  521   Wed May 11 04:01:14 2016 MaksatDRS4 Macro to save events

Dear Stefan,

I am trying to setup DRS inside radiation enclosure and would like to write a simple script that will automatically save certain number of events.

Could you please point to me an example that can I use for Mac OS? I saw there is drs_exam.cpp in the directory but was not able to get work in Mac OS. Any help would be greatly appreciated.

Thanks

 

 

  520   Mon May 2 14:31:28 2016 Dmitry Hitstwo DRS4 boards configuration with 2048 samples each

Hi Stefan

Any chance you have time to fix the software for multiboard configuration with 2048 samples each. I tried 5.0.5, but drsosc still shows only half of the waveform.

Dmitry

Stefan Ritt wrote:

The multi-board mode has never been tested with 2048 samples, so is very likely not to work. I don't know yet how much work this will be to fix, but I'm on a business trip the next three weeks and probably will only have time to look at it when I return.

Stefan

Dmitry Hits wrote:

Dear Stefan,

I daisy-chained two boards (master sn#: 2514 - slave sn#: 2513) each with 2048 samples. However, when I use drsosc and put check mark in "configure multi-board daisy-chain" I see only 1024 samples. Namely, the first 1024 samples, the last part is missing. When I remove this check mark, I see all 2048 samples. Is there a simple software fix for this or is it a more involved firmware limitation? 

Other parameters: software version: 5.0.4, firmware version 21305, configured for 0.7 GSPS, display at 500 ns/div

Thank you,

Dmitry Hits.

 

 

  519   Thu Apr 28 15:47:53 2016 Stefan RittNew software version and binary format

A new software version 5.0.5 has been released today. This fixes a few bugs in multi-board configurations, and adds saving of the scaler values into XML and binary files. Please note that the binary file format has been changed for that. The new format is described in an updated manual (page 25), and reflected in a new read_binary.cpp program contained in the distribution.

/Stefan

  518   Thu Apr 28 15:46:34 2016 Stefan RittBest settings for time measurements

The DRS4 chip has been designed to work best at high sampling speeds. At 700 MSPS, the chip is at it's limit and timing is very poorr (ns?). In order to get good timing, run it at least at 2 GSPS.

Stefan

Abaz Kryemadhi wrote:

I am studing some pulses that are about 200-300 ns wide and a rise time of few ns,    which settings would be best for coincidence time measurements?

In some preliminary work I found for 700 MegaS the time measurement is better without time calibration (in -0.05 to 1V) rather than with time calibration in -0.5 to 0.5,  my pulses are about 60 mV.   Is it expected that always with time calibration time accuracy would be better or depends?   

Also I use this code snippet to find time for channel 1 and the same idea for chan. 2.

// find peak in channel 1 above threshold
      for (i=0 ; i<1022 ; i++)
         if (waveform[0][i] < threshold1 && waveform[0][i+1] >= threshold1) {
            tt1 = (threshold1-waveform[0][i])/(waveform[0][i+1]-waveform[0][i])*(time[0][i+1]-time[0][i])+time[0][i];
            break;
         }

 

Thanks!

Abaz

 

  517   Wed Apr 27 20:04:12 2016 Abaz KryemadhiBest settings for time measurements

I am studing some pulses that are about 200-300 ns wide and a rise time of few ns,    which settings would be best for coincidence time measurements?

In some preliminary work I found for 700 MegaS the time measurement is better without time calibration (in -0.05 to 1V) rather than with time calibration in -0.5 to 0.5,  my pulses are about 60 mV.   Is it expected that always with time calibration time accuracy would be better or depends?   

Also I use this code snippet to find time for channel 1 and the same idea for chan. 2.

// find peak in channel 1 above threshold
      for (i=0 ; i<1022 ; i++)
         if (waveform[0][i] < threshold1 && waveform[0][i+1] >= threshold1) {
            tt1 = (threshold1-waveform[0][i])/(waveform[0][i+1]-waveform[0][i])*(time[0][i+1]-time[0][i])+time[0][i];
            break;
         }

 

Thanks!

