  Draft   Wed Oct 7 13:06:34 2015 Ilja BekmanVoltage Calibration with signal on the input
  678   Fri Apr 13 18:14:07 2018 Alessio BertiVoltage and Timing Calibration in drs_exam.cpp

Hi,

we were trying to implement an automatic way to calibrate our DRS4 both in voltage and in time (we have the V5 Evaluation Board). We started from drs_exam.cpp and tried with the following lines:

/* set input range to -0.5V ... +0.5V */

b->SetInputRange(0);

b->CalibrateVolt(NULL);
b->CalibrateTiming(NULL);

While the timing calibration seems to work (we checked with drsosc executable), the voltage calibration in our test program seems not to do the same as in drsosc when pressing the button "Execute Voltage Calibration". Specifically we think that no primary calibration, secondary calibration or spike removal is applied when calling CalibrateVolt(). It seems that the methods to perform those tasks are implemented in Osci.ccp/Osci.h, but drs_exam.cpp uses objects of the class DRS (i.e. defined in DRS.cpp and DRS.h).

Is there a way to execute the voltage calibration in drs_exam.cpp in the same way performed within drsosc?

Cheers,

Alessio

  687   Fri May 4 11:56:08 2018 Stefan RittVoltage and Timing Calibration in drs_exam.cpp

Have you set the sampling frequency 

b->SetFrequency(5, true);

before the calibration?

Note there is also the "drscl" program, which is a ocmmand linke interface to the evaluation board. Start it, and do the calibration there:

/drs4eb/software/drscl$ ./drscl
DRS command line tool, Revision 21435
Type 'help' for a list of available commands.

Found DRS4 board  0 on USB, serial #2400, firmware revision 30000
B0> freq 5
B0> calib
           Enter calibration frequency [GHz]: 5
                             Enter range [V]: 0
        Enter mode [1]024 or [2]048 bin mode: 1

Please make sure that no input signal are present then hit any key
Creating Calibration of Board on USB, serial #2400                
B0> ===============================================]
B0> 

then look at the code in drscl.cpp (around line 1097).

/Stefan

Alessio Berti wrote:

Hi,

we were trying to implement an automatic way to calibrate our DRS4 both in voltage and in time (we have the V5 Evaluation Board). We started from drs_exam.cpp and tried with the following lines:

/* set input range to -0.5V ... +0.5V */

b->SetInputRange(0);

b->CalibrateVolt(NULL);
b->CalibrateTiming(NULL);

While the timing calibration seems to work (we checked with drsosc executable), the voltage calibration in our test program seems not to do the same as in drsosc when pressing the button "Execute Voltage Calibration". Specifically we think that no primary calibration, secondary calibration or spike removal is applied when calling CalibrateVolt(). It seems that the methods to perform those tasks are implemented in Osci.ccp/Osci.h, but drs_exam.cpp uses objects of the class DRS (i.e. defined in DRS.cpp and DRS.h).

Is there a way to execute the voltage calibration in drs_exam.cpp in the same way performed within drsosc?

Cheers,

Alessio

 

  657   Tue Feb 27 13:17:00 2018 Steven BlockWIndows Connection problem with drs507 SOLVED

Hello All,

I too have been struggling with trying to get the drs4 (507) to work on my windows machine and I found it to be a problem with the libusb library. My solution is as follows and has worked on multiple PC's. I ran this solution after I first plugged in the drs4 and installed 507.

Go to http://zadig.akeo.ie/ and install the corresponding software.

After that, you will need to plug in the DRS4 to your computer. From there go to ‘Options’, and select ‘List all Devices’.

Finally, choose the DRS4 evaluation board from the list and press install driver and let it run. You should be fine after that. 

Best,

Steven

  658   Tue Feb 27 13:29:47 2018 Stefan RittWIndows Connection problem with drs507 SOLVED

Dear Steven, many thanks for this information, this is very useful. I know of people having problems on Windows 10, maybe this will also help them.

Stefan

Steven Block wrote:

Hello All,

I too have been struggling with trying to get the drs4 (507) to work on my windows machine and I found it to be a problem with the libusb library. My solution is as follows and has worked on multiple PC's. I ran this solution after I first plugged in the drs4 and installed 507.

Go to http://zadig.akeo.ie/ and install the corresponding software.

After that, you will need to plug in the DRS4 to your computer. From there go to ‘Options’, and select ‘List all Devices’.

Finally, choose the DRS4 evaluation board from the list and press install driver and let it run. You should be fine after that. 

Best,

Steven

 

  695   Wed May 9 14:07:10 2018 Alec ShacklefordWIndows Connection problem with drs507 SOLVED

Thank you for this fantastic solution. I had almost reinstalled windows 7 to see if that would solve the issue!

