  263   Thu Jul 4 08:32:11 2013 tmiron alonadd an average ability to the Scope

Stefan Ritt wrote:

tmiron alon wrote:

Stefan Ritt wrote:

tmiron alon wrote:

Hallo,
I'm using DRS4 Evaluation Board Rev 4.0 and I'm trying to change the output of the samples to be an average of  #  measurements (1000 or even more)
during the process I have encountered some difficulties I hope you will be able to help me  with:

1. the DRS chip have 8 channels but the Evaluation board have only 4 channels. does the default mode of the DRS in the Evaluation Board is 1024 bins for each channel or 2048?

2. in the readout mode, does it samples all the 1024 bins waveform from a channel and then move to the next one, or after each bin it move to the next channel?

3. In the file "drs4_eval4_app.vhd", I have a problem finding the names of the signals that represents the registers bits which tell me what is the number of the bin (1-1024) the ADC is reading from the DRS, and the signals

that represents registers A0-A3. can you send me their names? 

 

4. In another matter- is the -0.5V to 0.5V is the  upper and lower  limit of the input (or just a working range), and if not what is the limit for AC?  is there a fuse on the board in case of overload from the input? (I didn't see  it in the User's Manual, but I didn't know if you will mention it there in case there is one).

thanks in advance and have a nice day,

Tmiron

1. All 8 channels are read out, but only 4 are displayed in the oscilloscope.

2. It reads all 1024 bins from a channel, then switch to the next channel.

3. The ADC readout happens in lines 1576+. The register for the sample count is drs_sample_count, and the signal for the address is drs_addr.

4. The evaluation board manual explicitly mentions the maximum allowed input range on page 5.

/Stefan 

 Hi Stefan,

I'm trying to add the averaging feature to the oscilloscope instead. I have a basic knowledge in C++ so I'm having problem finding the initial place (places) of the data acquisition. furthermore, I saw you already made an averaging possibility for some of the features in the program at the measure window.
Can you please tell we in which of the programs of the Scope (and in what line if possible) can I find the initial\main data acquisition so I could follow it from there, and where is the location of the averaging function you already made?

have a nice week,

Tmiron

The averaging possibility in the measure window is just a "place holder" for future extensions, which are not yet implemented.

A good point to implement averaging is in the function Osci::ReadWaveforms() in Osci.cpp.

/Stefan 

Hi Stefan,

I have been trying to compile my changes to the program, but have a problem with the "wxWidgets". can you tell me which version of the wxWidgets and which c++ program did you use to built the drs4's scope program?

I'm using "microsoft visual c++ 2010 express", and the closest I got to compile something was with the wxWidgets 2.8.12 version (It's tell me I'm missing the file "wxmsw28d_core.lib" although I built it).

have a nice day,

Tmiron.

  264   Thu Jul 4 08:54:25 2013 Stefan Rittadd an average ability to the Scope

tmiron alon wrote:

Stefan Ritt wrote:

tmiron alon wrote:

Stefan Ritt wrote:

tmiron alon wrote:

Hallo,
I'm using DRS4 Evaluation Board Rev 4.0 and I'm trying to change the output of the samples to be an average of  #  measurements (1000 or even more)
during the process I have encountered some difficulties I hope you will be able to help me  with:

1. the DRS chip have 8 channels but the Evaluation board have only 4 channels. does the default mode of the DRS in the Evaluation Board is 1024 bins for each channel or 2048?

2. in the readout mode, does it samples all the 1024 bins waveform from a channel and then move to the next one, or after each bin it move to the next channel?

3. In the file "drs4_eval4_app.vhd", I have a problem finding the names of the signals that represents the registers bits which tell me what is the number of the bin (1-1024) the ADC is reading from the DRS, and the signals

that represents registers A0-A3. can you send me their names? 

