    Reply  Wed Dec 14 08:55:29 2011, Stefan Ritt, Synchronization Delay in the Firmware for 8051 Controller 

Hao Huan wrote:

Hi Stefan,

    I have a question regarding the DRS 4 evaluation board firmware for the 8051 controller embedded in the CY7C68013 USB chip: on the board the controller is running at 12 MHz and the FIFO interface of the USB chip is running at 30 MHz, so the number of delay cycles for synchronization as defined in fx2sdly.h should be (3*12000+5*30000-1)/(2*30000)=3, but the actual number used in drs_eval.c is (3*12000+5*48000-1)/(2*48000)=2, so there is a mismatch between the IFCLK frequency used in this calculation and the actual IFCLK frequency configured. Am I misunderstanding something or is there an explanation for that?

 

    Thanks,

Hao Huan

You are right. The SYNCDELAY should contain three wait cycles. I kind of remember that this delay is not very critical. That might be the reason why the system works even with the wrong delay reliably. 

    Reply  Fri Jan 20 08:09:38 2012, Stefan Ritt, drs_exam.cpp for evaluation board version 4 drs_exam.cpp

Heejong Kim wrote:

Hello,

I'm using DRS4 evaluation board version4 in Linux (Scientific Linux 5).

Version4 software (drs-4.0.0) was installed without any troubles.

The oscilloscope interfrace program (drsosc) is working fine with version4 software.

But when I tried drs_exam program, it doesn't work as expected.

(500 mV positive (width 50ns)  pulse is connected to Ch#1).

It keeps waiting trigger in the first event.

In the previous version (board/software drs-3.0.0), drs_exam program worked well.

I'm wondering if anybody is using drs_exam with V4 evaluation board.

The issue is that the V4 board has new trigger capabilities (such as coincidences between two channels) which require a slightly different configuration. Here it the new code:

 

   /* use following lines to enable hardware trigger on CH1 at 50 mV positive edge */
   if (b->GetBoardType() == 8) {     // Evaluaiton Board V4
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<0);        // set CH1 as source
   } else {                          // Evaluation Board V3
      b->EnableTrigger(0, 1);           // lemo off, analog trigger on
      b->SetTriggerSource(0);           // use CH1 as source
   }

The complete file is attached. Please try again with the new code. Probably next week I will make a new software release (including a Mac version of all programs) which will contain the new code. Sorry for any inconvenience.

Best regards,
Stefan

 

 

 

    Reply  Thu Jan 26 09:15:42 2012, Stefan Ritt, DRS4 Rev2.0 for analog pulse counting 

Ravindra Raghunath Shinde wrote:

Hello,

We are using DRS4 Rev.2.0 board.

We want to measure number of pulses generated  by charge particle detector. These negative going analog pulses are very fast having rise time about 2nS.

We want keep threshold level to -20mV. We expected pulse rate to be about 100 to 200 Hz.

I need help to implement this in  current DRS board with  dead time free operation.

 

If you just want to count pulses, you do not need a DRS board. Just use a discriminator and a counter, that's much simpler. The DRS board is not dead time free. 

    Reply  Thu Jan 26 09:49:38 2012, Stefan Ritt, DRS4 Rev2.0 for analog pulse counting 

Ravindra Raghunath Shinde wrote:

Stefan Ritt wrote:

Ravindra Raghunath Shinde wrote:

Hello,

We are using DRS4 Rev.2.0 board.

We want to measure number of pulses generated  by charge particle detector. These negative going analog pulses are very fast having rise time about 2nS.

We want keep threshold level to -20mV. We expected pulse rate to be about 100 to 200 Hz.

I need help to implement this in  current DRS board with  dead time free operation.

 

If you just want to count pulses, you do not need a DRS board. Just use a discriminator and a counter, that's much simpler. The DRS board is not dead time free. 

 Thanks for your prompt reply.

Along with pulse rate  we also want see pulse shape as well as charge measurements. That is why we are exploring this option with our existing DRS Set-up.

 

I understand that. But still the board has some dead time which is dominated by the USB data transfer speed.

There is a way to perform the counting dead time free, but that requires the V4 board, which has a hardware comparator on all four channels (The V2 board has only one comparator and a multiplexer). The output of these comparators go directly to the FPGA, which can then trigger on these signals. In principle one could implement a hardware counter in the FPGA, which works practically dead time free. But this requires a new firmware which has to be written. Either you do it yourself using the Xilinx development tools, or you wait until I find some time to implement this, which could take a couple of weeks or even months.

