DRS4 Forum
  DRS4 Discussion Forum, Page 17 of 46  Not logged in ELOG logo
New entries since:Thu Jan 1 01:00:00 1970
    Reply  Tue Oct 26 23:18:32 2021, Javier Caravaca, Trigger multiple boards independently 

Thank you Stefan. Actually I noticed that the source code of drs_exam was available after I started this thread, and that was the solution that occurred to me too. I'll give that a try.

A related question is: if the 4 boards are triggering at max rate (500Hz), would the total data throughtput (of the four boards together) be 2GHz (500Hz x 4)? Or is the limitation on the readout, rather than the triggering?

Best,

Javier.

Stefan Ritt wrote:

Unfortunately an independent operation from a single computer is not supported by the software. You can try to modify the drs_exam program and extend it. You can poll all boards in sequence and just read out that one which got a trigger, then start the loop again. But I don't know how good you are in programming. I needs a bit of experience to do that.

Stefan

Javier Caravaca wrote:

Hello,

I recently acquired 4 DRS4 boards and I wanted to ask if it was possible to trigger them independently from the same computer.

I know that you can daisy-chain boards and trigger them all at the same time, but in my case, each of my boards record independent events, so I want them to trigger when trigger conditions are met in each board. I currently use the provided DRSOSC software, that can trigger on only the board that is being displayed at that moment. I tried opening several instances of DRSOSC to associate each to each board, but that is not possible since the second instance already does not find the boards. I wonder if there is a way of triggering from each board independently without having to use four computers.

Thank you,

Javier.

 

 

    Reply  Wed Oct 27 08:11:42 2021, Stefan Ritt, Trigger multiple boards independently 

I'm not sure if the rate would go up to 2 kHz (not 2 GHz!). Depends how the USB hub is designed. What you can do however is to buy 4 RaspberryPis (total cost 150$) and run everythign in parallel. The evaluation boards works nicely with the Pi's.

Javier Caravaca wrote:

A related question is: if the 4 boards are triggering at max rate (500Hz), would the total data throughtput (of the four boards together) be 2GHz (500Hz x 4)? Or is the limitation on the readout, rather than the triggering?

Entry  Sun Sep 23 02:22:46 2018, Gerard Arino-Estrada, Trigger OUT pulse width variable from 100 us up to 100 ms 

Hello Stefan,

I am using the DRS4 board connected to a Raspberry PI and through the drsosc application. I am interested on using the "Trigger OUT" signal to do some extra data processing with NIM modules. According to the manual, for each hardware trigger a TTL pulse of 150 ns width should be send through the "trigger OUT". In my case I do see pulses with widths ranging from 100 microseconds up to hundreds of miliseconds. I am connecting the signal directly to an oscilloscope with 50 Ohm termination. I have tried two DRS4 boards in identical conditions and both show the same behavior.  Having such wide and variable pulses makes it complicated for me to do the extra post-processing. Have you any idea of what might be going wrong? Thank you very much.

Best regards,

Gerard

    Reply  Wed Sep 26 14:44:14 2018, Stefan Ritt, Trigger OUT pulse width variable from 100 us up to 100 ms 

The "Trigger OUT" has changed recently. It goes high on a new trigger, but then STAYS high until the board has been read out by the PC and re-started. This allows better synchronization with some external trigger, which can be re-armed with the falling edge of the trigger out signal. The signal can be quite long, since readout of an event via USB typically takes 2 ms, but can be more if the PC is busy. If you need back your 150 ns pulse, send the trigger out to an external pulse shaper with fixed shaping width.

Stefan

Gerard Arino-Estrada wrote:

Hello Stefan,

I am using the DRS4 board connected to a Raspberry PI and through the drsosc application. I am interested on using the "Trigger OUT" signal to do some extra data processing with NIM modules. According to the manual, for each hardware trigger a TTL pulse of 150 ns width should be send through the "trigger OUT". In my case I do see pulses with widths ranging from 100 microseconds up to hundreds of miliseconds. I am connecting the signal directly to an oscilloscope with 50 Ohm termination. I have tried two DRS4 boards in identical conditions and both show the same behavior.  Having such wide and variable pulses makes it complicated for me to do the extra post-processing. Have you any idea of what might be going wrong? Thank you very much.