 

4. In another matter- is the -0.5V to 0.5V is the  upper and lower  limit of the input (or just a working range), and if not what is the limit for AC?  is there a fuse on the board in case of overload from the input? (I didn't see  it in the User's Manual, but I didn't know if you will mention it there in case there is one).

thanks in advance and have a nice day,

Tmiron

1. All 8 channels are read out, but only 4 are displayed in the oscilloscope.

2. It reads all 1024 bins from a channel, then switch to the next channel.

3. The ADC readout happens in lines 1576+. The register for the sample count is drs_sample_count, and the signal for the address is drs_addr.

4. The evaluation board manual explicitly mentions the maximum allowed input range on page 5.

/Stefan 

 Hi Stefan,

I'm trying to add the averaging feature to the oscilloscope instead. I have a basic knowledge in C++ so I'm having problem finding the initial place (places) of the data acquisition. furthermore, I saw you already made an averaging possibility for some of the features in the program at the measure window.
Can you please tell we in which of the programs of the Scope (and in what line if possible) can I find the initial\main data acquisition so I could follow it from there, and where is the location of the averaging function you already made?

have a nice week,

Tmiron

The averaging possibility in the measure window is just a "place holder" for future extensions, which are not yet implemented.

A good point to implement averaging is in the function Osci::ReadWaveforms() in Osci.cpp.

/Stefan 

Hi Stefan,

I have been trying to compile my changes to the program, but have a problem with the "wxWidgets". can you tell me which version of the wxWidgets and which c++ program did you use to built the drs4's scope program?

I'm using "microsoft visual c++ 2010 express", and the closest I got to compile something was with the wxWidgets 2.8.12 version (It's tell me I'm missing the file "wxmsw28d_core.lib" although I built it).

have a nice day,

Tmiron.

I guess it should work with 2.8.12. Try to compile their examples, and consult their documentation if it does not work. I cannot give support for a package I have not written.

/Stefan 

  265   Thu Jul 4 09:07:24 2013 tmiron alonadd an average ability to the Scope

Stefan Ritt wrote:

tmiron alon wrote:

Stefan Ritt wrote:

tmiron alon wrote:

Stefan Ritt wrote:

tmiron alon wrote:

Hallo,
I'm using DRS4 Evaluation Board Rev 4.0 and I'm trying to change the output of the samples to be an average of  #  measurements (1000 or even more)
during the process I have encountered some difficulties I hope you will be able to help me  with:

1. the DRS chip have 8 channels but the Evaluation board have only 4 channels. does the default mode of the DRS in the Evaluation Board is 1024 bins for each channel or 2048?

2. in the readout mode, does it samples all the 1024 bins waveform from a channel and then move to the next one, or after each bin it move to the next channel?

3. In the file "drs4_eval4_app.vhd", I have a problem finding the names of the signals that represents the registers bits which tell me what is the number of the bin (1-1024) the ADC is reading from the DRS, and the signals

that represents registers A0-A3. can you send me their names? 

 

4. In another matter- is the -0.5V to 0.5V is the  upper and lower  limit of the input (or just a working range), and if not what is the limit for AC?  is there a fuse on the board in case of overload from the input? (I didn't see  it in the User's Manual, but I didn't know if you will mention it there in case there is one).

thanks in advance and have a nice day,

Tmiron

1. All 8 channels are read out, but only 4 are displayed in the oscilloscope.

2. It reads all 1024 bins from a channel, then switch to the next channel.

3. The ADC readout happens in lines 1576+. The register for the sample count is drs_sample_count, and the signal for the address is drs_addr.

4. The evaluation board manual explicitly mentions the maximum allowed input range on page 5.

/Stefan 

 Hi Stefan,

I'm trying to add the averaging feature to the oscilloscope instead. I have a basic knowledge in C++ so I'm having problem finding the initial place (places) of the data acquisition. furthermore, I saw you already made an averaging possibility for some of the features in the program at the measure window.
Can you please tell we in which of the programs of the Scope (and in what line if possible) can I find the initial\main data acquisition so I could follow it from there, and where is the location of the averaging function you already made?

have a nice week,

Tmiron

The averaging possibility in the measure window is just a "place holder" for future extensions, which are not yet implemented.

A good point to implement averaging is in the function Osci::ReadWaveforms() in Osci.cpp.

