Entry  Fri Nov 18 05:52:45 2016, Kurtis Nishimura, Channel offsets in GetTime() offsetInstructions.png

Hello,

I have a question about the GetTime() method in DRS.cpp.  I understand how the DT values are applied for all channels, and I also understand from the evaluation board manual that the timing of each channel is synchronized at sample 0, so samples should really be aligned from channel-to-channel relative to sample 0.

However, DRS.cpp has the following snippet in DrsBoard::GetTime():

   if (channelIndex > 0) {
      // correct all channels to channel 0 (Daniel's method)
      iend = tc >= 700 ? 700+1024 : 700;
      for (i=tc,gt0=0 ; i<iend ; i++)
         gt0 += fCellDT[chipIndex][0][i % 1024];
      
      for (i=tc,gt=0 ; i<iend ; i++)
         gt += fCellDT[chipIndex][channelIndex][i % 1024];
      
      for (i=0 ; i<fChannelDepth ; i++)
         time[i] += (float)(gt0 - gt);
   }

I can see what this is calculating and applying such an offset, but I don't understand why things seem to be referenced to sample 700.  Is there a particular reason why sample 700 is chosen here?  This does not seem like a straightforward application of the attached instructions from the evaluation board user's manual.

Any insight would be much appreciated!

Thanks so much,

-Kurtis

    Reply  Mon Nov 21 14:13:32 2016, Stefan Ritt, Channel offsets in GetTime() 

Cell 700 is arbitrary. You can choose any cell to align the channels to each other. The only requirement is that it's always the same cell for each event. Historically, Daniel chose cell #700 more or less arbitrary, but later we found out that this works with any cell. So for the publication we went with cell #0 (and that's why we have t_ch,0 in the paper), but cell #700 was left in the code because of lazyness. Feel free to replace 700 with any other number and you should get the same result. In a newer version of the software I use

// align cell#0 of all channels
float t1 = time[0][(1024-tc) % 1024];
for (int ch=1 ; ch<8 ; ch++) {
   float t2 = time[ch][(1024-tc) % 1024];
   float dt = t1 - t2;

   for (int i=0 ; i<1024 ; i++)
      time[ch][i] += dt;
   }

 

which is also a bit simpler. So time[ch] contains already the integrated time array (like 0.2 ns, 0.4 ns, 0.6 ns if at 5 GSPS, not the delta_t values as in the DRS.cpp code). Since the readout starts with cell # tc, the cell time[channel][1024-tc] is the physical cell #0 of the chip. The code makes sure that cell #0 in all 8 channels has the same time value.

Best regards,
Stefan

 

Kurtis Nishimura wrote:

Hello,

I have a question about the GetTime() method in DRS.cpp.  I understand how the DT values are applied for all channels, and I also understand from the evaluation board manual that the timing of each channel is synchronized at sample 0, so samples should really be aligned from channel-to-channel relative to sample 0.

However, DRS.cpp has the following snippet in DrsBoard::GetTime():

   if (channelIndex > 0) {
      // correct all channels to channel 0 (Daniel's method)
      iend = tc >= 700 ? 700+1024 : 700;
      for (i=tc,gt0=0 ; i<iend ; i++)
         gt0 += fCellDT[chipIndex][0][i % 1024];
      
      for (i=tc,gt=0 ; i<iend ; i++)
         gt += fCellDT[chipIndex][channelIndex][i % 1024];
      
      for (i=0 ; i<fChannelDepth ; i++)
         time[i] += (float)(gt0 - gt);
   }

I can see what this is calculating and applying such an offset, but I don't understand why things seem to be referenced to sample 700.  Is there a particular reason why sample 700 is chosen here?  This does not seem like a straightforward application of the attached instructions from the evaluation board user's manual.

Any insight would be much appreciated!

Thanks so much,

-Kurtis

 

Entry  Sat Oct 22 13:24:20 2022, Phan Van Chuan, Channel Cascading Option in the 2048-bin DRS4V51.png

Dear Stefan,
We are using DRS4 evaluation board version 5.1 and firmware version 30000 (as the picture attached). Now, I am in need one channel with length 2048 bin. However, I can't find the resistors R99, ... ,R106 on the hardware of evaluation board; it seems my DRS4 evaluation board doesn't use 2048 bins per channel.
Our question is, can we repair this hardware to read 2048 bins/channel? if that is possible please let me know what to add on hardware/software of DRS4 evaluation.
Best regards.
Phan Van Chuan.