Best regards,

Gerard

 

    Reply  Wed Sep 26 18:25:07 2018, Gerard Arino-Estrada, Trigger OUT pulse width variable from 100 us up to 100 ms 

Thank you very much for the answer, I really appreciate your help.

Thanks!

Gerard

Stefan Ritt wrote:

The "Trigger OUT" has changed recently. It goes high on a new trigger, but then STAYS high until the board has been read out by the PC and re-started. This allows better synchronization with some external trigger, which can be re-armed with the falling edge of the trigger out signal. The signal can be quite long, since readout of an event via USB typically takes 2 ms, but can be more if the PC is busy. If you need back your 150 ns pulse, send the trigger out to an external pulse shaper with fixed shaping width.

Stefan

Gerard Arino-Estrada wrote:

Hello Stefan,

I am using the DRS4 board connected to a Raspberry PI and through the drsosc application. I am interested on using the "Trigger OUT" signal to do some extra data processing with NIM modules. According to the manual, for each hardware trigger a TTL pulse of 150 ns width should be send through the "trigger OUT". In my case I do see pulses with widths ranging from 100 microseconds up to hundreds of miliseconds. I am connecting the signal directly to an oscilloscope with 50 Ohm termination. I have tried two DRS4 boards in identical conditions and both show the same behavior.  Having such wide and variable pulses makes it complicated for me to do the extra post-processing. Have you any idea of what might be going wrong? Thank you very much.

Best regards,

Gerard

 

 

    Reply  Wed Sep 26 18:28:20 2018, Gerard Arino-Estrada, Trigger OUT pulse width variable from 100 us up to 100 ms 

Thank you very much for the answer, I really appreciate your help.

Thanks!

Gerard

Stefan Ritt wrote:

The "Trigger OUT" has changed recently. It goes high on a new trigger, but then STAYS high until the board has been read out by the PC and re-started. This allows better synchronization with some external trigger, which can be re-armed with the falling edge of the trigger out signal. The signal can be quite long, since readout of an event via USB typically takes 2 ms, but can be more if the PC is busy. If you need back your 150 ns pulse, send the trigger out to an external pulse shaper with fixed shaping width.

Stefan

Gerard Arino-Estrada wrote:

Hello Stefan,

I am using the DRS4 board connected to a Raspberry PI and through the drsosc application. I am interested on using the "Trigger OUT" signal to do some extra data processing with NIM modules. According to the manual, for each hardware trigger a TTL pulse of 150 ns width should be send through the "trigger OUT". In my case I do see pulses with widths ranging from 100 microseconds up to hundreds of miliseconds. I am connecting the signal directly to an oscilloscope with 50 Ohm termination. I have tried two DRS4 boards in identical conditions and both show the same behavior.  Having such wide and variable pulses makes it complicated for me to do the extra post-processing. Have you any idea of what might be going wrong? Thank you very much.

Best regards,

Gerard

 

 

    Reply  Wed Sep 26 19:21:03 2018, Stefan Ritt, Trigger OUT pulse width variable from 100 us up to 100 ms 

In meantime I even updated the manual.

Stefan

Gerard Arino-Estrada wrote:

Thank you very much for the answer, I really appreciate your help.

Thanks!

Gerard

 

    Reply  Mon Dec 23 19:31:31 2024, Matias Henriquez, Trigger OUT pulse width variable from 100 us up to 100 ms 

Given this new scenario, what is the maximum rate of events that can be processed then (a rough estimation would be great, 1/2ms?)? is it mainly limited by the USB data transmission and the PC? How does the logic of the trigger and DRS4 data sampling works inside the FPGA in general terms? e.g: trigger activated -> dwrite ON -> ADC acquisition -> busy until data has been shipped off to the PC -> free to process new events. 

Is there a way to obtain some sort of timestamp for the trigger on each event? or is it better to use C++ time functions in the PC since the DRS4 is usually used in experiments with low rate of events so the long time it takes to the USB and PC is not a problem? (eg. particle physics).

Thanks for your help,

Matias H.

Stefan Ritt wrote:

The "Trigger OUT" has changed recently. It goes high on a new trigger, but then STAYS high until the board has been read out by the PC and re-started. This allows better synchronization with some external trigger, which can be re-armed with the falling edge of the trigger out signal. The signal can be quite long, since readout of an event via USB typically takes 2 ms, but can be more if the PC is busy. If you need back your 150 ns pulse, send the trigger out to an external pulse shaper with fixed shaping width.

