DRS4 Forum
  DRS4 Discussion Forum, Page 4 of 46  Not logged in ELOG logo
Entry  Mon Aug 19 23:01:22 2019, Bill Ashmanskas, should one deassert DENABLE while writing the write-shift register? 

Hi Stefan,

We have for some time now been using custom firmware on a custom board to read waveforms out of DRS4 chips.  Now we are working on cascaded readout mode, 4 channels @ 2048 samples, WSREG=0x55, in order to allow for longer trigger latency.

Doing a testbench simulation of the FPGA code raised a question for me:  Do I need to deassert DENABLE while I shift a new 8-bit value into the write-shift register?  What happens if, during the few-hundred nanoseconds it takes to shift 8 bits into the register, the domino wave crosses cell 768, thereby shifting the write-shift register left by one bit?  Is this shifting suppressed when A=0b1101?  Or does the update of the actual write-shift register occur only once, after the 8th SRCLK cycle?  (Maybe one is really shifting bits into a shadow register that is copied all at once into the actual register?)

I notice in simulating your drs4_eval5_app.vhd that if one sets bit 27 ("drs_ctl_dactive") of register 0 (do not deassert DENABLE on trigger), then starts the domino wave (set bit 0 of register 0), then issues a software trigger, then later writes to register 5 (config register, wsreg, etc.), DENABLE is not in fact deasserted during the time when A=0b1100 (conf_setup, conf_strobe) or when A=0b1101 (wsr_setup, wsr_strobe).

But my simulation testbench includes a simplified Verilog model of my interpretation of the DRS4 data sheet, and my simulated DRS4 happened to cause the write-shift register to shift (256 samples before DTAP toggled) during your "wsr_strobe" FSM state, thus corrupting the value that was being shifted into the WSREG via SRIN and SRCLK.

So I'm curious:  to be safe, should one deassert DENABLE before updating the write-shift register, or is it safe to update it even while the domino wave is active and looping?  It seems easy enough to be safe, since we should only need to write to the WSREG once during the setup phase and then let it loop forever.

Many thanks,

Bill

 

 

    Reply  Tue Aug 20 10:44:45 2019, Stefan Ritt, should one deassert DENABLE while writing the write-shift register? 

Hi Bill,

you keep DENABLE active all the time to keep the Domino Wave running, but you deassert DWRITE if you change any register via SRCLK. There is no shadow register, just a simple shift register, but with DWRITE being low, the domino circuitry does not touch it.

Best,
Stefan

Bill Ashmanskas wrote:

Hi Stefan,

We have for some time now been using custom firmware on a custom board to read waveforms out of DRS4 chips.  Now we are working on cascaded readout mode, 4 channels @ 2048 samples, WSREG=0x55, in order to allow for longer trigger latency.

Doing a testbench simulation of the FPGA code raised a question for me:  Do I need to deassert DENABLE while I shift a new 8-bit value into the write-shift register?  What happens if, during the few-hundred nanoseconds it takes to shift 8 bits into the register, the domino wave crosses cell 768, thereby shifting the write-shift register left by one bit?  Is this shifting suppressed when A=0b1101?  Or does the update of the actual write-shift register occur only once, after the 8th SRCLK cycle?  (Maybe one is really shifting bits into a shadow register that is copied all at once into the actual register?)

I notice in simulating your drs4_eval5_app.vhd that if one sets bit 27 ("drs_ctl_dactive") of register 0 (do not deassert DENABLE on trigger), then starts the domino wave (set bit 0 of register 0), then issues a software trigger, then later writes to register 5 (config register, wsreg, etc.), DENABLE is not in fact deasserted during the time when A=0b1100 (conf_setup, conf_strobe) or when A=0b1101 (wsr_setup, wsr_strobe).

But my simulation testbench includes a simplified Verilog model of my interpretation of the DRS4 data sheet, and my simulated DRS4 happened to cause the write-shift register to shift (256 samples before DTAP toggled) during your "wsr_strobe" FSM state, thus corrupting the value that was being shifted into the WSREG via SRIN and SRCLK.

