DRS4 Forum
  DRS4 Discussion Forum, Page 38 of 46  Not logged in ELOG logo
New entries since:Thu Jan 1 01:00:00 1970
Entry  Sun May 2 18:36:14 2010, Ignacio Diéguez Estremera, DRS4 chip model 

Hi all,

i'm an electronics engineering student at UCM (Madrid) working on my master's thesis within the CTA collaboration. I'm designing the readout electronics for the telescope's camera, and i'm focusing in using GAPDs instead of PMTs and using the domino chip for the sampling of the signal. I was wondering if there is a spice and/or RF model of the DRS4 chip available. It would be very useful to perform some simulations before deciding to use the chip as the sampling solution for our prototypes.

If the answer is negative, can you give me some advise for modelling the chip in spice? Have you done any simulations?

Thanks in advance,

Ignacio Diéguez Estremera.

    Reply  Mon May 3 11:09:12 2010, Stefan Ritt, DRS4 chip model 

Ignacio Diéguez Estremera wrote:

Hi all,

i'm an electronics engineering student at UCM (Madrid) working on my master's thesis within the CTA collaboration. I'm designing the readout electronics for the telescope's camera, and i'm focusing in using GAPDs instead of PMTs and using the domino chip for the sampling of the signal. I was wondering if there is a spice and/or RF model of the DRS4 chip available. It would be very useful to perform some simulations before deciding to use the chip as the sampling solution for our prototypes.

If the answer is negative, can you give me some advise for modelling the chip in spice? Have you done any simulations?

Thanks in advance,

Ignacio Diéguez Estremera.

Yes there is a transistor-level spice model, which I used to design the chip, but you won't be happy with it: Given the 500,000+ transistors on the chip, a 100 ns simulation takes a couple of weeks. We tried to make a simplified model just for the analog input using some measured S-parameters, but found that the RF behavior of the chip is almost impossible to describe to better than let's say 50%. In the end you have to fine-tune your analog front-end experimentally, to obtain optimal bandwidth. We are just working on a reference design with gives you 850 MHz bandwidth using an active input buffer.

    Reply  Mon May 3 17:06:02 2010, Ignacio Diéguez Estremera, DRS4 chip model 

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

Hi all,

i'm an electronics engineering student at UCM (Madrid) working on my master's thesis within the CTA collaboration. I'm designing the readout electronics for the telescope's camera, and i'm focusing in using GAPDs instead of PMTs and using the domino chip for the sampling of the signal. I was wondering if there is a spice and/or RF model of the DRS4 chip available. It would be very useful to perform some simulations before deciding to use the chip as the sampling solution for our prototypes.

If the answer is negative, can you give me some advise for modelling the chip in spice? Have you done any simulations?

Thanks in advance,

Ignacio Diéguez Estremera.

Yes there is a transistor-level spice model, which I used to design the chip, but you won't be happy with it: Given the 500,000+ transistors on the chip, a 100 ns simulation takes a couple of weeks. We tried to make a simplified model just for the analog input using some measured S-parameters, but found that the RF behavior of the chip is almost impossible to describe to better than let's say 50%. In the end you have to fine-tune your analog front-end experimentally, to obtain optimal bandwidth. We are just working on a reference design with gives you 850 MHz bandwidth using an active input buffer.

 Thanks for the information.

I would like to try the huge :-) model. Can you send it to my email address? Since the input signal are pulses of a few nanoseconds at FHWM, the simulation time may be reduced. I will post some feedback in the forum once i give it a try.

Kind regards.

    Reply  Mon May 3 17:10:29 2010, Stefan Ritt, DRS4 chip model 

Ignacio Diéguez Estremera wrote:

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

Hi all,

i'm an electronics engineering student at UCM (Madrid) working on my master's thesis within the CTA collaboration. I'm designing the readout electronics for the telescope's camera, and i'm focusing in using GAPDs instead of PMTs and using the domino chip for the sampling of the signal. I was wondering if there is a spice and/or RF model of the DRS4 chip available. It would be very useful to perform some simulations before deciding to use the chip as the sampling solution for our prototypes.

If the answer is negative, can you give me some advise for modelling the chip in spice? Have you done any simulations?

Thanks in advance,

Ignacio Diéguez Estremera.

