Entry  Wed Nov 21 08:34:52 2012, Gyuhee Kim, Question for using Multi board 

 Hi.

 

I have 2 DRS4 evaluation V4 boards, and I want to use these 2 board to multi board DAQ system for 4 ch vs 4 ch DAQ.

But there is no option for multi board use. I just only find the multi board trigger mode check button on DRS4 Oscilloscope program, but I couldn`t check. 

Is there any method to use multi board?

 

Best regards.

Gyuhee.

    Reply  Wed Nov 21 08:48:00 2012, Gyuhee Kim, Question for using Multi board 

Stefan Ritt wrote:

Gyuhee Kim wrote:

 Hi.

 

I have 2 DRS4 evaluation V4 boards, and I want to use these 2 board to multi board DAQ system for 4 ch vs 4 ch DAQ.

But there is no option for multi board use. I just only find the multi board trigger mode check button on DRS4 Oscilloscope program, but I couldn`t check. 

Is there any method to use multi board?

 

Best regards.

Gyuhee.

This mode is not yet implemented in firmware. Maybe I find some time towards the end of this year to add this. At the moment, you have to build and external trigger to synchronize the two boards. There are also 16-channel boards on the market where you would not need a multi-board mode. Just Google for "DT5742".

/Stefan

 Thanks Stefan.

I will build external trigger system. 

Entry  Mon Dec 3 08:32:28 2012, Gyuhee Kim, Another question about using multi boards. 

 Hi.

 

I asked about using multi boards some days ago, and I got answer to use external trigger. (Thanks Stefan!)

And here is another question. I made two external triggers and try to acquire coincidence data using two boards. but DRS Oscilloscope program can connect only one board and don`t acquire both of them simultaneously.

So I tried to use two computer for each board separately, but, well, you already know, I failed to acquire because two computers don`t promise to synchronize the two boards acquisition.

Is there any method to solve this problem?

 

1. I want to acquire coincidence data from the two DRS 4 Evaluation board V4 simultaneosly.

2. I have external trigger to provide the two boards at the same time.

3. How can I get data from the two boards?

 

Best regards.

Gyuhee.

    Reply  Mon Dec 3 11:40:35 2012, Gyuhee Kim, Another question about using multi boards. 

Stefan Ritt wrote:

Gyuhee Kim wrote:

 Hi.

 

I asked about using multi boards some days ago, and I got answer to use external trigger. (Thanks Stefan!)

And here is another question. I made two external triggers and try to acquire coincidence data using two boards. but DRS Oscilloscope program can connect only one board and don`t acquire both of them simultaneously.

So I tried to use two computer for each board separately, but, well, you already know, I failed to acquire because two computers don`t promise to synchronize the two boards acquisition.

Is there any method to solve this problem?

 

1. I want to acquire coincidence data from the two DRS 4 Evaluation board V4 simultaneosly.

2. I have external trigger to provide the two boards at the same time.

3. How can I get data from the two boards?

 

Best regards.

Gyuhee.

You have to write your own program. DRS Oscilloscope does not (yet) support this. Take drs_exam.cpp as a starting point and try to extend it to two boards. One tricky point is that the external trigger may only fire AFTER the two boards have been read out. So you need some means of re-enabling the external trigger after you read out both boards.

Stefan 

That`s very bad news for me. I don`t have much time to study & write C programming...

Anyway, Thank you very much Stefan.

 

Best regards.

Gyuhee.

Entry  Wed Sep 7 16:45:17 2011, Guillaume Blanchard, DRS4 and AD9222 

Hello,

I am designing a DAQ board with both DRS4 + AD9222 and a  FPGA to monitor.

Do I have to change the default value of O-OFS ?

Does a simple low-pass filter (series resistor + capacitor) on each AD9222 input is enough to limit the noise ?

I am planning to use the (DRS4,AD9222,FPGA) group as both a trigger and digitizing system (as shown in the DRS4 datasheet). The DRS4 will be working at 5Ghz with 8 active channels.
So each channel will have a time depth of 1/5Ghz x 1024 = 204.8ns. So, in order to miss nothing, the ADC latency + the trigger decision must be inferior to 204.8ns, am I correct ?
This leads me to implement on my board the 65Mhz version of the AD9222 as this converter has a 8 clock period latency, i.e. 123ns and it left me 81ns to perform a trigger decision ?

Cordially,

G.Blanchard

    Reply  Fri Sep 9 09:28:57 2011, Guillaume Blanchard, DRS4 and AD9222 

Thank you for your answers,

Another question : Have you ever tried to split the differential signal at the output of the DRS4 chip ? For example to feed both an AD9222 and a diff. amplifier (followed by discriminators) ?

 

Entry  Mon Apr 23 10:38:51 2012, Guillaume Blanchard, DRS4 Initialization 

Hello,

I am writing a VHDL code to drive a DRS4 chip.