Stefan

Gerard Arino-Estrada wrote:

Hello Stefan,

I am using the DRS4 board connected to a Raspberry PI and through the drsosc application. I am interested on using the "Trigger OUT" signal to do some extra data processing with NIM modules. According to the manual, for each hardware trigger a TTL pulse of 150 ns width should be send through the "trigger OUT". In my case I do see pulses with widths ranging from 100 microseconds up to hundreds of miliseconds. I am connecting the signal directly to an oscilloscope with 50 Ohm termination. I have tried two DRS4 boards in identical conditions and both show the same behavior.  Having such wide and variable pulses makes it complicated for me to do the extra post-processing. Have you any idea of what might be going wrong? Thank you very much.

Best regards,

Gerard

 

 

Entry  Thu Jul 18 01:03:44 2019, Ismael Garcia, Trace Impedance DRS4_Analog_IN.PNG

Hi Steffan,

              I'm an engineer at UCLA developing a board with the DRS4 chip. Our team has a question on what might be the required trace impedence for the analog inputs. Can that information be provided? 

Best Regards,
Ismael Garcia

    Reply  Thu Jul 18 11:37:56 2019, Stefan Ritt, Trace Impedance 

The requiremnet is the same as for any high speed analog board, there is othing special with the DRS4. If you want to terminate your line with 50 Ohms and you want a matched impedance layout, you route all lines with 50 Ohms impedance. Truth is however that nothing is perfect. The SMA connector is not exactly 50 Ohm, the PCB gets a 10-20% variation depending on the manufacturer. So even if you try hard, you will never have a 50 Ohm matched impedance. On the evaluation board we made some compromises as you have seen, but for us the board works satisfactory even with this compromises, and you can test it yourself with real hardware (namely the evaluation board). If you can do a better job, try it. But usually these compromises have only little influence on the signal quality.

Stefan

Ismael Garcia wrote:

Hi Steffan,

              I'm an engineer at UCLA developing a board with the DRS4 chip. Our team has a question on what might be the required trace impedence for the analog inputs. Can that information be provided? 

Best Regards,
Ismael Garcia

 

    Reply  Fri Jul 19 01:37:09 2019, Ismael Garcia, Trace Impedance 

When you're refering to laying a 50 Ohm trace, you're referring to the SMA input and not the interface between the output of the Op-AMP(THS4508) buffer 
and the inputs  of the DRS4(IN0-IN8). Is there a recommended diffential impedance for IN0-IN8? 

Ismael

Stefan Ritt wrote:

The requiremnet is the same as for any high speed analog board, there is othing special with the DRS4. If you want to terminate your line with 50 Ohms and you want a matched impedance layout, you route all lines with 50 Ohms impedance. Truth is however that nothing is perfect. The SMA connector is not exactly 50 Ohm, the PCB gets a 10-20% variation depending on the manufacturer. So even if you try hard, you will never have a 50 Ohm matched impedance. On the evaluation board we made some compromises as you have seen, but for us the board works satisfactory even with this compromises, and you can test it yourself with real hardware (namely the evaluation board). If you can do a better job, try it. But usually these compromises have only little influence on the signal quality.

Stefan

Ismael Garcia wrote:

Hi Steffan,

              I'm an engineer at UCLA developing a board with the DRS4 chip. Our team has a question on what might be the required trace impedence for the analog inputs. Can that information be provided? 

Best Regards,
Ismael Garcia

 

 

    Reply  Sat Jul 20 12:28:14 2019, Stefan Ritt, Trace Impedance 

The DRS4 input is high impedance. So if you like you can terminate it with 100 Ohm differentially and route it with 100 Ohm. But if you keep the lines short, the reflection is negligible. That’s what we made on the evaluation board.

Ismael Garcia wrote:

When you're refering to laying a 50 Ohm trace, you're referring to the SMA input and not the interface between the output of the Op-AMP(THS4508) buffer 
and the inputs  of the DRS4(IN0-IN8). Is there a recommended diffential impedance for IN0-IN8? 

