DRS4 Forum
  DRS4 Discussion Forum, Page 14 of 45  Not logged in ELOG logo
New entries since:Thu Jan 1 01:00:00 1970
ID Date Author Subject
  651   Wed Jan 17 09:51:16 2018 Tran Cong ThienThe input signals recorded are different with the signal showed in oscilloscope

Dear Stefan,

I am using an DRS4 board to record the signals from an plastic scintillator detector. It was working really good, yet a few day ago the signals became "not right". When I checked the signal using an oscilloscope it show the normal signals previously recorded. The signal amplitude are clearly reduced (from 0.3 in oscilloscope to lower than 0.1 in DRS4). Can you show me how to show this problem?

Thank you very much!

Best Regards,

Thien 

  650   Wed Dec 20 22:14:35 2017 Stefan Rittcascading -- DRS4 Osci.cpp & DRS.cpp

https://bitbucket.org/ritt/drs4eb

 

  649   Wed Dec 20 16:30:45 2017 Yoni Shercascading -- DRS4 Osci.cpp & DRS.cpp

Hi, 

The board is modified (and checks out with the DRSScope program). Could you please point me to the drs_exam_2048.cpp file? I can't seem to fine the most up-to-date git repository....

 

Thanks, 

Yoni

Stefan Ritt wrote:

First you need a board which is modified in hardware to support channel cascading. Basically there are internal resistors which connect each input connector to two channels. You have to specify this when you order the board. Then you can use the new drs_exam_2048.cpp file contains in the git repository which correctly configures and reads out the board in two-channel cascading mode. Putting all 8 channels together is not supported by the evaluation boards.

Stefan

Yoni Sher wrote:

Hi, 

I'm trying to do the same thing (get 1 channel with 8192 bins), but I'm having some trouble with it. When I call SetChannelConfig(0, 8, 1) as suggeted, I get output that looks like noise on all readouts. Could you please explain what is supposed to happen in this case? 

I will happily write the code to combine the channels correctly (and debug it) if I can understand what needs to be done. 

(I should mention that my primary concern is a MATLAB interface which I have already written and don't mind sharing when it's complete). 

 

Yoni Sher

Stefan Ritt wrote:

 

Jill Russek wrote:

 

Stefan Ritt wrote:

 

Jill Russek wrote:

 Stefan, thanks for your help so far. If I go with your plan A of just modifying drs_exam.cpp, is there a quick way to get it to save the data from the wave, like how osci.cpp spits out an xml file? (Ignoring the cascading aspect for now)

Thanks again :)

/Jill

Well, you have to learn C programming, I won't do it for you. drs_exam.cpp contains already code to write to the ASCII file data.txt, so you just can use that or modify it to your needs.

/Stefan

 Ha! So then the answer is no, there isn't a ready made function/method to pull out the timing and voltage,  like how it was done in osci.cpp. That's all I wanted to know. (Not whether you would write it for me! Only trying to save time!) Thanks!

/Jill

You misunderstood. The answer is yes. drs_exam.cpp contains already code to write to an ASCII file. If you actually look into the file, you see:

   f = fopen("data.txt", "w");
   ...
   b->GetTime(0, b->GetTriggerCell(0), time_array);
   ...
   b->GetWave(0, 0, wave_array[0]);
   ...
   fprintf(f, "%5.2f %6.2f\n", time_array[i], wave_array[0][i]);

which actually pulls out the timing and voltage and writes it to the file.

 

 

 

  648   Wed Dec 20 16:21:42 2017 Stefan Rittcascading -- DRS4 Osci.cpp & DRS.cpp

First you need a board which is modified in hardware to support channel cascading. Basically there are internal resistors which connect each input connector to two channels. You have to specify this when you order the board. Then you can use the new drs_exam_2048.cpp file contains in the git repository which correctly configures and reads out the board in two-channel cascading mode. Putting all 8 channels together is not supported by the evaluation boards.

Stefan

Yoni Sher wrote:

Hi, 

I'm trying to do the same thing (get 1 channel with 8192 bins), but I'm having some trouble with it. When I call SetChannelConfig(0, 8, 1) as suggeted, I get output that looks like noise on all readouts. Could you please explain what is supposed to happen in this case? 

