    Reply  Fri Jun 21 12:54:47 2019, Stefan Ritt, Evaluation firmware wait_vdd state 

Dear Andrew,

the posting you mention is still accurate. Any power supply will drop when you start the Domino wave, no matter how big your capacitor is. Unfortunately the output signal of the DRS4 scales with VDD. So if your VDD drops by 40 mV and you get a trigger and you immediately start the readout, the output baseline will also be shifted by about 40 mV. If you are sensitive to dead time, you can remove the wait_vdd state completely, but then you have to deal with varying baseline shifts. If you have narrow signals sitting on a broad baseline, you can correct for this by measuring the baseline outside your signal, then subtracting it before integrating your pulse. If you have lots of pile-up in your signals, it might sometimes be hard to evaluate the baseline on an event-by-event basis.

Stefan

Andrew Peck wrote:

Dear Stefan,

I am working with others at UCLA on a custom made board built around the DRS4. We are in the process of writing firmware so I am adapting the readout state machine from the evaluation board firmware.

I see in the state machine of the eval board firmware that after a trigger is received, the FPGA goes into the start readout state and then into "wait_vdd", where the FPGA waits "~120 us for vdd to stabilize" before reading out the ADC.

Our application is sensitive to deadtime and this wait_vdd state adds very significantly.  I am trying to find anything explaining the necessity of wait_vdd in the documentation / elog and have only found so far your old forum posting, https://elog.psi.ch/elogs/DRS4+Forum/12

Does this forum posting explain wait_vdd or is there a another purpose that I have missed?

If this post is relevant to wait_vdd, does the advice of large capacitance and an LDO with fast transient response still apply or are there any new recommendations?

Thank you,

Andrew Peck

 

    Reply  Wed Jun 26 13:08:42 2019, Stefan Ritt, drs_exam is always reading out a sin wave 

Sure, that’s correct. The example program turns on the internal sine wave generator in case people don’t have a real signal. That’s why it’s called „example“. Find the code which turns on the generator and change it. You will also have to change the trigger settings depending on your actual signal.

Stefan

 

Si Xie wrote:

We are using the drs_exam.cpp to read out waveforms, but it seems to be outputting only sin waves on all channels - as if it was reading out the simulated waveform from the oscilloscope program if we run it without the board plugged in. Does anyone know what is causing this?

We are taking data with a pulser plugged into channel 1, which produces a single pulse with width of 8ns, and nothing plugged into channel 2. 

Our board is as follows:

Found DRS4 evaluation board, serial #2567, firmware revision 21305
Board type: 9

The output is something like the following:

Event #0 ----------------------
  t1[ns]  u1[mV]  t2[ns] u2[mV]
  0.000  -452.7   0.026  -469.3
  0.289  -460.8   0.293  -469.8
  0.413  -477.3   0.400  -481.5
  0.642  -485.3   0.650  -482.4
  0.806  -486.9   0.821  -477.8
  1.086  -476.8   1.085  -457.2
  1.183  -467.3   1.162  -446.4
  1.450  -435.6   1.459  -405.1
  1.619  -410.1   1.630  -373.3
  1.843  -366.2   1.851  -323.9
  1.945  -342.9   1.948  -298.9
  2.221  -275.7   2.210  -229.3
  2.359  -237.6   2.357  -187.6
  2.602  -165.6   2.609  -111.2
  2.687  -141.1   2.697   -84.3
  2.976   -50.5   2.987     5.5
  3.164     8.4   3.144    53.3
  3.377    73.9   3.384   124.2
  3.503   111.4   3.506   158.0
  3.753   182.0   3.769   226.9
  3.924   227.5   3.929   265.8
 

 

    Reply  Mon Jul 8 14:29:12 2019, Stefan Ritt, drs_exam is always reading out a sin wave 

Actually in the original drs_exam.cpp the sine wave oscillator is turned off with this command

/* use following line to turn on the internal 100 MHz clock connected to all channels  */
//b->EnableTcal(1);

If you remove the "//" then the generator gets enabled. Probably you did this by accident. With this line commented out, you see the proper input like this:

Event #0 ----------------------
  t1[ns]  u1[mV]  t2[ns] u2[mV]
  0.000     1.9   0.000    -2.4
  0.195     0.5   0.195     0.3
  0.391     0.1   0.391    -1.4
  0.586    -0.7   0.586    -0.4
  0.781    -1.1   0.781    -2.4
  0.977    -0.6   0.977     0.0
  1.172    -1.5   1.172    -2.8
  1.367    -0.4   1.367    -0.6
  1.562    -1.2   1.562    -3.8
  1.758    -1.5   1.758    -1.7
  1.953    -1.0   1.953    -3.3
  2.148    -0.7   2.148    -1.8
  2.344    -1.6   2.344    -4.2
  2.539     0.5   2.539    -1.5
  2.734     0.2   2.734    -3.6
...

167.969    -3.4 167.969    -5.2
168.164    -3.7 168.164    -3.6
168.359     0.0 168.359    -2.0
168.555     1.9 168.555    -0.2
168.750     2.8 168.750    -2.8
168.945     5.4 168.945    -1.4
169.141    18.0 169.141     1.2
169.336    26.6 169.336     2.7
169.531    46.2 169.531     0.4
169.727    56.2 169.727     1.6
169.922    93.3 169.922     0.1
170.117   115.6 170.117     0.0
170.312   174.4 170.312    -1.5
170.508   206.9 170.508    -0.8
170.703   282.2 170.703    -2.4
170.898   328.4 170.898    -1.2
171.094   419.6 171.094    -3.2
171.289   465.8 171.289    -2.5
171.484   500.0 171.484    -2.0
171.680   500.0 171.680    -0.6
171.875   500.0 171.875    -4.0
172.070   500.0 172.070    -1.1
172.266   500.0 172.266    -3.7
172.461   500.0 172.461    -2.1
172.656   500.0 172.656    -5.0
172.852   500.0 172.852    -3.3
173.047   500.0 173.047    -4.8
173.242   500.0 173.242    -4.1
173.438   500.0 173.438    -5.1
173.633   500.0 173.633    -3.3
173.828   500.0 173.828    -6.4
174.023   500.0 174.023    -3.9
174.219   500.0 174.219    -5.5
174.414   500.0 174.414    -3.2
174.609   500.0 174.609    -3.6
174.805   500.0 174.805    -2.6
175.000   500.0 175.000    -5.2
175.195   500.0 175.195    -2.7
175.391   434.3 175.391    -3.9
175.586   391.7 175.586    -2.4
175.781   312.2 175.781    -4.1
175.977   275.7 175.977    -1.8
176.172   202.4 176.172    -3.8
176.367   167.6 176.367    -1.4
176.562   117.4 176.562    -2.9
176.758    96.1 176.758    -2.3
176.953    62.8 176.953    -3.3
177.148    49.1 177.148    -1.8
177.344    35.9 177.344    -4.3
177.539    33.4 177.539    -2.6
177.734    30.4 177.734    -4.2
...

 

Si Xie wrote:

I see. Where is the code that we can use to turn off the generator? I thought the example is taking data with CH1 as the trigger.

For our board, which is BoardType == 9, it is running these lines:

      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<0);        // set CH1 as source

Is that not using the hardware trigger with CH1 as the source?

 

Stefan Ritt wrote:

Sure, that’s correct. The example program turns on the internal sine wave generator in case people don’t have a real signal. That’s why it’s called „example“. Find the code which turns on the generator and change it. You will also have to change the trigger settings depending on your actual signal.

Stefan

 

Si Xie wrote:

We are using the drs_exam.cpp to read out waveforms, but it seems to be outputting only sin waves on all channels - as if it was reading out the simulated waveform from the oscilloscope program if we run it without the board plugged in. Does anyone know what is causing this?

We are taking data with a pulser plugged into channel 1, which produces a single pulse with width of 8ns, and nothing plugged into channel 2. 