Best regards,

Stefan 

Entry  Tue Jan 31 08:10:37 2012, Stefan Ritt, IEEE Real Time 2012 Call for Abstracts 

Hello,

I'm co-organizing the upcoming Real Time Conference, which covers also fields of waveform processing and sampling, so it might be interesting for people working with the DRS4 chip. If you have recent results, you could also consider to send an abstract to this conference.  It will be nicely located in Berkeley, California. We plan excursions to San Francisco and to Napa Valley.

Best regards,

Stefan Ritt


18th Real Time Conference
June 11 – 15, 2012
Berkeley, CA

We invite you to the Hotel Shattuck Plaza in downtown Berkeley, California for
the 2012 Real-Time Conference (RT2012).   It will take place Monday, June 11
through Friday, June 15, 2012, with optional pre-conference tutorials Saturday
and Sunday, June 9-10.

Like the previous editions, RT2012 will be a multidisciplinary conference
devoted to the latest developments on realtime techniques in the fields of
plasma and nuclear fusion, particle physics, nuclear physics and astrophysics,
space science, accelerators, medical physics, nuclear power instrumentation and
other radiation instrumentation.

Abstract submission is open as of 18 January (deadline 2 March). Please visit

http://www.npss-confs.org/rtc/welcome.asp?flag=44675.77&Retry=1 to submit an

abstract.

Call for Abstracts 

RT 2012 is an interdisciplinary conference on realtime data acquisition and
computing applications in the physical sciences. These applications include:

* High energy physics 
* Nuclear physics 
* Astrophysics and astroparticle physics 
* Nuclear fusion 
* Medical physics 
* Space instrumentation 
* Nuclear power instrumentation 
* Realtime security and safety 
* General Radiation Instrumentation 

Specific topics include (but are certainly not limited to) the list shown below.
We welcome correspondence to see how your research fits our venue.   

Key Dates

* Abstract submission opened:  January 18, 2012 
* Abstract deadline:  March 2, 2012 
* Program available: April 2 

Suggested Topics

* Realtime system architectures 
* Intelligent signal processing 
* Programmable devices 
* Fast data transfer links and networks 
* Trigger systems 
* Data acquisition 
* Processing farms 
* Control, monitoring, and test systems 
* Upgrades 
* Emerging realtime technologies 
* New standards 
* Realtime safety and security 
* Feedback on experiences 

Contact Information

If you have a question or wish to opt in for occasional e-mail updates about
RT2012, send us a message at RT2012@lbl.gov. To view full conference
information, visit http://rt2012.lbl.gov/index.html

    Reply  Mon Feb 6 08:15:38 2012, Stefan Ritt, what sort of detectors for physical experiment the DRS4 used? 

Zhongwei Du wrote:

Hello.

We are designing a waveform sampling board for Si strip array detector ,whose rise time is less than 10 ns, which makes we doubt whether the DRS4 can do more accurate than traditional charge integral circuit for charge measuring.

So we need to know what sort of detectors for physical experiment the DRS4 has been used in?

Can you give me some information? For example, Si strip array detector or CsI scintillator r ball detector ? PMT or APD ?

DRS4 is used for PMTs and APDs. The minimal rise/fall time which can be recorded with the DRS4 on the evaluation boards is about 0.8ns (see elog:84). Concerning charge measurement, it depends on your integration time. If your signal is for example 20ns and you sample with 5 GSPS, you get actually 100 samples. You digitize with 12 bits, but the S/N ratio is more like 11.5 bits. But since you have 100 samples, the accuracy of the measurement scales with sqrt(100)=10. So you have more like 11.5+3 bits = 14-15 bits, or a SNR of 20'000. Now your signal usually does not have an amplitude of 1V, will be more like 100mV or so, in which case your SNR goes to 2000. But this still gives you a resolution of 1/2000 = 0.5 per mille.

We did tests with a Ge detector for spectroscopy applications, where we used a shaping time of 2 micro seconds, both with traditional electronics and the DRS4, integrating the signal over 2 micro seconds. Both gave a resolution of about 0.4%, indicating that the resolution was not limited by the DRS but by the detector.