Ismael

Stefan Ritt wrote:

The requiremnet is the same as for any high speed analog board, there is othing special with the DRS4. If you want to terminate your line with 50 Ohms and you want a matched impedance layout, you route all lines with 50 Ohms impedance. Truth is however that nothing is perfect. The SMA connector is not exactly 50 Ohm, the PCB gets a 10-20% variation depending on the manufacturer. So even if you try hard, you will never have a 50 Ohm matched impedance. On the evaluation board we made some compromises as you have seen, but for us the board works satisfactory even with this compromises, and you can test it yourself with real hardware (namely the evaluation board). If you can do a better job, try it. But usually these compromises have only little influence on the signal quality.

Stefan

Ismael Garcia wrote:

Hi Steffan,

              I'm an engineer at UCLA developing a board with the DRS4 chip. Our team has a question on what might be the required trace impedence for the analog inputs. Can that information be provided? 

Best Regards,
Ismael Garcia

 

 

 

Entry  Wed Oct 21 15:03:13 2020, Seiya Nozaki, Timing diagram of SROUT/SRIN signal to write/read a write shift register drs4_srin_srout_srclk.pdf

Dear Stefan,

I have questions about the timing diagram of SROUT/SRIN signal to write/read a write shift register.
1) Value of SRIN signal is saved at the falling edge of SRCLK, correct? (It is written in datasheet, page12, "Bits are latched into the shift register on the falling edge of SRCLK")
2) When are 8-bits of write shift register shown through SROUT? At ridging edges of SRCLK? and with additional delay(~10ns)? or falling edges?
3) In my understanding, when SRCLK is sent to DRS4, we can read and write the values in parallel, right? If so, is it possible just to read the registers without updating the registers?

[Background]
We have two modes to set the write shift register, the first one is to always reconnect to the first line and another one is to reconnect to the same line as when DWRITE goes to Low.
We can read/write the write shift register with the first mode (channel reset, page1). But we rarely face the problem of write shift register, unexpected values are written, with the second mode. With this mode, SROUT signal is sent back to DRS from FPGA as SRIN to write the same value on the write shift register. So there is a small delay(~10ns) due to the routing (DRS->FPGA->DRS, page2). It seems SRIN signal is not stable around the falling edges of SRCLK, thus it could cause that unexpected values are written in write shifter register.
To understand the situation clearly, I'd like to know the answer to the above questions.

Thank you.

Best regards,
Seiya

    Reply  Tue Oct 27 13:37:23 2020, Stefan Ritt, Timing diagram of SROUT/SRIN signal to write/read a write shift register Screenshot_2020-10-27_at_13.45.39_.png

Dear Seiya,

1) That's correct. SRIN is ampled at the falling edge. Pleae make sure to obey the hold-time as written in the datasheet. P.12, Fig. 11: SRIN must be stable before the falling edge of SRCLK and tH after the falling clock. tH is 5ns according to table 1. 

2) The write shift register is a 8-bit shift register, with an input, output and clock. After the first clock pulse, the 7th bit is shown, after the second clock pulse the 6th bit and so on. You you need 8 clock pulses to read the whole register. At the same time you write to the register, so what ever is present at SRIN will replace the old 8 bits of that register.

3) No this is not possible. When you read the register, you write to it at the same time. One possibilty is to connect SROUT to SRIN during that (maybe via the FPGA). Then you have a circular register wich is the same after each 8 clock pulses.

For your reference, I posted a commercial D-Flip Flop (TI SNHCS72). The DRS4 write shift register is a simple array of 8 such registers, with no CLR or PRE, where SROUT is Q of the last Flip Flop.

Best,
Stefan

Seiya Nozaki wrote:

Dear Stefan,

I have questions about the timing diagram of SROUT/SRIN signal to write/read a write shift register.
1) Value of SRIN signal is saved at the falling edge of SRCLK, correct? (It is written in datasheet, page12, "Bits are latched into the shift register on the falling edge of SRCLK")
2) When are 8-bits of write shift register shown through SROUT? At ridging edges of SRCLK? and with additional delay(~10ns)? or falling edges?
3) In my understanding, when SRCLK is sent to DRS4, we can read and write the values in parallel, right? If so, is it possible just to read the registers without updating the registers?