I will happily write the code to combine the channels correctly (and debug it) if I can understand what needs to be done. 

(I should mention that my primary concern is a MATLAB interface which I have already written and don't mind sharing when it's complete). 

 

Yoni Sher

Stefan Ritt wrote:

 

Jill Russek wrote:

 

Stefan Ritt wrote:

 

Jill Russek wrote:

 Stefan, thanks for your help so far. If I go with your plan A of just modifying drs_exam.cpp, is there a quick way to get it to save the data from the wave, like how osci.cpp spits out an xml file? (Ignoring the cascading aspect for now)

Thanks again :)

/Jill

Well, you have to learn C programming, I won't do it for you. drs_exam.cpp contains already code to write to the ASCII file data.txt, so you just can use that or modify it to your needs.

/Stefan

 Ha! So then the answer is no, there isn't a ready made function/method to pull out the timing and voltage,  like how it was done in osci.cpp. That's all I wanted to know. (Not whether you would write it for me! Only trying to save time!) Thanks!

/Jill

You misunderstood. The answer is yes. drs_exam.cpp contains already code to write to an ASCII file. If you actually look into the file, you see:

   f = fopen("data.txt", "w");
   ...
   b->GetTime(0, b->GetTriggerCell(0), time_array);
   ...
   b->GetWave(0, 0, wave_array[0]);
   ...
   fprintf(f, "%5.2f %6.2f\n", time_array[i], wave_array[0][i]);

which actually pulls out the timing and voltage and writes it to the file.

 

 

  647   Wed Dec 20 15:30:38 2017 Yoni Shercascading -- DRS4 Osci.cpp & DRS.cpp

Hi, 

I'm trying to do the same thing (get 1 channel with 8192 bins), but I'm having some trouble with it. When I call SetChannelConfig(0, 8, 1) as suggeted, I get output that looks like noise on all readouts. Could you please explain what is supposed to happen in this case? 

I will happily write the code to combine the channels correctly (and debug it) if I can understand what needs to be done. 

(I should mention that my primary concern is a MATLAB interface which I have already written and don't mind sharing when it's complete). 

 

Yoni Sher

Stefan Ritt wrote:

 

Jill Russek wrote:

 

Stefan Ritt wrote:

 

Jill Russek wrote:

 Stefan, thanks for your help so far. If I go with your plan A of just modifying drs_exam.cpp, is there a quick way to get it to save the data from the wave, like how osci.cpp spits out an xml file? (Ignoring the cascading aspect for now)

Thanks again :)

/Jill

Well, you have to learn C programming, I won't do it for you. drs_exam.cpp contains already code to write to the ASCII file data.txt, so you just can use that or modify it to your needs.

/Stefan

 Ha! So then the answer is no, there isn't a ready made function/method to pull out the timing and voltage,  like how it was done in osci.cpp. That's all I wanted to know. (Not whether you would write it for me! Only trying to save time!) Thanks!

/Jill

You misunderstood. The answer is yes. drs_exam.cpp contains already code to write to an ASCII file. If you actually look into the file, you see:

   f = fopen("data.txt", "w");
   ...
   b->GetTime(0, b->GetTriggerCell(0), time_array);
   ...
   b->GetWave(0, 0, wave_array[0]);
   ...
   fprintf(f, "%5.2f %6.2f\n", time_array[i], wave_array[0][i]);

which actually pulls out the timing and voltage and writes it to the file.

 

  646   Tue Dec 12 13:58:06 2017 Stefan RittExternal trigger using Raspberry Pi

Indeed the code does not work for the current evaluation board, it has been written for a previous version and never been updated. Please use following code to enable the external trigger

   /* use following lines to enable the external trigger */
   if (b->GetBoardType() >= 8) {        // Evaluaiton Board V4&5
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerConfig(1<<4);        // set external trigger as source
   } else {                             // Evaluation Board V3
      b->EnableTrigger(1, 0);           // lemo on, analog trigger offf
   }

Please also make sure that the signal on the external trigger input is strong enough. You need at least 2.5V at 50 Ohms, and not every driver is capable of driving 50 Ohms.