Yes there is a transistor-level spice model, which I used to design the chip, but you won't be happy with it: Given the 500,000+ transistors on the chip, a 100 ns simulation takes a couple of weeks. We tried to make a simplified model just for the analog input using some measured S-parameters, but found that the RF behavior of the chip is almost impossible to describe to better than let's say 50%. In the end you have to fine-tune your analog front-end experimentally, to obtain optimal bandwidth. We are just working on a reference design with gives you 850 MHz bandwidth using an active input buffer.

 Thanks for the information.

I would like to try the huge :-) model. Can you send it to my email address? Since the input signal are pulses of a few nanoseconds at FHWM, the simulation time may be reduced. I will post some feedback in the forum once i give it a try.

Kind regards.

I just checked and realized that we are not allowed to give out the "huge" model since it contains parameters from the chip manufacturer's library which are confidentially. 

    Reply  Mon May 3 23:21:55 2010, Ignacio Diéguez Estremera, DRS4 chip model 

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

Hi all,

i'm an electronics engineering student at UCM (Madrid) working on my master's thesis within the CTA collaboration. I'm designing the readout electronics for the telescope's camera, and i'm focusing in using GAPDs instead of PMTs and using the domino chip for the sampling of the signal. I was wondering if there is a spice and/or RF model of the DRS4 chip available. It would be very useful to perform some simulations before deciding to use the chip as the sampling solution for our prototypes.

If the answer is negative, can you give me some advise for modelling the chip in spice? Have you done any simulations?

Thanks in advance,

Ignacio Diéguez Estremera.

Yes there is a transistor-level spice model, which I used to design the chip, but you won't be happy with it: Given the 500,000+ transistors on the chip, a 100 ns simulation takes a couple of weeks. We tried to make a simplified model just for the analog input using some measured S-parameters, but found that the RF behavior of the chip is almost impossible to describe to better than let's say 50%. In the end you have to fine-tune your analog front-end experimentally, to obtain optimal bandwidth. We are just working on a reference design with gives you 850 MHz bandwidth using an active input buffer.

 Thanks for the information.

I would like to try the huge :-) model. Can you send it to my email address? Since the input signal are pulses of a few nanoseconds at FHWM, the simulation time may be reduced. I will post some feedback in the forum once i give it a try.

Kind regards.

I just checked and realized that we are not allowed to give out the "huge" model since it contains parameters from the chip manufacturer's library which are confidentially. 

 Thank you for the effort anyway.

So i guess i won't be able to include drs4 in my simulations :-(. Any other suggestions? Maybe the S-params model you where working on? Anything is better than nothing :-)

    Reply  Tue May 4 11:26:21 2010, Stefan Ritt, DRS4 chip model DRS4_S-Parameter.pdf

Ignacio Diéguez Estremera wrote:

So i guess i won't be able to include drs4 in my simulations :-(. Any other suggestions? Maybe the S-params model you where working on? Anything is better than nothing :-)

Please find attached the S-parameters. 

    Reply  Tue May 4 16:23:16 2010, Ignacio Diéguez Estremera, DRS4 chip model 

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

So i guess i won't be able to include drs4 in my simulations :-(. Any other suggestions? Maybe the S-params model you where working on? Anything is better than nothing :-)

Please find attached the S-parameters. 

 Thanks :-)

    Reply  Wed May 12 11:47:39 2010, Jinhong Wang, DRS4 chip model 

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

So i guess i won't be able to include drs4 in my simulations :-(. Any other suggestions? Maybe the S-params model you where working on? Anything is better than nothing :-)

Please find attached the S-parameters. 

 Hi, we plan to do a time interpolating among the eight channels on a single chip to obtain a maximum 40 GSPS (or, maybe 30 GSPS ) sampling rate.  Hence RF behavior of the anlog input is very important for us.

Will you give us some advice on the modeling of  the anlog input circuit of the chip?  Perhaps just the Spice model of the analog input?

The attached S parameters I found  here is for fs =1 GSPS, what about fs=5GSPS?

thanks in advance,

                                                                               Jinhong Wang (wangjinh@mail.ustc.edu.cn  ;  alleyor.wang@gmail.com)

    Reply  Wed May 12 16:26:12 2010, Stefan Ritt, DRS4 chip model 

Jinhong Wang wrote:

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

So i guess i won't be able to include drs4 in my simulations :-(. Any other suggestions? Maybe the S-params model you where working on? Anything is better than nothing :-)

Please find attached the S-parameters. 