[Background]
We have two modes to set the write shift register, the first one is to always reconnect to the first line and another one is to reconnect to the same line as when DWRITE goes to Low.
We can read/write the write shift register with the first mode (channel reset, page1). But we rarely face the problem of write shift register, unexpected values are written, with the second mode. With this mode, SROUT signal is sent back to DRS from FPGA as SRIN to write the same value on the write shift register. So there is a small delay(~10ns) due to the routing (DRS->FPGA->DRS, page2). It seems SRIN signal is not stable around the falling edges of SRCLK, thus it could cause that unexpected values are written in write shifter register.
To understand the situation clearly, I'd like to know the answer to the above questions.

Thank you.

Best regards,
Seiya

 

    Reply  Tue Oct 27 15:02:09 2020, Seiya Nozaki, Timing diagram of SROUT/SRIN signal to write/read a write shift register 

Dear Stefan,

Thank you for your reply.
SRIN is directly connected to SROUT via FPGA for now, but it is unstable for the timing between clock and SRIN depending on the firmware logic.
We want to make our system more robust, so we are thinking to use a clock with a lower frequency (let's say 16.6 MHz) or change the duty cycle of a clock to keep more time between the rising edge and falling edge of a clock. This change is just for reading/writing the write shift register, we will use a 33 MHz clock for the analog readout in any case.
If we change like above, are there any concerns from the DRS4 side?

Best,
Seiya

Stefan Ritt wrote:

Dear Seiya,

1) That's correct. SRIN is ampled at the falling edge. Pleae make sure to obey the hold-time as written in the datasheet. P.12, Fig. 11: SRIN must be stable before the falling edge of SRCLK and tH after the falling clock. tH is 5ns according to table 1. 

2) The write shift register is a 8-bit shift register, with an input, output and clock. After the first clock pulse, the 7th bit is shown, after the second clock pulse the 6th bit and so on. You you need 8 clock pulses to read the whole register. At the same time you write to the register, so what ever is present at SRIN will replace the old 8 bits of that register.

3) No this is not possible. When you read the register, you write to it at the same time. One possibilty is to connect SROUT to SRIN during that (maybe via the FPGA). Then you have a circular register wich is the same after each 8 clock pulses.

For your reference, I posted a commercial D-Flip Flop (TI SNHCS72). The DRS4 write shift register is a simple array of 8 such registers, with no CLR or PRE, where SROUT is Q of the last Flip Flop.

Best,
Stefan

Seiya Nozaki wrote:

Dear Stefan,

I have questions about the timing diagram of SROUT/SRIN signal to write/read a write shift register.
1) Value of SRIN signal is saved at the falling edge of SRCLK, correct? (It is written in datasheet, page12, "Bits are latched into the shift register on the falling edge of SRCLK")
2) When are 8-bits of write shift register shown through SROUT? At ridging edges of SRCLK? and with additional delay(~10ns)? or falling edges?
3) In my understanding, when SRCLK is sent to DRS4, we can read and write the values in parallel, right? If so, is it possible just to read the registers without updating the registers?

[Background]
We have two modes to set the write shift register, the first one is to always reconnect to the first line and another one is to reconnect to the same line as when DWRITE goes to Low.
We can read/write the write shift register with the first mode (channel reset, page1). But we rarely face the problem of write shift register, unexpected values are written, with the second mode. With this mode, SROUT signal is sent back to DRS from FPGA as SRIN to write the same value on the write shift register. So there is a small delay(~10ns) due to the routing (DRS->FPGA->DRS, page2). It seems SRIN signal is not stable around the falling edges of SRCLK, thus it could cause that unexpected values are written in write shifter register.
To understand the situation clearly, I'd like to know the answer to the above questions.

Thank you.

Best regards,
Seiya

 

 

    Reply  Tue Oct 27 15:24:38 2020, Stefan Ritt, Timing diagram of SROUT/SRIN signal to write/read a write shift register 

This is a static shift register, so you can make the clock as slow as you want. Actually I don't use a "clock", I just use a data pin I control via a state machine in the VHDL code. This way I have more control over the edges. I need several (internal) clock cycles to produce one SRCLK clock cycle, but that does not matter for the DRS.