In order to configure the DRS4 chip, I have to set the "Config Register" and the "Write Shift Register" then ... (I do not plan to use simultaneous WR and R so I guess the Write Config Reg. is not needed)

My question is :

When do we have to perform a "Read Shift Register Initialization" ?

Every time before a full read-out, or juste once after a DRS4 reset ?

Further more, is this initialization needed for the ROI mode ?

And at last do the level of the DENABLE and DWRITE signals matter for the "Read Shift Register Initialization" ?

(To sum up : what is the purpose of the Read Shift Register and how does it work ?)

Cordially,

G.Blanchard.

Entry  Thu Jun 18 17:33:05 2015, Gregor Kramberger, drs 5.03 and windows 8.1 

I have problems with driver installation on windows 8.1 (software version 5.03). I have sen that that has been an issue before (driver signing) and I would like to know if this has been solved. We run several DRS4 evaluation boards on different PCs all running Win7 without any problems. Therefore we are almost confident that it is related to Win 8.1. Thanks.

 

    Reply  Fri Jun 19 12:32:10 2015, Gregor Kramberger, drs 5.03 and windows 8.1 

 

Gregor Kramberger wrote:

I have problems with driver installation on windows 8.1 (software version 5.03). I have sen that that has been an issue before (driver signing) and I would like to know if this has been solved. We run several DRS4 evaluation boards on different PCs all running Win7 without any problems. Therefore we are almost confident that it is related to Win 8.1. Thanks.

 

Solved. Need to restart Windows 8.1 (64 bit) in recovery mode and dissable driver signing as mandatory. Then it works fine.

Entry  Fri Jan 13 12:58:22 2017, Gregor Kramberger, DRS software doesn't work under Windows XP SP3 

Hi all

I have a problem with running the DRSOSC under windows XP SP3. We have some hardware which is not supported under newer versions of windows and we would like to use DRS boards along it, therefore we would higly appreciated any help in that direction. We have installed the software (V 5.03) to two different XP machines and got the same problem. The driver installs without any problem, but when the drsosc is run the system says " drsosc.exe is not a valid Win32 application". We have developed our own API for our software which also doesn't recognize the board. It says on the www page that it has been tested for windows XP, but I would appreciate if you can verify it? With best regards and thanks...

Entry  Mon Apr 10 08:50:11 2017, Giovanni Bruni, drs4 registers behaviour 

Hej everyone!
I have some questions regarding what happens to some DRS registers in some scenarios:
1. How are the registers affected by a RESET? According to the data sheet all the CONFIG REGISTER bits are initilialized to 1. But what about the WRITE SHIFT and the READ SHIFT registers? Are they affected somehow after a RESET has been applied?
2. Suppose the DRS is happily running and I have done some readouts in ROI mode, so that the only "1"-value bit in the READ SHIFT register is in a random position. If now I want to execute a FULL READOUT, should I use the procedure explained in the data sheet (figure 11) for the FULL READOUT mode? or is this procedure useless since my "1"-value bit is already set somewhere in the READ SHIFT register and therefore a ROI readout of 1024 cells would be the solution (and getting the initial position from the SROUT pin)?

Thanks a lot!
Have a nice day!

Giovanni

    Reply  Mon Apr 10 13:41:41 2017, Giovanni Bruni, drs4 registers behaviour 

Hej Stefan! Thank you for your answer!

Just to be sure to have understood properly:
1. Using the RESET line should be avoided. And in any case, the CONFIG register and the WRITE SHIFT register need to be initialized "by hand" using the A0-A3, SRCLK and SRIN pins. Is it correct?

2. Doing the procedure shown in Figure 11 will always inject a "1" in cell #0 of the READ SHIFT register, regardless if (before starting the procedure) there was a "1" in any other cell, right?

Thank you!
Giovanni
 

Stefan Ritt wrote:

Using the RESET line to reset registers is not a good idea since it can have some bad side-effects. The READ SHIFT register is NOT affected by RESET, so you have to inititialize these registers differently. To set a "1"-value at a defined position, you have to follow figure 11 in the data sheet. Once you executed that, your "1" is always at the same posiiton (namely cell #0), so after 1024 clock cycles you arrive at the same state, and do not have to re-do fig. 11 again.

Stefan

 

Giovanni Bruni wrote:

Hej everyone!
I have some questions regarding what happens to some DRS registers in some scenarios:
1. How are the registers affected by a RESET? According to the data sheet all the CONFIG REGISTER bits are initilialized to 1. But what about the WRITE SHIFT and the READ SHIFT registers? Are they affected somehow after a RESET has been applied?
2. Suppose the DRS is happily running and I have done some readouts in ROI mode, so that the only "1"-value bit in the READ SHIFT register is in a random position. If now I want to execute a FULL READOUT, should I use the procedure explained in the data sheet (figure 11) for the FULL READOUT mode? or is this procedure useless since my "1"-value bit is already set somewhere in the READ SHIFT register and therefore a ROI readout of 1024 cells would be the solution (and getting the initial position from the SROUT pin)?