Abaz

  516   Wed Apr 27 09:51:37 2016 Toshihiro Nonakaserial number problem

The serial number has been fixed by using drscl. Thank you!

Stefan Ritt wrote:

If dis- and reconnecting the board does not help, there is the (small) chance that the serial number got erased in the board. You can re-set it with the "drscl" command line tool:

$ drscl
Found DRS4 board 0 on USB, serial #0, firmware revision 21305
B0> serial 2172

 

Toshihiro Nonaka wrote:

Dear all,

I'm using 3 DRS boards simultaneously and their serial numbers are 2169, 2170, 2172 respectively.

Recently however,  I obtain serial number "0" by DRSBoard::GetBoardSerialNumber() for #2172 board.

Data taking can be done without any problems, but I'd like to know what is happening.

Any advice?

Thank you.

Toshihiro Nonaka

 

 

  515   Wed Apr 27 09:04:01 2016 Stefan Rittserial number problem

If dis- and reconnecting the board does not help, there is the (small) chance that the serial number got erased in the board. You can re-set it with the "drscl" command line tool:

$ drscl
Found DRS4 board 0 on USB, serial #0, firmware revision 21305
B0> serial 2172

 

Toshihiro Nonaka wrote:

Dear all,

I'm using 3 DRS boards simultaneously and their serial numbers are 2169, 2170, 2172 respectively.

Recently however,  I obtain serial number "0" by DRSBoard::GetBoardSerialNumber() for #2172 board.

Data taking can be done without any problems, but I'd like to know what is happening.

Any advice?

Thank you.

Toshihiro Nonaka

 

  514   Wed Apr 27 08:14:14 2016 Toshihiro Nonakaserial number problem

Dear all,

I'm using 3 DRS boards simultaneously and their serial numbers are 2169, 2170, 2172 respectively.

Recently however,  I obtain serial number "0" by DRSBoard::GetBoardSerialNumber() for #2172 board.

Data taking can be done without any problems, but I'd like to know what is happening.

Any advice?

Thank you.

Toshihiro Nonaka

  513   Tue Apr 26 13:42:42 2016 Stefan RittDRS4 purchase information

Just be patient. Anita is not at work this week.

Konstantin Gusev wrote:

    Hi,

 I can't contact with Anita Van Loon about DSR4 chip's price and delivery.

Did you still sell it? Can you provide me this information?

 

  512   Tue Apr 26 09:54:16 2016 Stefan RittNegative fCellDT values from GetTimeCalibration()

I just realized that the negative bin widht is not explicitly mentioned in the quoted paper. So let me explain it here:

The negative value of cell 498 is correct and "real" in the sense that the signal is first captured in cell 498 and later in cell 497. This is due to the exact layout of the cells on the chip and the input signal. Cell 498 is simply much closer to the input, so sees the signal earlier than cell 497, even if it's triggerd after cell 497. So nothing to worry about.

Stefan

Daniel Stricker-Shaver wrote:

Hi Kyle,

If I remember right the negative sampling width happens only for 498 and at high sampling speeds. It is described in a paper from Stefan:

http://arxiv.org/pdf/1405.4975.pdf

or

“Novel Calibration Method for Switched Capacitor Arrays Enables Time Measurements With Sub-Picosecond Resolution”( IEEE Transactions on Nuclear Science 61 (2014),Nr. 6, 3607–3617)

Kyle Weinfurther wrote:

Hello Stefan,

I am using four DRS4 v5 eval boards to digitize 16 channels of data. I have recently changed from saving the timing information of the waveform using GetTime() to GetTimeCalibration(). When changing over, I noticed that some values for fCellDT for cell 498 are negative. Over the 16 channels used, 4 of them have negative time bin widths for cell 498 while the other 12 channels are very close to 0 (in the ~10 ps range). One of the eval boards has no negative fCellDT whereas the other three boards have one or two channels with negative values.

Upon further inspection, I checked the time between samples of GetTime() and found the same results in cell 498. After finding this, I did a timing calibration again with CalibrateTiming() even though in a different post on the discussion forum you said it was valid for a wide range of temperatures and a long time (years). This still allowed the negative fCellDT values to persist.

