DRS4 Forum
  DRS4 Discussion Forum, Page 8 of 46  Not logged in ELOG logo
New entries since:Thu Jan 1 01:00:00 1970
Entry  Mon Apr 22 15:33:28 2013, Benjamin LeGeyt, effect of jitter/alignment between SRCLK and ADC clock 

Hello!
let me apologize in advance if this has already been covered somewhere and I missed it. 


I have a question about a statement made regarding the ADC clock in the evaluation board v4.0 manual.  At the bottom or page 23 there is a mention of jitter between the SRCLK signal and the ADC clock causing a baseline variation in the sampled output of up to a few mV.  Is there any more information out there about this?  I find this confusing for the following reason: If the DRS output has mostly settled after 28ns and the signal that is being sampled is a DC signal, I don't understand why an aperture jitter in the sampling ADC should cause a voltage error in the measured signal.  I already know about the possibility of noise spikes every 32 samples if these clocks are not properly aligned, though I don't know the origin of those spikes.  are these two things related?

 

Many Thanks!
 

    Reply  Mon Apr 22 15:52:53 2013, Stefan Ritt, effect of jitter/alignment between SRCLK and ADC clock adc_phase.jpg

Benjamin LeGeyt wrote:

Hello!
let me apologize in advance if this has already been covered somewhere and I missed it. 


I have a question about a statement made regarding the ADC clock in the evaluation board v4.0 manual.  At the bottom or page 23 there is a mention of jitter between the SRCLK signal and the ADC clock causing a baseline variation in the sampled output of up to a few mV.  Is there any more information out there about this?  I find this confusing for the following reason: If the DRS output has mostly settled after 28ns and the signal that is being sampled is a DC signal, I don't understand why an aperture jitter in the sampling ADC should cause a voltage error in the measured signal.  I already know about the possibility of noise spikes every 32 samples if these clocks are not properly aligned, though I don't know the origin of those spikes.  are these two things related?

Many Thanks!

Hi Benjamin,

In principle you are right, for a DC signal that should not matter. But in reality the DRS4 output signal is not constant even for a DC signal. When you switch from one sampling cell to another during readout, there is something called "charge injection". This causes the output to change up to several 10 mV. After 28 ns this is mostly settled, but not completely, since the DRS4 output driver has a relatively low bandwidth (~50 MHz). Furthermore, the signal line between the DRS4 and the ADC is not terminated, so you have some reflections going forth and back. In addition, you have some crosstalk from the SRCLK signal. So it's better that you sample on each cycle at exactly the same time. Here you see a plot of that (green: DRS4 output, blue: ADC clock, yellow SRCLK):

adc_phase.jpg 

Entry  Tue May 13 19:34:58 2014, Luka Pavelic, drsosc binary to cern ROOT file conversion 

Hi,

Does anybody have program for conversion from binary or xml to cern ROOT *.root file?
 

Thank you for any help you can provide,
Luka Pavelic


 

    Reply  Tue May 13 19:39:36 2014, Stefan Ritt, drsosc binary to cern ROOT file conversion 

Luka Pavelic wrote:

Hi,

Does anybody have program for conversion from binary or xml to cern ROOT *.root file?
 

Thank you for any help you can provide,
Luka Pavelic


 

You look here: elog:262

/Stefan

    Reply  Tue May 13 22:03:47 2014, Luka Pavelic, drsosc binary to cern ROOT file conversion 

Thank you for your fast and very helpful replay.

I made it work with drsosc version 4 but with version 5 i am getting weird results. Is it possible that they changed binary formatting?

 

    Reply  Tue May 13 23:08:50 2014, Stefan Ritt, drsosc binary to cern ROOT file conversion 

Luka Pavelic wrote:

Thank you for your fast and very helpful replay.

I made it work with drsosc version 4 but with version 5 i am getting weird results. Is it possible that they changed binary formatting?

 

Yes, but this is documented in the evaluation board manual. You have to modify the script slightly. I will update it myself in about 2-3 weeks.

Cheers,

Stefan

    Reply  Fri Jun 27 11:23:19 2014, ChengMing Du, drsosc binary to cern ROOT file conversion 

Stefan Ritt wrote:

Luka Pavelic wrote:

Thank you for your fast and very helpful replay.

I made it work with drsosc version 4 but with version 5 i am getting weird results. Is it possible that they changed binary formatting?

 

Yes, but this is documented in the evaluation board manual. You have to modify the script slightly. I will update it myself in about 2-3 weeks.

Cheers,

Stefan

 hi Stefan,can you update the code to convert binary to root for newest drsosc?Thanks.