 Hi, we plan to do a time interpolating among the eight channels on a single chip to obtain a maximum 40 GSPS (or, maybe 30 GSPS ) sampling rate.  Hence RF behavior of the anlog input is very important for us.

Will you give us some advice on the modeling of  the anlog input circuit of the chip?  Perhaps just the Spice model of the analog input?

The attached S parameters I found  here is for fs =1 GSPS, what about fs=5GSPS?

thanks in advance,

                                                                               Jinhong Wang (wangjinh@mail.ustc.edu.cn  ;  alleyor.wang@gmail.com)

To be honest, we never really succeeded to do any good simulation above let's say 500 MHz. We carefully tried to simulate the bond wire of the chip, the parasitic capacitances of the traces of the chip etc. but we always were off by a factor or two or so. Other groups reported the same problems. Some even did some 3D simulation model, without success. So our conclusion is that if you are interested in anything precise above 500 MHz, do a measurement.

So our current best design is with the THS4508. There is an AC coupled version going to 600 MHz, and a DC coupled version (uses more power) going to 800 MHz (-3dB). If you use passive inputs with a transformer for example, you can't go above 220 MHz. Next week I will publish both designs in this forum.

Entry  Wed Sep 7 16:45:17 2011, Guillaume Blanchard, DRS4 and AD9222 

Hello,

I am designing a DAQ board with both DRS4 + AD9222 and a  FPGA to monitor.

Do I have to change the default value of O-OFS ?

Does a simple low-pass filter (series resistor + capacitor) on each AD9222 input is enough to limit the noise ?

I am planning to use the (DRS4,AD9222,FPGA) group as both a trigger and digitizing system (as shown in the DRS4 datasheet). The DRS4 will be working at 5Ghz with 8 active channels.
So each channel will have a time depth of 1/5Ghz x 1024 = 204.8ns. So, in order to miss nothing, the ADC latency + the trigger decision must be inferior to 204.8ns, am I correct ?
This leads me to implement on my board the 65Mhz version of the AD9222 as this converter has a 8 clock period latency, i.e. 123ns and it left me 81ns to perform a trigger decision ?

Cordially,

G.Blanchard

    Reply  Wed Sep 7 16:56:43 2011, Stefan Ritt, DRS4 and AD9222 

Guillaume Blanchard wrote:

Hello,

I am designing a DAQ board with both DRS4 + AD9222 and a  FPGA to monitor.

Do I have to change the default value of O-OFS ?

Does a simple low-pass filter (series resistor + capacitor) on each AD9222 input is enough to limit the noise ?

I am planning to use the (DRS4,AD9222,FPGA) group as both a trigger and digitizing system (as shown in the DRS4 datasheet). The DRS4 will be working at 5Ghz with 8 active channels.
So each channel will have a time depth of 1/5Ghz x 1024 = 204.8ns. So, in order to miss nothing, the ADC latency + the trigger decision must be inferior to 204.8ns, am I correct ?
This leads me to implement on my board the 65Mhz version of the AD9222 as this converter has a 8 clock period latency, i.e. 123ns and it left me 81ns to perform a trigger decision ?

Cordially,

G.Blanchard

I'm not so sure about the temperature stability of the default (DRS4 internal O-OFS) value, that's why I used a precision DAC in my design. But actually I never tried without. Probably the default internal value is good enough if you calibrate each chip (which you do anyhow).

Most designs use no filter between DRS4 and AD9222. Since the DRS4 output is BW limited at around 50 MHz, there is not much you win if you put a 32.5 MHz low pass there. And the PCS gets just so much simpler.

You are right with the latency of the AD9222, this is an issue. From the remaining 81 ns you loose a few going out of the FPGA, you need one or two FPGA clock cycles to make the decision, the DRS4 has also some stopping latency (maybe 10 ns). So you are at the edge. Some people use hardware comparators which are faster than the AD9222, one guy uses even directly an LVDS input of the FPGA "mis-used" as a comparator, where the comparator level (=LVDS negative input) comes from a DAC.

 

- Stefan

    Reply  Wed Sep 7 17:28:25 2011, Hannes Friederich, DRS4 and AD9222 

Guillaume Blanchard wrote:

Hello,

I am designing a DAQ board with both DRS4 + AD9222 and a  FPGA to monitor.

Do I have to change the default value of O-OFS ?

Does a simple low-pass filter (series resistor + capacitor) on each AD9222 input is enough to limit the noise ?

