Entry  Fri Jun 9 04:11:40 2023, Javier Caravaca, Different sampling rates in multi-board configuration 

Hello,

Is it possible to have different sampling rates in multi-board configuration? I tried using the scope application but I am unable to change the sampling rate independently.

Best,

Javier.

Entry  Tue Nov 16 01:27:51 2021, Jacquelynne Vaughan, V3 board, only one channel works, all components at each channel input working 

Hi everyone,

I'm still looking through the forum for an answer to this question, but thought I'd go ahead and post anyway just in case it hasn't been answered yet. If it has I can take this post down.

I have a V3 board, and as far as I can tell only channel 2 gives an output. If I enable other channels using the DRS Oscilloscope software, they do show static but will not show a signal if I connect one to them (e.g. a series of subsequent square waves). A technician and I took the board out and tested all the components leading up to the microcontrollers for each channel, and everything seemed to be working fine. I thought maybe it was configured to only have one channel read an output, but I looked through the Config panel in the software and nothing seemed to indicate that.

I am a novice, and maybe I'm missing something that I didn't see in the manual. I can post screenshots if needed!

Thank you for your help!

Entry  Thu Sep 29 17:26:13 2016, Jacob Hwang, Output Timing Drifting Output_Drifting.jpg

Hello,

I have designed four DRS4 chips (36 channels) on my board running at 1GHz (REFCLK=488.28KHz) and ROI mode. All 4 chips' REFCLK, DWRITE, RSRLOAD, and SRCLK are buffer driven by the same source.  SRCLK is set to 40MHz to reduce the readout time.

If I injected a sine waveform, buffered and splitted into all 36 channels,I noticed all 9 channels on each DRS4 chip output almost the same as expected.  But the output phase from chip to chip is drifting as shown in attached picture which is from two different channels of different chips.  From the few boards I have built, I found few chips are drifting more than the others and is different on every board.

The sympton look like the DRS4 internal PLL is drifting, but I checked the DTAP output on every chip and found it's dead-lock steady even I used persistance setting on my oscilloscope.  Do you have any suggestion how to attack this problem?  Thank you.

Jacob Hwang
 

Entry  Tue Jan 12 17:57:03 2016, Jack Bargemann, Compiling DRS-exam 

I am trying to compile drs-exam, but am getting an error message I do not understand:

1>musbstd.obj : error LNK2019: unresolved external symbol _usb_open referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_close referenced in function _musb_close
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_get_descriptor referenced in function _musb_get_device
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_bulk_write referenced in function _musb_write
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_bulk_read referenced in function _musb_read
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_set_configuration referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_claim_interface referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_release_interface referenced in function _musb_close
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_set_altinterface referenced in function _musb_set_altinterface
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_reset referenced in function _musb_reset
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_init referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_set_debug referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_find_busses referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_find_devices referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_get_busses referenced in function _musb_open

I have tried redownloading a different version of libusb-1.0, but the problem was not solved.  What might I be doing wrong?

    Reply  Wed Jun 20 10:40:21 2012, Ivan Petrov, triger for measuring time between pulses in channels 

Stefan Ritt wrote:

Martin Petriska wrote:

Stefan Ritt wrote:

Martin Petriska wrote:

 I have two BaF2 detectors with PMT connected to Ch1 and Ch2. At this time Im using external triger module to start DRS4. My evalution board is version 3 so I have no possibility to trigger on two or more pulses occurence on different channels. But I have this idea, trigger with analog trigger on channel 1 (start detector) will start measurement on all channels. After that using FPGA inside EVM to look if some value in Ch2 is bigger as treshold value for example 0,5V and if yes then send data by USB to PC, if signal in Ch2 is lower then restart measurement and wait on triger in Ch1. This way I want to eliminate false data transfer throw USB. Is this possible to implement it into DRS4 evaluation board firmware ?

Thanks.

It is muuuuch easier to upgrade to a V4 board!

Modification of firmware is not so easy. You have to learn and understand VHDL. Then, you have to add additional registers for this thresholds, which requires modification of the C library as well. The data inside the evaluation boards is not yet calibrated (this is only done on the C library), so you have an uncertainty of 30-40mV in this data. 

