    Reply  Wed Jan 15 17:34:55 2014, Stefan Ritt, DRS4 v2.0 Eval Board running on higher versions of DRS Oscilloscope program 

Andrey Kuznetsov wrote:

Hi,

I have an old v2.0 board that I just upgraded firmware on using v4.0.0 download package which has a drs4_eval1.bit for v2.0 boards containing firmware 15158. So I would like to use the latest DRS Oscilloscope program, due to the faster acquisition speeds and advanced calibration techniques, however I seem to be running into a problem.

v4.0 DRS Oscilloscope program displays flat lines in any configuration instead of a pulse that I provide to it (I can't tell if calibration works because all my traces are flat and nothing triggers) but provides ~460Hz. Any idea why v4.0.0 DRS Oscilloscope program does not work with DRS4 Eval Board v2.0 fw:15158?

 

v3.0 DRS Oscilloscope program actually works and displays the pulse that I input (calibration also works), however it only has 64Hz speed due to v3.0.0 not having multithreading capability which is provided in v3.1.0 software version of the program.

 

Can you please upload and post on the website the latest software packages for v2 and v3? I would like to use v3.1.0 DRS Oscilloscope version. (Both Linux and Windows, but Linux preferably)

 

Also, I would like to report that for some reason, I don't know if it happens with v3.0 program used on v2.0 board only or a general issue, but after each calibration of voltage and timing, the trigger does not work. I have to exit the oscilloscope program, and run it again, then the trigger works fine, and the device is calibrated.

Thank you

Unfortunately I do not have the time to back-port any new software feature to the old V2 and V3 boards. That means you have to live with their software or try to get a new board. Right now we have the V5 board which allows you 1 ps time measurements. Maybe this is a good argument to upgrade the hardware.

/Stefan 

    Reply  Wed Jan 15 17:37:21 2014, Stefan Ritt, DRS4 installation on Windows 8 issues 

Andrey Kuznetsov wrote:

I'm also having trouble installing drivers and running DRSOsc program on another computer running Windows 8.


The issue with the driver is that it's not digitally signed.


The issue with the DRSOsc is that it's failing to find libusb0.dll. libusb-win32 seemed to have installed upon DRS4 software install, however the supplied version is Windows 7/8 incompatible, so on Windows 7 computer I had to download libusb_win32 v1.2.6.0 from the official website and install it directly, then everything worked fine. However in Windows 8, I am unable to install libusb-win32 because in  libusb-win32 Inf Wizard installation program when you select for which device the libusb should be used, it asks to install a driver, but when I point to DRS' driver, it says "Unknown Error: 1" and that's that. One way around the libusb issue is to copy the required dll and sys file directly where the .exe is stored.


I will attempt to disable signed driver signature requirement, and see if the driver installs then, but this should really be fixed instead.

Did you have any progress with that? Unfortunately I don't have a Windows 8 machine here at our institute, so I cannot reproduce your problem. At least I put the 1.2.6 libusb driver into the V5 software package. 

    Reply  Wed Feb 5 13:41:42 2014, Stefan Ritt, Repeated time calibration 

Hermann-Josef Mathes wrote:

Hi Stefan,

thanks for the quick reply, I know that this solution works with drscl but not within my code.

I tried to track it down, but gave up very soon. Seems as if AnalyzeWF() which is called by CalibrateTiming() finds to much zero-crossings when it is called the second time.

Regards

  -- Hermann-Josef

Time calibration has been changed completely in meantime. With the new V5 boards, we have a new oscillator on the board where on can calibrate each channel individually. This is necessary to obtain a good timing down to a few ps. With the current code the above problem has vanished. We also learned that the time calibration is very stable (less than a ps) over several months, so no need to repeat the calibration over and over again.

/Stefan 

    Reply  Tue Feb 18 14:12:37 2014, Stefan Ritt, Announcement of new Evaluation Board V5 scope.png

Stefan Ritt wrote:

Dear DRS community,

starting from this year, we ship the new evaluation board V5. This board has an improved internal timing calibration, with which one can measure the time with a precision down to a few ps. Following picture shows the time between two pulses, obtained with a function generator, a passive split and a delay cable. The single threshold time estimator of the DRSOsc program obtains with such signal a resolution of 2.5 ps (RMS).

Using more sophisticated algorithms such as cross-correlation, resolutions below 1 ps were already achieved.

The new board can now be ordered at the same price than the V4 board, delivery will start in March 2014.