    Reply  Mon Oct 24 12:50:24 2022, Stefan Ritt, Channel Cascading Option in the 2048-bin 

The board is delivered in one or the other mode and not meant to be changed by the user, since this requires very delicate soldering which is not easy. If you try anyhow, you loose the quarantee. You can send the board back to the manufacturer for the modification, but this costs quite some moeny.

Best regards,
Stefan

Phan Van Chuan wrote:

Dear Stefan,
We are using DRS4 evaluation board version 5.1 and firmware version 30000 (as the picture attached). Now, I am in need one channel with length 2048 bin. However, I can't find the resistors R99, ... ,R106 on the hardware of evaluation board; it seems my DRS4 evaluation board doesn't use 2048 bins per channel.
Our question is, can we repair this hardware to read 2048 bins/channel? if that is possible please let me know what to add on hardware/software of DRS4 evaluation.
Best regards.
Phan Van Chuan.

 

Entry  Mon Aug 31 16:44:12 2020, Hans Steiger, Channel Cascading 

Dear All,

I have a board with Channel Cascading Option. I have the problem, that it seems to be impossible to run all 4 Channels simultaneously for digitizing pulses. I can just run even or odd channels but not even and odd ones? If I run in combined option, My question: If a board comes with this combined option, is it still usable as a 4Ch Digitizer but with 1024bin traces?

 

All the best,

 

Hans

    Reply  Mon Aug 31 17:17:30 2020, Stefan Ritt, Channel Cascading Screenshot_2020-08-31_at_16.52.28_.png

If you have a board with cascading option, it should show the "combined" option in the 2048-bin option enabled (not grayed), as in the attached screen shot. If the 2048-bin option is all greyed out, the system does not recognize the cascading option. If your board has a sticker "2048 bin" and you still see the 2048-bin option greyed out, it might mean that a resistor on that board has been forgotten. If you do not see the "2048 bin" sticker on your board, you might not have a board with cascading option. So please check that. If the resistor is really missing, you can send us the board and we will add it.

Stefan

Hans Steiger wrote:

Dear All,

I have a board with Channel Cascading Option. I have the problem, that it seems to be impossible to run all 4 Channels simultaneously for digitizing pulses. I can just run even or odd channels but not even and odd ones? If I run in combined option, My question: If a board comes with this combined option, is it still usable as a 4Ch Digitizer but with 1024bin traces?

 

All the best,

 

Hans

 

Entry  Thu Nov 14 11:39:06 2013, Schablo, Cascading of channels  

 Hello,  I want use cascading of channels for 2048 cell - SetChannelConfig(0,8,4), but i can't understand how . Please, help me. Where i can dowload 2048_mode.ppt. (I found information about this file in DRS.cpp  (3445 line  "/ combine two halfs correctly, see 2048_mode.ppt")

Best regards,

Schablo Kostya 

    Reply  Thu Nov 14 12:51:56 2013, Stefan Ritt, Cascading of channels  2048_mode.pdf

Schablo wrote:

 Hello,  I want use cascading of channels for 2048 cell - SetChannelConfig(0,8,4), but i can't understand how . Please, help me. Where i can dowload 2048_mode.ppt. (I found information about this file in DRS.cpp  (3445 line  "/ combine two halfs correctly, see 2048_mode.ppt")

Best regards,

Schablo Kostya 

You have to combine two channels into one, and depending on where the domino wave stopped, things get a bit complicated. I attach 2048_mode.ppt for your reference, but am not sure if this will really help.

/Stefan 

    Reply  Thu Nov 21 14:35:57 2013, Schablo, Cascading of channels  

Stefan Ritt wrote:

Schablo wrote:

 Hello,  I want use cascading of channels for 2048 cell - SetChannelConfig(0,8,4), but i can't understand how . Please, help me. Where i can dowload 2048_mode.ppt. (I found information about this file in DRS.cpp  (3445 line  "/ combine two halfs correctly, see 2048_mode.ppt")

Best regards,

Schablo Kostya 

You have to combine two channels into one, and depending on where the domino wave stopped, things get a bit complicated. I attach 2048_mode.ppt for your reference, but am not sure if this will really help.

/Stefan 

Sorry for my question.

 I'm trying change "drs_exam.cpp" for read 2048 cell. 

I'm using SetChannnelConfig(0,8,4) and this code in "drs_exam.cpp" :

              ...

              float arrX[2048];

              float arrY[2048];

              ....

             b->GetTime(0, b->GetTriggerCell(0), arrX);

             b->GetWave(0, 0, arr.Y); 

             ....

Return 2048 values in arrX[2048] - correct values, but  in arrY[2048] -  not correct values.