Stefan

Diego Yankelevich wrote:

Dear Steffan:

We have been able to use the DRS4 using a Raspberry Pi but we have not been able to use the external trigger. What we are doing is basically comment out the code shown below (downloaded from PSI) to use the hardware trigger and uncomment the code to use the external trigger. We have not been able to get external trigger to work. Could you see what could be wrong?

Thanks

Diego

/* use following line to turn on the internal 100 MHz clock connected to all channels  */
   //b->EnableTcal(1);

   /* use following lines to enable hardware trigger on CH1 at 50 mV positive edge */

   /*
   if (b->GetBoardType() >= 8) {        // Evaluaiton Board V4&5
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<0);        // set CH1 as source
   } else if (b->GetBoardType() == 7) { // Evaluation Board V3
      b->EnableTrigger(0, 1);           // lemo off, analog trigger on
      b->SetTriggerSource(0);           // use CH1 as source
   }
   b->SetTriggerLevel(0.05);            // 0.05 V
   b->SetTriggerPolarity(false);        // positive edge
   */

   /* use following lines to set individual trigger elvels */
   //b->SetIndividualTriggerLevel(1, 0.1);
   //b->SetIndividualTriggerLevel(2, 0.2);
   //b->SetIndividualTriggerLevel(3, 0.3);
   //b->SetIndividualTriggerLevel(4, 0.4);
   //b->SetTriggerSource(15);

   b->SetTriggerDelayNs(0);             // zero ns trigger delay

   /* use following lines to enable the external trigger */
   if (b->GetBoardType() == 8) {     // Evaluaiton Board V4
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<4);        // set external trigger as source
   } else {                          // Evaluation Board V3
      b->EnableTrigger(1, 0);           // lemo on, analog trigger off
    }

 

  645   Tue Dec 12 00:25:50 2017 Diego YankelevichExternal trigger using Raspberry Pi

Dear Steffan:

We have been able to use the DRS4 using a Raspberry Pi but we have not been able to use the external trigger. What we are doing is basically comment out the code shown below (downloaded from PSI) to use the hardware trigger and uncomment the code to use the external trigger. We have not been able to get external trigger to work. Could you see what could be wrong?

Thanks

Diego

/* use following line to turn on the internal 100 MHz clock connected to all channels  */
   //b->EnableTcal(1);

   /* use following lines to enable hardware trigger on CH1 at 50 mV positive edge */

   /*
   if (b->GetBoardType() >= 8) {        // Evaluaiton Board V4&5
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<0);        // set CH1 as source
   } else if (b->GetBoardType() == 7) { // Evaluation Board V3
      b->EnableTrigger(0, 1);           // lemo off, analog trigger on
      b->SetTriggerSource(0);           // use CH1 as source
   }
   b->SetTriggerLevel(0.05);            // 0.05 V
   b->SetTriggerPolarity(false);        // positive edge
   */

   /* use following lines to set individual trigger elvels */
   //b->SetIndividualTriggerLevel(1, 0.1);
   //b->SetIndividualTriggerLevel(2, 0.2);
   //b->SetIndividualTriggerLevel(3, 0.3);
   //b->SetIndividualTriggerLevel(4, 0.4);
   //b->SetTriggerSource(15);

   b->SetTriggerDelayNs(0);             // zero ns trigger delay

   /* use following lines to enable the external trigger */
   if (b->GetBoardType() == 8) {     // Evaluaiton Board V4
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<4);        // set external trigger as source
   } else {                          // Evaluation Board V3
      b->EnableTrigger(1, 0);           // lemo on, analog trigger off
    }
  644   Wed Nov 22 14:52:31 2017 Stefan RittAveraging capabilities

This feature is not yet implemented. The (disabled) software swtich is more like a kind of a reminder to myself to work on that one day...

Diego Yankelevich wrote:

The Display window in the Oscilloscope software shows averaging capabilites but I have not been able to activate these. Is it possible to activate averaging with the existing oscilloscope software? Thanks

 

  643   Wed Nov 22 09:19:11 2017 chen wenjun using of the DRS Command Line Interface

Thank you very much !! All my fault for I thought it too comlicated. Thank you sincerely!

Stefan Ritt wrote:

Remove the check mark from the "Lock" box and enter a different value in the sampling speed box and hit return.

chen wenjun wrote:

OK!Thank you! One more question,when I use the Oscillocope ,I found that the actual speed is a constant value of 1.007G,how can change this speed.

