DRS4 Forum
  DRS4 Discussion Forum, Page 25 of 46  Not logged in ELOG logo
    Reply  Mon Jan 3 17:13:41 2022, Stefan Ritt, DRS4 request assistance 
1. fDOMINO is defined as fREFCLK * 2048

2. Good values can be derived from the evaluation board schematics: C1=4.7nF, C2=1nF, R=130 Ohm

3. A "1" means a logical high level. See Wikipedia: https://en.wikipedia.org/wiki/Logic_level
    Reply  Sat Jan 15 10:50:47 2022, Stefan Ritt, I want to know about the readout 



student_riku
wrote:



Am I right in thinking that inputting 1 to DMODE (Bit0) in the configuration
register will connect the 1024th cell to the 1st cell?
    Reply  Tue Jan 25 14:34:42 2022, Stefan Ritt, Regarding measuring for a set time 
drsosc is a graphical application contiously acquiring data from the board, and drscl is a command line tool for debugging, as written in the manual.

The drsosc application runs indefinitely, but I guess you refer to saving data (by hitting the "Save" button in the drsosc application).
Yes the save functionality has a number of events, since you cannot store data indefinitely, since your harddisk does not have indefinite space!
    Reply  Tue Feb 15 12:02:29 2022, Stefan Ritt, Cannot trigger on pulses, have to trigger on undershoot 
The trigger comparator is a ADCMP601 unit which requires a minimum pulse width of 3-4 ns. I see that your pulses are only 1-2 ns wide. You have to make
your pulses wider in order to trigger on them.

Stefan
    Reply  Thu Mar 3 13:47:26 2022, Stefan Ritt, How to convert samples to volt? 
The 'drscl' tool is more for experts, normal users are advised to use the DRSOsc oscilloscope.

The board has to be calibrated for a given sampling speed before calibrated data can be read out. Do that with the "calib" command,
specifying 5 for the sampling rate, 0 for the range (which is the middle between -0.5 and +0.5) and 1 for 1024 mode. If you then do "start",
    Reply  Thu Mar 3 16:14:16 2022, Stefan Ritt, Scaler issue to evaluate live time  
The scalers are read out 10x per seconds, so they have an accuracy of 10 Hz. I tried a 50 Hz pulser, and measured 40 Hz, I tried 52 Hz and measured 50
Hz. This is about what you can expect.

The scaler rate is measured after the discriminator of the trigger, so the trigger level also affects the scaler reading. If you have a 100 mV
    Reply  Mon Mar 7 08:45:32 2022, Stefan Ritt, Why does not trigger at higher sampling frequencies? 
Unfortunately I have not idea what the problem could be. In principle the trigger should be independent of the sampling speed, since the trigger is only
made with a discriminator and a flip-flop. The hardware must be ok since you see the trigger with the oscillocope app. All you can do is to go through
the sorce code of the oscilloscope app, especially drsosc/Osic.cpp::ScanBoards(), SetTriggerLevel(), SetTriggerPolariy() etc. to make sure you do the same
    Reply  Mon Mar 7 16:37:54 2022, Stefan Ritt, Scaler issue to evaluate live time  Screenshot_2022-03-07_at_16.37.32_.pngScreenshot_2022-03-07_at_16.35.44_.png
I tried your measurement with the DRSOscilloscope app (see below), and I measure a constant difference of 10 Hz among the whole range of 100 Hz to 3
kHz.

So I don't know what's wrong on your side. Did you try with the oscilloscope app as well? Have you checked your pulse generator? The
    Reply  Sat Mar 12 10:13:24 2022, Stefan Ritt, Time calibration and the C++ API 
DRSBoard::GetTime is declared in DRS.h line 720.