So I'm curious:  to be safe, should one deassert DENABLE before updating the write-shift register, or is it safe to update it even while the domino wave is active and looping?  It seems easy enough to be safe, since we should only need to write to the WSREG once during the setup phase and then let it loop forever.

Many thanks,

Bill

    Reply  Tue Aug 20 16:05:21 2019, Bill Ashmanskas, should one deassert DENABLE while writing the write-shift register? 

Aha -- many thanks.  I think what tripped up my test logic is that the "done" state in drs4_eval5_app.vhd that executes post-readout sets DWRITE back to 1 (drs_write_set).  If one then writes to FPGA register 5 while the FSM is in the "idle" state, the conf_strobe and wsr_strobe states occur with DWRITE and DENABLE both asserted.  This is if one sets the "dactive" bit in the FPGA app code, which is probably not the usual use case.  Maybe using the real DRS.cpp avoids this situation.  (I was simulating your FPGA code to test my understanding of what our FPGA code should do.)

Anyway, our own use case is fine: as you suggest, we leave DENABLE asserted, but we deassert DWRITE while reading out or while changing DRS4 register values.

Thanks again,

Bill

 

 

Stefan Ritt wrote:

Hi Bill,

you keep DENABLE active all the time to keep the Domino Wave running, but you deassert DWRITE if you change any register via SRCLK. There is no shadow register, just a simple shift register, but with DWRITE being low, the domino circuitry does not touch it.

Best,
Stefan

Bill Ashmanskas wrote:

Hi Stefan,

We have for some time now been using custom firmware on a custom board to read waveforms out of DRS4 chips.  Now we are working on cascaded readout mode, 4 channels @ 2048 samples, WSREG=0x55, in order to allow for longer trigger latency.

Doing a testbench simulation of the FPGA code raised a question for me:  Do I need to deassert DENABLE while I shift a new 8-bit value into the write-shift register?  What happens if, during the few-hundred nanoseconds it takes to shift 8 bits into the register, the domino wave crosses cell 768, thereby shifting the write-shift register left by one bit?  Is this shifting suppressed when A=0b1101?  Or does the update of the actual write-shift register occur only once, after the 8th SRCLK cycle?  (Maybe one is really shifting bits into a shadow register that is copied all at once into the actual register?)

I notice in simulating your drs4_eval5_app.vhd that if one sets bit 27 ("drs_ctl_dactive") of register 0 (do not deassert DENABLE on trigger), then starts the domino wave (set bit 0 of register 0), then issues a software trigger, then later writes to register 5 (config register, wsreg, etc.), DENABLE is not in fact deasserted during the time when A=0b1100 (conf_setup, conf_strobe) or when A=0b1101 (wsr_setup, wsr_strobe).

But my simulation testbench includes a simplified Verilog model of my interpretation of the DRS4 data sheet, and my simulated DRS4 happened to cause the write-shift register to shift (256 samples before DTAP toggled) during your "wsr_strobe" FSM state, thus corrupting the value that was being shifted into the WSREG via SRIN and SRCLK.

So I'm curious:  to be safe, should one deassert DENABLE before updating the write-shift register, or is it safe to update it even while the domino wave is active and looping?  It seems easy enough to be safe, since we should only need to write to the WSREG once during the setup phase and then let it loop forever.

Many thanks,

Bill

 

Entry  Wed Apr 27 08:14:14 2016, Toshihiro Nonaka, serial number problem  serial.png

Dear all,

I'm using 3 DRS boards simultaneously and their serial numbers are 2169, 2170, 2172 respectively.

Recently however,  I obtain serial number "0" by DRSBoard::GetBoardSerialNumber() for #2172 board.

Data taking can be done without any problems, but I'd like to know what is happening.

Any advice?

Thank you.