Best regards,
Stefan Ritt
 

The new software for the V5 evaluation board has been released today with following new features:

  • Hardware scalers for all four channels and the trigger working up to 200 MHz. With the trigger scaler one can measure for example coincidence rates between two channels.
  • New vertical and horizontal "slice" measurements. This allows to measure the amplitude of a signal at a certain time relative to the trigger point or the time when a signal crosses a certain level.
  • Gated charge measurement allowing to measure the charge of a signal between two time markers, like an old-fashioned charge integrating ADC.

The software is available at the the usual location http://www.psi.ch/drs/software-download for Linux, Windows and Mac OSX. I'm working right now to get it also into the Apple App Store.

/Stefan

    Reply  Thu Mar 6 11:12:44 2014, Stefan Ritt, Software drs-5.0.0 fails to compile (drsosc) 

Hermann-Josef Mathes wrote:

Hi,

the latest software drs-5.0.0.tar.gz fails to compile on my freshly installed SuSE 13.1 whereas the previous 4.0.1 is compiling out-of-the-box.

My system has the wxWidgets 2.8.12 which is probably together with gcc 4.8.1 the reason of the problem. I applied a number of corrections, mainly some sort of proper (?) typecasts, a patch file is attached.

Maybe you could consider to take them into account for the next patch release?

Thanks and best regards

Hermann-Josef

Thank you very much for the corrections. I know it in principle, but neither my Mac OSX nor the Windows compiler complains, so I usually don't see this errors. It's fixed now.

/Stefan 

    Reply  Wed Apr 16 08:20:36 2014, Stefan Ritt, drs_exam project fail to compile 

Carlo Stella wrote:

Hi,

when I try to compile drs_exam project my computer give me this output:

 

1>------ Rebuild All started: Project: drs_exam, Configuration: Debug Win32 ------
1>  averager.cpp
1>c:\users\daq\desktop\

original drs\drs5\src\averager.cpp(165): warning C4996: 'fopen': This function or variable may be unsafe. Consider using fopen_s instead. To disable deprecation, use _CRT_SECURE_NO_WARNINGS. See online help for details.
1>          c:\program files (x86)\microsoft visual studio 10.0\vc\include\stdio.h(234) : see declaration of 'fopen'
1>  DRS.cpp
1>c:\users\daq\desktop\original drs\drs5\src\drs.cpp(4597): warning C4244: '=' : conversion from 'double' to 'float', possible loss of data
1>  drs_exam.cpp
1>  Generating Code...
1>  musbstd.c
1>  mxml.c
1>  strlcpy.c
1>  Generating Code...
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_claim_interface referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_set_configuration referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_open referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_get_busses referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_set_debug referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_find_devices referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_find_busses referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_init referenced in function _musb_open
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_set_altinterface referenced in function _musb_set_altinterface
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_close referenced in function _musb_close
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_release_interface referenced in function _musb_close
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_bulk_write referenced in function _musb_write
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_bulk_read referenced in function _musb_read
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_reset referenced in function _musb_reset
1>musbstd.obj : error LNK2019: unresolved external symbol _usb_get_descriptor referenced in function _musb_get_device
1>.\Debug/drs_exam.exe : fatal error LNK1120: 15 unresolved externals
========== Rebuild All: 0 succeeded, 1 failed, 0 skipped ==========
 
Can anyone help me to solve the problem?

Have a look at the web site http://www.psi.ch/drs/software-download . Under the MS Windows section it says that you have to install the libusb-1.0 package first before you can compile the example program. This is also obvious from the missing _usb_* functions in the error listing above.

/Stefan

    Reply  Wed Apr 16 08:30:32 2014, Stefan Ritt, why is the first channel output error?  

Wang wrote:

 Hi,

 The diagram below is DRS4 output. Green is the output8+, blue is the output8-. Output8+ of the first channel is below  the baseline. It is not  right.

Others channel  is suitable. I check the circuit , Hardware is no problem, so I want to konw where the FPGA code  is wrong. what reason is caused? Thanks!

You are funny. Just by looking at a scope picture I should know what is wrong at your FPGA code. Unfortunately I'm not a magician. I looks to me like you have 11 channels in your diagram, although the chip has only 9. What I would recommend is to put some input to each channel one at a time, like a 10 MHz sine wave. You should then see that sine wave for the single channel at the output and can correlate input vs. output. Maybe your address bits are wrong or the chip has a soldering problem.