Stefan

Seiya Nozaki wrote:

Dear Stefan,

Thank you for your reply.
SRIN is directly connected to SROUT via FPGA for now, but it is unstable for the timing between clock and SRIN depending on the firmware logic.
We want to make our system more robust, so we are thinking to use a clock with a lower frequency (let's say 16.6 MHz) or change the duty cycle of a clock to keep more time between the rising edge and falling edge of a clock. This change is just for reading/writing the write shift register, we will use a 33 MHz clock for the analog readout in any case.
If we change like above, are there any concerns from the DRS4 side?

Best,
Seiya

Stefan Ritt wrote:

Dear Seiya,

1) That's correct. SRIN is ampled at the falling edge. Pleae make sure to obey the hold-time as written in the datasheet. P.12, Fig. 11: SRIN must be stable before the falling edge of SRCLK and tH after the falling clock. tH is 5ns according to table 1. 

2) The write shift register is a 8-bit shift register, with an input, output and clock. After the first clock pulse, the 7th bit is shown, after the second clock pulse the 6th bit and so on. You you need 8 clock pulses to read the whole register. At the same time you write to the register, so what ever is present at SRIN will replace the old 8 bits of that register.

3) No this is not possible. When you read the register, you write to it at the same time. One possibilty is to connect SROUT to SRIN during that (maybe via the FPGA). Then you have a circular register wich is the same after each 8 clock pulses.

For your reference, I posted a commercial D-Flip Flop (TI SNHCS72). The DRS4 write shift register is a simple array of 8 such registers, with no CLR or PRE, where SROUT is Q of the last Flip Flop.

Best,
Stefan

Seiya Nozaki wrote:

Dear Stefan,

I have questions about the timing diagram of SROUT/SRIN signal to write/read a write shift register.
1) Value of SRIN signal is saved at the falling edge of SRCLK, correct? (It is written in datasheet, page12, "Bits are latched into the shift register on the falling edge of SRCLK")
2) When are 8-bits of write shift register shown through SROUT? At ridging edges of SRCLK? and with additional delay(~10ns)? or falling edges?
3) In my understanding, when SRCLK is sent to DRS4, we can read and write the values in parallel, right? If so, is it possible just to read the registers without updating the registers?

[Background]
We have two modes to set the write shift register, the first one is to always reconnect to the first line and another one is to reconnect to the same line as when DWRITE goes to Low.
We can read/write the write shift register with the first mode (channel reset, page1). But we rarely face the problem of write shift register, unexpected values are written, with the second mode. With this mode, SROUT signal is sent back to DRS from FPGA as SRIN to write the same value on the write shift register. So there is a small delay(~10ns) due to the routing (DRS->FPGA->DRS, page2). It seems SRIN signal is not stable around the falling edges of SRCLK, thus it could cause that unexpected values are written in write shifter register.
To understand the situation clearly, I'd like to know the answer to the above questions.

Thank you.

Best regards,
Seiya

 

 

 

    Reply  Wed Oct 28 04:32:19 2020, Seiya Nozaki, Timing diagram of SROUT/SRIN signal to write/read a write shift register 

Dear Stefan,

OK, it's good to hear! Thank you!

Best,
Seiya

Stefan Ritt wrote:

This is a static shift register, so you can make the clock as slow as you want. Actually I don't use a "clock", I just use a data pin I control via a state machine in the VHDL code. This way I have more control over the edges. I need several (internal) clock cycles to produce one SRCLK clock cycle, but that does not matter for the DRS.

Stefan

Seiya Nozaki wrote:

Dear Stefan,

Thank you for your reply.
SRIN is directly connected to SROUT via FPGA for now, but it is unstable for the timing between clock and SRIN depending on the firmware logic.
We want to make our system more robust, so we are thinking to use a clock with a lower frequency (let's say 16.6 MHz) or change the duty cycle of a clock to keep more time between the rising edge and falling edge of a clock. This change is just for reading/writing the write shift register, we will use a 33 MHz clock for the analog readout in any case.
If we change like above, are there any concerns from the DRS4 side?

Best,
Seiya

Stefan Ritt wrote:

Dear Seiya,

1) That's correct. SRIN is ampled at the falling edge. Pleae make sure to obey the hold-time as written in the datasheet. P.12, Fig. 11: SRIN must be stable before the falling edge of SRCLK and tH after the falling clock. tH is 5ns according to table 1. 