    Reply  Wed Jul 30 17:05:38 2014, Stefan Ritt, drsosc binary to cern ROOT file conversion 

ChengMing Du wrote:

Stefan Ritt wrote:

Luka Pavelic wrote:

Thank you for your fast and very helpful replay.

I made it work with drsosc version 4 but with version 5 i am getting weird results. Is it possible that they changed binary formatting?

 

Yes, but this is documented in the evaluation board manual. You have to modify the script slightly. I will update it myself in about 2-3 weeks.

Cheers,

Stefan

 hi Stefan,can you update the code to convert binary to root for newest drsosc?Thanks.

See elog:361 

Entry  Tue Mar 28 21:53:12 2017, Jim Freeman, drscl doesn't find eval board but drsosc does (Windows 7) 

I cannot find the EVAL board using drscl version 5.06 while the drsosc works fine. I tried 2 different eval boards and 2 different computers and the same effect. I looked under device manager at the libusb and the drs4 was there, and checked the driver which was found to be up to date.

    Reply  Wed Apr 5 12:28:28 2017, Stefan Ritt, drscl doesn't find eval board but drsosc does (Windows 7) Screen_Shot_2017-04-05_at_12.27.46_.pngScreen_Shot_2017-04-05_at_11.45.07_.png

Two people report now this problem, while this works fine at our lab. So I'm puzzled right now.

I attach two screenshots from the device manager and the Command Line interface. Can you compare it with what you see? Which is the firmware version of your evaluaiton board?

Stefan

Jim Freeman wrote:

I cannot find the EVAL board using drscl version 5.06 while the drsosc works fine. I tried 2 different eval boards and 2 different computers and the same effect. I looked under device manager at the libusb and the drs4 was there, and checked the driver which was found to be up to date.

 

    Reply  Thu Jan 25 06:10:52 2018, chen wenjun, drscl doesn't find eval board but drsosc does (Windows 7) 

Hi! Jim:

  It seems that I meet the same question with you ,and I am confused ,have you find out the reason about this problem?Or can you tell me how you deal with it?

Thank you very much!

chen

Jim Freeman wrote:

I cannot find the EVAL board using drscl version 5.06 while the drsosc works fine. I tried 2 different eval boards and 2 different computers and the same effect. I looked under device manager at the libusb and the drs4 was there, and checked the driver which was found to be up to date.

 

Entry  Wed Mar 6 10:09:01 2019, Willy Chang, drscl "no board found" in some Win7 or Win8.X PCs 

Hi all, 

When connecting the board and running the Zadig program, some Windows PCs may return "driver installation failed." I coudn't find the solution from their download website. So I started the drscl first. Apparently it shows: Successfully scanned, but no boards found. Therefore I checked the Device Manager. A breakdown warning triangle appears under the serial port...

The possible solution may be found here.

Infact, the WinUsb driver has been in existence in your PC. One can just follow the instructions here: 

https://docs.microsoft.com/en-us/windows-hardware/drivers/usbcon/winusb-installation 

  1. Plug in your device to the host system.
  2. Open Device Manager and locate the device.
  3. Right-click the device and select Update driver software... from the context menu.
  4. In the wizard, select Browse my computer for driver software.
  5. Select Let me pick from a list of device drivers on my computer.
  6. From the list of device classes, select Universal Serial Bus devices.
  7. The wizard displays WinUsb Device. Select it to load the driver.

In the wizard, somehow the default setting displays Microsoft Device on the Top of the list and replaced the WinUsb Device. You can easily re-load the WinUsb Device. Just ignore the WARNING from the device manager. The board should work fine now. 

Willy

Entry  Thu Sep 16 19:04:06 2021, Patrick Moriishi Freeman, drs_exam_multi with non-v4 boards, default configuration 

Hello, 

I made a modified version drs_exam_multi.cpp, but ran into an issue when running.  When I ran it, it only found the two boards with lower serial numbers (2781 and 2879) and complained that the others (2880 and 2881) were not v4. Would there be a simple workaround for this type of thing? Also, would I be able to use the .dat format to keep the file sizes down. 

If not, I am curious if there is a way I can at least set a default configuration for the drsosc program. It seems the drsosc.cfg is written when drsosc starts? Does it load the configuration from somewhere else? It would be very helpful to keep the same settings between runs, in particular the trigger delays, levels, trigger mode, and voltage offsets. Maybe I can even do this with just a few of the CLI commands? I know this is for experts only, but I think I would just need a few commands (setTrig, setTrigMode,  setTrigDelay, that sort of thing) if they do exist. I would check the help now, but I'm running, and I'm pretty sure I saw some for trigger settings. 