 

All the best,

Alec

Stefan Ritt wrote:

Dear Steven, many thanks for this information, this is very useful. I know of people having problems on Windows 10, maybe this will also help them.

Stefan

Steven Block wrote:

Hello All,

I too have been struggling with trying to get the drs4 (507) to work on my windows machine and I found it to be a problem with the libusb library. My solution is as follows and has worked on multiple PC's. I ran this solution after I first plugged in the drs4 and installed 507.

Go to http://zadig.akeo.ie/ and install the corresponding software.

After that, you will need to plug in the DRS4 to your computer. From there go to ‘Options’, and select ‘List all Devices’.

Finally, choose the DRS4 evaluation board from the list and press install driver and let it run. You should be fine after that. 

Best,

Steven

 

 

  696   Mon May 14 09:21:29 2018 Alessio BertiWIndows Connection problem with drs507 SOLVED

Hi,

I have a machine with Windows 10 and the solution provided by Steven works fine. To give more details, the driver installed in my case is WinUSB (i.e. libusb, v6.1.7600.16385).

Cheers,

Alessio

Alec Shackleford wrote:

Thank you for this fantastic solution. I had almost reinstalled windows 7 to see if that would solve the issue!

 

All the best,

Alec

Stefan Ritt wrote:

Dear Steven, many thanks for this information, this is very useful. I know of people having problems on Windows 10, maybe this will also help them.

Stefan

Steven Block wrote:

Hello All,

I too have been struggling with trying to get the drs4 (507) to work on my windows machine and I found it to be a problem with the libusb library. My solution is as follows and has worked on multiple PC's. I ran this solution after I first plugged in the drs4 and installed 507.

Go to http://zadig.akeo.ie/ and install the corresponding software.

After that, you will need to plug in the DRS4 to your computer. From there go to ‘Options’, and select ‘List all Devices’.

Finally, choose the DRS4 evaluation board from the list and press install driver and let it run. You should be fine after that. 

Best,

Steven

 

 

 

  605   Sat Apr 15 03:48:31 2017 Strahinja LukicWave rotation during transfer from the board?

I don't know if this question is already documented elsewhere.

I am developing a DAQ code for the DRS evaluation board, v4 for a test beam experiment. I link parts of the existing DRS code as a library.

To understand the effect of various flags used in calls to the functions DRSBoard::GetTime() and DRSBoard::GetWave(), I performed several tests with the 100 MHz signal connected to the inputs of the chip (DRSBoard::EnableTcal()), and several tests with signals from scintillation counters.

My question is about the flag "adjustToClock" in the call to DRSBoard::GetWave(). From looking at the code, I expected it to cause the waveforms to be "rotated" to start from the trigger cell, in a similar way that the flag "rotated" in the call to DRSBoard::GetTime() does for the time array. However, "adjustToClock" seems to shift the waveforms wrongly. I.e., if I want both the time and the amplitude arrays "rotated"  to start from the trigger cell, I should set rotated=true for time and adjustToClock=false for the amplitude. This is also how these functions are called in e.g., Osci::ReadWaveforms().

Is this correct, and does this mean that the amplitude array is "rotated" already during the transfer from the board?

I am using DRS evaluation board serial #2733, firmware revision 30000.

Many thanks,

Strahinja

 

  606   Wed Apr 19 12:17:25 2017 Stefan RittWave rotation during transfer from the board?

This is correct. Actually the amplitude array is rotated already inside the DRS4 chip. So the readout starts with the stop cell plus one. If you do not do anything, the waveform is already "rotated". If you want the waveform to start with physical cell #0, you have to "unrotate" it.

Stefan

Strahinja Lukic wrote:

Is this correct, and does this mean that the amplitude array is "rotated" already during the transfer from the board?

  607   Thu Apr 20 06:30:13 2017 Strahinja LukicWave rotation during transfer from the board?

Thanks.

Strahinja

Stefan Ritt wrote:

This is correct. Actually the amplitude array is rotated already inside the DRS4 chip. So the readout starts with the stop cell plus one. If you do not do anything, the waveform is already "rotated". If you want the waveform to start with physical cell #0, you have to "unrotate" it.

Stefan

Strahinja Lukic wrote:

Is this correct, and does this mean that the amplitude array is "rotated" already during the transfer from the board?

 

  903   Wed Oct 25 19:44:25 2023 John WestmorelandWaveDREAM Design

Hello All,

Are there any design resources available for the WaveDREAM PCBA's?

Thanks In Advance,
John W.

  904   Wed Oct 25 19:47:23 2023 Stefan RittWaveDREAM Design

No. This is a proprietary design.

Best,
Stefan

  905   Wed Oct 25 19:52:33 2023 John WestmorelandWaveDREAM Design

Stefan,

Oh, didn't realize that.

Thanks!
John

Stefan Ritt wrote:

No. This is a proprietary design.