Ok, except this, I would have a question regarding to the new trigering posibility in V4 board. At this time, I am using Ztec ZT4612 which has some pattern triger posibility. Output from this card is used as an external trigger. Regarding this I have found a problem. Pulses from PMT have about 5-8 ns width. But I need to measure time diferences between pulses in range from 0-50ns. Problem is, that coincidence between pulses is working only on short pulse area (5-8ns) when they are overlapped. Additionaly the result histogram of time diferences is proportional to the pulse shapes. I solve this problem enabling 20MHz LPF filter in ZT4612, so the pulses are wider and overlaped on larger area. But, how it is with the V4 board? Will it trigger if I have for example one 5ns pulse on begiinning of CH1 and second pulse for example 50 ns later on Ch2 with the same probability when pulses are in the same time position?

No. 

Hello. I need to digitize pulses from two PMT. After pulse on first PMT I have to save all pulses from second PMT in range 200 ns (and starting pulse from the first, and time range between pulse from 1st PMT and pulses from 2nd PMT). Is it possible with DRS evaluation board?

    Reply  Wed Jun 20 14:36:01 2012, Ivan Petrov, triger for measuring time between pulses in channels 

Stefan Ritt wrote:

Ivan Petrov wrote:

 

Hello. I need to digitize pulses from two PMT. After pulse on first PMT I have to save all pulses from second PMT in range 200 ns (and starting pulse from the first, and time range between pulse from 1st PMT and pulses from 2nd PMT). Is it possible with DRS evaluation board?

If you run at 2 GSPS, you have a time window of 500 ns like on an oscilloscope. If you trigger on 1st PMT, you will get the traces of all 4 inputs for the next 500 ns. So I guess this is what you want. 

 Ok. And, If I understand correctly, the main bottleneck in data readout is USB. I.e., theoretically maximum readout rate is 500 Hz. Is it true?

Entry  Mon Jul 9 14:14:48 2012, Ivan Petrov, Problem compiling drs_exam.cpp on windows compile_log.txt

Hello again. I have not got evaluation board yet, but already faced some difficulties:) I'm trying to compile drs_exam.cpp on Windows 7 using dev-c++ with imagelib-2 and WxWindows 2.4.2 DevPaks installed, but nothing works. Compile log is attached. Honestly, I'm not very familiar with c++, so any suggestions will be helpful. Thank you.

    Reply  Wed Jul 11 10:04:51 2012, Ivan Petrov, Problem compiling drs_exam.cpp on windows 

Stefan Ritt wrote:

Ivan Petrov wrote:

Hello again. I have not got evaluation board yet, but already faced some difficulties:) I'm trying to compile drs_exam.cpp on Windows 7 using dev-c++ with imagelib-2 and WxWindows 2.4.2 DevPaks installed, but nothing works. Compile log is attached. Honestly, I'm not very familiar with c++, so any suggestions will be helpful. Thank you.

I have no experience with dev-c++, so I cannot be of help here. The supported systems are Linux and Windows with MS Visual C++. But it looks like the problems are related to compiling wxWidgets, which actually you do NOT need for drs_exam.cpp. The wxWidgets library is only needed for the DRSOsc application. If you want to compile it anyhow, first learn how to compile standard WxWidgets applications from

http://www.wxwidgets.org/docs/tutorials/devcpp.htm

 

Best regards,

Stefan 

Ok, this was my bad, I added some unnecessary files to project. Drs_exam compiles well with ms vc++. Thanks!

Entry  Thu Jul 18 01:03:44 2019, Ismael Garcia, Trace Impedance DRS4_Analog_IN.PNG

Hi Steffan,

              I'm an engineer at UCLA developing a board with the DRS4 chip. Our team has a question on what might be the required trace impedence for the analog inputs. Can that information be provided? 

Best Regards,
Ismael Garcia

    Reply  Fri Jul 19 01:37:09 2019, Ismael Garcia, Trace Impedance 

When you're refering to laying a 50 Ohm trace, you're referring to the SMA input and not the interface between the output of the Op-AMP(THS4508) buffer 
and the inputs  of the DRS4(IN0-IN8). Is there a recommended diffential impedance for IN0-IN8? 