Our board is as follows:

Found DRS4 evaluation board, serial #2567, firmware revision 21305
Board type: 9

The output is something like the following:

Event #0 ----------------------
  t1[ns]  u1[mV]  t2[ns] u2[mV]
  0.000  -452.7   0.026  -469.3
  0.289  -460.8   0.293  -469.8
  0.413  -477.3   0.400  -481.5
  0.642  -485.3   0.650  -482.4
  0.806  -486.9   0.821  -477.8
  1.086  -476.8   1.085  -457.2
  1.183  -467.3   1.162  -446.4
  1.450  -435.6   1.459  -405.1
  1.619  -410.1   1.630  -373.3
  1.843  -366.2   1.851  -323.9
  1.945  -342.9   1.948  -298.9
  2.221  -275.7   2.210  -229.3
  2.359  -237.6   2.357  -187.6
  2.602  -165.6   2.609  -111.2
  2.687  -141.1   2.697   -84.3
  2.976   -50.5   2.987     5.5
  3.164     8.4   3.144    53.3
  3.377    73.9   3.384   124.2
  3.503   111.4   3.506   158.0
  3.753   182.0   3.769   226.9
  3.924   227.5   3.929   265.8
 

 

 

 

    Reply  Mon Jul 15 17:26:50 2019, Stefan Ritt, Evaluation Board Test Functionality 

Have you set the trigger correctly to the channel with your signal, polarity and level? Do you undersand the difference between normal and auto trigger? Why don't you post a screendump. Are you ABSOLUTELY SURE that you have a signal on your cable? Have you tried with another oscilloscope? Are you sure that your SMA connector is good?

Stefan

 

Brendan Posehn wrote:

Hello, 

I have recently obtained a DRS4 Evaluation Board (V5), but I am unable to register signals when using the DRS Oscilloscope application. There seems to be some difference in noise when I have an input connected to a signal or not, but I am unable to view a simple, 0.2V amplitude square wave or other small signals. The only way I have been able to view a waveform is when connecting the reference clock to all channels. When running 'info' in the DRS Command Line Interface I am shown correct information. I was wondering if there is any way for me to test the functionality of the board (specifially ability to read signals on Ch 1-4) to ensure that it is indeed working as expected? 

Thanks, 

Brendan

 

    Reply  Thu Jul 18 11:37:56 2019, Stefan Ritt, Trace Impedance 

The requiremnet is the same as for any high speed analog board, there is othing special with the DRS4. If you want to terminate your line with 50 Ohms and you want a matched impedance layout, you route all lines with 50 Ohms impedance. Truth is however that nothing is perfect. The SMA connector is not exactly 50 Ohm, the PCB gets a 10-20% variation depending on the manufacturer. So even if you try hard, you will never have a 50 Ohm matched impedance. On the evaluation board we made some compromises as you have seen, but for us the board works satisfactory even with this compromises, and you can test it yourself with real hardware (namely the evaluation board). If you can do a better job, try it. But usually these compromises have only little influence on the signal quality.

Stefan

Ismael Garcia wrote:

Hi Steffan,

              I'm an engineer at UCLA developing a board with the DRS4 chip. Our team has a question on what might be the required trace impedence for the analog inputs. Can that information be provided? 

Best Regards,
Ismael Garcia

 

    Reply  Sat Jul 20 12:28:14 2019, Stefan Ritt, Trace Impedance 

The DRS4 input is high impedance. So if you like you can terminate it with 100 Ohm differentially and route it with 100 Ohm. But if you keep the lines short, the reflection is negligible. That’s what we made on the evaluation board.

Ismael Garcia wrote:

When you're refering to laying a 50 Ohm trace, you're referring to the SMA input and not the interface between the output of the Op-AMP(THS4508) buffer 
and the inputs  of the DRS4(IN0-IN8). Is there a recommended diffential impedance for IN0-IN8? 

Ismael

Stefan Ritt wrote:

The requiremnet is the same as for any high speed analog board, there is othing special with the DRS4. If you want to terminate your line with 50 Ohms and you want a matched impedance layout, you route all lines with 50 Ohms impedance. Truth is however that nothing is perfect. The SMA connector is not exactly 50 Ohm, the PCB gets a 10-20% variation depending on the manufacturer. So even if you try hard, you will never have a 50 Ohm matched impedance. On the evaluation board we made some compromises as you have seen, but for us the board works satisfactory even with this compromises, and you can test it yourself with real hardware (namely the evaluation board). If you can do a better job, try it. But usually these compromises have only little influence on the signal quality.

Stefan

Ismael Garcia wrote:

Hi Steffan,

              I'm an engineer at UCLA developing a board with the DRS4 chip. Our team has a question on what might be the required trace impedence for the analog inputs. Can that information be provided? 