Best,
Stefan

 

  1   Mon Dec 15 13:37:38 2008 Stefan RittWelcome

 Welcome to the DRS4 Discussion Forum. This forum contains information and discussions related to the DRS4 chip. Please subscribe to this forum to receive automatic email updates. If you have any technical questions, please feel free to post it here.

  871   Sun Mar 6 17:54:47 2022 Matias SengerWhy does not trigger at higher sampling frequencies?

I have connected 3 signals to the DRS4 Evaluation Board V5 which look like this in the drsosc app:

Note that here I am sampling at 5 GS/s. Using this app everything works perfect.

Now I want to repeat this using the C++ API (which I am actually wrapping to use within Python, see here if interested https://github.com/SengerM/pydrs ) but can only make this to work at lower sampling frequencies up to 3.9 GS/s. This is how I am configuring the board followint the `drs_exam.cpp` file:

```python
board.set_sampling_frequency(Hz=SAMPLING_FREQ)
board.set_transparent_mode('on')
board.set_input_range(center=0)
board.enable_trigger(True,False) # Don't know what this line does, it was in the example `drs_exam.cpp`.
board.set_trigger_source('ch4')
board.set_trigger_level(volts=-.1)
board.set_trigger_polarity(edge='falling')
board.set_trigger_delay(seconds=TRIGGER_DELAY)

```

The full code is here https://github.com/SengerM/pydrs/blob/master/tests/test_drs.py but anyway, my previous snippet can be considered as pseudo-code and if more details needed I can provide.

This is what I get as I increase the sampling frequency:

Up to 3.95 GS/s it works perfectly. At >= 4 GS/s it just never triggers. A few times I was able to make this work at 4 and 5 GS/s playing with the trigger delay, but this seemed to be some kind of random luck because I was not able to replicate it even with the same values.

Any help is appreciated.

Best,
Matías.

  872   Mon Mar 7 08:45:32 2022 Stefan RittWhy does not trigger at higher sampling frequencies?

Unfortunately I have not idea what the problem could be. In principle the trigger should be independent of the sampling speed, since the trigger is only made with a discriminator and a flip-flop. The hardware must be ok since you see the trigger with the oscillocope app. All you can do is to go through the sorce code of the oscilloscope app, especially drsosc/Osic.cpp::ScanBoards(), SetTriggerLevel(), SetTriggerPolariy() etc. to make sure you do the same calls as the oscilloscope app.

Stefan

  875   Tue Mar 8 00:25:56 2022 Matias SengerWhy does not trigger at higher sampling frequencies?

I have seen in the app that the trigger source buttons do something different than the "or" and "transparent trigger" buttons:

If I enable the setup from the right, i.e. OR in CH4 and "Enable Transparent Trigger" the app stops triggering. This is the configuration that seems to be applied in the `drs_exam.cpp` code if I am not mistaken. For some reason in that code it still triggers (I have modified the code to trigger on CH4 instead of CH1 and the trigger level, polarity, etc.).

What does the button in the left actually do? The circular checkbox with the "4" I mean. This is the trigger configuration I want to get in the C++ code.

I also don't know what the function `DRSBoard::EnableTrigger` does, what is the meaning of `flag1` and `flag2`? In my code there is a call to this function which I copied from the example.

Stefan Ritt wrote:

Unfortunately I have not idea what the problem could be. In principle the trigger should be independent of the sampling speed, since the trigger is only made with a discriminator and a flip-flop. The hardware must be ok since you see the trigger with the oscillocope app. All you can do is to go through the sorce code of the oscilloscope app, especially drsosc/Osic.cpp::ScanBoards(), SetTriggerLevel(), SetTriggerPolariy() etc. to make sure you do the same calls as the oscilloscope app.

Stefan

  876   Tue Mar 8 12:20:00 2022 Matias SengerWhy does not trigger at higher sampling frequencies?

Sorry for the spam. Just want to let you know that I was able to solve the problem, it was all due to a `float` being casted as `int` in the Python binding. Now it works like a charm.

Matias Senger wrote:

I have seen in the app that the trigger source buttons do something different than the "or" and "transparent trigger" buttons:

If I enable the setup from the right, i.e. OR in CH4 and "Enable Transparent Trigger" the app stops triggering. This is the configuration that seems to be applied in the `drs_exam.cpp` code if I am not mistaken. For some reason in that code it still triggers (I have modified the code to trigger on CH4 instead of CH1 and the trigger level, polarity, etc.).

What does the button in the left actually do? The circular checkbox with the "4" I mean. This is the trigger configuration I want to get in the C++ code.

I also don't know what the function `DRSBoard::EnableTrigger` does, what is the meaning of `flag1` and `flag2`? In my code there is a call to this function which I copied from the example.

