DRS4 Forum
  DRS4 Discussion Forum, Page 12 of 45  Not logged in ELOG logo
IDdown Date Author Subject Text Attachments
  683   Wed May 2 10:44:17 2018 Alessio BertiPeak at 0 mV in tracesHi,

we modified drs_exam.cpp to read
all 4 channels from the DRS4 and apply directly
 zero_peak_after_spike_removal_ch1.pngzero_peak_after_spike_removal_ch2.pngzero_peak_after_spike_removal_ch3.pngzero_peak_after_spike_removal_ch4.pngzero_peak_after_spike_removal_offset_correction_ch2.png 
  682   Wed May 2 09:24:53 2018 Stefan RittDRS4 using drs_exam.cpp to save as binary filesYou have to write the C/C++ code yourself
to write data in binary or any other format.
All information is present after the waveform
  
  681   Tue May 1 02:00:40 2018 Hyunseong KimDRS4 using drs_exam.cpp to save as binary filesHi, 

I would like to save the waveform
in a .dat binary file using drs_exam.cpp.
  
  680   Tue Apr 17 13:28:23 2018 Stefan RittDRS4 read_binary.cpp On the software download page at https://www.psi.ch/drs/software-download
you find a link to all versions of the DRS
software, which is located at: https://www.dropbox.com/sh/clqo7ekr0ysbrip/AACoWJzrQAbf3WiBJHG89bGGa?dl=0
  
  679   Mon Apr 16 21:21:29 2018 Sobimpe EniolaDRS4 read_binary.cpp Hello everyone, 

The new read_binary.cpp code 

I will be very glad if anyone can
  
  678   Fri Apr 13 18:14:07 2018 Alessio BertiVoltage and Timing Calibration in drs_exam.cppHi,

we were trying to implement an
automatic way to calibrate our DRS4 both
  
  677   Fri Mar 23 09:39:55 2018 Stefan RittRead the CalibrateWaveformYou don't have to read and calibrate
the waveforms in your user code, but can
rely on the DRS.cpp library to do that. Just
  
  676   Thu Mar 22 14:36:01 2018 Phan Van ChuanRead the CalibrateWaveformHelo
I'm building an application for
reading waveforms from the DRS4 board to
  
  675   Mon Mar 19 16:22:42 2018 Stefan RittROI The DRS4 has an internal storage of 1024
capacitors. They work as a ring buffer, so
at 5GSPS you can store 200ns wide signals.
  
  674   Mon Mar 19 15:12:02 2018 Stefan RittRunning drs_example.cppThe time channel is already calibrated
in ns. So for 5 GSPS, the time scale goes
from zero to 200. Concerning your other issues
  
  673   Fri Mar 16 14:00:06 2018 Stefan Rittconfusion about the description in drs.cppThe FPGA is very small, so it only has
an address space of 256 bytes. Look at the
definition in DRS.cpp
  
  672   Thu Mar 15 08:44:26 2018 Stefan Rittsub-ms precision timestamps?Putting sub-ms precision into the header
does not make sense, since the USB transfer
only happens in time-slots of about 2 ms.
  
  671   Wed Mar 14 09:13:39 2018 chen wenjunconfusion about the description in drs.cppHi,Stefan:

  recently,whtn I study the
drs.cpp code ,I found that  the buffer[1]
 20180314161201.jpg 
  668   Wed Mar 14 00:38:15 2018 Will Flanagansub-ms precision timestamps?Dear DRS4 community,

Is there a way to extract timestamps
with sub-ms precision? The milliseconds of
  
  667   Thu Mar 8 22:54:20 2018 Rodrigo Trindade de MenezesRunning drs_example.cppWe found a way to solve the previous problem,
but right now when we try to set the input
range only -0.5 to 0.5 is working. When we
  
  666   Wed Mar 7 22:49:38 2018 Rodrigo Trindade de MenezesRunning drs_example.cppHello,

We have been using the DRS4 evaluation
board (S/N 2636) that works with the scope
 drs_exam.cpp 
  665   Fri Mar 2 21:05:48 2018 Steven BlockROI Great! That is very helpful. 

One more question. If no signals
were detected in the 1024*200ps time frame
  
  664   Fri Mar 2 20:17:17 2018 Stefan RittROI N'/N is correct. The 2 us "from
the response you got from me" come from
the fact that after readout, you have to
  
  663   Fri Mar 2 18:08:55 2018 Steven BlockROI Hello,

I have a question about how ROI
works. From what I have read, it will only
  
  662   Tue Feb 27 18:12:32 2018 Stefan RittDRS4 Dead timesFor applications which are critical on
the dead time, one typically uses one ADC
per DRS4 channel, and thus the dead time
  
ELOG V3.1.5-fe60aaf