Anyhow, any help is appreciated in creating a more repeatable and automated data acquisition. Thanks!

 

    Reply  Sat Sep 18 15:48:30 2021, Stefan Ritt, drs_exam_multi with non-v4 boards, default configuration 

Hi,

please note the the evaluation board is what it says, a board to evaluate the chip, and is not meant for a full-blown shiny multi-board DAQ channel, so support for that is kind of limited.

Strange that you only find two out of four boards. What happens if you disconnect the two boards the system finds and then try again? Might be that your USB hub does not have enough power to supply four boards (each taking 2.5W, so you need 10W in total). Unplugging some board will show you if you have a power problem.

The drsosc.cfg stores the current configuration. For this to work, the drsosc program has to have write access to the directory where the drsosc.cfg program is stored, which is usually the directory from where the program is started. Maybe you have to adjust permissions. Yes you have commands to set everything, just look into drs_exam.cpp and you will find most of them.

Best,
Stefan

Patrick Moriishi Freeman wrote:

Hello, 

I made a modified version drs_exam_multi.cpp, but ran into an issue when running.  When I ran it, it only found the two boards with lower serial numbers (2781 and 2879) and complained that the others (2880 and 2881) were not v4. Would there be a simple workaround for this type of thing? Also, would I be able to use the .dat format to keep the file sizes down. 

If not, I am curious if there is a way I can at least set a default configuration for the drsosc program. It seems the drsosc.cfg is written when drsosc starts? Does it load the configuration from somewhere else? It would be very helpful to keep the same settings between runs, in particular the trigger delays, levels, trigger mode, and voltage offsets. Maybe I can even do this with just a few of the CLI commands? I know this is for experts only, but I think I would just need a few commands (setTrig, setTrigMode,  setTrigDelay, that sort of thing) if they do exist. I would check the help now, but I'm running, and I'm pretty sure I saw some for trigger settings. 

Anyhow, any help is appreciated in creating a more repeatable and automated data acquisition. Thanks!

 

 

Entry  Tue Mar 25 16:31:41 2025, Matías Tobar, drs_exam.cpp not compile DRS.hdrs_exam.cpp

Hi! i'm trying to compile drs_exam.cpp but it yields the following error lines:

I do have the "DRS.h" file on the same folder, which is clearly the file that's causing troubles. I also tried to run "DRS.cpp" but it yields a similar error lines. ('sin definir' means 'not defined').

Thanks a lot for the help!

$ g++ drs_exam.cpp  -o drs_exam
/usr/bin/ld: /tmp/ccroK16H.o: en la función `main':
drs_exam.cpp:(.text+0x48): referencia a `DRS::DRS()' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x135): referencia a `DRSBoard::Init()' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x155): referencia a `DRSBoard::SetFrequency(double, bool)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x169): referencia a `DRSBoard::SetTranspMode(int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x184): referencia a `DRSBoard::SetInputRange(double)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x1b6): referencia a `DRSBoard::EnableTrigger(int, int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x1ca): referencia a `DRSBoard::SetTriggerConfig(int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x1fe): referencia a `DRSBoard::EnableTrigger(int, int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x212): referencia a `DRSBoard::SetTriggerConfig(int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x22d): referencia a `DRSBoard::SetTriggerLevel(double)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x241): referencia a `DRSBoard::SetTriggerPolarity(bool)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x255): referencia a `DRSBoard::SetTriggerDelayNs(int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x2b6): referencia a `DRSBoard::StartDomino()' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x2e9): referencia a `DRSBoard::IsBusy()' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x30b): referencia a `DRSBoard::TransferWaves(int, int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x326): referencia a `DRSBoard::GetTriggerCell(unsigned int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x350): referencia a `DRSBoard::GetTime(unsigned int, int, int, float*, bool, bool)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x377): referencia a `DRSBoard::GetWave(unsigned int, unsigned char, float*)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x392): referencia a `DRSBoard::GetTriggerCell(unsigned int)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x3c3): referencia a `DRSBoard::GetTime(unsigned int, int, int, float*, bool, bool)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x3f1): referencia a `DRSBoard::GetWave(unsigned int, unsigned char, float*)' sin definir
/usr/bin/ld: drs_exam.cpp:(.text+0x52e): referencia a `DRS::~DRS()' sin definir
collect2: error: ld returned 1 exit status
 

 

    Reply  Wed Mar 26 08:42:08 2025, Stefan Ritt, drs_exam.cpp not compile 

You have to link against the DRS.cpp library, plus usblib, plus ... Note there is both a Makefile and a CMakeLists.txt for it. Google how to use "make" or "cmake".