Best regards,

Stefan 

    Reply  Fri Feb 24 15:52:43 2012, Stefan Ritt, DRS4- analog pulse counting 

sonal wrote:

Hello Sir,

Regarding to analog pulse counting by using DRS4 Rev.2.0 board, you have said that "There is a way to perform the counting dead time free, but that requires the V4 board, which has a hardware comparator on all four channels (The V2 board has only one comparator and a multiplexer). The output of these comparators go directly to the FPGA, which can then trigger on these signals. In principle one could implement a hardware counter in the FPGA, which works practically dead time free. But this requires a new firmware which has to be written. Either you do it yourself using the Xilinx development tools, or you wait until I find some time to implement this, which could take a couple of weeks or even months."

I am interested in it. I have DRS4 Rev.2.0 board. I have FPGA and Microcontroller firmware for this board. How can I implement this concept of counting pulses?

 


As I said above, you need a V4 board which has the four hardware comparators mounted. You cannot do that with the V2 board, even with a firmware upgrade.

The firmware upgrade for the V4 board is not yet ready and I won't have time for that in the next 3-4 weeks, but afterwards maybe.

Best regards,

Stefan 

    Reply  Thu Mar 1 19:22:26 2012, Stefan Ritt, DRS4- analog pulse counting 

Sonal wrote:

The DRS4 Rev.2.0 with one Multiplexer and a Comparator I think may be is ok for me. Since I wanted to implement counter in FPGA, I have gone through firmware but its very difficult for me to get the flow of data from DRS to USB. Could you help me to get familiar with or to get flow of the data from DRS to the USB?

Hi Sonai,

unfortunately I cannot teach you VHDL, you should take a course somewhere. Actually you don't have to touch the USB part, you just have to implement a new register in the firmware, and connect it to a VHDL counter attached to the trigger input. I will do this in a couple of weeks myself, so you can also just wait. I will however only do this for the V4 board, so you have to port it back to the V2 firmware.

Best regards,

Stefan

    Reply  Tue Mar 20 16:33:50 2012, Stefan Ritt, triger for measuring time between pulses in channels 

Martin Petriska wrote:

 I have two BaF2 detectors with PMT connected to Ch1 and Ch2. At this time Im using external triger module to start DRS4. My evalution board is version 3 so I have no possibility to trigger on two or more pulses occurence on different channels. But I have this idea, trigger with analog trigger on channel 1 (start detector) will start measurement on all channels. After that using FPGA inside EVM to look if some value in Ch2 is bigger as treshold value for example 0,5V and if yes then send data by USB to PC, if signal in Ch2 is lower then restart measurement and wait on triger in Ch1. This way I want to eliminate false data transfer throw USB. Is this possible to implement it into DRS4 evaluation board firmware ?

Thanks.

It is muuuuch easier to upgrade to a V4 board!

Modification of firmware is not so easy. You have to learn and understand VHDL. Then, you have to add additional registers for this thresholds, which requires modification of the C library as well. The data inside the evaluation boards is not yet calibrated (this is only done on the C library), so you have an uncertainty of 30-40mV in this data. 

    Reply  Wed Mar 21 09:39:33 2012, Stefan Ritt, triger for measuring time between pulses in channels 

Martin Petriska wrote:

Stefan Ritt wrote:

Martin Petriska wrote:

 I have two BaF2 detectors with PMT connected to Ch1 and Ch2. At this time Im using external triger module to start DRS4. My evalution board is version 3 so I have no possibility to trigger on two or more pulses occurence on different channels. But I have this idea, trigger with analog trigger on channel 1 (start detector) will start measurement on all channels. After that using FPGA inside EVM to look if some value in Ch2 is bigger as treshold value for example 0,5V and if yes then send data by USB to PC, if signal in Ch2 is lower then restart measurement and wait on triger in Ch1. This way I want to eliminate false data transfer throw USB. Is this possible to implement it into DRS4 evaluation board firmware ?

Thanks.

It is muuuuch easier to upgrade to a V4 board!

Modification of firmware is not so easy. You have to learn and understand VHDL. Then, you have to add additional registers for this thresholds, which requires modification of the C library as well. The data inside the evaluation boards is not yet calibrated (this is only done on the C library), so you have an uncertainty of 30-40mV in this data. 

