Thu Dec 13 19:49:47 2012 |
Evgeni | DRS-4 trigger |
Stefan Ritt wrote: |
Evgeni wrote: |
How to configure DRS oscilloscope for the oscillations with an amplitude greater than the value of the exposed
in the trigger (internal).
|
Sorry, I don't understand that question. The DRS4 Evaluation board input signal range is 1V. If you have larger signals, you have to attenuante them externally.
/Stefan
|
Can I adjust the internal trigger DRS oscilloscope for signal extraction 0.5 volts (for example) from any of the four channels. In our case, there is a lot of noise with low amplitude, which must be removed during the registration. We need to record the individual pulses of higher amplitude than the noise. So we want to use the internal trigger DRS oscilloscope to cut this noise by setting up its threshold amplitude noise.
|
206
|
Thu Dec 13 12:14:35 2012 |
Stefan Ritt | DRS-4 trigger |
Evgeni wrote: |
How to configure DRS oscilloscope for the oscillations with an amplitude greater than the value of the exposed
in the trigger (internal).
|
Sorry, I don't understand that question. The DRS4 Evaluation board input signal range is 1V. If you have larger signals, you have to attenuante them externally.
/Stefan |
205
|
Thu Dec 13 12:03:29 2012 |
Evgeni | DRS-4 trigger | How to configure DRS oscilloscope for the oscillations with an amplitude greater than the value of the exposed
in the trigger (internal). |
204
|
Thu Dec 6 09:23:36 2012 |
Martin Petriska | EVM rev4 board trigger change and drs_example | I switched from rev 3 to rev 4 board, but have some problems with triggering, board is now waiting for trigger (rev.3 is working). How to do in drs_exam.cpp for example triggering on Ch0 && CH1 ?
Software 4.0.0, windows version.
Here is old trigger initialisation:
b->EnableTrigger(0,1);
b->SetTriggerSource(0);
b->SetTriggerLevel(0.25, false);
b->SetTriggerDelayNs(0);
Btw. Is it possible to set up different trigger Levels for each channel ?
(If there is some interest here is my code in Qt, still aplha) http://sourceforge.net/p/qtpals/code |
203
|
Tue Dec 4 09:55:43 2012 |
Stefan Ritt | Question of drs4 using |
Zhongwei Du wrote: |
Stefan Ritt wrote: |
Zhongwei Du wrote: |
When Denable and Dwrite is high , the voltage of PLLOUT is 0 V. And the Dtap is turn high with no delay when the Denable turns high.
After power up and configuration(the WSR,WCR,CR are all set to 11111111), the readout data is no change whenever the input analog signal and rofs,bias,oofs changes. I have test useing the DAC to supply the Dspeed voltage, and change a new DRS4 chip, but all is the same. The readout data is strange : the first about 100 cells is rise or fall and the last 900 cells is out of the range of ADC.
So how should I do for debugging the drs4 now.
|
The first thing to make work is to have DTAP oscillating with fsamp/2048. Keep Denable and Dwrite low (required during power-on, see elog:10), set Dspeed to 2.5V, then rise Denable and Dwrite. You should see Dtap toggling at about 2.4 MHz. If not, double check all supply voltages, and especially all soldering points. The QFN package is a bit hard to solder.
/Stefan
|
"Keep Denable and Dwrite low (required during power-on, see elog:10), set Dspeed to 2.5V, then rise Denable and Dwrite. You should see Dtap toggling at about 2.4 MHz. "
In this process , should i config any registers( WSR,WCR,CR ) ?
|
After power-up reset, these registers are all set to "1", which should be ok to start.
BTW, Jinhong Wang <wangjinh@mail.ustc.edu.cn> from your institute hast the chip correctly working. Maybe he can help you in a more direct way than I can.
|
202
|
Tue Dec 4 09:50:11 2012 |
Zhongwei Du | Question of drs4 using |
Stefan Ritt wrote: |
Zhongwei Du wrote: |
When Denable and Dwrite is high , the voltage of PLLOUT is 0 V. And the Dtap is turn high with no delay when the Denable turns high.
After power up and configuration(the WSR,WCR,CR are all set to 11111111), the readout data is no change whenever the input analog signal and rofs,bias,oofs changes. I have test useing the DAC to supply the Dspeed voltage, and change a new DRS4 chip, but all is the same. The readout data is strange : the first about 100 cells is rise or fall and the last 900 cells is out of the range of ADC.
