Entry  Thu Mar 5 14:37:48 2026, QICHAOWU, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode figure1.pngfigure2.pngfigure3.png
    Reply  Thu Mar 5 14:52:07 2026, Stefan Ritt, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 
       Reply  Thu Mar 5 15:39:30 2026, QICHAOWU, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 
Message ID: 932     Entry time: Thu Mar 5 15:39:30 2026     In reply to: 931
Author: QICHAOWU 
Subject: Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 

Thank you for your prompt reply. We have developed our own DRS4-based hardware and successfully read out all 1024 bins using ROI mode. However, we are encountering the waveform truncation issue mentioned earlier. If possible, we would also like to purchase an evaluation board for reference.

Stefan Ritt wrote:

Does this happen with your own hardware or with the DRS4 evaluation board? If it's your own hardware, do you read out all 1024 bins in the ROI mode or only a partial waveforem?

QICHAOWU wrote:

Dear Dr. Stefan Ritt

I am writing to seek your advice regarding an issue I encountered while using the DRS4 chip in ROI (Region of Interest) readout mode to sample waveforms. My name is Qichao Wu, and I have been working with the DRS4 for waveform sampling.

In my setup, I input two identical sine waves (with a period of 250 ns and the same phase) into two different channels of the same DRS4 chip. Ideally, the waveforms acquired from both channels should completely overlap, as shown in Figure 1. However, I occasionally observe that the waveform from one of the channels is truncated. The truncated portion appears to originate from a different segment of the input signal, resulting in only partial overlap between the two channels. This phenomenon is illustrated in Figure 1. Additionally, there are instances where both channels exhibit truncation simultaneously, as depicted in Figure 2.

I have confirmed this issue by observing the truncated waveforms directly on an oscilloscope when the DRS4 is operated in ROI mode. Notably, the truncation points always occur near cell 0 and cell 1023 of the DRS4 switch array.

Could you kindly provide guidance on how to resolve or mitigate this issue? Any suggestions or insights you could offer would be greatly appreciated.

I look forward to your positive response and thank you in advance for your time and assistance.

Best regards,
Qichao Wu

 

 

 