Best Regards,
Ismael Garcia

 

 

 

    Reply  Tue Aug 20 10:44:45 2019, Stefan Ritt, should one deassert DENABLE while writing the write-shift register? 

Hi Bill,

you keep DENABLE active all the time to keep the Domino Wave running, but you deassert DWRITE if you change any register via SRCLK. There is no shadow register, just a simple shift register, but with DWRITE being low, the domino circuitry does not touch it.

Best,
Stefan

Bill Ashmanskas wrote:

Hi Stefan,

We have for some time now been using custom firmware on a custom board to read waveforms out of DRS4 chips.  Now we are working on cascaded readout mode, 4 channels @ 2048 samples, WSREG=0x55, in order to allow for longer trigger latency.

Doing a testbench simulation of the FPGA code raised a question for me:  Do I need to deassert DENABLE while I shift a new 8-bit value into the write-shift register?  What happens if, during the few-hundred nanoseconds it takes to shift 8 bits into the register, the domino wave crosses cell 768, thereby shifting the write-shift register left by one bit?  Is this shifting suppressed when A=0b1101?  Or does the update of the actual write-shift register occur only once, after the 8th SRCLK cycle?  (Maybe one is really shifting bits into a shadow register that is copied all at once into the actual register?)

I notice in simulating your drs4_eval5_app.vhd that if one sets bit 27 ("drs_ctl_dactive") of register 0 (do not deassert DENABLE on trigger), then starts the domino wave (set bit 0 of register 0), then issues a software trigger, then later writes to register 5 (config register, wsreg, etc.), DENABLE is not in fact deasserted during the time when A=0b1100 (conf_setup, conf_strobe) or when A=0b1101 (wsr_setup, wsr_strobe).

But my simulation testbench includes a simplified Verilog model of my interpretation of the DRS4 data sheet, and my simulated DRS4 happened to cause the write-shift register to shift (256 samples before DTAP toggled) during your "wsr_strobe" FSM state, thus corrupting the value that was being shifted into the WSREG via SRIN and SRCLK.

So I'm curious:  to be safe, should one deassert DENABLE before updating the write-shift register, or is it safe to update it even while the domino wave is active and looping?  It seems easy enough to be safe, since we should only need to write to the WSREG once during the setup phase and then let it loop forever.

Many thanks,

Bill

    Reply  Tue Aug 27 09:14:03 2019, Stefan Ritt, DRS4 

Is a 5 GSPS oscilloscope suitable for use with Silicon surface barier detectors?

chinmay basu wrote:

Is DRS4 suitable for use with Silicon surface barrier detectors?

 

    Reply  Mon Oct 14 10:14:46 2019, Stefan Ritt, how to acquire the stop position with channel cascading 

You first set A3-A0, on the next clock cycle you issue pulses on srclk, and about 10ns after each clock pulse the output shows up at srout. Best is to verity this with an oscilloscope.

The radout of the shift register is independent of the readout mode, so you can use with with MUXOUT as well.

Stefan

Danyang wrote:

Hi Steffan,

       In DSR4 DATASHEET Rev.0.9 Page13,  there is a paragraph "If the DRS4 is configured for channel cascading or daisy chaining, it is necessary to know which the current channel is where the sampling has been stopped. This can be
determined by addressing the Write Shift Register withA3-A0 = 1101b and by applying clock pulses to the SRCLK input ...".

       My question is the timing details about srclk, srout, A3-A0 in the above control and its timing relation with stop shift register (Figure 15).  And can this configuration be used in the full readout mode with output MUXOUT?               

Best Regards,
Danyang (sun2222@mail.ustc.edu.cn)

 

    Reply  Mon Oct 14 12:56:13 2019, Stefan Ritt, how to acquire the stop position with channel cascading 

Note that you have to read out the Write Shift Register only if you do channel cascading, e.g. configuring the chip with 4x2048 bins by setting the Write Shift Register to 01010101b. Then the Write Shift Register tells you in which 1024-bin segment the Domino Wave has been stopped. If you use the normal 8x1024 bin mode, you don't have to read out the Write Shift Register since it continas only 1's. 