So how should I do for debugging the drs4 now.
|
The first thing to make work is to have DTAP oscillating with fsamp/2048. Keep Denable and Dwrite low (required during power-on, see elog:10), set Dspeed to 2.5V, then rise Denable and Dwrite. You should see Dtap toggling at about 2.4 MHz. If not, double check all supply voltages, and especially all soldering points. The QFN package is a bit hard to solder.
/Stefan
|
"Keep Denable and Dwrite low (required during power-on, see elog:10), set Dspeed to 2.5V, then rise Denable and Dwrite. You should see Dtap toggling at about 2.4 MHz. "
In this process , should i config any registers( WSR,WCR,CR ) ? |
201
|
Tue Dec 4 09:39:44 2012 |
Stefan Ritt | Question of drs4 using |
Zhongwei Du wrote: |
When Denable and Dwrite is high , the voltage of PLLOUT is 0 V. And the Dtap is turn high with no delay when the Denable turns high.
After power up and configuration(the WSR,WCR,CR are all set to 11111111), the readout data is no change whenever the input analog signal and rofs,bias,oofs changes. I have test useing the DAC to supply the Dspeed voltage, and change a new DRS4 chip, but all is the same. The readout data is strange : the first about 100 cells is rise or fall and the last 900 cells is out of the range of ADC.
So how should I do for debugging the drs4 now.
|
The first thing to make work is to have DTAP oscillating with fsamp/2048. Keep Denable and Dwrite low (required during power-on, see elog:10), set Dspeed to 2.5V, then rise Denable and Dwrite. You should see Dtap toggling at about 2.4 MHz. If not, double check all supply voltages, and especially all soldering points. The QFN package is a bit hard to solder.
/Stefan |
200
|
Tue Dec 4 09:24:22 2012 |
Zhongwei Du | Question of drs4 using | When Denable and Dwrite is high , the voltage of PLLOUT is 0 V. And the Dtap is turn high with no delay when the Denable turns high.
After power up and configuration(the WSR,WCR,CR are all set to 11111111), the readout data is no change whenever the input analog signal and rofs,bias,oofs changes. I have test useing the DAC to supply the Dspeed voltage, and change a new DRS4 chip, but all is the same. The readout data is strange : the first about 100 cells is rise or fall and the last 900 cells is out of the range of ADC.
So how should I do for debugging the drs4 now. |
199
|
Mon Dec 3 11:40:35 2012 |
Gyuhee Kim | Another question about using multi boards. |
Stefan Ritt wrote: |
Gyuhee Kim wrote: |
Hi.
I asked about using multi boards some days ago, and I got answer to use external trigger. (Thanks Stefan!)
And here is another question. I made two external triggers and try to acquire coincidence data using two boards. but DRS Oscilloscope program can connect only one board and don`t acquire both of them simultaneously.
So I tried to use two computer for each board separately, but, well, you already know, I failed to acquire because two computers don`t promise to synchronize the two boards acquisition.
Is there any method to solve this problem?
1. I want to acquire coincidence data from the two DRS 4 Evaluation board V4 simultaneosly.
2. I have external trigger to provide the two boards at the same time.
3. How can I get data from the two boards?
Best regards.
Gyuhee.
|
You have to write your own program. DRS Oscilloscope does not (yet) support this. Take drs_exam.cpp as a starting point and try to extend it to two boards. One tricky point is that the external trigger may only fire AFTER the two boards have been read out. So you need some means of re-enabling the external trigger after you read out both boards.
Stefan
|
That`s very bad news for me. I don`t have much time to study & write C programming...
Anyway, Thank you very much Stefan.
Best regards.
Gyuhee. |
198
|
Mon Dec 3 09:18:09 2012 |
Stefan Ritt | Another question about using multi boards. |
Gyuhee Kim wrote: |
Hi.
I asked about using multi boards some days ago, and I got answer to use external trigger. (Thanks Stefan!)
And here is another question. I made two external triggers and try to acquire coincidence data using two boards. but DRS Oscilloscope program can connect only one board and don`t acquire both of them simultaneously.
So I tried to use two computer for each board separately, but, well, you already know, I failed to acquire because two computers don`t promise to synchronize the two boards acquisition.
Is there any method to solve this problem?
1. I want to acquire coincidence data from the two DRS 4 Evaluation board V4 simultaneosly.
2. I have external trigger to provide the two boards at the same time.
3. How can I get data from the two boards?
Best regards.