Toshihiro Nonaka

    Reply  Wed Apr 27 09:04:01 2016, Stefan Ritt, serial number problem  

If dis- and reconnecting the board does not help, there is the (small) chance that the serial number got erased in the board. You can re-set it with the "drscl" command line tool:

$ drscl
Found DRS4 board 0 on USB, serial #0, firmware revision 21305
B0> serial 2172

 

Toshihiro Nonaka wrote:

Dear all,

I'm using 3 DRS boards simultaneously and their serial numbers are 2169, 2170, 2172 respectively.

Recently however,  I obtain serial number "0" by DRSBoard::GetBoardSerialNumber() for #2172 board.

Data taking can be done without any problems, but I'd like to know what is happening.

Any advice?

Thank you.

Toshihiro Nonaka

 

    Reply  Wed Apr 27 09:51:37 2016, Toshihiro Nonaka, serial number problem  

The serial number has been fixed by using drscl. Thank you!

Stefan Ritt wrote:

If dis- and reconnecting the board does not help, there is the (small) chance that the serial number got erased in the board. You can re-set it with the "drscl" command line tool:

$ drscl
Found DRS4 board 0 on USB, serial #0, firmware revision 21305
B0> serial 2172

 

Toshihiro Nonaka wrote:

Dear all,

I'm using 3 DRS boards simultaneously and their serial numbers are 2169, 2170, 2172 respectively.

Recently however,  I obtain serial number "0" by DRSBoard::GetBoardSerialNumber() for #2172 board.

Data taking can be done without any problems, but I'd like to know what is happening.

Any advice?

Thank you.

Toshihiro Nonaka

 

 

Entry  Tue Nov 26 15:36:39 2013, Dmitry Hits, reducing sampling speed 

Dear Stefan

Is there an easy way to reduce sampling speed below 0.7 GSPS? I would like to record traces up to 5 usec long.

Thank you

Dmitry 

    Reply  Tue Nov 26 15:38:13 2013, Stefan Ritt, reducing sampling speed 

Dmitry Hits wrote:

Dear Stefan

Is there an easy way to reduce sampling speed below 0.7 GSPS? I would like to record traces up to 5 usec long.

Thank you

Dmitry 

No. See the DRS4 datasheet: http://www.psi.ch/drs/DocumentationEN/DRS4_rev09.pdf 

Minimum sampling speed is 700 MSPS.

 

/Stefan

Entry  Tue May 4 21:18:28 2021, Abaz Kryemadhi, recording only timestamp and amplitude and/or filesize maximum 

Hi,

I have been collecting some date using the DRS4 board at a trigger rate of 10-20 Hz,    I only need the timestamp and the amplitude, is there anyway to select only these two live as the data comes in to be stored. 

Alternatively,  What's the maximum file size or maximum number of events I can store in one binary file in linux. 

Thanks,

Best,

Abaz

    Reply  Wed May 5 10:12:44 2021, Stefan Ritt, recording only timestamp and amplitude and/or filesize maximum 

The maximum file size depends on the underlying linux file system. Common values are 4-16 GBytes.

Stefan

Abaz Kryemadhi wrote:

Hi,

I have been collecting some date using the DRS4 board at a trigger rate of 10-20 Hz,    I only need the timestamp and the amplitude, is there anyway to select only these two live as the data comes in to be stored. 

Alternatively,  What's the maximum file size or maximum number of events I can store in one binary file in linux. 

Thanks,

Best,

Abaz

 

Entry  Wed Jun 1 22:29:01 2016, Dominik Neise, problems when stop cell >= 767 ?? stop_cell_distribution.png

Hello Stefan,

some colleages told me a story, I was neither able to confirm nor find anything in the datsheet about. According to them:

For some internal reason of the DRS4, if the “stop capacitor” of the DRS4 is >= 767, the true stop channel is one before the stop channel read from the DRS4. In other words, the stop channel which returns the DRS4 shifts after sampling to the capacitor ID 766.

Can you confirm that, or even say a few words about that matter?

I wanted to confirm this by plotting the stop cell distribution for random triggered data, taken with one of the FACT boards. I assumed (possibly misunderstanding the matter), that this would lead to missing values in the area of stop cell 767, but cannot see any significant excess or lack of entries in that area.