Ok, except this, I would have a question regarding to the new trigering posibility in V4 board. At this time, I am using Ztec ZT4612 which has some pattern triger posibility. Output from this card is used as an external trigger. Regarding this I have found a problem. Pulses from PMT have about 5-8 ns width. But I need to measure time diferences between pulses in range from 0-50ns. Problem is, that coincidence between pulses is working only on short pulse area (5-8ns) when they are overlapped. Additionaly the result histogram of time diferences is proportional to the pulse shapes. I solve this problem enabling 20MHz LPF filter in ZT4612, so the pulses are wider and overlaped on larger area. But, how it is with the V4 board? Will it trigger if I have for example one 5ns pulse on begiinning of CH1 and second pulse for example 50 ns later on Ch2 with the same probability when pulses are in the same time position?

No. 

    Reply  Wed Apr 25 13:42:37 2012, Stefan Ritt, DRS4 Initialization 

Guillaume Blanchard wrote:

Hello,

I am writing a VHDL code to drive a DRS4 chip.

In order to configure the DRS4 chip, I have to set the "Config Register" and the "Write Shift Register" then ... (I do not plan to use simultaneous WR and R so I guess the Write Config Reg. is not needed)

My question is :

When do we have to perform a "Read Shift Register Initialization" ?

Every time before a full read-out, or juste once after a DRS4 reset ?

Further more, is this initialization needed for the ROI mode ?

And at last do the level of the DENABLE and DWRITE signals matter for the "Read Shift Register Initialization" ?

(To sum up : what is the purpose of the Read Shift Register and how does it work ?)

Cordially,

G.Blanchard.

There are two readout modes "Full Readout Mode" and  "ROI mode". 

In the Full Readout Mode, the Read Shift Register has to be initialized before the first readout by applying the sequence shown in Figure 11 in the data sheet. This clears the full shift register and sets the first cell to "1". In principle in the following events one applies each time 1024 clocks. Since the shift register is circula, the single "1" rotates through the shift register and is at the same position after 1024 clocks. So in principle the register does not have to be re-initialized. To be hones I have never tried this myself, so I'm not completely sure if that works.

In the ROI mode, you initialize the Read Shift Register by a single RSRLOAD pulse as shown in Figure 15. Since the inverter chain stops at different positions in each event, this pulse has to be applied before each event. The SROUT bits will then tell you where the inverter chain has been stopped.

Most people I know of use the ROI mode, since the initialization is much simpler (just a single pulse).

Best regards,

Stefan

    Reply  Wed Jun 20 12:45:05 2012, Stefan Ritt, triger for measuring time between pulses in channels 

Ivan Petrov wrote:

 

Hello. I need to digitize pulses from two PMT. After pulse on first PMT I have to save all pulses from second PMT in range 200 ns (and starting pulse from the first, and time range between pulse from 1st PMT and pulses from 2nd PMT). Is it possible with DRS evaluation board?

If you run at 2 GSPS, you have a time window of 500 ns like on an oscilloscope. If you trigger on 1st PMT, you will get the traces of all 4 inputs for the next 500 ns. So I guess this is what you want. 

    Reply  Wed Jun 20 14:44:38 2012, Stefan Ritt, triger for measuring time between pulses in channels 

Ivan Petrov wrote:

Stefan Ritt wrote:

Ivan Petrov wrote:

 

Hello. I need to digitize pulses from two PMT. After pulse on first PMT I have to save all pulses from second PMT in range 200 ns (and starting pulse from the first, and time range between pulse from 1st PMT and pulses from 2nd PMT). Is it possible with DRS evaluation board?

If you run at 2 GSPS, you have a time window of 500 ns like on an oscilloscope. If you trigger on 1st PMT, you will get the traces of all 4 inputs for the next 500 ns. So I guess this is what you want. 

 Ok. And, If I understand correctly, the main bottleneck in data readout is USB. I.e., theoretically maximum readout rate is 500 Hz. Is it true?

On the evaluation board, yes. This board is not optimized for high readout rate. If you do your own electronics, like GBit Ethernet, you could be much faster. 

    Reply  Mon Jun 25 14:21:13 2012, Stefan Ritt, triger for measuring time between pulses in channels 

Andrey Kuznetsov wrote:

Stefan Ritt wrote:

On the evaluation board, yes. This board is not optimized for high readout rate. If you do your own electronics, like GBit Ethernet, you could be much faster. 

 What is the readout rate via GBit Ethernet that you have achieved?

Where is the bottleneck in ethernet?

