
    b->SetInputRange(0);

    /* use following line to turn on the internal 100 MHz clock connected to all channels  */
    b->EnableTcal(1);

    printf("Number of channels: ");
    fgets(str, sizeof(str), stdin);
    csel = atoi(str); //Number of channels
    printf("DRS4 configured for %d channels\n", atoi(str));
    
    b->EnableTrigger(1, 0);
    b->SetTriggerSource(15); // 15 is an OR on CH1-4, 3 is OR on CH1-2, apparently 768 is the AND value for CH1 and CH2 
    b->SetTriggerPolarity(true);        // positive edge -> false, negative edge -> true

    std::vector<int> avn = { 1, 2, 3, 4 };  // Available number of channels: 0, 2, 4, 6
    std::vector<double> triglist = { 0, 0, 0, 0 }; //Store trigger thresholds
    for (i = 0;i < csel;i++) {
        printf("Channel %d trigger level (V): ", avn[i]);
        fgets(str, sizeof(str), stdin);
        b->SetIndividualTriggerLevel((avn[i] - 1), atof(str));
        triglist[i] = atof(str) * -1000; //store as mV
        printf("Trigger level set to %1.3lf Volt\n", atof(str));
    }
    b->SetTriggerDelayNs(0);             // zero ns trigger delay

    fph = fopen("fph.txt", "w");
    if (fph == NULL) {
        perror("ERROR: Cannot open file \"fph.txt\"");
        return 1;
    }
    
    numC = 0; //Event counter

    for (i = 0;i < csel;i++) {
        fprintf(fph, "%7.1f,", triglist[i]);
    }

    fprintf(fph, "%d\n", 0);
    printf("hi\n");
    while (true) {
        // Check if a key is pressed
        if (_kbhit()) {
            char ch = _getch();  // Read the pressed key without waiting for Enter
            if (ch == 'q')       // If 'q' is pressed, exit the loop
                break;
        }

        k = 0;
        l = 0;
        m = 0;
        n = 0;

        std::vector<double> outh(csel, 0); //Create output for pulse heights as 0s
        printf("Before startdomino\n");
        b->StartDomino();

        /* wait for trigger */
        printf("Before fflushstdout\n");
        fflush(stdout);
        while (b->IsBusy());

        /*Take time stamp*/

        auto now = std::chrono::system_clock::now();
        auto duration = std::chrono::duration<double>(now.time_since_epoch());
        double timeT = duration.count();
        
        /*Read waveforms*/

        b->TransferWaves(0, 8);

        for (p = 0;p < csel;p++) {
            b->GetWave(0, p * 2, wave_array[p]); // this reads channel i*2 to array index i
            float minVali = wave_array[p][0], maxVali = wave_array[p][0]; //creates floats for min and max values
            for (i = 0; i < 1024; i++) {
                // find the min and max values from each channel
                if (wave_array[p][i] < minVali) minVali = wave_array[p][i];
                if (wave_array[p][i] > maxVali) maxVali = wave_array[p][i];
            }
            float heighti = maxVali - minVali;
            outh[p] = heighti;
        }

        for (double value : outh) {
            fprintf(fph, "%7.1f,", value);
        }

        fprintf(fph, "%.6f\n", timeT);
        printf("%d\n", numC);

      
        if (numC % 1000 == 0) {
            fflush(fph);
        }
        numC++;



    }

    fclose(fph);
    std::getchar();

    /* delete DRS object -> close USB connection */
    delete drs;
}
  401   Sun Apr 5 22:16:48 2015 Julien WulfDRS4 Evaluation Board Baseline/Voltage Calibration

Hi,

I`m trying to calibrate my DRS4 evoluation board to an input range of 0-1V but it doesn`t work.

1) First I tried to calibrate it with the drsosc (version 5.0.3) Software. The -0.5V - 0.5V calibration works, but during the 0 - 1V calibration the Software crashes.

2) I also tried to calibrate the input range with a C++ DAQ Package (based on drs_exam). Here the code of the calibration:

    ....

    b->SetInputRange(0.) (Center at 0 V )

    b->CalibrateVolt(NULL);

   ....

   Calibration Works

   ....

   b->SetInputRange(0.5) (Center at 0.5 V )

   b->CalibrateVolt(NULL);

  ....

   Results in: Found 4096 stuck pixels on this board.

Did I do a mistake or is this a normal behaviour of the board? Also the board often crashes and I get a magic number 0000 after restarting the DAQ. Then the board needs to be restarted via pulling the plug. ( I ensured that I terminate the USB connection before I close the program with "delete drs"). Is there a possibility to avoid this error?

My OS: Ubuntu 14.04 LTS.

Ciao,

Julien

  700   Wed Jun 13 13:23:17 2018 Julian KempMaximum analog input voltage

Dear all,

I have been wondering what the maximum analog input voltage for the DRS4 V5 evaluation board is. It came with a sticker indicating that it is "2.5V pk Max". On the other hand, when checking the manual (https://www.psi.ch/drs/DocumentationEN/manual_rev50.pdf), it says maximum allowed is 10V DC or even 30V for short pulses. I foresee an application where I cannot make sure that pulses stay below 2.5V, so the correct value will be quite important for me.

Best,
Julian

  702   Wed Jun 13 16:34:28 2018 Julian KempMaximum analog input voltage

Thank you! That solves my problem.

Stefan Ritt wrote:

In principle the numbers in the manual are correct. But they relate to pulses of a certain length, because the input protection only works for DC voltage and for pulses which are not too long. Since we could not write this all on the label of the board, we decided to put there 100% safe value as a "warning" to people, meaning that if pulses are above 2.5V, they should look into the manual and read the details. 

Stefan

Julian Kemp wrote:

Dear all,

I have been wondering what the maximum analog input voltage for the DRS4 V5 evaluation board is. It came with a sticker indicating that it is "2.5V pk Max". On the other hand, when checking the manual (https://www.psi.ch/drs/DocumentationEN/manual_rev50.pdf), it says maximum allowed is 10V DC or even 30V for short pulses. I foresee an application where I cannot make sure that pulses stay below 2.5V, so the correct value will be quite important for me.

Best,
Julian

 

 

ELOG V3.1.5-3fb85fa6