Gyuhee.
|
You have to write your own program. DRS Oscilloscope does not (yet) support this. Take drs_exam.cpp as a starting point and try to extend it to two boards. One tricky point is that the external trigger may only fire AFTER the two boards have been read out. So you need some means of re-enabling the external trigger after you read out both boards.
Stefan |
197
|
Mon Dec 3 08:32:28 2012 |
Gyuhee Kim | Another question about using multi boards. | Hi.
I asked about using multi boards some days ago, and I got answer to use external trigger. (Thanks Stefan!)
And here is another question. I made two external triggers and try to acquire coincidence data using two boards. but DRS Oscilloscope program can connect only one board and don`t acquire both of them simultaneously.
So I tried to use two computer for each board separately, but, well, you already know, I failed to acquire because two computers don`t promise to synchronize the two boards acquisition.
Is there any method to solve this problem?
1. I want to acquire coincidence data from the two DRS 4 Evaluation board V4 simultaneosly.
2. I have external trigger to provide the two boards at the same time.
3. How can I get data from the two boards?
Best regards.
Gyuhee. |
196
|
Wed Nov 28 16:54:46 2012 |
Stefan Ritt | DRS Oscilloscope for Raspberry Pi and Mac OSX 10.8 | I made a pre-compiled package for Mac OSX 10.8 (Mountain Lion), so one should be able to install the DRS Oscilloscope software with one mouse click on a recent Mac.
The Makefile in the tar ball now also supports OSX 10.8, so one could even compile it from the sources on a Mac, after libusb-1.0 and wxWidgets have been installed. That might then also work on older versions of OSX.
In addition, the software has been adaptes so that it compiles nicely on a Raspberry Pi (http://www.raspberrypi.org), a screenshot has been attached. Together with a Raspberry Pi and an old screen, the evaluation board is one of the cheapest available oscilloscopes.
|
195
|
Wed Nov 21 08:48:00 2012 |
Gyuhee Kim | Question for using Multi board |
Stefan Ritt wrote: |
Gyuhee Kim wrote: |
Hi.
I have 2 DRS4 evaluation V4 boards, and I want to use these 2 board to multi board DAQ system for 4 ch vs 4 ch DAQ.
But there is no option for multi board use. I just only find the multi board trigger mode check button on DRS4 Oscilloscope program, but I couldn`t check.
Is there any method to use multi board?
Best regards.
Gyuhee.
|
This mode is not yet implemented in firmware. Maybe I find some time towards the end of this year to add this. At the moment, you have to build and external trigger to synchronize the two boards. There are also 16-channel boards on the market where you would not need a multi-board mode. Just Google for "DT5742".
/Stefan
|
Thanks Stefan.
I will build external trigger system. |
194
|
Wed Nov 21 08:38:26 2012 |
Stefan Ritt | Question for using Multi board |
Gyuhee Kim wrote: |
Hi.
I have 2 DRS4 evaluation V4 boards, and I want to use these 2 board to multi board DAQ system for 4 ch vs 4 ch DAQ.
But there is no option for multi board use. I just only find the multi board trigger mode check button on DRS4 Oscilloscope program, but I couldn`t check.
Is there any method to use multi board?
Best regards.
Gyuhee.
|
This mode is not yet implemented in firmware. Maybe I find some time towards the end of this year to add this. At the moment, you have to build and external trigger to synchronize the two boards. There are also 16-channel boards on the market where you would not need a multi-board mode. Just Google for "DT5742".
/Stefan |
193
|
Wed Nov 21 08:34:52 2012 |
Gyuhee Kim | Question for using Multi board | Hi.
I have 2 DRS4 evaluation V4 boards, and I want to use these 2 board to multi board DAQ system for 4 ch vs 4 ch DAQ.
But there is no option for multi board use. I just only find the multi board trigger mode check button on DRS4 Oscilloscope program, but I couldn`t check.
Is there any method to use multi board?
Best regards.