/Stefan 

Hi Stefan,

I have been trying to compile my changes to the program, but have a problem with the "wxWidgets". can you tell me which version of the wxWidgets and which c++ program did you use to built the drs4's scope program?

I'm using "microsoft visual c++ 2010 express", and the closest I got to compile something was with the wxWidgets 2.8.12 version (It's tell me I'm missing the file "wxmsw28d_core.lib" although I built it).

have a nice day,

Tmiron.

I guess it should work with 2.8.12. Try to compile their examples, and consult their documentation if it does not work. I cannot give support for a package I have not written.

/Stefan 

My problem is not with wxWidgets itself but with links your program is doing to the wxWidgets. if you can please tell me the wxWidget version you used, I'll use this one instead the version I'm using and I think it will solve the problem.

Tmiron.

  266   Thu Jul 4 09:17:31 2013 Stefan Rittadd an average ability to the Scope

tmiron alon wrote:

 

My problem is not with wxWidgets itself but with links your program is doing to the wxWidgets. if you can please tell me the wxWidget version you used, I'll use this one instead the version I'm using and I think it will solve the problem.

Tmiron.

I currently use version 2.9.4, but chances are high that it will work with 2.8.12.

/Stefan 

  267   Thu Jul 4 10:01:06 2013 tmiron alonadd an average ability to the Scope

Stefan Ritt wrote:

tmiron alon wrote:

 

My problem is not with wxWidgets itself but with links your program is doing to the wxWidgets. if you can please tell me the wxWidget version you used, I'll use this one instead the version I'm using and I think it will solve the problem.

Tmiron.

I currently use version 2.9.4, but chances are high that it will work with 2.8.12.

/Stefan 

 but was version 2.9.4 the one you used when you built  Version 4.0.0 of the DRS4 oscilloscope program, or was it an erilar version, and if so, do you remeber what version it was?

tmiron.

  268   Thu Jul 4 10:14:32 2013 Stefan Rittadd an average ability to the Scope

tmiron alon wrote:

Stefan Ritt wrote:

tmiron alon wrote:

 

My problem is not with wxWidgets itself but with links your program is doing to the wxWidgets. if you can please tell me the wxWidget version you used, I'll use this one instead the version I'm using and I think it will solve the problem.

Tmiron.

I currently use version 2.9.4, but chances are high that it will work with 2.8.12.

/Stefan 

 but was version 2.9.4 the one you used when you built  Version 4.0.0 of the DRS4 oscilloscope program, or was it an erilar version, and if so, do you remeber what version it was?

tmiron.

I used 2.8.12 for Version 4.0.0.

/Stefan 

  274   Tue Jul 16 10:02:28 2013 tmiron alonadd an average ability to the Scope

Stefan Ritt wrote:

tmiron alon wrote:

Stefan Ritt wrote:

tmiron alon wrote:

 

My problem is not with wxWidgets itself but with links your program is doing to the wxWidgets. if you can please tell me the wxWidget version you used, I'll use this one instead the version I'm using and I think it will solve the problem.

Tmiron.

I currently use version 2.9.4, but chances are high that it will work with 2.8.12.

/Stefan 

 but was version 2.9.4 the one you used when you built  Version 4.0.0 of the DRS4 oscilloscope program, or was it an erilar version, and if so, do you remeber what version it was?

tmiron.

I used 2.8.12 for Version 4.0.0.

/Stefan 

 Hi Stefan.

I managed to add the averaging possibility to the oscilloscope program, but it runs too slow, so now I'm tring to add this option to the example file you attached to the program (drs_exam.cpp).
My problem is that when I run the file is not passing  the stage of getting the trigger. (it just wait after writing the line "Waiting for trigger...")
When testing the software, I work with a signal of up to 10 MHz, limited in voltage to the DRS Eval 4 ability, and I connect the signal to channel 1,from it by default is supposed to get the trigger. I also tried to change the channel settings to other channels or external trigger, and use them,  but it still not working.