If you want to measure timing down to ps, you need some basic knowledge, especially about signal-to-noise and risetime. This cannot be taught
in a few sentenses, needs a full lecture. As a starting point please read that papat:
    Reply  Mon Mar 14 08:59:51 2022, Stefan Ritt, Time calibration and the C++ API Screenshot_2022-03-14_at_9.04.07_.pngScreenshot_2022-03-14_at_9.03.47_.png
Looks like you have the some time calibration, not sure if it's the correct one. Sample the sine wave from the calibration clock, once with and once
without the timing calibration, then you will see if all points lie on a smooth line. Left: without timing calibration, right: with proper timing calibration:

    Reply  Tue Apr 12 10:49:27 2022, Stefan Ritt,  
A3-A0 = 1001 should be all you need to activate OUT0-OUT7. It works in our designs. Maybe double check the address lines with an oscilloscope.

Stefan




LynseyShun
    Reply  Fri Jul 29 14:09:35 2022, Stefan Ritt, Increase event rate, use ROI mode, and install sw from source in Mac 
The firmware from the website always reads 1024 bins. You have to modify it to stop before that, like reading only 128 samples or so. For compiling under
MacOSX, this should work, since I do it myself. 

Regards,
    Reply  Fri Jul 29 17:23:43 2022, Stefan Ritt, Spikes/noise sensitive to clock settings? 
Look at the DRS4 data sheet, Figure 12. You see there the rising SRCLK pulse which outputs the next analog value. You also see tSAMP which describes
the sampling piont (strobe or clock sent to your ADC). The value of tSAMP must be such that the values is sampled at the point where it flattens out, just
2-3 ns BEFORE the next analog sample is clocked out, as written in the text. So you have to phase shift your clock going to SRCLK and the one going to
    Reply  Tue Sep 27 10:37:11 2022, Stefan Ritt, Required Firmware for DRS4 Evaluation Board Version 2.0 
You find each software version at the usual download location at

https://www.dropbox.com/home/drs/drs4/distribution/Download/Linux

The one you need is probably drs-2.1.3.tar.gz which was the last version for the 2.0 board which is now more than 10 years old.
    Reply  Tue Sep 27 15:20:55 2022, Stefan Ritt, Required Firmware for DRS4 Evaluation Board Version 2.0 
Sorry, got the wrong link. Here the right one: https://www.dropbox.com/sh/clqo7ekr0ysbrip/AACoWJzrQAbf3WiBJHG89bGGa?dl=0

If you untar the archive, you will find a "firmware" subdirectory with all VHDL code.

Stefan
    Reply  Mon Oct 24 12:50:24 2022, Stefan Ritt, Channel Cascading Option in the 2048-bin 
The board is delivered in one or the other mode and not meant to be changed by the user, since this requires very delicate soldering which is not easy.
If you try anyhow, you loose the quarantee. You can send the board back to the manufacturer for the modification, but this costs quite some moeny.

Best regards,
    Reply  Mon Feb 6 13:28:28 2023, Stefan Ritt, DRS4 installation via tar in ubuntu not working 
I fixed the described error. Can you try the new version from https://bitbucket.org/ritt/drs4eb/commits/80b3af753ed32eb365725f0f3244a4109347c01b




Sebastian
Infante wrote:



Hello i cant install any the last versions that i downloaded from
    Reply  Mon Jun 12 14:22:04 2023, Stefan Ritt, Different sampling rates in multi-board configuration 
No, that's unfortunately not possible.

Stefan




Javier
    Reply  Wed Sep 13 13:18:45 2023, Stefan Ritt, Input range switch added in Version 2.1.3 
To achieve an input range of -1V to 0V, you need an external buffer which can shift this range into the DRS4 range of -0.5V to +0.5V. This external buffer
has then to operate with bipolar power supplies, like -2.5V to +2.5V, which are not present on the evaluation board.

Best regards,
    Reply  Wed Oct 25 19:47:23 2023, Stefan Ritt, WaveDREAM Design 
No. This is a proprietary design.

Best,
Stefan
ELOG V3.1.5-3fb85fa6