Entry  Thu Mar 5 14:37:48 2026, QICHAOWU, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode figure1.pngfigure2.pngfigure3.png
    Reply  Thu Mar 5 14:52:07 2026, Stefan Ritt, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 
       Reply  Thu Mar 5 15:39:30 2026, QICHAOWU, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 
    Reply  Fri Mar 27 09:56:06 2026, Stefan Ritt, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 
       Reply  Fri Apr 17 10:54:20 2026, QICHAOWU, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 
          Reply  Fri Apr 17 12:39:11 2026, Stefan Ritt, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 
             Reply  Tue Apr 28 04:40:38 2026, QICHAOWU, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 
                Reply  Tue Apr 28 08:15:35 2026, Stefan Ritt, Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 
Message ID: 935     Entry time: Fri Apr 17 12:39:11 2026     In reply to: 934     Reply to this: 936
Author: Stefan Ritt 
Subject: Waveform Truncation Near the Cell0-Cell1023 Boundary in DRS4 ROI Readout Mode 

Not much I can say here. I know that the DRS4 works reliably, since we have an experiment here with 500 chips running since 5 years and we did not observer what you described. Therefore, it must be some subtle detail of your FPGA code or the voltage levels on your board. As I said, what I would do is to get the evaluation abord and copare signals 1:1.

Stefan

QICHAOWU wrote:

Dear Dr.Stefan

Thank you for your detailed response. Following your suggestion, I used an oscilloscope to observe the RSLOAD and SRCLK signals, and they conform to the readout timing described in the datasheet.

I can now confirm that the DRS4 occasionally fails to correctly update the stop position to cell 1023, or the stored waveform from cell 0 to the stop position is corrupted. As a result, part of the waveform comes from the most recent acquisition while the other part comes from an earlier one, causing the truncation I observed. This issue occurs with a certain probability—more than half of the acquisitions still yield complete, continuous waveforms.

Interestingly, when I use the same FPGA code and host software on the DRS4 evaluation board, no truncation occurs at all. I am currently comparing the behavior of the two setups from both hardware and software perspectives. Could you kindly provide some suggestions on further tests I could perform?

Thank you very much for your help.

Best regards,
Qichao Wu

 

Stefan Ritt wrote:

Actually looking at your waveforms again, I think I understand maybe your problem. You stop the DRS4 randomly, but your ROI readout mode does not really work. Seems like you always read from the first cell, or just from a wrong cell. Since your sine wave period is not a multiple of the readout window, you see a phase jump between the beginning of your sampling window and the end of your sampling window. The little spikes before and after the jump are normal, since the first and last samples of the DRS4 readout might have some offset. Check that your FPGA really uses RSLOAD to start the readout, and not the "full readout".

The evaluation board might really help you, since you can compare 1:1 all control signals with yours.

Stefan

QICHAOWU wrote:

Dear Dr. Stefan Ritt

I am writing to seek your advice regarding an issue I encountered while using the DRS4 chip in ROI (Region of Interest) readout mode to sample waveforms. My name is Qichao Wu, and I have been working with the DRS4 for waveform sampling.

In my setup, I input two identical sine waves (with a period of 250 ns and the same phase) into two different channels of the same DRS4 chip. Ideally, the waveforms acquired from both channels should completely overlap, as shown in Figure 1. However, I occasionally observe that the waveform from one of the channels is truncated. The truncated portion appears to originate from a different segment of the input signal, resulting in only partial overlap between the two channels. This phenomenon is illustrated in Figure 1. Additionally, there are instances where both channels exhibit truncation simultaneously, as depicted in Figure 2.

I have confirmed this issue by observing the truncated waveforms directly on an oscilloscope when the DRS4 is operated in ROI mode. Notably, the truncation points always occur near cell 0 and cell 1023 of the DRS4 switch array.

Could you kindly provide guidance on how to resolve or mitigate this issue? Any suggestions or insights you could offer would be greatly appreciated.

I look forward to your positive response and thank you in advance for your time and assistance.

Best regards,
Qichao Wu

 

 

 

 

ELOG V3.1.5-3fb85fa6