Thanks a lot!
Have a nice day!

Giovanni

 

 

    Reply  Tue Apr 11 09:07:33 2017, Giovanni Bruni, drs4 registers behaviour 

Thank you Stefan for replying!
I have still the RESET issue in mind: how would you suggest to reset properly the DRS? Is there a particular procedure to follow instead of just sending a negative pulse to the RESET pin? Is it preferable to turn the DRS off and then restart?

Thanks!

Giovanni

 

Stefan Ritt wrote:

1. WRITE SHIFT register and CONFIG registers are initialized to "1" on power up, but if you want to change that, use A0-A3 etc. as you indicated.

2. If you address the READ SHIFT register by applyin "1011" to A0-A3, the input of the register is connected to SRIN. So in fig. 11, you apply 1023x"0" plus 1x"1", which effectively clears the register and keeps one "1" at the last position, so on the next rising clock this gets shifted into position #0. If you do the readout, and NOT addresing the READ SHIFT register, then the input of that register is connected to it's output internally. Therefore the single "1" keep rotating on every 1024 clock cycles.

Giovanni Bruni wrote:

Hej Stefan! Thank you for your answer!

Just to be sure to have understood properly:
1. Using the RESET line should be avoided. And in any case, the CONFIG register and the WRITE SHIFT register need to be initialized "by hand" using the A0-A3, SRCLK and SRIN pins. Is it correct?

2. Doing the procedure shown in Figure 11 will always inject a "1" in cell #0 of the READ SHIFT register, regardless if (before starting the procedure) there was a "1" in any other cell, right?

Thank you!
Giovanni
 

Stefan Ritt wrote:

Using the RESET line to reset registers is not a good idea since it can have some bad side-effects. The READ SHIFT register is NOT affected by RESET, so you have to inititialize these registers differently. To set a "1"-value at a defined position, you have to follow figure 11 in the data sheet. Once you executed that, your "1" is always at the same posiiton (namely cell #0), so after 1024 clock cycles you arrive at the same state, and do not have to re-do fig. 11 again.

Stefan

 

Giovanni Bruni wrote:

Hej everyone!
I have some questions regarding what happens to some DRS registers in some scenarios:
1. How are the registers affected by a RESET? According to the data sheet all the CONFIG REGISTER bits are initilialized to 1. But what about the WRITE SHIFT and the READ SHIFT registers? Are they affected somehow after a RESET has been applied?
2. Suppose the DRS is happily running and I have done some readouts in ROI mode, so that the only "1"-value bit in the READ SHIFT register is in a random position. If now I want to execute a FULL READOUT, should I use the procedure explained in the data sheet (figure 11) for the FULL READOUT mode? or is this procedure useless since my "1"-value bit is already set somewhere in the READ SHIFT register and therefore a ROI readout of 1024 cells would be the solution (and getting the initial position from the SROUT pin)?

Thanks a lot!
Have a nice day!

Giovanni

 

 

 

 

    Reply  Fri Nov 18 16:38:42 2016, Gerard Montarou, LabView 

Hello,

Did you start to write some VI to interface DRS4board with labview ?

i also have in mind to do that.I am surprised that nobody alraedy did it since there is no answer toyour question

gerard

Christian D wrote:

Hi,

I would like to use the DRS4 board with LabView for fast readout.
Do you know anyone who has written a VI for that?

Thanks,
Christian

 

Entry  Sun Sep 23 02:22:46 2018, Gerard Arino-Estrada, Trigger OUT pulse width variable from 100 us up to 100 ms 

Hello Stefan,

I am using the DRS4 board connected to a Raspberry PI and through the drsosc application. I am interested on using the "Trigger OUT" signal to do some extra data processing with NIM modules. According to the manual, for each hardware trigger a TTL pulse of 150 ns width should be send through the "trigger OUT". In my case I do see pulses with widths ranging from 100 microseconds up to hundreds of miliseconds. I am connecting the signal directly to an oscilloscope with 50 Ohm termination. I have tried two DRS4 boards in identical conditions and both show the same behavior.  Having such wide and variable pulses makes it complicated for me to do the extra post-processing. Have you any idea of what might be going wrong? Thank you very much.

Best regards,

Gerard

    Reply  Wed Sep 26 18:25:07 2018, Gerard Arino-Estrada, Trigger OUT pulse width variable from 100 us up to 100 ms 

Thank you very much for the answer, I really appreciate your help.

Thanks!