What is the proposed scheme by which the GBit Ethernet will be implemented, will the DRS4 Eval Board have to wait for the computer to respond before sending the data (wouldn't this make the readout much slower?), or will the DRS4 Eval Board keep sending the data to the computer?

With GBit Ethernet you get close to 100 MB/sec, which is the maximal line speed. The protocol to be implemented will achieve that rate. What one usually does is to send events in large blocks upon request from the PC. The trick is to do the request in a clever way, like using a high water mark on the receiving event buffer. So as long as the PC can digest the data quickly enough, the board just keeps sending, which means no overhead.

 

    Reply  Tue Jul 10 13:15:00 2012, Stefan Ritt, Problem compiling drs_exam.cpp on windows 

Ivan Petrov wrote:

Hello again. I have not got evaluation board yet, but already faced some difficulties:) I'm trying to compile drs_exam.cpp on Windows 7 using dev-c++ with imagelib-2 and WxWindows 2.4.2 DevPaks installed, but nothing works. Compile log is attached. Honestly, I'm not very familiar with c++, so any suggestions will be helpful. Thank you.

I have no experience with dev-c++, so I cannot be of help here. The supported systems are Linux and Windows with MS Visual C++. But it looks like the problems are related to compiling wxWidgets, which actually you do NOT need for drs_exam.cpp. The wxWidgets library is only needed for the DRSOsc application. If you want to compile it anyhow, first learn how to compile standard WxWidgets applications from

http://www.wxwidgets.org/docs/tutorials/devcpp.htm

 

Best regards,

Stefan 

    Reply  Mon Aug 6 02:44:00 2012, Stefan Ritt, Calculation of loop filter parameters (R,C1and C1) for 1 GHz 

Mayank S. Rajguru wrote:

 Hi,

we are planning to use the DRS4 in our board for 1 GHz sampling and digitization.

I have seen in the data sheet that "For the PLL to work, an external loop filter is required. This filter ensures quick locking and stable operation at the desired sampling frequency".

What formula do you use to calculate the values of R, C1 and C2?

Can we use the same given value for different frequencies?

Thanks,

Mayak

I never worked out an exact formula for the parameters. It also depends if you want a fast locking time, or a low phase jitter of the PLL. A good starting point are the values from the evaluation board:

R = 130 Ohm

C1 = 4.7 nF

C2 = 1 uF

They should work from 800 MHz to 5 GHz.

- Stefan 

    Reply  Wed Aug 29 10:52:44 2012, Stefan Ritt, DRS-4.0.0 DOScreen.cpp 

Zach Miller wrote:

Hi,

I found an old thread regarding a fix for DOScreen.cpp for DRS-3.1.0, that fixes an "ambiguous overload problem." Currently when I attempt to build the drs-4.0.0, I get this similar error:

src/DOScreen.cpp:332:39: error: call of overloaded ‘Append(int)’ is ambiguous
 

This section of code is different than what the previous thread was correcting, and though I attempted to apply the logic of the old thread, I haven't fixed this yet.

The following is the code for that section:

-----

329  for (int i=0 ; i<5 ; i++)
330       if (tc & (1<<(i+8))) {
331            if (i < 4)
332              wxst1.Append(wxT('1'+i));
333            else
334              wxst1.Append(wxT('E'));
335           wxst1.Append(wxT('&'));
336        }
337      if (wxst1.Length() > 0)
338         wxst1 = wxst1.Left(wxst1.Length()-1);
339      wxst1.Append(wxT(')'));

------

I've attempted a few fixes, but unfortunately, my understanding of the code is not great, and I haven't managed to fix this yet. Any help would be appreciated.

Thanks,

Zach Miller

Just put (char) in front of wxT(...), like

   if (tc > 0) {

      wxString wxst1, wxst2;

      wxst1.Append((char)wxT('('));

      for (int i=0 ; i<5 ; i++)

         if (tc & (1<<i)) {

            if (i < 4)

               wxst1.Append((char) (wxT('1'+i)));

            else

               wxst1.Append((char) wxT('E'));

            wxst1.Append((char) wxT('|'));

         }

      for (int i=0 ; i<5 ; i++)

         if (tc & (1<<(i+8))) {

            if (i < 4)

               wxst1.Append((char) (wxT('1'+i)));

            else

               wxst1.Append((char) wxT('E'));

            wxst1.Append((char) wxT('&'));