I am planning to use the (DRS4,AD9222,FPGA) group as both a trigger and digitizing system (as shown in the DRS4 datasheet). The DRS4 will be working at 5Ghz with 8 active channels.
So each channel will have a time depth of 1/5Ghz x 1024 = 204.8ns. So, in order to miss nothing, the ADC latency + the trigger decision must be inferior to 204.8ns, am I correct ?
This leads me to implement on my board the 65Mhz version of the AD9222 as this converter has a 8 clock period latency, i.e. 123ns and it left me 81ns to perform a trigger decision ?

Cordially,

G.Blanchard

 Like Stefan pointed out, your time constraints are quite tight. In those 81 ns, you also need to deserialize the AD9222 output. Unless you implement some really fancy input comparison logic, this will consume another 1-2 ADC clock cycles. Perhaps you should first verify that your FPGA design actually can do its job within those 81 ns. In our system, we sample at only 1-2 GHz and have enough margin to implement really complex triggers in FPGA. But the total latency (ADC + FPGA deserialization) takes 250 ns.

Depending on the application, you do need a low-pass filter. Not only because of the noise, but also in order to be able to trigger reliably. Using fast PMTs for example, you will not be able to see all pulses in full size if the bandwidth is 50 MHz and you're only sampling at 65 MSPS.

Hannes

    Reply  Fri Sep 9 09:28:57 2011, Guillaume Blanchard, DRS4 and AD9222 

Thank you for your answers,

Another question : Have you ever tried to split the differential signal at the output of the DRS4 chip ? For example to feed both an AD9222 and a diff. amplifier (followed by discriminators) ?

 

    Reply  Fri Sep 9 09:31:33 2011, Stefan Ritt, DRS4 and AD9222 

Guillaume Blanchard wrote:

Thank you for your answers,

Another question : Have you ever tried to split the differential signal at the output of the DRS4 chip ? For example to feed both an AD9222 and a diff. amplifier (followed by discriminators) ?

 

Yes. Just have a look at the schematics of the evaluation board. This is exactly what is done there.

Actually in the newest version we went one step further and put the comparator at the input of the DRS chip. This way it is active even during the readout of the DRS4 chip and we can use this as a counter to count the overall hit rate at the input.

 

- Stefan

Entry  Mon Oct 21 14:43:21 2013, Stephane Debieux, DRS4 analog outputs - interfacing DRS4 to AD9222 ADC 

Hi,

I wish to interface the DRS4 with the 8-channel ADC AD9222 (or AD9637).

I'm reading from the DRS4 datasheet that "the analog output of the DRS4 chip has been designed to match directly the input of the AD9222". OUT+ output of DRS4 is in the range from 0.8V to 1.8V and OUT- output is shifted by the voltage applied to the O-OFS pin.

The span of the AD9222 ADC core is defined by REFT and REFB which are resp. 1.4V and 0.4V in a typical case (AVDD=1.8V, VREF=1V). My understanding is that the ADC analog inputs must be within the voltage range defined by REFT and REFB and so I don't quite see how this matches the DRS4 outputs.

Can we use the full-scale range indeed? Do we have to use AC-coupling with mid-supply bias? What is the point I missed?

Thank you for your help.

 

Entry  Wed May 11 04:01:14 2016, Maksat, DRS4 Macro to save events 

Dear Stefan,

I am trying to setup DRS inside radiation enclosure and would like to write a simple script that will automatically save certain number of events.

Could you please point to me an example that can I use for Mac OS? I saw there is drs_exam.cpp in the directory but was not able to get work in Mac OS. Any help would be greatly appreciated.

Thanks

 

 

    Reply  Thu May 12 12:38:17 2016, Stefan Ritt, DRS4 Macro to save events 

Dear Maksat,

If your car does not run, and you call the car dealer and tell him "my car does not run", what will the car dealer ask you? Eh... ? Right ! He will ask "what are the symptoms, what did you try, what did and what did not work". Here it's the same. "was not able to get it work" is not a valid statement, since I have absolutely no idea what did not work and what you did try.

The official way is to follow the instruction in the evlauation board manual on section 2.4 - Installation under Linux. If that does not work, please be a bit more precise what errors you get.

Cheers,
Stefan

Maksat wrote:

Dear Stefan,

I am trying to setup DRS inside radiation enclosure and would like to write a simple script that will automatically save certain number of events.