 

    Reply  Wed Jun 1 23:16:01 2016, Stefan Ritt, problems when stop cell >= 767 ?? 

I cannot confirm the story with the "stop capacitor > 767". It can be seen from your plots that the distribution of stop cells are even, no holes or bins with double height.

There is an issue with cell 767, but this is when one tries to do simultaneous reading/writing to the chip. This does not really work as writen in the data sheet. Waveforms sometimgs get cut off at cell 767. But the stop cell is always correct, otherwise one could not calibrate the data. If you use the evaluation board for example, which is perfectly calibrated, and introduce an "artifical" shift like

if stop cell > 767 then
  stop cell = stop cell + 1

then you would see that the voltage calibration would become wrong and very noisy.

Stefan

Dominik Neise wrote:

Hello Stefan,

some colleages told me a story, I was neither able to confirm nor find anything in the datsheet about. According to them:

For some internal reason of the DRS4, if the “stop capacitor” of the DRS4 is >= 767, the true stop channel is one before the stop channel read from the DRS4. In other words, the stop channel which returns the DRS4 shifts after sampling to the capacitor ID 766.

Can you confirm that, or even say a few words about that matter?

I wanted to confirm this by plotting the stop cell distribution for random triggered data, taken with one of the FACT boards. I assumed (possibly misunderstanding the matter), that this would lead to missing values in the area of stop cell 767, but cannot see any significant excess or lack of entries in that area.

 

 

Entry  Sun Jun 12 08:45:52 2016, Michael, problems of DRS4 

Hi

I want to use DRS4 to digitize 16 channels of signals. The width of signal is about 20 ns, with frequency of 50Hz. The time differences between these 16 signals are not constant, arranging from 3us to 0. I am confused about this in some aspects.

  1. Can I use SIMULTANEOUS WRITINT AND READING to realize this? I saw the VHDL program, and if I understand it correctly, it did not work at this state.
  2. Or sampling at 1GSPS, using CASCADING OF CHANNELS, I can sample signal at most 4us or 8us, then digitizing all signals of one chip. Have you tested 4 or more channels cascading before?

Besides, any advice will be helpful!

Thank you.

Entry  Sun Jun 12 08:49:54 2016, Michael, problems of DRS4 

Hi

I want to use DRS4 to digitize 16 channels of signals. The width of signal is about 20 ns, with frequency of 50Hz. The time differences between these 16 signals are not constant, arranging from 3us to 0. I am confused about this in some aspects.

  1. Can I use SIMULTANEOUS WRITINT AND READING to realize this? I saw the VHDL program, and if I understand it correctly, it did not work at this state.
  2. Or sampling at 1GSPS, using CASCADING OF CHANNELS, I can sample signal at most 4us or 8us, then digitizing all signals of one chip. Have you tested 4 or more channels cascading before?

Besides, any advice will be helpful!

Thank you.

    Reply  Wed Jun 15 14:49:00 2016, Stefan Ritt, problems of DRS4 

1. Simultaneous writing and reading is not possible with the DRS4 chip. The manual says differently on p. 14, but due to a bug in the chip waveforms get clipped at the end if one does that. We hopt to fix this problem in a future version of the chip.

2. You can cascade 2,4 or 8 channels. If you cascade 8 channels and run at 1 GSPS, you digitize a window of 8 us. If you have 16 signals, you then need 16 chips.

/Stefan

Michael wrote:

Hi

I want to use DRS4 to digitize 16 channels of signals. The width of signal is about 20 ns, with frequency of 50Hz. The time differences between these 16 signals are not constant, arranging from 3us to 0. I am confused about this in some aspects.

  1. Can I use SIMULTANEOUS WRITINT AND READING to realize this? I saw the VHDL program, and if I understand it correctly, it did not work at this state.
  2. Or sampling at 1GSPS, using CASCADING OF CHANNELS, I can sample signal at most 4us or 8us, then digitizing all signals of one chip. Have you tested 4 or more channels cascading before?

Besides, any advice will be helpful!

Thank you.