I found out that the problem is related to the line "while (b->IsBusy());" in the code, but I have a problem understanding what the function "IsBusy" do and how it's work.

Can you please explain me a bit what this function does?

have a nice day,

Tmiron

  275   Tue Jul 16 16:25:43 2013 Stefan Rittadd an average ability to the Scope

tmiron alon wrote:

Hi Stefan.

I managed to add the averaging possibility to the oscilloscope program, but it runs too slow, so now I'm tring to add this option to the example file you attached to the program (drs_exam.cpp).
My problem is that when I run the file is not passing  the stage of getting the trigger. (it just wait after writing the line "Waiting for trigger...")
When testing the software, I work with a signal of up to 10 MHz, limited in voltage to the DRS Eval 4 ability, and I connect the signal to channel 1,from it by default is supposed to get the trigger. I also tried to change the channel settings to other channels or external trigger, and use them,  but it still not working.

I found out that the problem is related to the line "while (b->IsBusy());" in the code, but I have a problem understanding what the function "IsBusy" do and how it's work.

Can you please explain me a bit what this function does?

have a nice day,

Tmiron

The function b->IsBusy() returns true if there has been no trigger. So "while (b->IsBusy());" is simply an endless loop until the board triggered. Please check your signals and trigger configuration. Maybe you have to adjust the trigger level or polarity via b->SetTriggerLevel(). Try signals with lower event rates. Verify that the board triggers with the DRSOsc program, you can then read off the optimal trigger level from there. 

/Stefan

  280   Sun Jul 28 09:52:25 2013 tmiron alonadd an average ability to the Scope

Stefan Ritt wrote:

tmiron alon wrote:

Hi Stefan.

I managed to add the averaging possibility to the oscilloscope program, but it runs too slow, so now I'm tring to add this option to the example file you attached to the program (drs_exam.cpp).
My problem is that when I run the file is not passing  the stage of getting the trigger. (it just wait after writing the line "Waiting for trigger...")
When testing the software, I work with a signal of up to 10 MHz, limited in voltage to the DRS Eval 4 ability, and I connect the signal to channel 1,from it by default is supposed to get the trigger. I also tried to change the channel settings to other channels or external trigger, and use them,  but it still not working.

I found out that the problem is related to the line "while (b->IsBusy());" in the code, but I have a problem understanding what the function "IsBusy" do and how it's work.

Can you please explain me a bit what this function does?

have a nice day,

Tmiron

The function b->IsBusy() returns true if there has been no trigger. So "while (b->IsBusy());" is simply an endless loop until the board triggered. Please check your signals and trigger configuration. Maybe you have to adjust the trigger level or polarity via b->SetTriggerLevel(). Try signals with lower event rates. Verify that the board triggers with the DRSOsc program, you can then read off the optimal trigger level from there. 

/Stefan

Hi satefan,

I did some debug on the DRSOsc program and saw that everywhere you used the function "IsBusy()", you used it with "SoftTrigger()", which (to my understanding) create a software trigger. did you used another function, other than "ISBusy", that check the hardware trigger?

have a nice day,

Tmiron.

  281   Mon Jul 29 06:04:45 2013 Stefan Rittadd an average ability to the Scope

Hi satefan,

I did some debug on the DRSOsc program and saw that everywhere you used the function "IsBusy()", you used it with "SoftTrigger()", which (to my understanding) create a software trigger. did you used another function, other than "ISBusy", that check the hardware trigger?

have a nice day,

Tmiron.

No. IsBusy() is the only way the hardware trigger is checked. If there is no hardware trigger, you can issue a "fake" trigger (called software trigger) to stop the acquisition. This happens for example if you switch DRSOsc to "Auto" trigger instead of "Normal" trigger, very similar than a normal oscilloscope does.

/Stefan 

  285   Mon Aug 12 15:08:17 2013 tmiron alonadd an average ability to the Scope

Stefan Ritt wrote:

Hi satefan,

I did some debug on the DRSOsc program and saw that everywhere you used the function "IsBusy()", you used it with "SoftTrigger()", which (to my understanding) create a software trigger. did you used another function, other than "ISBusy", that check the hardware trigger?

have a nice day,

Tmiron.