Stefan

    Reply  Tue Apr 1 16:24:33 2025, Matías Tobar, drs_exam.cpp not compile 

Thanks! I solved it by running ./drs_exam after execute Makefile (there was an inconsistency in the files directory).

Matías

Stefan Ritt wrote:

You have to link against the DRS.cpp library, plus usblib, plus ... Note there is both a Makefile and a CMakeLists.txt for it. Google how to use "make" or "cmake".

Stefan

 

Entry  Thu Jan 19 23:26:26 2012, Heejong Kim, drs_exam.cpp for evaluation board version 4 

Hello,

I'm using DRS4 evaluation board version4 in Linux (Scientific Linux 5).

Version4 software (drs-4.0.0) was installed without any troubles.

The oscilloscope interfrace program (drsosc) is working fine with version4 software.

But when I tried drs_exam program, it doesn't work as expected.

(500 mV positive (width 50ns)  pulse is connected to Ch#1).

It keeps waiting trigger in the first event.

In the previous version (board/software drs-3.0.0), drs_exam program worked well.

I'm wondering if anybody is using drs_exam with V4 evaluation board.

Any comments/help would be welcomed.

 

Thanks,

Heejong

    Reply  Fri Jan 20 08:09:38 2012, Stefan Ritt, drs_exam.cpp for evaluation board version 4 drs_exam.cpp

Heejong Kim wrote:

Hello,

I'm using DRS4 evaluation board version4 in Linux (Scientific Linux 5).

Version4 software (drs-4.0.0) was installed without any troubles.

The oscilloscope interfrace program (drsosc) is working fine with version4 software.

But when I tried drs_exam program, it doesn't work as expected.

(500 mV positive (width 50ns)  pulse is connected to Ch#1).

It keeps waiting trigger in the first event.

In the previous version (board/software drs-3.0.0), drs_exam program worked well.

I'm wondering if anybody is using drs_exam with V4 evaluation board.

The issue is that the V4 board has new trigger capabilities (such as coincidences between two channels) which require a slightly different configuration. Here it the new code:

 

   /* use following lines to enable hardware trigger on CH1 at 50 mV positive edge */
   if (b->GetBoardType() == 8) {     // Evaluaiton Board V4
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<0);        // set CH1 as source
   } else {                          // Evaluation Board V3
      b->EnableTrigger(0, 1);           // lemo off, analog trigger on
      b->SetTriggerSource(0);           // use CH1 as source
   }

The complete file is attached. Please try again with the new code. Probably next week I will make a new software release (including a Mac version of all programs) which will contain the new code. Sorry for any inconvenience.

Best regards,
Stefan

 

 

 

    Reply  Fri Jan 20 23:50:39 2012, Heejong Kim, drs_exam.cpp for evaluation board version 4 

Stefan Ritt wrote:

Heejong Kim wrote:

Hello,

I'm using DRS4 evaluation board version4 in Linux (Scientific Linux 5).

Version4 software (drs-4.0.0) was installed without any troubles.

The oscilloscope interfrace program (drsosc) is working fine with version4 software.

But when I tried drs_exam program, it doesn't work as expected.

(500 mV positive (width 50ns)  pulse is connected to Ch#1).

It keeps waiting trigger in the first event.

In the previous version (board/software drs-3.0.0), drs_exam program worked well.

I'm wondering if anybody is using drs_exam with V4 evaluation board.

The issue is that the V4 board has new trigger capabilities (such as coincidences between two channels) which require a slightly different configuration. Here it the new code:

 

   /* use following lines to enable hardware trigger on CH1 at 50 mV positive edge */
   if (b->GetBoardType() == 8) {     // Evaluaiton Board V4
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<0);        // set CH1 as source
   } else {                          // Evaluation Board V3
      b->EnableTrigger(0, 1);           // lemo off, analog trigger on
      b->SetTriggerSource(0);           // use CH1 as source
   }

The complete file is attached. Please try again with the new code. Probably next week I will make a new software release (including a Mac version of all programs) which will contain the new code. Sorry for any inconvenience.

Best regards,
Stefan

 

 

 

Hello Stefan,

Thanks for your prompt reply.

drs_exam is working now after modification as above.

By some trials, I found that external trigger is possible by 'b->EnableTrigger(1,0); b->SetTriggerSource(1<<4);'

Best,

Heejong

 

 