   I can't understand what values return "GetWave" function. Please, say me how make, that GetWave function return correct values.  " not - correct values "(i mean that i give signal in drs bord and values not true.) 

Best regards,

Schablo Kostya 

 

    Reply  Thu Nov 21 14:45:56 2013, Stefan Ritt, Cascading of channels  

Schablo wrote:

Stefan Ritt wrote:

Schablo wrote:

 Hello,  I want use cascading of channels for 2048 cell - SetChannelConfig(0,8,4), but i can't understand how . Please, help me. Where i can dowload 2048_mode.ppt. (I found information about this file in DRS.cpp  (3445 line  "/ combine two halfs correctly, see 2048_mode.ppt")

Best regards,

Schablo Kostya 

You have to combine two channels into one, and depending on where the domino wave stopped, things get a bit complicated. I attach 2048_mode.ppt for your reference, but am not sure if this will really help.

/Stefan 

Sorry for my question.

 I'm trying change "drs_exam.cpp" for read 2048 cell. 

I'm using SetChannnelConfig(0,8,4) and this code in "drs_exam.cpp" :

              ...

              float arrX[2048];

              float arrY[2048];

              ....

             b->GetTime(0, b->GetTriggerCell(0), arrX);

             b->GetWave(0, 0, arr.Y); 

             ....

Return 2048 values in arrX[2048] - correct values, but  in arrY[2048] -  not correct values.

   I can't understand what values return "GetWave" function. Please, say me how make, that GetWave function return correct values.  " not - correct values "(i mean that i give signal in drs bord and values not true.) 

Best regards,

Schablo Kostya 

 

The evaluation board V4 does not support cascading by default. You have to connect each input to two channels, which can in principle be made by soldering some zero Ohm resistors on the board, but these are tiny parts and not everybody can do it. Note that the 2048 cell mode is an option when ordering the board. So first make sure that you can modify the board before trying to run the software. 

Entry  Sat Feb 12 13:06:56 2022, Matias Senger, Cannot trigger on pulses, have to trigger on undershoot 

I am using the DRS4 board trying to measure pulses produced by an LGAD. I have no prior experience with this board, have just installed the `drsosc` application and am exploring. I am experiencing some strange trigger behavior. Consider the following screenshot:

 

Here nothing is strange, the board is triggering on the undershoot and it is working fine, I can trigger on rising/falling edge, different levels, etc.

Now, the strange thing is that if I pull the trigger up to trigger on the pulse itself it stops triggering:

I have tried many different setups for the trigger (rising, falling edge, different levels, etc) and nothing works. In the undershoot, everything works.

I have tried with the internal test signal and it works fine:

What could be the problem?

I have run the voltage and time calibrations as suggested in the manual.

    Reply  Tue Feb 15 12:02:29 2022, Stefan Ritt, Cannot trigger on pulses, have to trigger on undershoot 

The trigger comparator is a ADCMP601 unit which requires a minimum pulse width of 3-4 ns. I see that your pulses are only 1-2 ns wide. You have to make your pulses wider in order to trigger on them.

Stefan

Entry  Thu May 29 04:22:43 2014, Toshihiro Nonaka, CalibrationWaveform offset.png

I'm writing the drs_exam.cpp to use multi-boards(v3, firmware:4.0.0), and taking data succeeded. But I have several questions about function written in DRS.cpp.

 

  1. I wrote following code in drs_exam.cpp to set input range -0.4~0.6

                                     b1->SetInputRange(0.1);

            And the 100mV offset appeared(I attached a picture). I think this is due to the voltage calibration isn't done.(Calibrated to -0.5~0.5mV in DRS Oscilloscope)

            If so, could you show me a simple usage of "CalibrationWaveform()" function in DRS.cpp? (or other function?)

 

       2. Although this question might be the almost same with above, is there any way to execute voltage and timing calibration in drs_exam.cpp?

           Now I start DAQ by executing drs_exam.cpp after I execute voltage and timing calibration to each board by DRS Oscilloscope program.

 

      3. Which command is right to use external trigger?

                                   b1->SetTriggerSource(4);   or  b1->SetTriggerSource(1<<4);

 

Best regards,

Toshihiro Nonaka

    Reply  Thu Jun 12 17:16:13 2014, Stefan Ritt, CalibrationWaveform 

Toshihiro Nonaka wrote:

I'm writing the drs_exam.cpp to use multi-boards(v3, firmware:4.0.0), and taking data succeeded. But I have several questions about function written in DRS.cpp.