/Stefan 

    Reply  Wed Apr 16 10:24:55 2014, Stefan Ritt, DRS4 Evalboard V5 with Windows7Pro64bit 
> 
> Dear Stefan
> 
> I am trying to use the DRS4 eval board on a Windows7 machine. Unfortunately I get an error message saying "No DRS 
> board found". But I can see the DRS board in the device manager with the proper driver loaded. Is there any known 
> problem with win7?
> 
> I am using windows7 professional (SP1) with the drs software 5.0.1.
> 
> Cheers,
> Roman
> 
> PS: Everything is working on my mac. But not under windows7.

Hi Roman,

please read section 2.3. (page 13)  from the Evaluation Board manual: http://www.psi.ch/drs/DocumentationEN/manual_rev50.pdf

You have to update the USB driver in your Computer Management.

Cheers,
Stefan 
    Reply  Tue May 13 19:39:36 2014, Stefan Ritt, drsosc binary to cern ROOT file conversion 

Luka Pavelic wrote:

Hi,

Does anybody have program for conversion from binary or xml to cern ROOT *.root file?
 

Thank you for any help you can provide,
Luka Pavelic


 

You look here: elog:262

/Stefan

    Reply  Tue May 13 23:08:50 2014, Stefan Ritt, drsosc binary to cern ROOT file conversion 

Luka Pavelic wrote:

Thank you for your fast and very helpful replay.

I made it work with drsosc version 4 but with version 5 i am getting weird results. Is it possible that they changed binary formatting?

 

Yes, but this is documented in the evaluation board manual. You have to modify the script slightly. I will update it myself in about 2-3 weeks.

Cheers,

Stefan

    Reply  Mon May 19 08:04:57 2014, Stefan Ritt, simultaneous writing and reading with region of interest mode? 

Benjamin LeGeyt wrote:

Hello!

We're developing electronics based on the DRS4 to read out a breast PET scanner and our event rate will be quite high so we're concerned about dead-time.  with that in mind, I have a question regarding the mode of simultaneous writing and reading that is described in the DRS4 data sheet.  I think the description there is quite clear but I'd like to ask for a few clarifications.

1) Are the channels required to be read out via the channel multiplexer when doing the simultaneous write/read or is it ok to read out all channels in parallel (even the ones still sampling) and just throw away the ones you don't want?

2) If one wanted to use region of interest mode along with the simultaneous write/read, how would that work?  Here is what I would think - please tell me if I'm missing some important detail:

-upon trigger, deassert dwrite.

-strobe RSRLOAD

-increment write config register

-reassert dwrite

-start the readout (reading out stop shift register value on SROUT as data comes out)

3) now to add even more complexity - I would actually like to use simultaneous write/read along with region of interest mode and also with pairs of cascaded channels as we need >500ns latency and 2Gsps is too slow for our signals.  the combination of cascading and simultaneous write/read is addressed in the data sheet but I still have one question.  In normal circumstances when cascading channels, one would read out the value in the write shift register to know which channel was active when the domino wave stopped.  I assume that this is not possible when dwrite is enabled as the write shift register is then advanced by the domino wave, so I see three possibilities:

-accept more dead-time and read out the write-shift-register each time (adds ~240ns to deadtime)

-just read out both channels every time and figure out later where is the data you want

-attempt to keep track of the expected state of write-shift-register in firmware.

is there a better option that I have not thought of?

 

many thanks!

Benjamin LeGeyt

Unfortunately the simultaneous writing/reading does not work as described in the data sheet. Just recently we found out that due to a bug in the chip a part of the waveform is missing if you read and write at the same time. The only clean solution is to use two DRS4 chips in parallel. You read one chip while the other samples, then you switch over between them. In that case all the ROI scheme and channel cascading works normally. The dead time will be addressed by the DRS5 chip, which will be dead time free, but will not be available until in maybe 2-3 years.

/Stefan 

    Reply  Tue May 27 16:07:17 2014, Stefan Ritt, Spikes in DRS4 data on custom baord. 

Dominik Neise wrote:

We see quite some spikes in our DRS4 sampled data in FACT.  We see different types of spikes:

  • single cell spikes, usually showing a large amplitude of 200mV
  • double cell spikes, usually only in the order of 20mV.
  • Even triple and quadro cell spikes are rarely seen.

The double cell spikes often occur as symmetrical double cell spikes mirrored at cell 512. quadro cell spikes seem to be nothing else, than connected symmetrical double cell spikes. For the triple cell spikes we have no idea.
Currently we use simple filters to get rid of these spikes, this workes rather well for the large single cell spikes, but with the occurance of tripples and quadros we started to worry about higher multiples and revived our DRS4 spike investigations.

