| ID  | Date | Author | Subject | Text |  | 
| 705 | Tue Jun 19 12:54:51 2018 | Phan Van Chuan | The data acquisition speed | Thank Stefan Ritt, I added the SoftTrigger() just after StartDomino(), so now, The data
 acquisition speed the same speed as in the
 |  | 
| 704 | Tue Jun 19 10:05:50 2018 | Stefan Ritt | The data acquisition speed | How do you tigger the board? In your code below you start the board (StartDomino())
 and then wait for a trigger. Setting the
 |  | 
| 703 | Tue Jun 19 06:42:23 2018 | Phan Van Chuan | The data acquisition speed | Dear Stefan, 
 We are using an DRS4 board V5.1
 for building a metering system for the scintillator
 |   | 
| 702 | Wed Jun 13 16:34:28 2018 | Julian Kemp | Maximum analog input voltage | Thank you! That solves my problem. 
 
 
 
 
 |  | 
| 701 | Wed Jun 13 13:42:47 2018 | Stefan Ritt | Maximum analog input voltage | In principle the numbers in the manual are correct. But they relate to pulses of
 a certain length, because the input protection
 |  | 
| 700 | Wed Jun 13 13:23:17 2018 | Julian Kemp | Maximum analog input voltage | Dear all, 
 I have been wondering what the
 maximum analog input voltage for the DRS4
 |  | 
| 699 | Fri Jun  8 08:11:05 2018 | Stefan Ritt |  | Several people reported this problem, but we cannot reproduce it at our lab. Both the
 oscilloscope and the command line interface
 |  | 
| 698 | Thu Jun  7 16:27:21 2018 | Phan Van Chuan |  | Dear Stefan, 
 I am using an DRS4 board to test
 the signal from an scintillator detector;
 |    | 
| 697 | Thu May 17 13:29:34 2018 | Stefan Ritt | "Symmetric spikes" fixed | Good news for all DRS4 users. After many years, I finally understand where the "symmetric
 spikes" come from and how to fix them.
 |     | 
| 696 | Mon May 14 09:21:29 2018 | Alessio Berti | WIndows Connection problem with drs507 SOLVED | Hi, 
 I have a machine with Windows 10
 and the solution provided by Steven works
 |  | 
| 695 | Wed May  9 14:07:10 2018 | Alec Shackleford | WIndows Connection problem with drs507 SOLVED | Thank you for this fantastic solution. I had almost reinstalled windows 7 to see
 if that would solve the issue!
 |  | 
| 694 | Wed May  9 09:03:52 2018 | Stefan Ritt | Manual Rev5.1 Figure 1, optional components | I updated the picture in the manual with a current picture of a Rev5.1 board, and
 also added a picture of the bottom side.
 |  | 
| 693 | Tue May  8 23:58:35 2018 | Sean Quinn | Manual Rev5.1 Figure 1, optional components | Dear All, 
 
 
 I'm troubleshooting a board
 |  | 
| 692 | Tue May  8 14:43:03 2018 | Stefan Ritt | Peak at 0 mV in traces | The DRS chip is read out with a 12 bit ADC, thus the phyical resolution is roughly
 1V/4096 = 0.24 mV. I say roughly since the
 |  | 
| 691 | Tue May  8 12:15:54 2018 | Alessio Berti | Peak at 0 mV in traces | Hi Stefan, 
 following your example, we tried
 to perform the same measurement, using drs_exam
 |      | 
| 690 | Sun May  6 11:45:09 2018 | Stefan Ritt | confusion about the description in drs.cpp | The locbus_addr is indeed 32 bits wide, since the firmware was originally derived
 from some firmware running in a VME crate,
 |  | 
| 689 | Sun May  6 08:13:37 2018 | chen wenjun | confusion about the description in drs.cpp | Hi Stefan: 
 I'm still confused that
 althought the 8 bits buffer is enough,the
 |  | 
| 688 | Fri May  4 12:11:57 2018 | Stefan Ritt | Running drs_example.cpp | And here is the second part of your answer: When you change the input range, you have
 to redo the voltage calibration. Best is
 |  | 
| 687 | Fri May  4 11:56:08 2018 | Stefan Ritt | Voltage and Timing Calibration in drs_exam.cpp | Have you set the sampling frequency 
 b->SetFrequency(5, true);
 
 before the calibration?
 |  | 
| 686 | Fri May  4 11:35:20 2018 | Stefan Ritt | Peak at 0 mV in traces | I tried the following: 
 - trigger on a 10 MHz sine wave
 on CH0, CH1 was open
 |   |