    Reply  Mon Jul 15 19:34:25 2019, Brendan Posehn, Evaluation Board Test Functionality 
Hello Stefan, 

Thanks for the quick reply. The issue was a faulty SMA connector, should have checked this first. Signal looks good now.

Thanks for your time, 
Entry  Tue Jul 23 22:31:08 2013, alonzi, Evaluation Board Behavior Screenshot.pngdata_problem.png
Working with the DRS evaluation board we noticed some funny behavior: See attatchment 1. In about 1% of scope traces we see the first and last bin take
on a value substantially different from the baseline, note the small spikes on the end of the traces. These spikes occur across all channels and either
appear in all channels or in none. Attachment two shows what several thousand scope traces look like. You can clearly see that some of the traces are offset
    Reply  Tue Jul 23 22:35:08 2013, Stefan Ritt, Evaluation Board Behavior 


    
        
            alonzi wrote:
        
        
            

           Working with the DRS evaluation board we noticed some funny behavior: See attatchment 1. In about 1% of scope traces we see the first and
    Reply  Tue Jul 23 22:42:31 2013, alonzi, Evaluation Board Behavior 


    
        
            Stefan Ritt wrote:
        
        
            
            
            
    Reply  Thu Jul 25 01:31:29 2013, Andrey Kuznetsov, Evaluation Board Behavior 


    
        
            alonzi wrote:
        
        
            

           
            
                
           
Entry  Tue May 21 18:13:08 2024, Rebecca Hicks, Error when running drsosc 
Hi, I'm a student trying to figure out the DRS4 board. I cloned the github repo, but when I run drsosc, I get an error: Gtk-Message: 10:06:38.376:
Failed to load module "canberra-gtk-module". I'm not sure what that means. The oscilloscope window does open up for me though. Thanks for
any help!
    Reply  Fri Jun 28 23:33:51 2024, Patricia Lecomti, Error when running drsosc 
Salut !

Je vois que tu rencontres un petit problème avec ton installation. Le message "Gtk-Message: Failed to load module 'canberra-gtk-module'"
indique que ton système essaie de charger un module GTK spécifique qui n'est pas installé. Heureusement, ce n'est pas un problème
Entry  Mon Jul 16 19:39:35 2018, Woon-Seng Choong, Effect of interpolation on timing 
Using a test pulse split into two channels of the DRS4 Evaluation Board v5, I looked at the time resolution using a leading edge threshold.
The voltage and timing calibration was performed. One method (1) is to linearly interpolate between two points of the raw waveform that
is above and below the threshold (this is exactly the algorithm given in read_binary.c in the drs4 source distribution); and another (2) is to
    Reply  Fri Jul 20 00:44:13 2018, Woon-Seng Choong, Effect of interpolation on timing 
Just a follow-up update.

It turns out that I was using a cubic spline interpolation with smoothing. If I required the cubic spline to go through the sampled points, then
I obtained similar time resolution as the simple linear interpolation.
Entry  Thu Dec 6 09:23:36 2012, Martin Petriska, EVM rev4 board trigger change and drs_example 
 I switched from rev 3 to rev 4 board, but have some problems with triggering, board is now waiting for trigger (rev.3 is working). How to do in
drs_exam.cpp for example triggering on Ch0 && CH1 ?
Software 4.0.0, windows version.
    Reply  Fri Dec 14 21:49:29 2012, Stefan Ritt, EVM rev4 board trigger change and drs_example 


    
        
            Martin Petriska wrote:
        
        
            
             I switched from rev 3 to rev 4 board, but have some problems with triggering, board is now waiting
Entry  Sat Aug 29 22:00:30 2020, Hans Steiger, Dynamic Range Evaluation Board and Software 
Dear Evaluation Board Team,

 

currently I am facing the problem of digitizing pulses with an amplitude of -0.6V to -0.8V. As the dynamic range of the board is 1Vpp, this should
    Reply  Mon Aug 31 10:52:42 2020, Stefan Ritt, Dynamic Range Evaluation Board and Software 
You cannot go below -0.5V for the inputs, since the board does not have an internal negative power supply, which would be necessary for that. If you
have -0.8V pulses, the easiest is to use a passive inverter at the input to convert it to a 0.8V pulse.

Stefan
Entry  Wed Dec 23 15:38:14 2015, mony orbach, Dtap stops toggling after 40msec 
Hi

the drs4 start to generate Dtap signal after reset and standard configuration.

while in reset Denable and  Dwrite are low
    Reply  Wed Dec 23 15:48:42 2015, Stefan Ritt, Dtap stops toggling after 40msec 
No idea what you do wrong. I need to see oscilloscope traces for all your inputs and voltages. What is your REFCLK input?




mony
orbach wrote:



Hi
    Reply  Thu Dec 24 10:51:31 2015, mony orbach, Dtap stops toggling after 40msec 
my refclk is 1.25Mhz

what are the inputs and voltage you need to see?

Avdd and Dvdd are 2.5v
    Reply  Thu Dec 24 12:45:41 2015, Stefan Ritt, Dtap stops toggling after 40msec 
I want to see the trace on the scope for the DTAP, the REFCLK, the DENABLE and the DWRITE.

Probably (but it's just a guess), you have a problem with the soldering of the DRS chip, maybe to the PLL loop filter. Or you chose the wrong
capacitor/resistor combination for the loop filter. There are ~10 other groupsl who did the same and it works for all of them, so there must be a problem
    Reply  Sun Dec 27 15:06:59 2015, mony orbach, Dtap stops toggling after 40msec 
Hi

We have some meesurs to show (attached)


Dtap and Denable
Dtap+Denable in zoom
Dtap + Ref+
Dtap + Dspeed


From the screen shots
    Reply  Sun Dec 27 15:41:32 2015, mony orbach, Dtap stops toggling after 40msec Dtap-Denable.gifdtap-Danable2.gifDtap-refck.gifDtap-Dspeed.gif
Hi

We have some measures to show (attached)


Dtap and Denable
Dtap+Denable in zoom
Dtap + Refck+
Dtap + Dspeed


From the screen
    Reply  Mon Dec 28 11:05:15 2015, Stefan Ritt, Dtap stops toggling after 40msec 
Thanks for posting the plots. It really looks like the PLL is not working. I see two possible reasons: 1) The PLLEN bit in the configuration register
is not set and 2) The REFCLK signal does not reach the chip. We had cases whrere people had a hard time to solder the DRS4 correctly due to the small pins.
So if the REFCLK+ and REFCLK- signals have a poor connection, then the PLL of course won't work. Putting some more tin at the pins manually usually