Now I was told, that you Stefan know already where these spikes come from and even a paper exisits. Unfortunately so far I was unable to find it.
I wonder if it is possible to predict the occurance of these spikes, so one does not have to search for them anymore and can get rid of the filters.

Best regards

Dominik

All I know is that the  "20mV" spikes are always symmetrical around cell #512, that they are typically 17.4 mV in height, and that they occur always in all 9 channels simultaneously. They cannot occur in all locations, but there only like 32 possible locations where they can occur. With this information it should be easy to fix them by filtering.

200 mV spikes are new to me. I do not see them in our boards, so it must be related to the board readout and not to the chip.

Best regards,
Stefan
 

    Reply  Wed Jun 11 11:13:50 2014, Stefan Ritt, Announcement of new Evaluation Board V5 

Osip Lishilin wrote:

Stefan Ritt wrote:

Hardware scalers for all four channels and the trigger working up to 200 MHz. With the trigger scaler one can measure for example coincidence rates between two channels.

 Does it give the ability to measure triggering rate? I'm talking again about possibility of use DRS4 as pulse counter for PMT's. If yes, do I need new v5 board or it is possible to use v4 board?

Yes it is possible to measure the raw trigger rate, with a resolution of 10 Hz. You need a new V5 board for that. 

    Reply  Thu Jun 12 12:46:00 2014, Stefan Ritt, DRS eval bord v5 Timing 
> a) Calibration:
> I am using 4 boards daisy chained. To achieve optimal time resolution I did first a voltage calibration and right afterwards a time calibration. For all 
> boards after the master I am not sure how to do it.
> After setting the flag "Configure multi-board daisy-chain" in the config menu, all the slave boards set the flag "use external reference clock".  By 
> hitting the voltage calibration button, the slave boards unset this flag. Is it true, that I have to re-set this before doing the time-calibration right 
> afterwards?

Please do NOT do any calibration in multi-board mode. This will not work. Calibrate the boards separately, then activate the multi-board mode. Please note that the timing between the boards is not better 
than ~50 ps. This is a limitation of the FPGA clock generators. If you need better timing, you have to feed an external clock into one channel of each board (leaving only 3 channels for DAQ). The upcoming 
WaveDREAM board will have 16 channels per board, so building bigger DAQ systems will be much easier (and more precise).

> b) getting the right times in binary format:
> To get the time out of the time width (i.e. the t_ch[i]) you sum up in your documentation from j=0 to j=i (see attachment). In your example code 
> read_binary.cpp (line 113) you sum from j=0 to j=i-1. Since you get the the bin with in the binary file, I guess that the example code is correct one?

Yes, I will correct the documentation.

Cheers,
Stefan
    Reply  Thu Jun 12 17:16:13 2014, Stefan Ritt, CalibrationWaveform 

Toshihiro Nonaka wrote:

I'm writing the drs_exam.cpp to use multi-boards(v3, firmware:4.0.0), and taking data succeeded. But I have several questions about function written in DRS.cpp.

 

  1. I wrote following code in drs_exam.cpp to set input range -0.4~0.6

                                     b1->SetInputRange(0.1);

            And the 100mV offset appeared(I attached a picture). I think this is due to the voltage calibration isn't done.(Calibrated to -0.5~0.5mV in DRS Oscilloscope)

            If so, could you show me a simple usage of "CalibrationWaveform()" function in DRS.cpp? (or other function?)

 

       2. Although this question might be the almost same with above, is there any way to execute voltage and timing calibration in drs_exam.cpp?

           Now I start DAQ by executing drs_exam.cpp after I execute voltage and timing calibration to each board by DRS Oscilloscope program.

 

      3. Which command is right to use external trigger?

                                   b1->SetTriggerSource(4);   or  b1->SetTriggerSource(1<<4);

 

Best regards,

Toshihiro Nonaka

1. b->CalibrateVolt(NULL);

2. see 1.

3. For the V3 boards use b->SetTriggerSource(4), for V4 and V5 boards, use b->SetTriggerSource(1<<4). I had to change that because from V4 on we can have logical combinations between channels (like channel 1 AND channel 2).

Best regards,

Stefan 

    Reply  Wed Jul 16 12:10:19 2014, Stefan Ritt, change cascading from 1024 to 2048 bins for each input channel 

Yves Bianga wrote:

Hello,

 
I want to ask whether it is possible to modify a Evaluation Board 5.0 from 1024 to 2048 cells for each of the 4 input channels.
On the rev50 manual at page 31 I found an option to connect the 4 unused channels by setting 8 solder bridges.
The source code for controlling the board seems already prepared for 2048 bins, since version 5.0.2.
 