No. IsBusy() is the only way the hardware trigger is checked. If there is no hardware trigger, you can issue a "fake" trigger (called software trigger) to stop the acquisition. This happens for example if you switch DRSOsc to "Auto" trigger instead of "Normal" trigger, very similar than a normal oscilloscope does.

/Stefan 

 hallo stefan,

I managed to fixed the problem. The problem was with the number I inserted to the function "b->SetTriggerSource" th get external trigger (it suppose to be "16" and not "4" or "12" as I understood from the function's comments).

Right now I'm trying to speed up the number of wavefrom  per second. I'm using your  drs_exam.cpp program you wrote as my basic program. When you wrote it you used the function "b->StartDomino()" inside the loop, which means that before every trigger you gave him a command to start the domino-wave.

From  the "DRS4 datasheet" (page 8) I understand that I can bypass the need to restart the  domino-wave by using "b->SetDominoActive(1)" and "b->SetDominoMode(1)", but when I tried it didn't work (the waveform I got every readout remain the same which means the dominowave froze after the first readout).

Am I understanding wrong or do I need to add somthing more\else so the domino-wave will not stop after each readout? is there any hazard by doing that, as mentiond in page 8 of the the datasheet?

thanks again,

Tmiron.

  286   Mon Aug 12 22:18:39 2013 Stefan Rittadd an average ability to the Scope

tmiron alon wrote:

Right now I'm trying to speed up the number of wavefrom  per second. I'm using your  drs_exam.cpp program you wrote as my basic program. When you wrote it you used the function "b->StartDomino()" inside the loop, which means that before every trigger you gave him a command to start the domino-wave.

From  the "DRS4 datasheet" (page 8) I understand that I can bypass the need to restart the  domino-wave by using "b->SetDominoActive(1)" and "b->SetDominoMode(1)", but when I tried it didn't work (the waveform I got every readout remain the same which means the dominowave froze after the first readout).

Am I understanding wrong or do I need to add somthing more\else so the domino-wave will not stop after each readout? is there any hazard by doing that, as mentiond in page 8 of the the datasheet?

The Domino wave anyhow is always active in mode 1, but you have to enable the writing to the memory cells (via the DENABLE signal). This is done with the b->StartDomino(). You are right that this call takes about 1ms over USB 2.0, so avoiding it would speed up the DAQ by about a factor of two. I have some plans to implement an automatic restart in firmware of the evaluation board, but I won't have time for that until fall of this year.

/Stefan

  863   Tue Feb 15 11:59:22 2022 Alex Myczkoapt install drs4eb

drs4b is now officially on these distributions:

https://repology.org/project/drs4eb/versions

enjoy

  479   Mon Feb 29 12:58:17 2016 Dmitry Philippovbaseline shift

Hello! My name is Dmitry. I am from SiPM Lab is NRNU MEPhI (Russia, Moscow). We bought DRS4 evaluation board V5 with firmware 21305. We use 5.0.4 build 21911 2015-11-23 software version (and before that we used 5.0.3 build 21508, 2014-10-15) with Windows 7 32bit.

We observe some strange behaviour. When we save waveforms (in xml or binary data) we see that some of them have the baseline shifted of about -5 mV.

The first picture (pic1) is 1000 waveforms which were glued in one. It is clearly see that baseline quite often has the shift.

The same effect can be seen without saving (writting): rarely when we use normal or auto trigger mode (pic3), and always in single trigger mode (pic2).

The images are attached.

Do you have any idea how it can be fixed?

 

Thanks, Dmitry.

 

  480   Mon Feb 29 13:09:29 2016 Stefan Rittbaseline shift

The baseline shift comes from some instable power supply inside the evaluation board which cannot be controlled to the mV level. In a real measurement, you usually get an additional baseline shift due to some environmental electromagnetic interferences, such as a 50 Hz signal. People fix this shifting baseline by always aquiring a small portion (10-20 samples) of the baseline before any signal from a particle detector. The signal is then corrected event-by-event by subtracting the baseline from each waveform. By doing that, you fix not only the 50 Hz noise, but also the shifting baseline you mention.