Stefan

Danyang wrote:

I tried the logic in my designed board.  The results are shown in the picture: Srout keeps low when A3-A0  is set to 1101 and srclk is set as you mentioned. And the drs4 chip does not output sine wave in such configuration.

Srout signal only reacts after the rsrload signal is pulled high and A3-A0 is not 1101.  

The number of srclk is not enough?  Is there any recommended time to configure the command?

 

Best Regards,
Danyang

 

Stefan Ritt wrote:

You first set A3-A0, on the next clock cycle you issue pulses on srclk, and about 10ns after each clock pulse the output shows up at srout. Best is to verity this with an oscilloscope.

The radout of the shift register is independent of the readout mode, so you can use with with MUXOUT as well.

Stefan

Danyang wrote:

Hi Steffan,

       In DSR4 DATASHEET Rev.0.9 Page13,  there is a paragraph "If the DRS4 is configured for channel cascading or daisy chaining, it is necessary to know which the current channel is where the sampling has been stopped. This can be
determined by addressing the Write Shift Register withA3-A0 = 1101b and by applying clock pulses to the SRCLK input ...".

       My question is the timing details about srclk, srout, A3-A0 in the above control and its timing relation with stop shift register (Figure 15).  And can this configuration be used in the full readout mode with output MUXOUT?               

Best Regards,
Danyang (sun2222@mail.ustc.edu.cn)

 

 

 

    Reply  Mon Oct 14 15:27:09 2019, Stefan Ritt, how to acquire the stop position with channel cascading 

If you configure the Write Shift Register with 01010101b, then all you have to do after a trigger is to set A3-A0 to 1101. The WSROUT pin shows you then either ther state 01010101b or 10101010b, you the pin should be 1 or 0, and that's all you need. The Write Shift Register is NOT routed to the SROUT pin, you only see it at the WSROUT pin.

Stefan

Danyang wrote:

Yes, firstly I configured the chip with 4x2048 bins by setting the Write Shift Register to 01010101b, A3-A0 keeps 1101----> secondly I enabled the domino wave, wait  some time for stable,  A3-A0 keeps 1111  ---->thirdly stops the domino wave when the trigger comes, A3-A0 keeps 1101 (or 1010, 0000)----> forthly send the clock pulse to the srclk pin, A3-A0 keeps 1101,  srout pin keeps low----> fifthly enable rsrload, A3-A0 (0000-1000),  srout pin reacts nomally.   I think the cascading is worked when I checked the waveform on the oscilloscope. Is there any step I missed?

Best Regards,
Danyang

Stefan Ritt wrote:

Note that you have to read out the Write Shift Register only if you do channel cascading, e.g. configuring the chip with 4x2048 bins by setting the Write Shift Register to 01010101b. Then the Write Shift Register tells you in which 1024-bin segment the Domino Wave has been stopped. If you use the normal 8x1024 bin mode, you don't have to read out the Write Shift Register since it continas only 1's. 

Stefan

Danyang wrote:

I tried the logic in my designed board.  The results are shown in the picture: Srout keeps low when A3-A0  is set to 1101 and srclk is set as you mentioned. And the drs4 chip does not output sine wave in such configuration.

Srout signal only reacts after the rsrload signal is pulled high and A3-A0 is not 1101.  

The number of srclk is not enough?  Is there any recommended time to configure the command?

 

Best Regards,
Danyang

 

Stefan Ritt wrote:

You first set A3-A0, on the next clock cycle you issue pulses on srclk, and about 10ns after each clock pulse the output shows up at srout. Best is to verity this with an oscilloscope.

The radout of the shift register is independent of the readout mode, so you can use with with MUXOUT as well.

Stefan

Danyang wrote:

Hi Steffan,

       In DSR4 DATASHEET Rev.0.9 Page13,  there is a paragraph "If the DRS4 is configured for channel cascading or daisy chaining, it is necessary to know which the current channel is where the sampling has been stopped. This can be
determined by addressing the Write Shift Register withA3-A0 = 1101b and by applying clock pulses to the SRCLK input ...".

       My question is the timing details about srclk, srout, A3-A0 in the above control and its timing relation with stop shift register (Figure 15).  And can this configuration be used in the full readout mode with output MUXOUT?               

