Entry  Wed Mar 3 17:36:31 2010, Hao Huan, Initialization of the Domino Circuit 

Hi Stefan,

    I read in the datasheet that every time after power up the Domino wave in DRS4 needs to be started and stopped once to initialize the Domino circuit. However in your firmware it seems the chip immediately goes into the idle state after reset. Is that Domino circuit initialization really necessary?

    Also an aside question: in your firmware the readout process has the SRCLK sent to DRS4 only about 200ns later after RSRLOAD gets asserted instead of immediately following RSRLOAD. Is there any reason for that?

    Thanks a lot!

 

    Reply  Wed Mar 3 17:49:30 2010, Stefan Ritt, Initialization of the Domino Circuit 

Hao Huan wrote:

Hi Stefan,

    I read in the datasheet that every time after power up the Domino wave in DRS4 needs to be started and stopped once to initialize the Domino circuit. However in your firmware it seems the chip immediately goes into the idle state after reset. Is that Domino circuit initialization really necessary?

    Also an aside question: in your firmware the readout process has the SRCLK sent to DRS4 only about 200ns later after RSRLOAD gets asserted instead of immediately following RSRLOAD. Is there any reason for that?

    Thanks a lot!

 

The start/stop requirement is obsolete and has been replaced by  elog:10. I need to update this in the datasheet. The delay between the RSRLOAD and the SRCLK has the following reason: On the RSRLOAD the first sampling cell is output to the chip and to the ADC. This can sometimes be a rather high swing, which needs some time to settle, and some warm-up for the output driver. But actually I never really measured it, so it's there just as a safety margin. But I would encourage you to try to reduce this time and see it the first few bins of the readout change in offset.

Entry  Tue Jul 19 02:35:04 2022, Jingyu Zhang, Increase event rate, use ROI mode, and install sw from source in Mac 

Dear experts, 

We are trying to increase the event rate of the DRS4. We looked into the ROI but couldn’t figure out how to run in ROI mode. We are wondering if there is pre-existing firmware for this? We also tried to download and build the software from source on MacOS 12.4 but we were not successful. Can you kindly help us with these?

Best regards, 

Jingyu

    Reply  Fri Jul 29 14:09:35 2022, Stefan Ritt, Increase event rate, use ROI mode, and install sw from source in Mac 

The firmware from the website always reads 1024 bins. You have to modify it to stop before that, like reading only 128 samples or so. For compiling under MacOSX, this should work, since I do it myself. 

Regards,
Stefan

Jingyu Zhang wrote:

Dear experts, 

We are trying to increase the event rate of the DRS4. We looked into the ROI but couldn’t figure out how to run in ROI mode. We are wondering if there is pre-existing firmware for this? We also tried to download and build the software from source on MacOS 12.4 but we were not successful. Can you kindly help us with these?

Best regards, 

Jingyu

 

Entry  Tue Jan 31 08:10:37 2012, Stefan Ritt, IEEE Real Time 2012 Call for Abstracts 

Hello,

I'm co-organizing the upcoming Real Time Conference, which covers also fields of waveform processing and sampling, so it might be interesting for people working with the DRS4 chip. If you have recent results, you could also consider to send an abstract to this conference.  It will be nicely located in Berkeley, California. We plan excursions to San Francisco and to Napa Valley.

Best regards,

Stefan Ritt


18th Real Time Conference
June 11 – 15, 2012
Berkeley, CA

We invite you to the Hotel Shattuck Plaza in downtown Berkeley, California for
the 2012 Real-Time Conference (RT2012).   It will take place Monday, June 11
through Friday, June 15, 2012, with optional pre-conference tutorials Saturday
and Sunday, June 9-10.

Like the previous editions, RT2012 will be a multidisciplinary conference
devoted to the latest developments on realtime techniques in the fields of
plasma and nuclear fusion, particle physics, nuclear physics and astrophysics,
space science, accelerators, medical physics, nuclear power instrumentation and
other radiation instrumentation.

Abstract submission is open as of 18 January (deadline 2 March). Please visit

http://www.npss-confs.org/rtc/welcome.asp?flag=44675.77&Retry=1 to submit an

abstract.

Call for Abstracts 

RT 2012 is an interdisciplinary conference on realtime data acquisition and
computing applications in the physical sciences. These applications include:

* High energy physics 
* Nuclear physics 
* Astrophysics and astroparticle physics 
* Nuclear fusion 
* Medical physics 
* Space instrumentation 
* Nuclear power instrumentation 
* Realtime security and safety 
* General Radiation Instrumentation 

Specific topics include (but are certainly not limited to) the list shown below.
We welcome correspondence to see how your research fits our venue.   