Is this a common occurance? If so, is there a method to fix this issue? Is there a reason for cell 498 to have a small value for fCellDT? I searched the discussion forum and did not find anything relating to this issue.

Attached are a couple waveform traces using GetTime() zoomed in on cell 498.

Thanks,

Kyle Weinfurther

 

 

  511   Sat Apr 23 12:33:17 2016 Daniel Stricker-ShaverNegative fCellDT values from GetTimeCalibration()

Hi Kyle,

If I remember right the negative sampling width happens only for 498 and at high sampling speeds. It is described in a paper from Stefan:

http://arxiv.org/pdf/1405.4975.pdf

or

“Novel Calibration Method for Switched Capacitor Arrays Enables Time Measurements With Sub-Picosecond Resolution”( IEEE Transactions on Nuclear Science 61 (2014),Nr. 6, 3607–3617)

Kyle Weinfurther wrote:

Hello Stefan,

I am using four DRS4 v5 eval boards to digitize 16 channels of data. I have recently changed from saving the timing information of the waveform using GetTime() to GetTimeCalibration(). When changing over, I noticed that some values for fCellDT for cell 498 are negative. Over the 16 channels used, 4 of them have negative time bin widths for cell 498 while the other 12 channels are very close to 0 (in the ~10 ps range). One of the eval boards has no negative fCellDT whereas the other three boards have one or two channels with negative values.

Upon further inspection, I checked the time between samples of GetTime() and found the same results in cell 498. After finding this, I did a timing calibration again with CalibrateTiming() even though in a different post on the discussion forum you said it was valid for a wide range of temperatures and a long time (years). This still allowed the negative fCellDT values to persist.

Is this a common occurance? If so, is there a method to fix this issue? Is there a reason for cell 498 to have a small value for fCellDT? I searched the discussion forum and did not find anything relating to this issue.

Attached are a couple waveform traces using GetTime() zoomed in on cell 498.

Thanks,

Kyle Weinfurther

 

  509   Thu Apr 21 22:16:43 2016 Kyle WeinfurtherNegative fCellDT values from GetTimeCalibration()

Hello Stefan,

I am using four DRS4 v5 eval boards to digitize 16 channels of data. I have recently changed from saving the timing information of the waveform using GetTime() to GetTimeCalibration(). When changing over, I noticed that some values for fCellDT for cell 498 are negative. Over the 16 channels used, 4 of them have negative time bin widths for cell 498 while the other 12 channels are very close to 0 (in the ~10 ps range). One of the eval boards has no negative fCellDT whereas the other three boards have one or two channels with negative values.

Upon further inspection, I checked the time between samples of GetTime() and found the same results in cell 498. After finding this, I did a timing calibration again with CalibrateTiming() even though in a different post on the discussion forum you said it was valid for a wide range of temperatures and a long time (years). This still allowed the negative fCellDT values to persist.

Is this a common occurance? If so, is there a method to fix this issue? Is there a reason for cell 498 to have a small value for fCellDT? I searched the discussion forum and did not find anything relating to this issue.

Attached are a couple waveform traces using GetTime() zoomed in on cell 498.

Thanks,

Kyle Weinfurther

  508   Fri Apr 15 12:58:46 2016 Konstantin GusevDRS4 purchase information

    Hi,

 I can't contact with Anita Van Loon about DSR4 chip's price and delivery.

Did you still sell it? Can you provide me this information?

  507   Wed Apr 6 09:46:10 2016 Daniel DribinDRS Oscilloscope freezing after a long run
Martin Petriska wrote:

 

Stefan Ritt wrote:

I tried this night to run the board at a 10 Hz rate with an external pulser, without writing, and it did not freeze after ~14 hours of running on Mac OSX. This night I will try again with writing.