 

Entry  Thu Jan 25 05:24:05 2018, chen wenjun, problem with the drscl(drs507) 

Hi! Stefan:

  when I change a new computer(win7,64bit),I meet a problem that the drscl app cannot found the board! It shows"USB successfully scanned,but no boards found",but the drsosc runs well . when I connect to other win7*64bits computer,only one of them runs property! Is there any driver else I need to install? 

Thank you!

Chen

    Reply  Thu Jan 25 08:00:16 2018, Stefan Ritt, problem with the drscl(drs507) 

This problem has been reported by several people, like elog:551

So far I could not solve it. On the computers at our lab it works find so I cannot reproduce and fix the problem. One suspicion I have is that the underlying libusb library needs to be updated. You can try to install the newest version from their website at http://libusb.info/, but I haven't tried it myself.

Stefan

 

chen wenjun wrote:

Hi! Stefan:

  when I change a new computer(win7,64bit),I meet a problem that the drscl app cannot found the board! It shows"USB successfully scanned,but no boards found",but the drsosc runs well . when I connect to other win7*64bits computer,only one of them runs property! Is there any driver else I need to install? 

Thank you!

Chen

 

    Reply  Thu Jan 25 08:07:32 2018, chen wenjun, problem with the drscl(drs507) 

I have tried about 4 computers,only one worked fine.I truly want to know how others get this fixed,can you get in touch with them?

Stefan Ritt wrote:

This problem has been reported by several people, like elog:551

So far I could not solve it. On the computers at our lab it works find so I cannot reproduce and fix the problem. One suspicion I have is that the underlying libusb library needs to be updated. You can try to install the newest version from their website at http://libusb.info/, but I haven't tried it myself.

Stefan

 

chen wenjun wrote:

Hi! Stefan:

  when I change a new computer(win7,64bit),I meet a problem that the drscl app cannot found the board! It shows"USB successfully scanned,but no boards found",but the drsosc runs well . when I connect to other win7*64bits computer,only one of them runs property! Is there any driver else I need to install? 

Thank you!

Chen

 

 

Entry  Mon Oct 19 11:26:29 2009, Jinhong Wang, output common mode voltage of DRS4 
Hello Mr. Stifan.Ritt
       In the DSR4 datasheet, it is mentioned that there is an additional buffer at each analog output, this buffer shifts the the differential range of -0.5V~0.5V to 0.8V~1.8V. Does it mean that this buffer shifts a voltage of about 1.3V for the primary differential range? 
       Again for the differential range of -0.5V~0.5V, can the common mode voltage of the analog output at OUT+/OUT-  be chaned? In the example presented in the datasheet, OUT+ is 0.8V~1.8V and OUT- is 1.8V~0.8V. So for an output swing of 2V p-p, can the common mode voltage be modified to the desired value? Supposed that the input ranges from -0.5V~0.5V.
                                                      
     Thank you!
                                             Jinhong Wang
    Reply  Mon Oct 19 12:46:12 2009, Stefan Ritt, output common mode voltage of DRS4 

Jinhong Wang wrote:
Does it mean that this buffer shifts a voltage of about 1.3V for the primary differential range? 

No. It shifts about ROFS-0.25V. So only if ROFS=1.55V, the shift will be 1.3V.

Jinhong Wang wrote:
Again for the differential range of -0.5V~0.5V, can the common mode voltage of the analog output at OUT+/OUT-  be chaned?

Just read the datasheet under "ANALOG OUTPUTS". I'm sorry if I did not describe this clearly, but the U+ voltage is fixed (only dependent on ROFS), and U- can be calculated using Uofs as written in the datasheet. 

Jinhong Wang wrote:
In the example presented in the datasheet, OUT+ is 0.8V~1.8V and OUT- is 1.8V~0.8V. So for an output swing of 2V p-p, can the common mode voltage be modified to the desired value? Supposed that the input ranges from -0.5V~0.5V.

OUT+ is 0.8V~1.8V, OUT- is 2*Uofs-OUT+. So you can only change the OUT- level, not the OUT+ level. 

ELOG V3.1.5-3fb85fa6