Stefan Ritt wrote:

Unfortunately I have not idea what the problem could be. In principle the trigger should be independent of the sampling speed, since the trigger is only made with a discriminator and a flip-flop. The hardware must be ok since you see the trigger with the oscillocope app. All you can do is to go through the sorce code of the oscilloscope app, especially drsosc/Osic.cpp::ScanBoards(), SetTriggerLevel(), SetTriggerPolariy() etc. to make sure you do the same calls as the oscilloscope app.

Stefan

 

  260   Mon Jun 10 14:09:13 2013 tmiron alonadd an average ability to the Scope

Stefan Ritt wrote:

tmiron alon wrote:

Hallo,
I'm using DRS4 Evaluation Board Rev 4.0 and I'm trying to change the output of the samples to be an average of  #  measurements (1000 or even more)
during the process I have encountered some difficulties I hope you will be able to help me  with:

1. the DRS chip have 8 channels but the Evaluation board have only 4 channels. does the default mode of the DRS in the Evaluation Board is 1024 bins for each channel or 2048?

2. in the readout mode, does it samples all the 1024 bins waveform from a channel and then move to the next one, or after each bin it move to the next channel?

3. In the file "drs4_eval4_app.vhd", I have a problem finding the names of the signals that represents the registers bits which tell me what is the number of the bin (1-1024) the ADC is reading from the DRS, and the signals

that represents registers A0-A3. can you send me their names? 

 

4. In another matter- is the -0.5V to 0.5V is the  upper and lower  limit of the input (or just a working range), and if not what is the limit for AC?  is there a fuse on the board in case of overload from the input? (I didn't see  it in the User's Manual, but I didn't know if you will mention it there in case there is one).

thanks in advance and have a nice day,

Tmiron

1. All 8 channels are read out, but only 4 are displayed in the oscilloscope.

2. It reads all 1024 bins from a channel, then switch to the next channel.

3. The ADC readout happens in lines 1576+. The register for the sample count is drs_sample_count, and the signal for the address is drs_addr.

4. The evaluation board manual explicitly mentions the maximum allowed input range on page 5.

/Stefan 

 Hi Stefan,

I'm trying to add the averaging feature to the oscilloscope instead. I have a basic knowledge in C++ so I'm having problem finding the initial place (places) of the data acquisition. furthermore, I saw you already made an averaging possibility for some of the features in the program at the measure window.
Can you please tell we in which of the programs of the Scope (and in what line if possible) can I find the initial\main data acquisition so I could follow it from there, and where is the location of the averaging function you already made?

have a nice week,

Tmiron

  261   Mon Jun 10 16:24:21 2013 Stefan Rittadd an average ability to the Scope

tmiron alon wrote:

Stefan Ritt wrote:

tmiron alon wrote:

Hallo,
I'm using DRS4 Evaluation Board Rev 4.0 and I'm trying to change the output of the samples to be an average of  #  measurements (1000 or even more)
during the process I have encountered some difficulties I hope you will be able to help me  with:

1. the DRS chip have 8 channels but the Evaluation board have only 4 channels. does the default mode of the DRS in the Evaluation Board is 1024 bins for each channel or 2048?

2. in the readout mode, does it samples all the 1024 bins waveform from a channel and then move to the next one, or after each bin it move to the next channel?

3. In the file "drs4_eval4_app.vhd", I have a problem finding the names of the signals that represents the registers bits which tell me what is the number of the bin (1-1024) the ADC is reading from the DRS, and the signals

that represents registers A0-A3. can you send me their names? 

 

4. In another matter- is the -0.5V to 0.5V is the  upper and lower  limit of the input (or just a working range), and if not what is the limit for AC?  is there a fuse on the board in case of overload from the input? (I didn't see  it in the User's Manual, but I didn't know if you will mention it there in case there is one).

thanks in advance and have a nice day,

Tmiron

1. All 8 channels are read out, but only 4 are displayed in the oscilloscope.

2. It reads all 1024 bins from a channel, then switch to the next channel.

3. The ADC readout happens in lines 1576+. The register for the sample count is drs_sample_count, and the signal for the address is drs_addr.

4. The evaluation board manual explicitly mentions the maximum allowed input range on page 5.

/Stefan 

 Hi Stefan,

I'm trying to add the averaging feature to the oscilloscope instead. I have a basic knowledge in C++ so I'm having problem finding the initial place (places) of the data acquisition. furthermore, I saw you already made an averaging possibility for some of the features in the program at the measure window.
Can you please tell we in which of the programs of the Scope (and in what line if possible) can I find the initial\main data acquisition so I could follow it from there, and where is the location of the averaging function you already made?

have a nice week,

Tmiron

The averaging possibility in the measure window is just a "place holder" for future extensions, which are not yet implemented.

A good point to implement averaging is in the function Osci::ReadWaveforms() in Osci.cpp.

/Stefan 