Best Regards,
Danyang (sun2222@mail.ustc.edu.cn)

 

 

 

 

 

    Reply  Fri Oct 25 16:39:07 2019, Stefan Ritt, Computing corrected time from binary data...what is t_0,0? 

t0,0 refers to the time of cell #0 of channel #0. So basically you keep channel 0 fixed, calculate the difference of each channel's cell #0 in respect to channel 0, and align all channels except channel 0 so that their cell #0 has the same value. There is an inconsistency between the channel numbering. The formula uses 0...3 and the manual says "channel 1" but it means actually the first channel, which uses index "0".

Stefan

John Jendzurski wrote:

In the equations for computing the corrected time for channels other than channel 1, does anyone know what the term t0,0 refers to?  This is the last term in the last equation on page 24 of DRS4 Evaluation Board User’s Manual, Board Revision 5 as of January 2014, Last revised: April 27, 2016.

Screenshot from User's Manual is attached below.

Thank you!

 

    Reply  Fri May 22 12:53:33 2020, Stefan Ritt, DRS4 Evaluation board control tool 'drscl' with macro file 

There is an example program in the distribution under software/drscl/drs_exam.cpp which is a stand-alone program to do what you need. It uses the C library coming with the distribution. It configureres the board, defines a trigger, and then writes a few waveforms into a file. You can use it as a starting point for your development. If you need any other language, you have to develop bindings to the C library.

Stefan

Keita Mizukoshi wrote:

Dear experts,

 

I would like to use DRS4 evaluation board as DAQ system for small, table-top experiment.

I need waveforms capture as binary file on some trigger based on command line without GUI.

I found `drscl` tool in official software, but it require interactive command. I'd rather use static macro or so on to control DAQ as same behaviour in each time.

I guess, experts are thinking users should develop DAQ code by themselves for their experiment specifically, but my request is very common so someone has already developed these tool.

 

Best regards,

Keita

 

    Reply  Fri May 22 13:24:51 2020, Stefan Ritt, Type check at DOFrame.h in official software 

The software is a bit outdated, I will soon make a new release. 

In meantime, you can replace that like with

bool GetRefclk(int board) { return m_refClk[board]; }

Best,
Stefan 

Keita Mizukoshi wrote:

Hi,

 

I've failured to compile official software. The cause is the following line.

DOFrame.h L.111    bool GetRefclk()        { return m_refClk > 0; }

 

m_refClk is pointer to bool. I guess these line is for null-check of the pointer.

Can I replace the following line as 

bool GetRefclk()        { return m_refClk != nullptr; }

?

The latest compilers may not accept C-style check.

 

My compiler version is

Apple clang version 11.0.3 (clang-1103.0.32.59)
Target: x86_64-apple-darwin19.4.0
Thread model: posix
InstalledDir: /Applications/Xcode.app/Contents/Developer/Toolchains/XcodeDefault.xctoolchain/usr/bin

Best regards,

Keita

 

    Reply  Tue May 26 12:44:16 2020, Stefan Ritt, Domino wave Screenshot_2020-05-26_at_12.43.40_.png

Look at the attached picture. For simplicity, only 4 cells are open and tracking the input signal. Time is flowing from top to bottom. So initially, a train of 4 cells is open. When it's stopped, the train stops not immediately, but kind of "runs against a wall" at the stop cell. So each cell is open for four time ticks effectively, and you can use all 1024 cells. 

 

xggg wrote:

Hi Stefan,

According to the datasheet DRS_rev09, the write signal is always 16 cells wide. So when the domino wave runs in infinite mode and be stopped by setting DENABLE low , there are always 16 cells capicitors tracking the input signal . It means that the effective sample cells is 1024-16=1008? That's confusing. 

 

    Reply  Mon Aug 31 10:52:42 2020, Stefan Ritt, Dynamic Range Evaluation Board and Software 

You cannot go below -0.5V for the inputs, since the board does not have an internal negative power supply, which would be necessary for that. If you have -0.8V pulses, the easiest is to use a passive inverter at the input to convert it to a 0.8V pulse.

Stefan

Hans Steiger wrote:

Dear Evaluation Board Team,

 

currently I am facing the problem of digitizing pulses with an amplitude of -0.6V to -0.8V. As the dynamic range of the board is 1Vpp, this should be feasible. However, I do not know how to set in the software a correct range. I see only -0.5V/0.5V, and the two positive options. Normally I would use -0.5V/0.5V and give the thing an offset of 0.4V or so? Is this possible? Where can I set such a offset?