Ismael

Stefan Ritt wrote:

The requiremnet is the same as for any high speed analog board, there is othing special with the DRS4. If you want to terminate your line with 50 Ohms and you want a matched impedance layout, you route all lines with 50 Ohms impedance. Truth is however that nothing is perfect. The SMA connector is not exactly 50 Ohm, the PCB gets a 10-20% variation depending on the manufacturer. So even if you try hard, you will never have a 50 Ohm matched impedance. On the evaluation board we made some compromises as you have seen, but for us the board works satisfactory even with this compromises, and you can test it yourself with real hardware (namely the evaluation board). If you can do a better job, try it. But usually these compromises have only little influence on the signal quality.

Stefan

Ismael Garcia wrote:

Hi Steffan,

              I'm an engineer at UCLA developing a board with the DRS4 chip. Our team has a question on what might be the required trace impedence for the analog inputs. Can that information be provided? 

Best Regards,
Ismael Garcia

 

 

Entry  Tue May 19 14:14:45 2015, Ilja Bekman, DRS4 firmware UCF constraints  
Hello, I'm using two DRS4 rev.5 boards for 8ch readout and triggering.

I needed to modify the trigger logic and implement some tweaks in the firmware, and noticed that 
the firmware source in drs-5.0.2 (and 5.0.3, SVN:5339) while still compiling fine with Xilinx ISE 10, stops
doing so in the ISE 14.7 (also already in 13.2)

While the Synthesis is running through (in the new ISE it complains about using more than 100% of resources.)
The Mapping fails due to constraints (firmware/ucf/drs4_eval5.ucf) complaining about an illegal IOSTANDARD
for P_IO_PMC_USR<55> (LVDS_25).

In the newer version the wild-cards (lines 67..69 ) are not properly dealt with, it seems, and if I move the
property by hand to all the wild-carded NETs it start to recognise the LVDS_25.
But after that Place&Route fails with messages about locked Banks due to incompatible VCCO.

I'm trying to adapt the ucf file and am reading about the changes in the ISE software and constraints files, but
want to ask if some of you guys have seen the same issue and resolved it out "officially".
Entry  Wed Oct 7 13:06:34 2015, Ilja Bekman, Voltage Calibration with signal on the input 
 
Entry  Sun May 2 18:36:14 2010, Ignacio Diéguez Estremera, DRS4 chip model 

Hi all,

i'm an electronics engineering student at UCM (Madrid) working on my master's thesis within the CTA collaboration. I'm designing the readout electronics for the telescope's camera, and i'm focusing in using GAPDs instead of PMTs and using the domino chip for the sampling of the signal. I was wondering if there is a spice and/or RF model of the DRS4 chip available. It would be very useful to perform some simulations before deciding to use the chip as the sampling solution for our prototypes.

If the answer is negative, can you give me some advise for modelling the chip in spice? Have you done any simulations?

Thanks in advance,

Ignacio Diéguez Estremera.

    Reply  Mon May 3 17:06:02 2010, Ignacio Diéguez Estremera, DRS4 chip model 

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

Hi all,

i'm an electronics engineering student at UCM (Madrid) working on my master's thesis within the CTA collaboration. I'm designing the readout electronics for the telescope's camera, and i'm focusing in using GAPDs instead of PMTs and using the domino chip for the sampling of the signal. I was wondering if there is a spice and/or RF model of the DRS4 chip available. It would be very useful to perform some simulations before deciding to use the chip as the sampling solution for our prototypes.

If the answer is negative, can you give me some advise for modelling the chip in spice? Have you done any simulations?

Thanks in advance,

Ignacio Diéguez Estremera.

Yes there is a transistor-level spice model, which I used to design the chip, but you won't be happy with it: Given the 500,000+ transistors on the chip, a 100 ns simulation takes a couple of weeks. We tried to make a simplified model just for the analog input using some measured S-parameters, but found that the RF behavior of the chip is almost impossible to describe to better than let's say 50%. In the end you have to fine-tune your analog front-end experimentally, to obtain optimal bandwidth. We are just working on a reference design with gives you 850 MHz bandwidth using an active input buffer.

 Thanks for the information.

I would like to try the huge :-) model. Can you send it to my email address? Since the input signal are pulses of a few nanoseconds at FHWM, the simulation time may be reduced. I will post some feedback in the forum once i give it a try.

Kind regards.