Stefan

Stefan Ritt wrote:

Then it seems that there is some USB communication problem. I heard this also from other people, that the USB data transfer under Windows has sometimes problems. I develop and run the board under Mac OSX, and there the same software runs for days without problem. So I guess it's related to the underlying libusb lib which is used by the DRS oscilloscope, on which I have no influence. So the only advice I can give is to take shorter series of data. Anyhow the board is not considered a full DAQ system, just an "evaluation board" which means one can try the DRS4 chip and play with it. For serious business one should build own electronics with the chip. Anyhow we are currently developping an Ethernet board which allows much faster acquisition rates, so USB will be obsolete some day. Nevertheless I will try to reproduce your problem and see if I can do anything. At what trigger rate does it show up most prominently?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

Yes I use Windows 7, If the DRS Oscilloscope program stays on for a couple of hours without saving the data, the problem will occur. It seems it happens more often when there is data writing and when the rate of events is slow, about 100 events per second, at high rates it almost doesn't happen. Can it be temperature related?

Daniel

Stefan Ritt wrote:

Dear Daniel,

sorry my late reply, I'm pretty busy these days. The behavior you report has not been seen before, but I guess no one tried to take such long runs of data yet. Can you confirm that the problem also occurs without writing data to disk, or is it disk-related? I guess you use it under Windows 7, right?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

I am using a DRS4 v5 to do timing measurements of Positron lifetime. I use the DRS Oscilloscope with triggering on 2 channels when I have a coincidence. Attached is a picture with all the setting that I use. When I use the DRS4 for a long measurements of 5 million events for a couple of hours, the DRS Oscilloscope stops showing any signal .After the first restart of the program I get a strange signal which is at the bottom of the scope range of voltage picture below(in the picture I changed the vertical positions of the channels for better viewing). Only after a couple of DRS Oscilloscope restarts and USB reconnections do I get the results again.

I currently am using another DRS4 v5 and the same situation occurs again although with lower frequency.

What can I do to solve this problem?

thank you very much,

Daniel

 

 

 

 

 

Hi I have also positron annihilation system based on DRS4v4 cards. Its running several weeks, sometimes months, without freezing in windows7 64bit system (Pentium Core Quad 2Ghz, 4GRam).  Problem was when widows was trying to install updates and restarted PC. In beginning I had some problem with memory leak in my application, but it was simple seen in task manager that application memory was rising and was need to find memory leak in application code. Now I remember card was sometimes freezing when room air conditioning with 2kW was starting and high electricity pulses were reason of USB problems, it helped to put air conditioner and PC in different power line input. Hope it help to solve zour problems.

Martin

I too have air conditioning in the room, not sure to which power line it is connected, but I'll try to check this aswell.

Thank you very much.

  506   Wed Apr 6 09:43:52 2016 Daniel DribinDRS Oscilloscope freezing after a long run

At hight rates I worked with files of up to 20 GB so I don't think this is the problem.

I will try to run it under Ubuntu and see if i can recreate the problem.

Thank you very much for the quick responses and help.

Stefan Ritt wrote:

Even with writing for one night no problem (see below). Have you checked how big your data file is? I guess there is a limit under Windows of 2 GB. If that's the case, you have to write shorter files.

 

 

  505   Wed Apr 6 09:01:28 2016 Martin PetriskaDRS Oscilloscope freezing after a long run

 

Stefan Ritt wrote:

I tried this night to run the board at a 10 Hz rate with an external pulser, without writing, and it did not freeze after ~14 hours of running on Mac OSX. This night I will try again with writing.

Stefan

Stefan Ritt wrote:

Then it seems that there is some USB communication problem. I heard this also from other people, that the USB data transfer under Windows has sometimes problems. I develop and run the board under Mac OSX, and there the same software runs for days without problem. So I guess it's related to the underlying libusb lib which is used by the DRS oscilloscope, on which I have no influence. So the only advice I can give is to take shorter series of data. Anyhow the board is not considered a full DAQ system, just an "evaluation board" which means one can try the DRS4 chip and play with it. For serious business one should build own electronics with the chip. Anyhow we are currently developping an Ethernet board which allows much faster acquisition rates, so USB will be obsolete some day. Nevertheless I will try to reproduce your problem and see if I can do anything. At what trigger rate does it show up most prominently?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

Yes I use Windows 7, If the DRS Oscilloscope program stays on for a couple of hours without saving the data, the problem will occur. It seems it happens more often when there is data writing and when the rate of events is slow, about 100 events per second, at high rates it almost doesn't happen. Can it be temperature related?