 

  1. I wrote following code in drs_exam.cpp to set input range -0.4~0.6

                                     b1->SetInputRange(0.1);

            And the 100mV offset appeared(I attached a picture). I think this is due to the voltage calibration isn't done.(Calibrated to -0.5~0.5mV in DRS Oscilloscope)

            If so, could you show me a simple usage of "CalibrationWaveform()" function in DRS.cpp? (or other function?)

 

       2. Although this question might be the almost same with above, is there any way to execute voltage and timing calibration in drs_exam.cpp?

           Now I start DAQ by executing drs_exam.cpp after I execute voltage and timing calibration to each board by DRS Oscilloscope program.

 

      3. Which command is right to use external trigger?

                                   b1->SetTriggerSource(4);   or  b1->SetTriggerSource(1<<4);

 

Best regards,

Toshihiro Nonaka

1. b->CalibrateVolt(NULL);

2. see 1.

3. For the V3 boards use b->SetTriggerSource(4), for V4 and V5 boards, use b->SetTriggerSource(1<<4). I had to change that because from V4 on we can have logical combinations between channels (like channel 1 AND channel 2).

Best regards,

Stefan 

Entry  Wed Aug 1 17:42:32 2012, Mayank S. Rajguru, Calculation of loop filter parameters (R,C1and C1) for 1 GHz 

 Hi,

we are planning to use the DRS4 in our board for 1 GHz sampling and digitization.

I have seen in the data sheet that "For the PLL to work, an external loop filter is required. This filter ensures quick locking and stable operation at the desired sampling frequency".

What formula do you use to calculate the values of R, C1 and C2?

Can we use the same given value for different frequencies?

Thanks,

Mayak

    Reply  Mon Aug 6 02:44:00 2012, Stefan Ritt, Calculation of loop filter parameters (R,C1and C1) for 1 GHz 

Mayank S. Rajguru wrote:

 Hi,

we are planning to use the DRS4 in our board for 1 GHz sampling and digitization.

I have seen in the data sheet that "For the PLL to work, an external loop filter is required. This filter ensures quick locking and stable operation at the desired sampling frequency".

What formula do you use to calculate the values of R, C1 and C2?

Can we use the same given value for different frequencies?

Thanks,

Mayak

I never worked out an exact formula for the parameters. It also depends if you want a fast locking time, or a low phase jitter of the PLL. A good starting point are the values from the evaluation board:

R = 130 Ohm

C1 = 4.7 nF

C2 = 1 uF

They should work from 800 MHz to 5 GHz.

- Stefan 

Entry  Wed Jul 14 14:55:09 2021, Mehrpad Monajem, C code to read the 4 channel with external trigger 

Hi there,

Recently I bought a 5GSPS evaluation board with 2048 sampling points.
I want to read 4 inputs of the evaluation bord ar 5 GSPS or 2.5GSP and use an external trigger.
I've checked your website and download drs-5.0.5 which contains the source code in C. It seems that the file drs_exam.cpp can do what I am looking for.
So far I could make and compile the project in Linux Ubuntu, but I couldn't compile it in Windows 10.  I've used Cygwin64 to compile the project in windows 10.

I have the following questions:

1- Since I only need to compile the drs_exam.cpp file, could please help me with how can I compile it directly(without making the entire project). Or tell me which version of Wxwidget and libusb I have to install.

2- If you have any sample code that can read 4 inputs with an external trigger, please tell me where can I find it.

In the end, I want to write a wrapper on this C file(which returns digitized data) and run it from my python program. Thank you in advance.
Best regards,
Mehrpad

 

    Reply  Mon Aug 9 12:50:31 2021, Stefan Ritt, C code to read the 4 channel with external trigger 

Sorry the late reply, I was on vacation. 

Here are some answers:

1. I'm sorry I can't help much here, since I currently don't have a Windows 10 computer here to compile any code. I moved now completely to MacOSX, being very similar to Linux. I'm not allowed to run a Windows 7 computer any more for security reasons. Last time this worked for me was with Wxwidget version 3.0 and libusb 1.0, but I guess libusb is not critical so you can use a newer version. If you just compile drs_exam.cpp, you don't need any Wxwidget library. That one is only used for the oscilloscope program.

2. The program drs_exam_2048.cpp is meant to read channels in 2048-bin mode.

3. To adjust the delay between the trigger and the readout, use the function b->SetTriggerDelayNs(xxx)

Best,
Stefan

Mehrpad Monajem wrote:

Hi there,

Recently I bought a 5GSPS evaluation board with 2048 sampling points.
I want to read 4 inputs of the evaluation bord ar 5 GSPS or 2.5GSP and use an external trigger.
I've checked your website and download drs-5.0.5 which contains the source code in C. It seems that the file drs_exam.cpp can do what I am looking for.
So far I could make and compile the project in Linux Ubuntu, but I couldn't compile it in Windows 10.  I've used Cygwin64 to compile the project in windows 10.