Could you please point to me an example that can I use for Mac OS? I saw there is drs_exam.cpp in the directory but was not able to get work in Mac OS. Any help would be greatly appreciated.

Thanks

 

 

 

Entry  Wed Nov 30 17:48:39 2016, samridha kunwar, DRS4 Initiation 

I am having a general problem getting read back using the ROI mode.  In the transparent mode everything looks good. These are the steps that I take:

1) configure register (b"11111111",addr = "1100")

2) configure write shift register (b"11111111", addr = "1101")

3)  assert DENABLE and DWRITE

4) wait for trigger

5) on trigger deassert DWRITE

6) Strobe RSRLOAD

7)Set drs4 address to enable all channels (address = "1001")

8)give n SRCLK pulses

9) goto 3 and repeat.

 

Am I missing something? Everything looks straight forward based on the manual yet in the readout mode I only get noise. I do get the stop position on SROUT and the refclk is at 475 KHz as desired and I get the desired behaviour  for DTAP toggling at the same frequency as refclk.

    Reply  Wed Nov 30 19:05:24 2016, Stefan Ritt, DRS4 Initiation 

Uhh, there are 1000 things which might be wrong. A bit like "my car is not working, it makes strange noise". Without having a look under the hood, there is just some wild guessing:

- Is your ROFS input at the right value? Your O-OFS?

- All VDD voltages there? Input voltage outside the rails?

- Your RSLOAD pulse long enough (>10ns)

- What happens if you put a really big sinal at the input, like 100 MHz sine wave with 2V p-p

The easiest is to have a look at the evaluation board and copy your new board like 1:1, also copy the VHDL readout code. Much easier that to start from scratch.

Stefan

 

 

samridha kunwar wrote:

I am having a general problem getting read back using the ROI mode.  In the transparent mode everything looks good. These are the steps that I take:

1) configure register (b"11111111",addr = "1100")

2) configure write shift register (b"11111111", addr = "1101")

3)  assert DENABLE and DWRITE

4) wait for trigger

5) on trigger deassert DWRITE

6) Strobe RSRLOAD

7)Set drs4 address to enable all channels (address = "1001")

8)give n SRCLK pulses

9) goto 3 and repeat.

 

Am I missing something? Everything looks straight forward based on the manual yet in the readout mode I only get noise. I do get the stop position on SROUT and the refclk is at 475 KHz as desired and I get the desired behaviour  for DTAP toggling at the same frequency as refclk.

 

    Reply  Fri Dec 2 15:32:52 2016, samridha kunwar, DRS4 Initiation 

Thanks for replying Stefan.

I was more so just concerned with the steps in the firmware when I had asked. However, yes the ROFS (1.05V) and O-OFS (0.9 V was 1.3 V earlier but, changed this becasue of ADC input requirements) are per spec, the VDD voltages are all there and input voltages are within the rails and finally the RSLOAD  (16 ns) too is ok. Looking at your eval board firmware , on appearance it looks exactly like what I am doing. I thought maybe I was/ still am missing some intermediate addressing stage. What I wrote earlier is what I still have.

Stefan Ritt wrote:

Uhh, there are 1000 things which might be wrong. A bit like "my car is not working, it makes strange noise". Without having a look under the hood, there is just some wild guessing:

- Is your ROFS input at the right value? Your O-OFS?

- All VDD voltages there? Input voltage outside the rails?

- Your RSLOAD pulse long enough (>10ns)

- What happens if you put a really big sinal at the input, like 100 MHz sine wave with 2V p-p

The easiest is to have a look at the evaluation board and copy your new board like 1:1, also copy the VHDL readout code. Much easier that to start from scratch.

Stefan

 

 

samridha kunwar wrote:

I am having a general problem getting read back using the ROI mode.  In the transparent mode everything looks good. These are the steps that I take:

1) configure register (b"11111111",addr = "1100")

2) configure write shift register (b"11111111", addr = "1101")

3)  assert DENABLE and DWRITE

4) wait for trigger

5) on trigger deassert DWRITE

6) Strobe RSRLOAD

7)Set drs4 address to enable all channels (address = "1001")

8)give n SRCLK pulses

9) goto 3 and repeat.

 

Am I missing something? Everything looks straight forward based on the manual yet in the readout mode I only get noise. I do get the stop position on SROUT and the refclk is at 475 KHz as desired and I get the desired behaviour  for DTAP toggling at the same frequency as refclk.

 

 

ELOG V3.1.5-3fb85fa6