2) The write shift register is a 8-bit shift register, with an input, output and clock. After the first clock pulse, the 7th bit is shown, after the second clock pulse the 6th bit and so on. You you need 8 clock pulses to read the whole register. At the same time you write to the register, so what ever is present at SRIN will replace the old 8 bits of that register.

3) No this is not possible. When you read the register, you write to it at the same time. One possibilty is to connect SROUT to SRIN during that (maybe via the FPGA). Then you have a circular register wich is the same after each 8 clock pulses.

For your reference, I posted a commercial D-Flip Flop (TI SNHCS72). The DRS4 write shift register is a simple array of 8 such registers, with no CLR or PRE, where SROUT is Q of the last Flip Flop.

Best,
Stefan

Seiya Nozaki wrote:

Dear Stefan,

I have questions about the timing diagram of SROUT/SRIN signal to write/read a write shift register.
1) Value of SRIN signal is saved at the falling edge of SRCLK, correct? (It is written in datasheet, page12, "Bits are latched into the shift register on the falling edge of SRCLK")
2) When are 8-bits of write shift register shown through SROUT? At ridging edges of SRCLK? and with additional delay(~10ns)? or falling edges?
3) In my understanding, when SRCLK is sent to DRS4, we can read and write the values in parallel, right? If so, is it possible just to read the registers without updating the registers?

[Background]
We have two modes to set the write shift register, the first one is to always reconnect to the first line and another one is to reconnect to the same line as when DWRITE goes to Low.
We can read/write the write shift register with the first mode (channel reset, page1). But we rarely face the problem of write shift register, unexpected values are written, with the second mode. With this mode, SROUT signal is sent back to DRS from FPGA as SRIN to write the same value on the write shift register. So there is a small delay(~10ns) due to the routing (DRS->FPGA->DRS, page2). It seems SRIN signal is not stable around the falling edges of SRCLK, thus it could cause that unexpected values are written in write shifter register.
To understand the situation clearly, I'd like to know the answer to the above questions.

Thank you.

Best regards,
Seiya

 

 

 

 

Entry  Thu Nov 8 11:44:35 2018, Davide Depaoli, Timing Issue 
Hi,

We are using the DRS4 Evaluation Board as a digitizer in our laboratory.
We found a strange behavior in the saved files, more specifically the time difference between two consecutive points is not constant, also after the Timing Calibration.
As an example, I paste a piece of a xml file saved using the drsosc program, acquiring CH1 (open):
	
	---------------------------
	---[ START XML EXAMPLE ]---
	---------------------------
	
	<?xml version="1.0" encoding="ISO-8859-1"?>
	<!-- created by MXML on Thu Nov  8 11:13:27 2018 -->
	<DRSOSC>
	<Event>
	<Serial>1</Serial>
	<Time>2018/11/08 11:13:27.163</Time>
	<HUnit>ns</HUnit>
	<VUnit>mV</VUnit>
	<Board_2796>
	<Trigger_Cell>216</Trigger_Cell>
	<Scaler0>0</Scaler0>
	<CHN1>
	<Data>0.000,-1.0</Data>
	<Data>1.083,-1.0</Data>
	<Data>2.143,-1.0</Data>
	<Data>2.926,-1.0</Data>
	<Data>4.249,-0.1</Data>
	<Data>4.929,-0.6</Data>
	<Data>6.075,-0.4</Data>
	<Data>7.042,0.0</Data>
	<Data>8.299,0.2</Data>
	
	[...]
	
	-------------------------
	---[ END XML EXAMPLE ]---
	-------------------------

We found the same behavior saving events in the binary format, and then reading them with the read_binary.cpp

Is there a way to fix our issue?

Thanks a lot

Davide and Alessio
    Reply  Thu Nov 8 11:54:33 2018, Stefan Ritt, Timing Issue 
That's not a bug, but a feature of the DRS4 chip. The time bins have different values by the properties of the chip. They are generated by a chain of inverters, which all have different propagation times. This delay is measured by the time calibration and then applied. If you want equidistant bins, 
you have to interpolate your data points (linearly or by splines) and resample the signal. You can find more details in the DRS4 data sheet.

