DRS4 Forum
  DRS4 Discussion Forum, Page 18 of 46  Not logged in ELOG logo
New entries since:Thu Jan 1 01:00:00 1970
ID Date Author Subjectdown Text Attachments
  624   Wed Jul 12 20:16:05 2017 Stefan RittTime resolution between boardsYes this should  be possible.

Stefan

  
  626   Fri Jul 21 09:16:02 2017 Volodymyr RodinTime outputHello Stefan

I tried to convert binary to a
simple txt file and found next problem -
  
  627   Tue Jul 25 14:47:05 2017 Volodymyr RodinTime outputHi again.

Okay, it works with 5.05 version very
good and it is enough for me.
  
  633   Tue Oct 17 14:58:58 2017 Vadym DenysenkoTime offset Hello.

 

I have a simple question, can I
  
  634   Wed Oct 18 09:12:26 2017 Stefan RittTime offset No this is not possible. But you can delay
your signal externally (like with a delay
cable or electronically) and then send the
  
  635   Wed Oct 18 11:48:14 2017 Vadym DenysenkoTime offset Thank you for your reply!




  
  877   Fri Mar 11 17:26:15 2022 Matias SengerTime calibration and the C++ APII am using the V5 board at a fixed sampling
frequency. With the `drsosc` app I have executed
the time calibration at 5 GS/s (actually
  
  878   Sat Mar 12 10:13:24 2022 Stefan RittTime calibration and the C++ APIDRSBoard::GetTime is declared in DRS.h
line 720.

If you want to measure timing down
  
  879   Sat Mar 12 16:52:36 2022 Matias SengerTime calibration and the C++ APIDear Stefan,

For the time of each bin I am using
the values returend by `GetTime` without
  
  880   Mon Mar 14 08:59:51 2022 Stefan RittTime calibration and the C++ APILooks like you have the some time calibration,
not sure if it's the correct one. Sample
the sine wave from the calibration clock,
 Screenshot_2022-03-14_at_9.04.07_.pngScreenshot_2022-03-14_at_9.03.47_.png 
  881   Tue Mar 15 13:07:50 2022 Matias SengerTime calibration and the C++ APIThanks for your help. If I look into the
app the behavior for the 4 channels is exactly
as you show:
  
  651   Wed Jan 17 09:51:16 2018 Tran Cong ThienThe input signals recorded are different with the signal showed in oscilloscope Dear Stefan,

I am using an DRS4 board to record
the signals from an plastic scintillator
  
  652   Wed Jan 17 10:09:09 2018 Stefan RittThe input signals recorded are different with the signal showed in oscilloscope First thing is to do another voltage calibration.
Disconnect input, "Config", "Execute
Voltage Calibration". If this does not
  
  340   Thu Apr 17 12:02:28 2014 Wang The first channel is wrong. Hi,  QQ??20140417174309.jpg 
  703   Tue Jun 19 06:42:23 2018 Phan Van ChuanThe data acquisition speedDear Stefan,

We are using an DRS4 board V5.1
for building a metering system for the scintillator
 wavech0.png 
  704   Tue Jun 19 10:05:50 2018 Stefan RittThe data acquisition speedHow do you tigger the board? In your code
below you start the board (StartDomino())
and then wait for a trigger. Setting the
  
  705   Tue Jun 19 12:54:51 2018 Phan Van ChuanThe data acquisition speedThank Stefan Ritt, I added the SoftTrigger()
just after StartDomino(), so now, The data
acquisition speed the same speed as in the
  
  140   Wed Dec 14 00:44:37 2011 Hao HuanSynchronization Delay in the Firmware for 8051 ControllerHi Stefan,
    I have a question
regarding the DRS 4 evaluation board firmware
  
  141   Wed Dec 14 08:55:29 2011 Stefan RittSynchronization Delay in the Firmware for 8051 Controller

    

       
            
  
  600   Thu Apr 13 16:42:21 2017 Christian FarinaStand-alone Time Calibration for PSI BoardHello everybody,

I was trying to create a stand-alone
program that would perform a time calibration
  
ELOG V3.1.5-3fb85fa6