    Reply  Mon May 3 23:21:55 2010, Ignacio Diéguez Estremera, DRS4 chip model 

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

Hi all,

i'm an electronics engineering student at UCM (Madrid) working on my master's thesis within the CTA collaboration. I'm designing the readout electronics for the telescope's camera, and i'm focusing in using GAPDs instead of PMTs and using the domino chip for the sampling of the signal. I was wondering if there is a spice and/or RF model of the DRS4 chip available. It would be very useful to perform some simulations before deciding to use the chip as the sampling solution for our prototypes.

If the answer is negative, can you give me some advise for modelling the chip in spice? Have you done any simulations?

Thanks in advance,

Ignacio Diéguez Estremera.

Yes there is a transistor-level spice model, which I used to design the chip, but you won't be happy with it: Given the 500,000+ transistors on the chip, a 100 ns simulation takes a couple of weeks. We tried to make a simplified model just for the analog input using some measured S-parameters, but found that the RF behavior of the chip is almost impossible to describe to better than let's say 50%. In the end you have to fine-tune your analog front-end experimentally, to obtain optimal bandwidth. We are just working on a reference design with gives you 850 MHz bandwidth using an active input buffer.

 Thanks for the information.

I would like to try the huge :-) model. Can you send it to my email address? Since the input signal are pulses of a few nanoseconds at FHWM, the simulation time may be reduced. I will post some feedback in the forum once i give it a try.

Kind regards.

I just checked and realized that we are not allowed to give out the "huge" model since it contains parameters from the chip manufacturer's library which are confidentially. 

 Thank you for the effort anyway.

So i guess i won't be able to include drs4 in my simulations :-(. Any other suggestions? Maybe the S-params model you where working on? Anything is better than nothing :-)

    Reply  Tue May 4 16:23:16 2010, Ignacio Diéguez Estremera, DRS4 chip model 

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

So i guess i won't be able to include drs4 in my simulations :-(. Any other suggestions? Maybe the S-params model you where working on? Anything is better than nothing :-)

Please find attached the S-parameters. 

 Thanks :-)

Entry  Wed May 5 22:30:50 2010, Ignacio Diéguez Estremera, Random noise spec in datasheet 

Hi,

According to DRS4's datasheet, the random noise is 0.35mVrms. Is this the input equivalent noise voltage? It is computed over the 0-950MHz frequency band?

Regards.

Entry  Fri Nov 3 12:11:14 2017, Håkan Wennlöf, Triggering using AND 

Hi!

I'm using the DRSOsc program, and I have a question that I need a bit clarified;

When triggering using AND between two channels, am I then triggering on rising/falling edge of both channels, or on the actual values?

That is, is it the change in value that it triggers on, or when the actual value goes above a certain threshold?

Using AND, it seems like I get a trigger when one (CH2) is above its trigger value (sometimes quite far above), and the other (CH1) changes to go above its trigger value. This works for my purposes, which is nice, but I just want to be sure I understand why it works. Ideally, I'd like to trigger when one of my channels is above a certain value, and the other has a rising edge above a certain value.

I'm sorry if it's a silly question! I've just noticed that the Keysight oscilloscopes I've used can only do one channel with a rising edge at a time when doing a logic trigger, and I thought the same thing might be going on in the background here (which would be ideal for my purposes).

 

Kind regards,

Håkan Wennlöf

Entry  Tue May 1 02:00:40 2018, Hyunseong Kim, DRS4 using drs_exam.cpp to save as binary files 

Hi, 

I would like to save the waveform in a .dat binary file using drs_exam.cpp.

I know the distributed software allows us to save as binary files with the save button, but I currently need to save multiple runs using a script.

I've seen that drs_exam.cpp can save the waveform as .txt files.

Is there any .cpp file or function that allows us to save the waveforms in binary format (.dat)?

Thank you for your help. 

Entry  Fri Jul 5 12:46:45 2013, Hermann-Josef Mathes, Missing methods in drs-4.0.1.tar.gz 

Hi,

while trying to create python bindings for the DRS stuff using SWIG 2.0.4, two undefined methods prevent the python interpreter from loading the generated shared library. These methods are:

  • int DRSBoard::SetADCActive(unsigned char)
  • bool ResponseCalibration::Calibrate(unsigned int,unsigned int,float *,float *,float,bool)

Can we safely removed those methods from the DRS header files or (what I have actually done) is it better to fake some empty implementation in the input file to SWIG?

Another minor issue is that the python interpreter always terminates with a SegFault after script termination. I had not yet time to track that down...

Thanks & regards

Hermann-Josef

 