 

All the best,

Hans

 

    Reply  Mon Aug 31 17:17:30 2020, Stefan Ritt, Channel Cascading Screenshot_2020-08-31_at_16.52.28_.png

If you have a board with cascading option, it should show the "combined" option in the 2048-bin option enabled (not grayed), as in the attached screen shot. If the 2048-bin option is all greyed out, the system does not recognize the cascading option. If your board has a sticker "2048 bin" and you still see the 2048-bin option greyed out, it might mean that a resistor on that board has been forgotten. If you do not see the "2048 bin" sticker on your board, you might not have a board with cascading option. So please check that. If the resistor is really missing, you can send us the board and we will add it.

Stefan

Hans Steiger wrote:

Dear All,

I have a board with Channel Cascading Option. I have the problem, that it seems to be impossible to run all 4 Channels simultaneously for digitizing pulses. I can just run even or odd channels but not even and odd ones? If I run in combined option, My question: If a board comes with this combined option, is it still usable as a 4Ch Digitizer but with 1024bin traces?

 

All the best,

 

Hans

 

    Reply  Wed Oct 7 10:56:03 2020, Stefan Ritt, External triggering 

The trigger is there only to trigger the chip, but cannot be used as a precise time reference. If you want to measure precise timing, do this always BETWEEN two inputs, never between an input and the trigger. You might want to split and delay your trigger signal and feed one copy to another input of the evaluation board as your reference.

Stefan

Elmer Grundeman wrote:

Dear all,

I had a question about timing jitter and external triggering.

I trigger the board externally with a 3V pulse from a DG645 delay generator and as a test I use the gated charge function to integrate another pulse of the DG which goes into channel 1 (the timing jitter between different outputs of the DG is on the order of ~25 picoseconds).

The issue I’m encountering is that the signal on channel 1 is jittering in time with ~1 ns, which means the signal is jittering with respect to my integration gate (point A and B). If I look at the data it always starts at t = 0.000 but my signal (pulse) moves around in time.

If I don’t use the external trigger but trigger on channel 1 directly the signal does not move with respect to the gate, but I can see the start and end of the trace move in time. If I look at the data the first data point is not at t = 0.000 but some other time, which jitters with ~1 ns.

I did repeat the voltage and timing calibration, but that did not help either.

Do you know where this jitter comes from and if I can get rid of it?

Best regards,

 

Elmer

 

    Reply  Tue Oct 27 13:37:23 2020, Stefan Ritt, Timing diagram of SROUT/SRIN signal to write/read a write shift register Screenshot_2020-10-27_at_13.45.39_.png

Dear Seiya,

1) That's correct. SRIN is ampled at the falling edge. Pleae make sure to obey the hold-time as written in the datasheet. P.12, Fig. 11: SRIN must be stable before the falling edge of SRCLK and tH after the falling clock. tH is 5ns according to table 1. 

2) The write shift register is a 8-bit shift register, with an input, output and clock. After the first clock pulse, the 7th bit is shown, after the second clock pulse the 6th bit and so on. You you need 8 clock pulses to read the whole register. At the same time you write to the register, so what ever is present at SRIN will replace the old 8 bits of that register.

3) No this is not possible. When you read the register, you write to it at the same time. One possibilty is to connect SROUT to SRIN during that (maybe via the FPGA). Then you have a circular register wich is the same after each 8 clock pulses.

For your reference, I posted a commercial D-Flip Flop (TI SNHCS72). The DRS4 write shift register is a simple array of 8 such registers, with no CLR or PRE, where SROUT is Q of the last Flip Flop.

Best,
Stefan

Seiya Nozaki wrote:

Dear Stefan,

I have questions about the timing diagram of SROUT/SRIN signal to write/read a write shift register.
1) Value of SRIN signal is saved at the falling edge of SRCLK, correct? (It is written in datasheet, page12, "Bits are latched into the shift register on the falling edge of SRCLK")
2) When are 8-bits of write shift register shown through SROUT? At ridging edges of SRCLK? and with additional delay(~10ns)? or falling edges?
3) In my understanding, when SRCLK is sent to DRS4, we can read and write the values in parallel, right? If so, is it possible just to read the registers without updating the registers?