I have the following questions:

1- Since I only need to compile the drs_exam.cpp file, could please help me with how can I compile it directly(without making the entire project). Or tell me which version of Wxwidget and libusb I have to install.

2- If you have any sample code that can read 4 inputs with an external trigger, please tell me where can I find it.

In the end, I want to write a wrapper on this C file(which returns digitized data) and run it from my python program. Thank you in advance.
Best regards,
Mehrpad

 

 

    Reply  Tue Aug 10 13:57:09 2021, Mehrpad Monajem, C code to read the 4 channel with external trigger 

Thank you for the reply.

In the version that I have, I cannot find drs_exam_2048.cpp file. Could you please send me the link to download the software folder, which contain this file.

Best,

Mehrpad

Stefan Ritt wrote:

Sorry the late reply, I was on vacation. 

Here are some answers:

1. I'm sorry I can't help much here, since I currently don't have a Windows 10 computer here to compile any code. I moved now completely to MacOSX, being very similar to Linux. I'm not allowed to run a Windows 7 computer any more for security reasons. Last time this worked for me was with Wxwidget version 3.0 and libusb 1.0, but I guess libusb is not critical so you can use a newer version. If you just compile drs_exam.cpp, you don't need any Wxwidget library. That one is only used for the oscilloscope program.

2. The program drs_exam_2048.cpp is meant to read channels in 2048-bin mode.

3. To adjust the delay between the trigger and the readout, use the function b->SetTriggerDelayNs(xxx)

Best,
Stefan

Mehrpad Monajem wrote:

Hi there,

Recently I bought a 5GSPS evaluation board with 2048 sampling points.
I want to read 4 inputs of the evaluation bord ar 5 GSPS or 2.5GSP and use an external trigger.
I've checked your website and download drs-5.0.5 which contains the source code in C. It seems that the file drs_exam.cpp can do what I am looking for.
So far I could make and compile the project in Linux Ubuntu, but I couldn't compile it in Windows 10.  I've used Cygwin64 to compile the project in windows 10.

I have the following questions:

1- Since I only need to compile the drs_exam.cpp file, could please help me with how can I compile it directly(without making the entire project). Or tell me which version of Wxwidget and libusb I have to install.

2- If you have any sample code that can read 4 inputs with an external trigger, please tell me where can I find it.

In the end, I want to write a wrapper on this C file(which returns digitized data) and run it from my python program. Thank you in advance.
Best regards,
Mehrpad

 

 

 

Entry  Thu Nov 10 04:41:24 2016, Abhishek Rajput, Break Statements in DRS4 Binary to ROOT Macro 

Hello,

I recently modified the binary to ROOT convertor written by Stefan (https://midas.psi.ch/elogs/DRS4+Forum/361) so it can decode data taken with any channel or set of channels  on the DRS4. In the process of testing this modifed version for data taken on all 4 channels, I encountered problems with decoding some of the event data. More specifically, upon hitting a certain event in some channel, the histograms for that channel would no longer be filled and the histograms for subsequent channels would not be filled with any event at all. 

After considerable bug hunting, I discovered the source of this problem was due to the break statement in the following code extract from the ROOT to binary macro:

 for (n=0 ; n<5 ; n++) {
      // read event header
      i = fread(&eh, sizeof(eh), 1, f);
      if (i < 1)
         break;

For some events apparently, the event header fails to be read properly (fread line returns 0 in this case). Moreover, when I used the feof and ferror functions on a particular file I was testing, the feof function returned a value of 1.

So my questions deal with two scenarios.

Firstly, in the event of an fread error, is a break statement is necessary? Is it not possible to skip the voltage data for those events whose event header fails to be read properly? Or is it the case that when some "corrupted" event header is encountered, all waveform data subsequent to that event is likewise corrupted? If the former is the case, is it advisable to replace the break condition with an fseek line that advances the position indicator of the stream by an additional 2052*n_channels + 32 bytes (in accordance with the binary file specifications of page 25 on the DRS4 manual) so that the next set of voltage data can be read? 

Secondly, in the case of an end of file error, does there exist any possible solution? Or is such an error an indication of a faulty drs4 channel or corrupted binary file? 

Any help with the aforementioned issues would be greatly appreciated.

 

Abhishek

 