         }

      if (wxst1.Length() > 0)

         wxst1 = wxst1.Left(wxst1.Length()-1);

      wxst1.Append((char) wxT(')')); 

    Reply  Wed Aug 29 16:45:36 2012, Stefan Ritt, DRS-4.0.0 DOScreen.cpp 

Zach Miller wrote:

Stefan Ritt wrote:

Zach Miller wrote:

Hi,

I found an old thread regarding a fix for DOScreen.cpp for DRS-3.1.0, that fixes an "ambiguous overload problem." Currently when I attempt to build the drs-4.0.0, I get this similar error:

src/DOScreen.cpp:332:39: error: call of overloaded ‘Append(int)’ is ambiguous
 

This section of code is different than what the previous thread was correcting, and though I attempted to apply the logic of the old thread, I haven't fixed this yet.

The following is the code for that section:

-----

329  for (int i=0 ; i<5 ; i++)
330       if (tc & (1<<(i+8))) {
331            if (i < 4)
332              wxst1.Append(wxT('1'+i));
333            else
334              wxst1.Append(wxT('E'));
335           wxst1.Append(wxT('&'));
336        }
337      if (wxst1.Length() > 0)
338         wxst1 = wxst1.Left(wxst1.Length()-1);
339      wxst1.Append(wxT(')'));

------

I've attempted a few fixes, but unfortunately, my understanding of the code is not great, and I haven't managed to fix this yet. Any help would be appreciated.

Thanks,

Zach Miller

Just put (char) in front of wxT(...), like

   if (tc > 0) {

      wxString wxst1, wxst2;

      wxst1.Append((char)wxT('('));

      for (int i=0 ; i<5 ; i++)

         if (tc & (1<<i)) {

            if (i < 4)

               wxst1.Append((char) (wxT('1'+i)));

            else

               wxst1.Append((char) wxT('E'));

            wxst1.Append((char) wxT('|'));

         }

      for (int i=0 ; i<5 ; i++)

         if (tc & (1<<(i+8))) {

            if (i < 4)

               wxst1.Append((char) (wxT('1'+i)));

            else

               wxst1.Append((char) wxT('E'));

            wxst1.Append((char) wxT('&'));

         }

      if (wxst1.Length() > 0)

         wxst1 = wxst1.Left(wxst1.Length()-1);

      wxst1.Append((char) wxT(')')); 

 Hi Stefan,

Thanks for the response. I have tried that and it fixes all the lines except the ones that have:

wxst1.Append((char) (wxT('1'+i)));

Those lines still don't compile for me. I've also tried: wxst1.Append((char) (wxT('1'+(char)i))); and wxst1.Append((char) (wxT((char)'1'+i))); as well as a few other combinations of (char) and Append.

I get the same error of: "Call of overloaded "Append(int)" is ambiguous."

Any other help would be greatly appreciated. Thanks!

-Zach

wxst1.Append((char) (wxT((char)('1'+i))));     maybe ??? 

    Reply  Thu Oct 4 20:59:18 2012, Stefan Ritt, DRS5 

Zach Miller wrote:

Hi,

Our group had previously heard that a "DRS-5.0" might be on the horizon and that it may have ethernet capabilities as well as 16-input channels (we heard this when ordering the DRS-4). Is this still in the works and accurate information? If so, is there a rough estimate to the "release date?"

Thanks for your time,

Zach Miller

You mix up two things: The DRS5 chip is a new device with improved samling speed (10 GSPS) and lower dead time. This chip might come in 2-3 years. The 16-input board you mentioned is a DAQ board based on the DRS4 chip. This board well be operational beginning of 2013 as a prototype. It is not clear however at this point in which way this board will be made available for public. Maybe we will license this to industry. The design is however pretty much defined: 16 channels with gain 0.1-100, 1 GHz Bandwidth, Gigabit Ethernet output, and multi-board capabilities. Trigger on each channel with logical combinations. 80 MSPS continuous sampling (in addition to the DRS4 sampling). Each channel can be biased 0-210 V for SiPMT or APD power. A 19" 3 HE crate will host 16 boards with 256 channels.

/Stefan

    Reply  Fri Oct 12 14:09:37 2012, Stefan Ritt, DRS abbreviation 

Moritz von Witzleben wrote:

Hello,

what is the abbreviation of DRS?

Thanks and kind Regards,

Moritz

Domino Ring Sampler. 