[Background]
We have two modes to set the write shift register, the first one is to always reconnect to the first line and another one is to reconnect to the same line as when DWRITE goes to Low.
We can read/write the write shift register with the first mode (channel reset, page1). But we rarely face the problem of write shift register, unexpected values are written, with the second mode. With this mode, SROUT signal is sent back to DRS from FPGA as SRIN to write the same value on the write shift register. So there is a small delay(~10ns) due to the routing (DRS->FPGA->DRS, page2). It seems SRIN signal is not stable around the falling edges of SRCLK, thus it could cause that unexpected values are written in write shifter register.
To understand the situation clearly, I'd like to know the answer to the above questions.

Thank you.

Best regards,
Seiya

 

    Reply  Tue Oct 27 15:24:38 2020, Stefan Ritt, Timing diagram of SROUT/SRIN signal to write/read a write shift register 

This is a static shift register, so you can make the clock as slow as you want. Actually I don't use a "clock", I just use a data pin I control via a state machine in the VHDL code. This way I have more control over the edges. I need several (internal) clock cycles to produce one SRCLK clock cycle, but that does not matter for the DRS.

Stefan

Seiya Nozaki wrote:

Dear Stefan,

Thank you for your reply.
SRIN is directly connected to SROUT via FPGA for now, but it is unstable for the timing between clock and SRIN depending on the firmware logic.
We want to make our system more robust, so we are thinking to use a clock with a lower frequency (let's say 16.6 MHz) or change the duty cycle of a clock to keep more time between the rising edge and falling edge of a clock. This change is just for reading/writing the write shift register, we will use a 33 MHz clock for the analog readout in any case.
If we change like above, are there any concerns from the DRS4 side?

Best,
Seiya

Stefan Ritt wrote:

Dear Seiya,

1) That's correct. SRIN is ampled at the falling edge. Pleae make sure to obey the hold-time as written in the datasheet. P.12, Fig. 11: SRIN must be stable before the falling edge of SRCLK and tH after the falling clock. tH is 5ns according to table 1. 

2) The write shift register is a 8-bit shift register, with an input, output and clock. After the first clock pulse, the 7th bit is shown, after the second clock pulse the 6th bit and so on. You you need 8 clock pulses to read the whole register. At the same time you write to the register, so what ever is present at SRIN will replace the old 8 bits of that register.

3) No this is not possible. When you read the register, you write to it at the same time. One possibilty is to connect SROUT to SRIN during that (maybe via the FPGA). Then you have a circular register wich is the same after each 8 clock pulses.

For your reference, I posted a commercial D-Flip Flop (TI SNHCS72). The DRS4 write shift register is a simple array of 8 such registers, with no CLR or PRE, where SROUT is Q of the last Flip Flop.

Best,
Stefan

Seiya Nozaki wrote:

Dear Stefan,

I have questions about the timing diagram of SROUT/SRIN signal to write/read a write shift register.
1) Value of SRIN signal is saved at the falling edge of SRCLK, correct? (It is written in datasheet, page12, "Bits are latched into the shift register on the falling edge of SRCLK")
2) When are 8-bits of write shift register shown through SROUT? At ridging edges of SRCLK? and with additional delay(~10ns)? or falling edges?
3) In my understanding, when SRCLK is sent to DRS4, we can read and write the values in parallel, right? If so, is it possible just to read the registers without updating the registers?

[Background]
We have two modes to set the write shift register, the first one is to always reconnect to the first line and another one is to reconnect to the same line as when DWRITE goes to Low.
We can read/write the write shift register with the first mode (channel reset, page1). But we rarely face the problem of write shift register, unexpected values are written, with the second mode. With this mode, SROUT signal is sent back to DRS from FPGA as SRIN to write the same value on the write shift register. So there is a small delay(~10ns) due to the routing (DRS->FPGA->DRS, page2). It seems SRIN signal is not stable around the falling edges of SRCLK, thus it could cause that unexpected values are written in write shifter register.
To understand the situation clearly, I'd like to know the answer to the above questions.

Thank you.

Best regards,