Gyuhee. |
192
|
Tue Nov 13 11:26:32 2012 |
Stefan Ritt | GetWave |
Martin Petriska wrote: |
I have some question according to GetWave function. In drs_exam.cpp simple GetWave(0,0,wave_array[]) etc...is used. Is there primary (cell) calibration, secondary calibration (Readout) and remove Spikes used, as in DRS Oscilloscope application?
|
Yes, yes, no. To get spike removals, you need the function RemoveSpikes from Osci.cpp in the DRSOsc project. |
191
|
Thu Nov 1 20:46:53 2012 |
hongwei yang | DRS4 firmware |
Stefan Ritt wrote: |
hongwei yang wrote: |
hongwei yang wrote: |
Stefan Ritt wrote: |
hongwei yang wrote: |
Hi,
We are using drs4 board, but oscilloscope app will somehow stop to work if we config trigger into "or and", When I look into the drs4 firmware file drs4_eval3_app.vhd, I couldn't find the trigger_config value assignment which is mentioned at(#7 offset 0x1E from 31 downto 16) in manual_version 4.
could you help me find this trigger_config access point?
thanks
Hongwei
|
The "and" in the trigger section means now "coincidence". So the V4 board can trigger on a coincidence between two or more channels. If there is no pulse at the same time on the coincidence channels, the board will of course not trigger. The according firmware was introduced in V4, so please look at drs4_eval4_app.vhd (not eval3).
I just realized that the V4 firmware might be missing in the distribution, so I have attached it here. Look for drs_ctl_trigger_config.
Best regards,
Stefan
|
Ah, great, that helps, Thank you!
Hongwei
|
By the way, will there be a drs4_eval4.vhd as well?
|
Here it is.
|
Thanks. have a good day |
190
|
Thu Nov 1 20:32:03 2012 |
Stefan Ritt | DRS4 firmware |
hongwei yang wrote: |
hongwei yang wrote: |
Stefan Ritt wrote: |
hongwei yang wrote: |
Hi,
We are using drs4 board, but oscilloscope app will somehow stop to work if we config trigger into "or and", When I look into the drs4 firmware file drs4_eval3_app.vhd, I couldn't find the trigger_config value assignment which is mentioned at(#7 offset 0x1E from 31 downto 16) in manual_version 4.
could you help me find this trigger_config access point?
thanks
Hongwei
|
The "and" in the trigger section means now "coincidence". So the V4 board can trigger on a coincidence between two or more channels. If there is no pulse at the same time on the coincidence channels, the board will of course not trigger. The according firmware was introduced in V4, so please look at drs4_eval4_app.vhd (not eval3).
I just realized that the V4 firmware might be missing in the distribution, so I have attached it here. Look for drs_ctl_trigger_config.
Best regards,
Stefan
|
Ah, great, that helps, Thank you!
Hongwei
|
By the way, will there be a drs4_eval4.vhd as well?
|
Here it is. |
189
|
Thu Nov 1 20:25:53 2012 |
hongwei yang | DRS4 firmware |
hongwei yang wrote: |
Stefan Ritt wrote: |
hongwei yang wrote: |
Hi,
We are using drs4 board, but oscilloscope app will somehow stop to work if we config trigger into "or and", When I look into the drs4 firmware file drs4_eval3_app.vhd, I couldn't find the trigger_config value assignment which is mentioned at(#7 offset 0x1E from 31 downto 16) in manual_version 4.
could you help me find this trigger_config access point?
thanks
Hongwei
|
The "and" in the trigger section means now "coincidence". So the V4 board can trigger on a coincidence between two or more channels. If there is no pulse at the same time on the coincidence channels, the board will of course not trigger. The according firmware was introduced in V4, so please look at drs4_eval4_app.vhd (not eval3).
I just realized that the V4 firmware might be missing in the distribution, so I have attached it here. Look for drs_ctl_trigger_config.
Best regards,
Stefan
|
Ah, great, that helps, Thank you!
Hongwei
|
By the way, will there be a drs4_eval4.vhd as well? |
188
|
Thu Nov 1 20:21:44 2012 |
hongwei yang | DRS4 firmware |
Stefan Ritt wrote: |
hongwei yang wrote: |
Hi,
We are using drs4 board, but oscilloscope app will somehow stop to work if we config trigger into "or and", When I look into the drs4 firmware file drs4_eval3_app.vhd, I couldn't find the trigger_config value assignment which is mentioned at(#7 offset 0x1E from 31 downto 16) in manual_version 4.
could you help me find this trigger_config access point?
thanks
Hongwei
|
The "and" in the trigger section means now "coincidence". So the V4 board can trigger on a coincidence between two or more channels. If there is no pulse at the same time on the coincidence channels, the board will of course not trigger. The according firmware was introduced in V4, so please look at drs4_eval4_app.vhd (not eval3).
I just realized that the V4 firmware might be missing in the distribution, so I have attached it here. Look for drs_ctl_trigger_config.
Best regards,
Stefan
|
Ah, great, that helps, Thank you!
Hongwei |
|