Stefan Ritt wrote:

The command line interface is more a debugging tool for experts, and you are not supposed to use it except to test the connection to the evaluation board. The programs for the user are the DRS Oscilloscope and the drs_exam.cpp example program to read out the board with your own program.

Stefan

chen wenjun wrote:

Hello! I'm using DRS4 evaluation board V5 with the drs command line interface,but the mannal only explained the meaning of the command--"info".And I can't get the hang of the use of other commands through "help",so is there anywhere can I learn more about other commands?Or I can only learn it through the datasheet of DRS4 chip.

Thanks!

 

 

 

 

 

 

  642   Wed Nov 22 09:14:18 2017 Stefan Rittusing of the DRS Command Line Interface

Remove the check mark from the "Lock" box and enter a different value in the sampling speed box and hit return.

chen wenjun wrote:

OK!Thank you! One more question,when I use the Oscillocope ,I found that the actual speed is a constant value of 1.007G,how can change this speed.

Stefan Ritt wrote:

The command line interface is more a debugging tool for experts, and you are not supposed to use it except to test the connection to the evaluation board. The programs for the user are the DRS Oscilloscope and the drs_exam.cpp example program to read out the board with your own program.

Stefan

chen wenjun wrote:

Hello! I'm using DRS4 evaluation board V5 with the drs command line interface,but the mannal only explained the meaning of the command--"info".And I can't get the hang of the use of other commands through "help",so is there anywhere can I learn more about other commands?Or I can only learn it through the datasheet of DRS4 chip.

Thanks!

 

 

 

 

 

  641   Wed Nov 22 08:58:33 2017 chen wenjun using of the DRS Command Line Interface

OK!Thank you! One more question,when I use the Oscillocope ,I found that the actual speed is a constant value of 1.007G,how can change this speed.

Stefan Ritt wrote:

The command line interface is more a debugging tool for experts, and you are not supposed to use it except to test the connection to the evaluation board. The programs for the user are the DRS Oscilloscope and the drs_exam.cpp example program to read out the board with your own program.

Stefan

chen wenjun wrote:

Hello! I'm using DRS4 evaluation board V5 with the drs command line interface,but the mannal only explained the meaning of the command--"info".And I can't get the hang of the use of other commands through "help",so is there anywhere can I learn more about other commands?Or I can only learn it through the datasheet of DRS4 chip.

Thanks!

 

 

 

 

Attachment 1: ΢ÐÅͼƬ_20171122160245.png
΢ÐÅͼƬ_20171122160245.png
  640   Wed Nov 22 08:48:36 2017 Stefan Rittusing of the DRS Command Line Interface

The command line interface is more a debugging tool for experts, and you are not supposed to use it except to test the connection to the evaluation board. The programs for the user are the DRS Oscilloscope and the drs_exam.cpp example program to read out the board with your own program.

Stefan

chen wenjun wrote:

Hello! I'm using DRS4 evaluation board V5 with the drs command line interface,but the mannal only explained the meaning of the command--"info".And I can't get the hang of the use of other commands through "help",so is there anywhere can I learn more about other commands?Or I can only learn it through the datasheet of DRS4 chip.

Thanks!

 

 

 

  639   Wed Nov 22 08:31:03 2017 chen wenjun using of the DRS Command Line Interface

Hello! I'm using DRS4 evaluation board V5 with the drs command line interface,but the mannal only explained the meaning of the command--"info".And I can't get the hang of the use of other commands through "help",so is there anywhere can I learn more about other commands?Or I can only learn it through the datasheet of DRS4 chip.

Thanks!

 

 

Attachment 1: ΢ÐÅͼƬ_20171122153834.png
΢ÐÅͼƬ_20171122153834.png
  638   Thu Nov 16 02:55:44 2017 Diego YankelevichAveraging capabilities

The Display window in the Oscilloscope software shows averaging capabilites but I have not been able to activate these. Is it possible to activate averaging with the existing oscilloscope software? Thanks

  637   Fri Nov 3 13:28:04 2017 Stefan RittTriggering using AND

Think about: How would you make a coincidence (AND) between two edges? Since an edge is infinitesimally small, there is no way to make a meaningful coincidence between edges. Therefore, the DRS4 EB firmware makes a simple AND of levels. If you trigger on rising signals and do an AND, then you get a trigger if both values are above their threshold. For falling edge trigger (arrow goes down in the trigger configuration) the board triggers when both signals are BELOW the trigger threshold. So actually the singnal which crosses the threshold last determines the timing of the trigger. I see no other way how to implement an AND.