Gerard

Stefan Ritt wrote:

The "Trigger OUT" has changed recently. It goes high on a new trigger, but then STAYS high until the board has been read out by the PC and re-started. This allows better synchronization with some external trigger, which can be re-armed with the falling edge of the trigger out signal. The signal can be quite long, since readout of an event via USB typically takes 2 ms, but can be more if the PC is busy. If you need back your 150 ns pulse, send the trigger out to an external pulse shaper with fixed shaping width.

Stefan

Gerard Arino-Estrada wrote:

Hello Stefan,

I am using the DRS4 board connected to a Raspberry PI and through the drsosc application. I am interested on using the "Trigger OUT" signal to do some extra data processing with NIM modules. According to the manual, for each hardware trigger a TTL pulse of 150 ns width should be send through the "trigger OUT". In my case I do see pulses with widths ranging from 100 microseconds up to hundreds of miliseconds. I am connecting the signal directly to an oscilloscope with 50 Ohm termination. I have tried two DRS4 boards in identical conditions and both show the same behavior.  Having such wide and variable pulses makes it complicated for me to do the extra post-processing. Have you any idea of what might be going wrong? Thank you very much.

Best regards,

Gerard

 

 

    Reply  Wed Sep 26 18:28:20 2018, Gerard Arino-Estrada, Trigger OUT pulse width variable from 100 us up to 100 ms 

Thank you very much for the answer, I really appreciate your help.

Thanks!

Gerard

Stefan Ritt wrote:

The "Trigger OUT" has changed recently. It goes high on a new trigger, but then STAYS high until the board has been read out by the PC and re-started. This allows better synchronization with some external trigger, which can be re-armed with the falling edge of the trigger out signal. The signal can be quite long, since readout of an event via USB typically takes 2 ms, but can be more if the PC is busy. If you need back your 150 ns pulse, send the trigger out to an external pulse shaper with fixed shaping width.

Stefan

Gerard Arino-Estrada wrote:

Hello Stefan,

I am using the DRS4 board connected to a Raspberry PI and through the drsosc application. I am interested on using the "Trigger OUT" signal to do some extra data processing with NIM modules. According to the manual, for each hardware trigger a TTL pulse of 150 ns width should be send through the "trigger OUT". In my case I do see pulses with widths ranging from 100 microseconds up to hundreds of miliseconds. I am connecting the signal directly to an oscilloscope with 50 Ohm termination. I have tried two DRS4 boards in identical conditions and both show the same behavior.  Having such wide and variable pulses makes it complicated for me to do the extra post-processing. Have you any idea of what might be going wrong? Thank you very much.

Best regards,

Gerard

 

 

Entry  Wed Feb 27 13:47:32 2013, Georg Winner, Chip Test - Cell Error 

When starting Chip Test in DRS Command Line Interface, I receive the following message:

Cell error on channel 1, cell 5: -154.4 mV instead 0 mV

Chip Error!

What does this mean? The maximal peak-to-peak Amplitude given to channel was for a short time 10V.

The graphical interface shows no artefacts when using channel 1.

 

Entry  Mon Mar 25 11:12:53 2013, Georg Winner, Differences in Source Code 

I have noticed some differences in the source code between Windows (4.0.0) and Linux (4.0.1) Version.

drs_exam.cpp: In the windows version when setting the trigger there is no part "if (b->GetBoardType() == 8) {...} else {...}" like in Linux version. So under Windows drs_exam does not start readout of DRS 4 Evalutation Board V4, because it does not get the trigger, under linux the board can be read out succesfull. I have found out, that adding the missing part solves the problem for the windows version.

drs.cpp (Windows Version), line 2101, function "int DRSBoard::SetTriggerDelayNs(int delay)":

There is no operation which  calculates the variable "fTriggerDelayNs" out of variable "ticks" like in function "int DRSBoard::SetTriggerDelayPercent(int delay)" (Line 2073). So "fTriggerDelayNs" can get diverse values when using one of the Trigger Setting Functions. Was this intended?

 

Entry  Thu Apr 9 11:46:33 2015, Felix Bachmair, DRSBoard::SetTriggerSource 

Hi

I have a question about the function SetTriggerSource in the class DRSBoard (DRS.h/DRS.cpp)

In the implementation there is the following comment:

// Set trigger configuration

// OR 0=CH1, 1=CH2, 2=CH3, 3=CH4, 4=EXT

// AND 8=CH1, 9=CH2, 10=CH3, 11=CH4, 12=EXT

 

What does this exactly mean? I am assuming that this are the bits which are set?

e.g

source = 1 ==> CH1

source = 4352 = 0x1100 ==> CH1 and ext

How is the AND/Or logic implemented?

When i have:

source = 0x1803 (bit 12,11,1,0)

what is the right way to set the brackets to expalin the logic?

(EXT and CH4 ) or CH2 or CH1 ?

 

Cheers Felix Bachmair

ETH Zurich