Best,
Stefan


> Hi,
> 
> We are using the DRS4 Evaluation Board as a digitizer in our laboratory.
> We found a strange behavior in the saved files, more specifically the time difference between two consecutive points is not constant, also after the Timing Calibration.
> As an example, I paste a piece of a xml file saved using the drsosc program, acquiring CH1 (open):
> 	
> 	---------------------------
> 	---[ START XML EXAMPLE ]---
> 	---------------------------
> 	
> 	<?xml version="1.0" encoding="ISO-8859-1"?>
> 	<!-- created by MXML on Thu Nov  8 11:13:27 2018 -->
> 	<DRSOSC>
> 	<Event>
> 	<Serial>1</Serial>
> 	<Time>2018/11/08 11:13:27.163</Time>
> 	<HUnit>ns</HUnit>
> 	<VUnit>mV</VUnit>
> 	<Board_2796>
> 	<Trigger_Cell>216</Trigger_Cell>
> 	<Scaler0>0</Scaler0>
> 	<CHN1>
> 	<Data>0.000,-1.0</Data>
> 	<Data>1.083,-1.0</Data>
> 	<Data>2.143,-1.0</Data>
> 	<Data>2.926,-1.0</Data>
> 	<Data>4.249,-0.1</Data>
> 	<Data>4.929,-0.6</Data>
> 	<Data>6.075,-0.4</Data>
> 	<Data>7.042,0.0</Data>
> 	<Data>8.299,0.2</Data>
> 	
> 	[...]
> 	
> 	-------------------------
> 	---[ END XML EXAMPLE ]---
> 	-------------------------
> 
> We found the same behavior saving events in the binary format, and then reading them with the read_binary.cpp
> 
> Is there a way to fix our issue?
> 
> Thanks a lot
> 
> Davide and Alessio
    Reply  Thu Nov 8 12:02:34 2018, Davide Depaoli, Timing Issue 
Thanks a lot for the quick response.
We will do as you suggest.

Best regards

Davide and Alessio

> That's not a bug, but a feature of the DRS4 chip. The time bins have different values by the properties of the chip. They are generated by a chain of inverters, which all have different 
propagation times. This delay is measured by the time calibration and then applied. If you want equidistant bins, 
> you have to interpolate your data points (linearly or by splines) and resample the signal. You can find more details in the DRS4 data sheet.
> 
> Best,
> Stefan
> 
> 
> > Hi,
> > 
> > We are using the DRS4 Evaluation Board as a digitizer in our laboratory.
> > We found a strange behavior in the saved files, more specifically the time difference between two consecutive points is not constant, also after the Timing Calibration.
> > As an example, I paste a piece of a xml file saved using the drsosc program, acquiring CH1 (open):
> > 	
> > 	---------------------------
> > 	---[ START XML EXAMPLE ]---
> > 	---------------------------
> > 	
> > 	<?xml version="1.0" encoding="ISO-8859-1"?>
> > 	<!-- created by MXML on Thu Nov  8 11:13:27 2018 -->
> > 	<DRSOSC>
> > 	<Event>
> > 	<Serial>1</Serial>
> > 	<Time>2018/11/08 11:13:27.163</Time>
> > 	<HUnit>ns</HUnit>
> > 	<VUnit>mV</VUnit>
> > 	<Board_2796>
> > 	<Trigger_Cell>216</Trigger_Cell>
> > 	<Scaler0>0</Scaler0>
> > 	<CHN1>
> > 	<Data>0.000,-1.0</Data>
> > 	<Data>1.083,-1.0</Data>
> > 	<Data>2.143,-1.0</Data>
> > 	<Data>2.926,-1.0</Data>
> > 	<Data>4.249,-0.1</Data>
> > 	<Data>4.929,-0.6</Data>
> > 	<Data>6.075,-0.4</Data>
> > 	<Data>7.042,0.0</Data>
> > 	<Data>8.299,0.2</Data>
> > 	
> > 	[...]
> > 	
> > 	-------------------------
> > 	---[ END XML EXAMPLE ]---
> > 	-------------------------
> > 
> > We found the same behavior saving events in the binary format, and then reading them with the read_binary.cpp
> > 
> > Is there a way to fix our issue?
> > 
> > Thanks a lot
> > 
> > Davide and Alessio
ELOG V3.1.5-3fb85fa6