DRS4 Forum
  DRS4 Discussion Forum, Page 30 of 46  Not logged in ELOG logo
Entry  Thu Mar 11 21:37:32 2010, Hao Huan, Input Bandwidth of the DRS Chip 
Hi Stefan,
    I read in the DRS datasheet that the input bandwidth if 950MHz. However, it also says the output bandwidth in the transparent
mode is 50MHz. Since in the transparent mode the input is routed to the output, does it mean the input bandwidth also gets reduced in the transparent mode?
    Reply  Fri Mar 12 08:04:44 2010, Stefan Ritt, Input Bandwidth of the DRS Chip 


    
        
            Hao Huan wrote:
        
        
            
            I read in the DRS datasheet that the input bandwidth if 950MHz. However, it also says the output bandwidth
Entry  Wed Mar 3 17:36:31 2010, Hao Huan, Initialization of the Domino Circuit 
Hi Stefan,
    I read in the datasheet that every time after power up the Domino wave in DRS4 needs to be started and stopped once to initialize
the Domino circuit. However in your firmware it seems the chip immediately goes into the idle state after reset. Is that Domino circuit initialization
    Reply  Wed Mar 3 17:49:30 2010, Stefan Ritt, Initialization of the Domino Circuit 


    
        
            Hao Huan wrote:
        
        
            
            Hi Stefan,
Entry  Tue Jul 19 02:35:04 2022, Jingyu Zhang, Increase event rate, use ROI mode, and install sw from source in Mac 
Dear experts, 

We are trying to increase the event rate of the DRS4. We looked into the ROI
but couldn’t figure out how to run in ROI mode. We are wondering if there is pre-existing firmware for this? We also tried to download and build
    Reply  Fri Jul 29 14:09:35 2022, Stefan Ritt, Increase event rate, use ROI mode, and install sw from source in Mac 
The firmware from the website always reads 1024 bins. You have to modify it to stop before that, like reading only 128 samples or so. For compiling under
MacOSX, this should work, since I do it myself. 

Regards,
Entry  Tue Jan 31 08:10:37 2012, Stefan Ritt, IEEE Real Time 2012 Call for Abstracts 
Hello,
I'm co-organizing the upcoming Real Time Conference, which covers also fields of waveform processing and sampling, so it might be interesting for
people working with the DRS4 chip. If you have recent results, you could also consider to send an abstract to this conference.  It will be nicely
Entry  Sat Jan 15 09:13:42 2022, student_riku, I want to know about the readout 
Hello, everyone.
I'm a student in Japan.
Please forgive me if this is a very rudimentary question.
    Reply  Sat Jan 15 10:50:47 2022, Stefan Ritt, I want to know about the readout 



student_riku
wrote:



Am I right in thinking that inputting 1 to DMODE (Bit0) in the configuration
register will connect the 1024th cell to the 1st cell?
    Reply  Wed Jan 26 06:44:11 2022, student_riku, I want to know about the readout 
Dear Stefan

Thanks a lot.

I solved it.
Entry  Mon Oct 31 09:15:02 2011, Zhongwei Du, How to link PMT 
I want to measure the signal from PMT . But it is a current signal, should i just put a series resistance, or use a amplifier to convert it to voltage
signal before drs4?  
Can you give me some advice ? 
    Reply  Tue Nov 1 11:07:02 2011, Stefan Ritt, How to link PMT 


    
        
            Zhongwei Du wrote:
        
        
            
            I want to measure the signal from PMT . But it is a current signal, should i just put a series resistance,
Entry  Wed Mar 2 17:25:10 2022, Matias Senger, How to convert samples to volt? 
I am using the `drscl` app. My prior experience is practically zero, sorry if this is a very naive question. When I read using `read 0 1` (channel 0,
with calibration) I get this:

```
    Reply  Thu Mar 3 13:47:26 2022, Stefan Ritt, How to convert samples to volt? 
The 'drscl' tool is more for experts, normal users are advised to use the DRSOsc oscilloscope.

The board has to be calibrated for a given sampling speed before calibrated data can be read out. Do that with the "calib" command,
specifying 5 for the sampling rate, 0 for the range (which is the middle between -0.5 and +0.5) and 1 for 1024 mode. If you then do "start",
Entry  Thu Mar 14 03:43:49 2019, Deepak Samuel, How to buy DRS evaluation kit 
Dear Stefan,

I have emailed drs4@psi.ch a couple of times regarding the pricing of the evaluation kits for academic use in India and have not received any
reply and hence writing in this forum. Could you please help me in this?
Entry  Wed May 26 19:18:09 2010, Hao Huan, High Frequency Input for DRS 
Hi Stefan,
    I read in the DRS datasheet that the bandwidth for the transparent mode OUT+ is only 200MHz which I think cannot be improved
by any active input buffer; so if you want to operate the chip for really high frequency input, would it be better to feed on-board discriminators not
    Reply  Tue Jun 1 13:36:18 2010, Stefan Ritt, High Frequency Input for DRS 


    
        
            Hao Huan wrote:
        
        
            
            Hi Stefan,
Entry  Wed Feb 10 02:57:55 2010, pepe sanchez lopez, Hello 
hello i am an student and i want to do my final project with drs4 board and i really can´t find how to open waveform file and how can i save or
opened many of them quickly.
if you can tell me how i will be very grateful.
    Reply  Wed Feb 10 15:35:09 2010, Stefan Ritt, Hello 


    
        
            pepe sanchez lopez wrote:
        
        
            
            hello i am an student and i want to do my final project with drs4 board and i really can´t find
Entry  Thu May 8 23:41:03 2025, Jonathan Bradshaw, Handling of Write Shift Register and Write Config Register 
Hi all

We're building a product which will use two different operating modes; firstly a long capcture using channel daisy chaining (2048 samples)
and secondly a segmented capture (2 separate captures of 1024 samples each). 
ELOG V3.1.5-3fb85fa6