Key Dates

* Abstract submission opened:  January 18, 2012 
* Abstract deadline:  March 2, 2012 
* Program available: April 2 

Suggested Topics

* Realtime system architectures 
* Intelligent signal processing 
* Programmable devices 
* Fast data transfer links and networks 
* Trigger systems 
* Data acquisition 
* Processing farms 
* Control, monitoring, and test systems 
* Upgrades 
* Emerging realtime technologies 
* New standards 
* Realtime safety and security 
* Feedback on experiences 

Contact Information

If you have a question or wish to opt in for occasional e-mail updates about
RT2012, send us a message at RT2012@lbl.gov. To view full conference
information, visit http://rt2012.lbl.gov/index.html

Entry  Sat Jan 15 09:13:42 2022, student_riku, I want to know about the readout 

Hello, everyone.
I'm a student in Japan.
Please forgive me if this is a very rudimentary question.

Am I right in thinking that inputting 1 to DMODE (Bit0) in the configuration register will connect the 1024th cell to the 1st cell?

Also, let's assume that after sampling 1024 cells on channel 0, 200 cells are sampled in the second week.
Does the readout of 1024 cells start from cell 0?
Or does it start from the 200th cell?

    Reply  Sat Jan 15 10:50:47 2022, Stefan Ritt, I want to know about the readout 
student_riku wrote:

Am I right in thinking that inputting 1 to DMODE (Bit0) in the configuration register will connect the 1024th cell to the 1st cell?

Yes, as desceribed in the manual .

student_riku wrote:

Also, let's assume that after sampling 1024 cells on channel 0, 200 cells are sampled in the second week.
Does the readout of 1024 cells start from cell 0?
Or does it start from the 200th cell?

I don't understand what you mean by "second week". 

Normally, you run the chip continously (DMODE=1) until you receive a trigger. Then, you typically read out the chip from the stop position by using a RSRLOAD pulse (as described unter "REGION-OF-INTEREST READOUT MODE".

Stefan

    Reply  Wed Jan 26 06:44:11 2022, student_riku, I want to know about the readout 

Dear Stefan

Thanks a lot.

I solved it.

 

Stefan Ritt wrote:
student_riku wrote:

Am I right in thinking that inputting 1 to DMODE (Bit0) in the configuration register will connect the 1024th cell to the 1st cell?

Yes, as desceribed in the manual .

student_riku wrote:

Also, let's assume that after sampling 1024 cells on channel 0, 200 cells are sampled in the second week.
Does the readout of 1024 cells start from cell 0?
Or does it start from the 200th cell?

I don't understand what you mean by "second week". 

Normally, you run the chip continously (DMODE=1) until you receive a trigger. Then, you typically read out the chip from the stop position by using a RSRLOAD pulse (as described unter "REGION-OF-INTEREST READOUT MODE".

Stefan

 

Entry  Mon Oct 31 09:15:02 2011, Zhongwei Du, How to link PMT 

I want to measure the signal from PMT . But it is a current signal, should i just put a series resistance, or use a amplifier to convert it to voltage signal before drs4?  

Can you give me some advice ? 

    Reply  Tue Nov 1 11:07:02 2011, Stefan Ritt, How to link PMT 

Zhongwei Du wrote:

I want to measure the signal from PMT . But it is a current signal, should i just put a series resistance, or use a amplifier to convert it to voltage signal before drs4?  

Can you give me some advice ? 

The evaluation board has a 50 Ohm termination resistor, which already converts your current into a voltage signal. If the resulting signal is too low (<20 mV) you can put in an amplifier or raise the HV of your PMT (inside the valid range given by its datasheet of course).

- Stefan 

Entry  Wed Mar 2 17:25:10 2022, Matias Senger, How to convert samples to volt? 

I am using the `drscl` app. My prior experience is practically zero, sorry if this is a very naive question. When I read using `read 0 1` (channel 0, with calibration) I get this:

```
Calibration not valid for board #2946
  10    3    7    4   10    8   14    5    5    9    3    4    9    8    9    4
   3    3   12    5    5   13    3    8    1    5    0    4    8    6    6    3
...etc...
```

Why does it says that the calibration is not valid? How am I supposed to go from integers to volts?

If I run the `info` command I get this:

```
==============================
Mezz. Board index:    0
DRS type:             DRS4
Board type:           9
Serial number:        2946
Firmware revision:    30000
Temperature:          43.4 C
Input range:          -0.5V...0.5V
Calibrated range:     -0.5V...0.5V
Calibrated frequency: 0.000 GHz
Status reg.:          0000009A
Control reg.:         00000000
  DMODE circular
Trigger bus:          00000000
Frequency:            1.007 GHz
```

    Reply  Thu Mar 3 13:47:26 2022, Stefan Ritt, How to convert samples to volt? 