Daniel

Stefan Ritt wrote:

Dear Daniel,

sorry my late reply, I'm pretty busy these days. The behavior you report has not been seen before, but I guess no one tried to take such long runs of data yet. Can you confirm that the problem also occurs without writing data to disk, or is it disk-related? I guess you use it under Windows 7, right?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

I am using a DRS4 v5 to do timing measurements of Positron lifetime. I use the DRS Oscilloscope with triggering on 2 channels when I have a coincidence. Attached is a picture with all the setting that I use. When I use the DRS4 for a long measurements of 5 million events for a couple of hours, the DRS Oscilloscope stops showing any signal .After the first restart of the program I get a strange signal which is at the bottom of the scope range of voltage picture below(in the picture I changed the vertical positions of the channels for better viewing). Only after a couple of DRS Oscilloscope restarts and USB reconnections do I get the results again.

I currently am using another DRS4 v5 and the same situation occurs again although with lower frequency.

What can I do to solve this problem?

thank you very much,

Daniel

 

 

 

 

 

Hi I have also positron annihilation system based on DRS4v4 cards. Its running several weeks, sometimes months, without freezing in windows7 64bit system (Pentium Core Quad 2Ghz, 4GRam).  Problem was when widows was trying to install updates and restarted PC. In beginning I had some problem with memory leak in my application, but it was simple seen in task manager that application memory was rising and was need to find memory leak in application code. Now I remember card was sometimes freezing when room air conditioning with 2kW was starting and high electricity pulses were reason of USB problems, it helped to put air conditioner and PC in different power line input. Hope it help to solve zour problems.

Martin

  504   Wed Apr 6 08:41:08 2016 Stefan RittDRS Oscilloscope freezing after a long run

Even with writing for one night no problem (see below). Have you checked how big your data file is? I guess there is a limit under Windows of 2 GB. If that's the case, you have to write shorter files.

 

  503   Tue Apr 5 16:08:59 2016 Stefan RittDRS Oscilloscope freezing after a long run

I tried this night to run the board at a 10 Hz rate with an external pulser, without writing, and it did not freeze after ~14 hours of running on Mac OSX. This night I will try again with writing.

Stefan

Stefan Ritt wrote:

Then it seems that there is some USB communication problem. I heard this also from other people, that the USB data transfer under Windows has sometimes problems. I develop and run the board under Mac OSX, and there the same software runs for days without problem. So I guess it's related to the underlying libusb lib which is used by the DRS oscilloscope, on which I have no influence. So the only advice I can give is to take shorter series of data. Anyhow the board is not considered a full DAQ system, just an "evaluation board" which means one can try the DRS4 chip and play with it. For serious business one should build own electronics with the chip. Anyhow we are currently developping an Ethernet board which allows much faster acquisition rates, so USB will be obsolete some day. Nevertheless I will try to reproduce your problem and see if I can do anything. At what trigger rate does it show up most prominently?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

Yes I use Windows 7, If the DRS Oscilloscope program stays on for a couple of hours without saving the data, the problem will occur. It seems it happens more often when there is data writing and when the rate of events is slow, about 100 events per second, at high rates it almost doesn't happen. Can it be temperature related?

Daniel

Stefan Ritt wrote:

Dear Daniel,

sorry my late reply, I'm pretty busy these days. The behavior you report has not been seen before, but I guess no one tried to take such long runs of data yet. Can you confirm that the problem also occurs without writing data to disk, or is it disk-related? I guess you use it under Windows 7, right?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

I am using a DRS4 v5 to do timing measurements of Positron lifetime. I use the DRS Oscilloscope with triggering on 2 channels when I have a coincidence. Attached is a picture with all the setting that I use. When I use the DRS4 for a long measurements of 5 million events for a couple of hours, the DRS Oscilloscope stops showing any signal .After the first restart of the program I get a strange signal which is at the bottom of the scope range of voltage picture below(in the picture I changed the vertical positions of the channels for better viewing). Only after a couple of DRS Oscilloscope restarts and USB reconnections do I get the results again.