So my first question: Are there any implementations in the VHDL Code to control the write shift register in 2048 mode? / Is there a necessity for a newer/other VHDL Code or is it already implemented?
 
And the second: Are there any other modifications except the eight zero Ohm resistors and maybe changes in the FPGA code?
 
My board info output:
 
Mezz. Board index:    0
DRS type:             DRS4
Board type:           9
Serial number:        2451
Firmware revision:    21260
 
 
Thanks a lot!
 
Yves Bianga

Indeed you only need R99-R106 to be installed. Unfortunately the firm/software cannot know if the resistors are there, that's why we introduced R142/R143, which connect J44 of the FPGA optionally to low. So if J44 is low (R143 installed), this tells the system that we are in 2048 bin mode. Unfortunately you need firmware revision 21305 or later to support this bit, which you apparently do not have. So you can either upgrade the firmware (if you have a download cable) or "fake" the 2048 bin mode in software. Go to line 4345 of DRS.cpp and look for DRSBoard::Is2048ModeCapable(). This function just returns the status of this bit. If you installed R99-R106, you could modify this function to always return "1" instead of "0". Then the DRSOsc program will display 2048 bins for each of the four channels.

Best regards,

Stefan 

    Reply  Wed Jul 30 17:05:06 2014, Stefan Ritt, ROOT program to decode binary data from DRSOsc read_binary.Cread_binary.cpp

Stefan Ritt wrote:

Please find attached a simple ROOT based program (http://root.cern.ch) to decode binary data from the DRSOsc program. It assumes that all four channels were recorded. If this is not the case, the program can be adjusted accordingly.

To use it, simply type (assuming that you have written a data file "test.dat" with DRSOsc):

root [0] .L decode.C+
Info in <TUnixSystem::ACLiC>: creating shared library /tmp/./decode_C.so
root [1] decode("test");
Info in <TCanvas::MakeDefCanvas>:  created default TCanvas with name c1
1927 events processed
"test.root" written
root [2] 

If you have turned on the clock on channel4 of the DRS4 evaluation board, it will produce a plot like this:
 
c1.gif 

 

/Stefan

I updated this ROOT program for the new format used with the V5 boards. It's now called "read_binary.C". Usage stays the same. There is also a standalone C program "read_binary.cpp". Both are attached. 

    Reply  Wed Jul 30 17:05:38 2014, Stefan Ritt, drsosc binary to cern ROOT file conversion 

ChengMing Du wrote:

Stefan Ritt wrote:

Luka Pavelic wrote:

Thank you for your fast and very helpful replay.

I made it work with drsosc version 4 but with version 5 i am getting weird results. Is it possible that they changed binary formatting?

 

Yes, but this is documented in the evaluation board manual. You have to modify the script slightly. I will update it myself in about 2-3 weeks.

Cheers,

Stefan

 hi Stefan,can you update the code to convert binary to root for newest drsosc?Thanks.

See elog:361 

    Reply  Tue Aug 26 12:32:21 2014, Stefan Ritt, 10GSps on DRS4 Evm with delay cables 

Martin Petriska wrote:

 Hi, I read its possible to use channels 2,4,6 to extend 200ns to 400ns (1024bins to 2048).

Is it possible to use same channels to double sampling rate with paralel feeding, one channel delayed by Ts/2, for 5,12GS/s is it cca 3cm delay cable?

 

Martin

In principle yes (you could split your signal externally and add some cable delay to one side), but it is not supported by the software. You would have to combine the data from the two channels yourself. But it won't help much. The analog bandwidth of the evaluation board is about 700 MHz. So sampling at 10 GSPS vs. 5 GSPS won't give you any additional information, since the highest frequencies in your signal will be only 700 MHz. You could as well take your 5 GSPS measurement and interpolate it with some sinc function to get exactly the same result. See here for details: http://en.wikipedia.org/wiki/Whittaker%E2%80%93Shannon_interpolation_formula 

    Reply  Fri Sep 12 13:00:04 2014, Stefan Ritt, synchronizing two DRS4 evaluation boards readout with one computer 

Dmitry Hits wrote:

 Hi everyone,

Has anyone tried to synchronize 2 (two) DRS4 evaluation boards readout by the same computer? I have read about some attempts on this board in the past, but I do not know if they have succeeded. If yes, could you share your experience and/or software.

Thank you very much,

Dmitry.

 

Please read the manual http://www.psi.ch/drs/DocumentationEN/manual_rev50.pdf page 25 where this is described in detail.

/Stefan

ELOG V3.1.5-3fb85fa6