The 'drscl' tool is more for experts, normal users are advised to use the DRSOsc oscilloscope.

The board has to be calibrated for a given sampling speed before calibrated data can be read out. Do that with the "calib" command, specifying 5 for the sampling rate, 0 for the range (which is the middle between -0.5 and +0.5) and 1 for 1024 mode. If you then do "start", "stop", "read 0 1" you get calibrated data in mV.

Stefan

Matias Senger wrote:

I am using the `drscl` app. My prior experience is practically zero, sorry if this is a very naive question. When I read using `read 0 1` (channel 0, with calibration) I get this:

```
Calibration not valid for board #2946
  10    3    7    4   10    8   14    5    5    9    3    4    9    8    9    4
   3    3   12    5    5   13    3    8    1    5    0    4    8    6    6    3
...etc...
```

Why does it says that the calibration is not valid? How am I supposed to go from integers to volts?

If I run the `info` command I get this:

```
==============================
Mezz. Board index:    0
DRS type:             DRS4
Board type:           9
Serial number:        2946
Firmware revision:    30000
Temperature:          43.4 C
Input range:          -0.5V...0.5V
Calibrated range:     -0.5V...0.5V
Calibrated frequency: 0.000 GHz
Status reg.:          0000009A
Control reg.:         00000000
  DMODE circular
Trigger bus:          00000000
Frequency:            1.007 GHz
```

 

Entry  Thu Mar 14 03:43:49 2019, Deepak Samuel, How to buy DRS evaluation kit 

Dear Stefan,

I have emailed drs4@psi.ch a couple of times regarding the pricing of the evaluation kits for academic use in India and have not received any reply and hence writing in this forum. Could you please help me in this?

 

Thanks and regards,
Deepak Samuel.

Entry  Wed May 26 19:18:09 2010, Hao Huan, High Frequency Input for DRS 

Hi Stefan,

    I read in the DRS datasheet that the bandwidth for the transparent mode OUT+ is only 200MHz which I think cannot be improved by any active input buffer; so if you want to operate the chip for really high frequency input, would it be better to feed on-board discriminators not from the output of DRS but from the input end?

    Thanks!

 

    Reply  Tue Jun 1 13:36:18 2010, Stefan Ritt, High Frequency Input for DRS 

Hao Huan wrote:

Hi Stefan,

    I read in the DRS datasheet that the bandwidth for the transparent mode OUT+ is only 200MHz which I think cannot be improved by any active input buffer; so if you want to operate the chip for really high frequency input, would it be better to feed on-board discriminators not from the output of DRS but from the input end?

    Thanks!

 

First, the 200 MHz is not correct. Table 1 clearly states ANALOG OUTPUTS - Bandwidth (-3dB): 50 MHz. This is also shown in plot 11 (revision 0.9). But that does not necessarily mean that you have to drive your discriminators from the input of the DRS4 chip. If you use this for triggering, a 1-2 ns timing jitter does not matter, since stopping the domino wave anyhow has a 3-4 cell jitter. If you send a very fast signal though a 50 MHz low pass filter, the timing anyhow doe not change, only the slope of your signal gets lower, so you are more sensitive to noise, which in turn causes the 1-2 ns timing jitter. But I personally would not worry about that too much. Putting any signal splitting components in the input path would reduce the input bandwidth, which would be much more of an issue.

Entry  Wed Feb 10 02:57:55 2010, pepe sanchez lopez, Hello 

hello i am an student and i want to do my final project with drs4 board and i really can´t find how to open waveform file and how can i save or opened many of them quickly.

if you can tell me how i will be very grateful.

thanks,

kind regards.

    Reply  Wed Feb 10 15:35:09 2010, Stefan Ritt, Hello 

pepe sanchez lopez wrote:

hello i am an student and i want to do my final project with drs4 board and i really can´t find how to open waveform file and how can i save or opened many of them quickly.

if you can tell me how i will be very grateful.

thanks,

kind regards.

There is no built-in possibility to open waveform files, you have to write your own programs to do that. 

Entry  Thu May 8 23:41:03 2025, Jonathan Bradshaw, Handling of Write Shift Register and Write Config Register 

Hi all

We're building a product which will use two different operating modes; firstly a long capcture using channel daisy chaining (2048 samples) and secondly a segmented capture (2 separate captures of 1024 samples each). 

For the long capture, I'm looking to capture 2048 samples for 4 channels.  Therefore I configure the Write Shift Register to 0b01010101 and the Write Config Register to 0b11111111.  During capture with DWRITE=1 the Write Shift Register will update.  Am I correct that once the capture is done and DWRITE=0, I can set A3..0 to 0b1101 and simply read the value of WSROUT to tell the difference?