I currently am using another DRS4 v5 and the same situation occurs again although with lower frequency.

What can I do to solve this problem?

thank you very much,

Daniel

 

 

 

 

 

  502   Mon Apr 4 12:08:15 2016 Stefan RittDRS Oscilloscope freezing after a long run

Then it seems that there is some USB communication problem. I heard this also from other people, that the USB data transfer under Windows has sometimes problems. I develop and run the board under Mac OSX, and there the same software runs for days without problem. So I guess it's related to the underlying libusb lib which is used by the DRS oscilloscope, on which I have no influence. So the only advice I can give is to take shorter series of data. Anyhow the board is not considered a full DAQ system, just an "evaluation board" which means one can try the DRS4 chip and play with it. For serious business one should build own electronics with the chip. Anyhow we are currently developping an Ethernet board which allows much faster acquisition rates, so USB will be obsolete some day. Nevertheless I will try to reproduce your problem and see if I can do anything. At what trigger rate does it show up most prominently?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

Yes I use Windows 7, If the DRS Oscilloscope program stays on for a couple of hours without saving the data, the problem will occur. It seems it happens more often when there is data writing and when the rate of events is slow, about 100 events per second, at high rates it almost doesn't happen. Can it be temperature related?

Daniel

Stefan Ritt wrote:

Dear Daniel,

sorry my late reply, I'm pretty busy these days. The behavior you report has not been seen before, but I guess no one tried to take such long runs of data yet. Can you confirm that the problem also occurs without writing data to disk, or is it disk-related? I guess you use it under Windows 7, right?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

I am using a DRS4 v5 to do timing measurements of Positron lifetime. I use the DRS Oscilloscope with triggering on 2 channels when I have a coincidence. Attached is a picture with all the setting that I use. When I use the DRS4 for a long measurements of 5 million events for a couple of hours, the DRS Oscilloscope stops showing any signal .After the first restart of the program I get a strange signal which is at the bottom of the scope range of voltage picture below(in the picture I changed the vertical positions of the channels for better viewing). Only after a couple of DRS Oscilloscope restarts and USB reconnections do I get the results again.

I currently am using another DRS4 v5 and the same situation occurs again although with lower frequency.

What can I do to solve this problem?

thank you very much,

Daniel

 

 

 

 

  501   Mon Apr 4 11:41:26 2016 Daniel DribinDRS Oscilloscope freezing after a long run

Dear Stefan Ritt,

Yes I use Windows 7, If the DRS Oscilloscope program stays on for a couple of hours without saving the data, the problem will occur. It seems it happens more often when there is data writing and when the rate of events is slow, about 100 events per second, at high rates it almost doesn't happen. Can it be temperature related?

Daniel

Stefan Ritt wrote:

Dear Daniel,

sorry my late reply, I'm pretty busy these days. The behavior you report has not been seen before, but I guess no one tried to take such long runs of data yet. Can you confirm that the problem also occurs without writing data to disk, or is it disk-related? I guess you use it under Windows 7, right?

Stefan

Daniel Dribin wrote:

Dear Stefan Ritt,

I am using a DRS4 v5 to do timing measurements of Positron lifetime. I use the DRS Oscilloscope with triggering on 2 channels when I have a coincidence. Attached is a picture with all the setting that I use. When I use the DRS4 for a long measurements of 5 million events for a couple of hours, the DRS Oscilloscope stops showing any signal .After the first restart of the program I get a strange signal which is at the bottom of the scope range of voltage picture below(in the picture I changed the vertical positions of the channels for better viewing). Only after a couple of DRS Oscilloscope restarts and USB reconnections do I get the results again.

I currently am using another DRS4 v5 and the same situation occurs again although with lower frequency.

What can I do to solve this problem?

thank you very much,

Daniel

 

 

 

ELOG V3.1.5-3fb85fa6