Thu Jun 7 16:27:21 2018, Phan Van Chuan, 
|
Dear Stefan,
I am using an DRS4 board to test the signal from an scintillator detector; It has connected well to the computer on DRS Oscilloscope (Figure
1). Now, I am having a problem of developing from the code of the drs_exam program, because the DRS4 board has not connected to the computer when translation |
Fri Jun 8 08:11:05 2018, Stefan Ritt,
|
Several people reported this problem, but we cannot reproduce it at our lab. Both the oscilloscope and the command line interface use exactly the same
code to connect to the board. Have you tried the solution reported here: elog:657 ?
Best, |
Tue Apr 12 10:40:36 2022, LynseyShun,
|
Hello, I am Lynsey. now I set A3-A0 to 1001 in ROI mode, but only OUT0 has output, and the other seven channels(OUT1-OUT7) do not output corresponding
waveforms.
In ROI mode, can OUT0-OUT7 output sampled waveforms at the same time? |
Tue Apr 12 10:49:27 2022, Stefan Ritt,
|
A3-A0 = 1001 should be all you need to activate OUT0-OUT7. It works in our designs. Maybe double check the address lines with an oscilloscope.
Stefan
LynseyShun |
Thu Jun 16 05:31:25 2022, LynseyShun,
|
Thank you very much for your help!
Stefan
Ritt wrote:
A3-A0 = 1001 should be all you need to activate OUT0-OUT7. It works |