Stefan

Dmitry Philippov wrote:

Hello! My name is Dmitry. I am from SiPM Lab is NRNU MEPhI (Russia, Moscow). We bought DRS4 evaluation board V5 with firmware 21305. We use 5.0.4 build 21911 2015-11-23 software version (and before that we used 5.0.3 build 21508, 2014-10-15) with Windows 7 32bit.

We observe some strange behaviour. When we save waveforms (in xml or binary data) we see that some of them have the baseline shifted of about -5 mV.

The first picture (pic1) is 1000 waveforms which were glued in one. It is clearly see that baseline quite often has the shift.

The same effect can be seen without saving (writting): rarely when we use normal or auto trigger mode (pic3), and always in single trigger mode (pic2).

The images are attached.

Do you have any idea how it can be fixed?

 

Thanks, Dmitry.

 

 

  363   Wed Aug 13 20:17:19 2014 Roman Gredigbinary files time calibration header in drs-5.0.2
Dear Stefan

I have a problem considering binary data files.
Usually binary files start with TIME... (the time calibration header). But I observed the following reproducible problem.
1. Start to save a binary file (e.g. 001.dat) with 1000000 events. 
2. Hit the close button before this limit has reached. (so far the binary files seems to be ok) 
3. Start to save again a file with the SAME filename (and agree to replace the already existing it)
If I do it like this, the file has no time information anymore and starts directly with EHDR.

could it be that the m_evSerial counter is not reset in this specific situation?

Cheers,
Roman
  391   Thu Oct 16 16:16:12 2014 Stefan Rittbinary files time calibration header in drs-5.0.2
> Dear Stefan
> 
> I have a problem considering binary data files.
> Usually binary files start with TIME... (the time calibration header). But I observed the following reproducible problem.
> 1. Start to save a binary file (e.g. 001.dat) with 1000000 events. 
> 2. Hit the close button before this limit has reached. (so far the binary files seems to be ok) 
> 3. Start to save again a file with the SAME filename (and agree to replace the already existing it)
> If I do it like this, the file has no time information anymore and starts directly with EHDR.
> 
> could it be that the m_evSerial counter is not reset in this specific situation?
> 
> Cheers,
> Roman

This problem has also been fixed in version 5.0.3

/Stefan
  366   Tue Aug 26 14:16:26 2014 Roman Gredigbinary files with more than 4 drs board ver. 5.0.2
Dear Stefan

after having some problems with writing binary files with more than 4 drs boards (in multiboard-mode) I might 
have found the solution.
In the file src/Osci.cpp at line 838 is: 
unsigned char buffer[100000];

If I understand the binary format right, this works only with up to four boards. With the maximum number of 
boards in your specification (16 boards) and all channels switched on on all boards this array needs to have 
about 400000 entries (for the first event, where the time information is written too).

Could you please cross-check that?

Thank you very much!
Cheers,
Roman
  390   Thu Oct 16 16:15:16 2014 Stefan Rittbinary files with more than 4 drs board ver. 5.0.2
> Dear Stefan
> 
> after having some problems with writing binary files with more than 4 drs boards (in multiboard-mode) I might 
> have found the solution.
> In the file src/Osci.cpp at line 838 is: 
> unsigned char buffer[100000];
> 
> If I understand the binary format right, this works only with up to four boards. With the maximum number of 
> boards in your specification (16 boards) and all channels switched on on all boards this array needs to have 
> about 400000 entries (for the first event, where the time information is written too).
> 
> Could you please cross-check that?
> 
> Thank you very much!
> Cheers,
> Roman

This problem has been fixed in software version 5.0.3

/Stefan
  234   Mon Apr 8 18:11:02 2013 Dmitry Hitsbinary to root

 Hi,

 

Does anyone has a program that converts a binary file from drsosc  output to a ROOT tree format?

 

Thank you,

 

Dmitry.