For the segmented capture, I'm looking to capture 1024 samples for 4 channels on a first tirgger pulse, followed by 1024 samples for 4 channels on a second pulse.  Therefore I configure the Write Shift Register to 0b11111111 and the Write Config Register to 0b01010101 and set DWRITE=0 to capture.  After the first trigger I set DWRITE=0 and need to update the Write Config Register.  Do I need to write in a whole 8 bits to the Write Config Register (i.e. 0b10101010), or can I just shift in a single new bit (value 0b0)?

    Reply  Fri May 9 08:17:50 2025, Stefan Ritt, Handling of Write Shift Register and Write Config Register 

This is correct. Setting A0-A3 to 0b1101 multiplexes the Shift Write Register to SROUT, so you will either a "0" or a "1" depending on which of the two channels was written last.

Your segmented capture does unfortunately not work. Due to a bug in the silicon, the first (e.g. even) written channel gets half overwritten when you start sampling the second (odd) channel. I should remove that from the documentation.

Furthermore, reading the chip while writing on the "other side" introduces quite some additional noise. The recommended way to do simultaneous reading and writing is therefore to use two separate DRS4 chips and split the input signals to both chips, then read from one chip while writing to the other chip. This keeps the crosstalk at a minimum and both chips run at full performance.

Stefan

Jonathan Bradshaw wrote:

Hi all

We're building a product which will use two different operating modes; firstly a long capcture using channel daisy chaining (2048 samples) and secondly a segmented capture (2 separate captures of 1024 samples each). 

For the long capture, I'm looking to capture 2048 samples for 4 channels.  Therefore I configure the Write Shift Register to 0b01010101 and the Write Config Register to 0b11111111.  During capture with DWRITE=1 the Write Shift Register will update.  Am I correct that once the capture is done and DWRITE=0, I can set A3..0 to 0b1101 and simply read the value of WSROUT to tell the difference?

For the segmented capture, I'm looking to capture 1024 samples for 4 channels on a first tirgger pulse, followed by 1024 samples for 4 channels on a second pulse.  Therefore I configure the Write Shift Register to 0b11111111 and the Write Config Register to 0b01010101 and set DWRITE=0 to capture.  After the first trigger I set DWRITE=0 and need to update the Write Config Register.  Do I need to write in a whole 8 bits to the Write Config Register (i.e. 0b10101010), or can I just shift in a single new bit (value 0b0)?

 

    Reply  Tue May 13 04:10:30 2025, Jonathan Bradshaw, Handling of Write Shift Register and Write Config Register 

Hi Stefan

Just so I'm 100% clear; is there no reliable way to perform 2 segmented captures with a single DRS4 IC?

While not a showstopper, this is a bit disappointing.

 

Stefan Ritt wrote:

This is correct. Setting A0-A3 to 0b1101 multiplexes the Shift Write Register to SROUT, so you will either a "0" or a "1" depending on which of the two channels was written last.

Your segmented capture does unfortunately not work. Due to a bug in the silicon, the first (e.g. even) written channel gets half overwritten when you start sampling the second (odd) channel. I should remove that from the documentation.

Furthermore, reading the chip while writing on the "other side" introduces quite some additional noise. The recommended way to do simultaneous reading and writing is therefore to use two separate DRS4 chips and split the input signals to both chips, then read from one chip while writing to the other chip. This keeps the crosstalk at a minimum and both chips run at full performance.

Stefan

Jonathan Bradshaw wrote:

Hi all

We're building a product which will use two different operating modes; firstly a long capcture using channel daisy chaining (2048 samples) and secondly a segmented capture (2 separate captures of 1024 samples each). 

For the long capture, I'm looking to capture 2048 samples for 4 channels.  Therefore I configure the Write Shift Register to 0b01010101 and the Write Config Register to 0b11111111.  During capture with DWRITE=1 the Write Shift Register will update.  Am I correct that once the capture is done and DWRITE=0, I can set A3..0 to 0b1101 and simply read the value of WSROUT to tell the difference?

For the segmented capture, I'm looking to capture 1024 samples for 4 channels on a first tirgger pulse, followed by 1024 samples for 4 channels on a second pulse.  Therefore I configure the Write Shift Register to 0b11111111 and the Write Config Register to 0b01010101 and set DWRITE=0 to capture.  After the first trigger I set DWRITE=0 and need to update the Write Config Register.  Do I need to write in a whole 8 bits to the Write Config Register (i.e. 0b10101010), or can I just shift in a single new bit (value 0b0)?

 

 

ELOG V3.1.5-3fb85fa6