Stefan

Håkan Wennlöf wrote:

Hi!

I'm using the DRSOsc program, and I have a question that I need a bit clarified;

When triggering using AND between two channels, am I then triggering on rising/falling edge of both channels, or on the actual values?

That is, is it the change in value that it triggers on, or when the actual value goes above a certain threshold?

Using AND, it seems like I get a trigger when one (CH2) is above its trigger value (sometimes quite far above), and the other (CH1) changes to go above its trigger value. This works for my purposes, which is nice, but I just want to be sure I understand why it works. Ideally, I'd like to trigger when one of my channels is above a certain value, and the other has a rising edge above a certain value.

I'm sorry if it's a silly question! I've just noticed that the Keysight oscilloscopes I've used can only do one channel with a rising edge at a time when doing a logic trigger, and I thought the same thing might be going on in the background here (which would be ideal for my purposes).

 

Kind regards,

Håkan Wennlöf

 

  636   Fri Nov 3 12:11:14 2017 Håkan WennlöfTriggering using AND

Hi!

I'm using the DRSOsc program, and I have a question that I need a bit clarified;

When triggering using AND between two channels, am I then triggering on rising/falling edge of both channels, or on the actual values?

That is, is it the change in value that it triggers on, or when the actual value goes above a certain threshold?

Using AND, it seems like I get a trigger when one (CH2) is above its trigger value (sometimes quite far above), and the other (CH1) changes to go above its trigger value. This works for my purposes, which is nice, but I just want to be sure I understand why it works. Ideally, I'd like to trigger when one of my channels is above a certain value, and the other has a rising edge above a certain value.

I'm sorry if it's a silly question! I've just noticed that the Keysight oscilloscopes I've used can only do one channel with a rising edge at a time when doing a logic trigger, and I thought the same thing might be going on in the background here (which would be ideal for my purposes).

 

Kind regards,

Håkan Wennlöf

  635   Wed Oct 18 11:48:14 2017 Vadym DenysenkoTime offset

Thank you for your reply!

Stefan Ritt wrote:

No this is not possible. But you can delay your signal externally (like with a delay cable or electronically) and then send the dealyed signal to the evaluation board for triggering.

Stefan

Vadym Denysenko wrote:

Hello.

 

I have a simple question, can I set SetTriggerDelayNs() more than 1631 ns?

I need to set this delay to about 5 us... Can I do this? 

 

Thank you in advance! 

 

With best regards, 

Vadym

 

 

  634   Wed Oct 18 09:12:26 2017 Stefan RittTime offset

No this is not possible. But you can delay your signal externally (like with a delay cable or electronically) and then send the dealyed signal to the evaluation board for triggering.

Stefan

Vadym Denysenko wrote:

Hello.

 

I have a simple question, can I set SetTriggerDelayNs() more than 1631 ns?

I need to set this delay to about 5 us... Can I do this? 

 

Thank you in advance! 

 

With best regards, 

Vadym

 

  633   Tue Oct 17 14:58:58 2017 Vadym DenysenkoTime offset

Hello.

 

I have a simple question, can I set SetTriggerDelayNs() more than 1631 ns?

I need to set this delay to about 5 us... Can I do this? 

 

Thank you in advance! 

 

With best regards, 

Vadym

  632   Mon Oct 16 15:35:22 2017 Stefan RittRaspberry Pi Connection Failure

Have you tried as root? Maybe you miss some permissions.

Stefan

Jonathan Wapman wrote:

I am currently attempting to use a raspberry pi to connect to the DRS 4 board. I whenever I try to use the DRS Command Line TOol, Revision 21435 to connect to the drs board, I get the error

"musb_open: libusb_open() error -3"

"USB successfully scanned, but no boards found"

"No DRS Boards Found".

I successfully compiled the libusb driver before compiling the drs software 5.0.6, and installed all other listed packages in the install instructions.

 

ELOG V3.1.5-3fb85fa6