  540   Thu Oct 6 11:18:05 2016 Stefan RittTimestamp for each DRS4 waveform

In the mentioned read_binary.cpp file you have the line where you read the event header

i = fread(&eh, sizeof(eh), 1, f);

The C structure eh now contains the full timestamp, and you can access it with 

eh.year
eh.month
eh.day
eh.hour
eh.minute
eh.second
eh.millisecond

Cheers,
Stefan

Will Flanagan wrote:

Hi DRS4 Experts,

I have been analyzing DRS4 binary data with scripts based on Stefan's (very helpful!) macro:

https://midas.psi.ch/elogs/DRS4+Forum/361

I would now like to look at the stability of my waveforms over a long period of time. In order to do this, I would need a timestamp encoded with each waveform. Are there timestamps within default DRS4 binary data? If so, does anyone have sample code for extracting them?

Best Regards,

Will

 

  539   Wed Oct 5 22:43:29 2016 Will FlanaganTimestamp for each DRS4 waveform

Hi DRS4 Experts,

I have been analyzing DRS4 binary data with scripts based on Stefan's (very helpful!) macro:

https://midas.psi.ch/elogs/DRS4+Forum/361

I would now like to look at the stability of my waveforms over a long period of time. In order to do this, I would need a timestamp encoded with each waveform. Are there timestamps within default DRS4 binary data? If so, does anyone have sample code for extracting them?

Best Regards,

Will

  538   Fri Sep 30 17:03:38 2016 Stefan RittOutput Timing Drifting

Hi Jacob,

you are missing the timing calibration. Each sampling cell has not the same width. Running at 5 GSPS, cell widths scatter from 150 ps to 250 ps. If you integrate these widhts, you get a time scale which can be off by a few ns between chips, something you see in your plot. Here is a paper which explains in detail how to do a timing calibration: https://arxiv.org/abs/1405.4975

Cheers,
Stefan

Jacob Hwang wrote:

Hello,

I have designed four DRS4 chips (36 channels) on my board running at 1GHz (REFCLK=488.28KHz) and ROI mode. All 4 chips' REFCLK, DWRITE, RSRLOAD, and SRCLK are buffer driven by the same source.  SRCLK is set to 40MHz to reduce the readout time.

If I injected a sine waveform, buffered and splitted into all 36 channels,I noticed all 9 channels on each DRS4 chip output almost the same as expected.  But the output phase from chip to chip is drifting as shown in attached picture which is from two different channels of different chips.  From the few boards I have built, I found few chips are drifting more than the others and is different on every board.

The sympton look like the DRS4 internal PLL is drifting, but I checked the DTAP output on every chip and found it's dead-lock steady even I used persistance setting on my oscilloscope.  Do you have any suggestion how to attack this problem?  Thank you.

Jacob Hwang
 

 

  537   Thu Sep 29 17:26:13 2016 Jacob HwangOutput Timing Drifting

Hello,

I have designed four DRS4 chips (36 channels) on my board running at 1GHz (REFCLK=488.28KHz) and ROI mode. All 4 chips' REFCLK, DWRITE, RSRLOAD, and SRCLK are buffer driven by the same source.  SRCLK is set to 40MHz to reduce the readout time.

If I injected a sine waveform, buffered and splitted into all 36 channels,I noticed all 9 channels on each DRS4 chip output almost the same as expected.  But the output phase from chip to chip is drifting as shown in attached picture which is from two different channels of different chips.  From the few boards I have built, I found few chips are drifting more than the others and is different on every board.

The sympton look like the DRS4 internal PLL is drifting, but I checked the DTAP output on every chip and found it's dead-lock steady even I used persistance setting on my oscilloscope.  Do you have any suggestion how to attack this problem?  Thank you.

Jacob Hwang
 

  536   Mon Aug 29 12:51:48 2016 Stefan Rittincrement write config register on the fly?

The problem is when you change the write config register from 11111111 to 01111111, or from 00001111 to 00000111, then the last 256 sampels of the previous channel (in the first case #0, in the scond #4) would be overwritten as soon as dwrite =1 again. So you loose 1/4 ef each channel.

Concerning the readout, indeed you can keep track in the FPGA, but only with a certainty of a few cells. This gives some timing inacccuracy of maybe 10-20 ns, which certainly would be disturbing you.

 

benjamin legeyt wrote:

If I may trouble you for a little more information, the critical point then is that there should not be any zeroes in the write config register while the sampling is active?  In case it was unclear I would only be reading out once sampling was stopped (dwrite = 0).  

As for the readout, I know that I would have to read out all 1024 samples each time, and keep track of where each channel stopped in the FPGA.  I would never know the exact cell where sampling stopped but I hoped that if I discard some number of cells on each side of the expected stopping point that I would be OK.  

Thanks again

Stefan Ritt wrote:

The issue with "stopping at cell 767" would also affect this mode of operation. Furthermore, the DRS4 chip has only 10 bit register which records in which cell the event has occured, and where the readout must be started. If you record 8 separate events, you don't know where to start the readout.

The DRS5 chip will have all this possibilitied, but unfortunately it won't be ready before 2-3 years from now.

Stefan

 

benjamin legeyt wrote:

Hello,

I have a question about using the write config register to enable/disable sampling on the fly.  I am looking to instrument an experiment at EPFL where multiple short events need to be captured during a 20us period followed by an 80us quiet period during which we could read out the chip.  Would it be possible to start an acquisition with all channels seeing the same signal and the write config register set to 111111111 and then shift a zero into the write config reg after each event is detected to freeze the channels in time one-by-one?  In this way we could measure up to 8 different events during the active period and then read them all out together during the quiet period.  I have read the posts about the simultaneous read-write mode and the issue with waveforms stopping at cell 767.  not knowing the exact details of what causes this issue I wonder if it would effect this sort of operation?  Also, I would like to know if dwrite must be de-asserted while the write config register is being updated or if it could be done while the sampling is active?  The latter would obviously be preferable as we would not incur any dead-time during the active period.

Thanks in advance for the information,

Benjamin LeGeyt

 

 

 

  535   Mon Aug 29 12:18:49 2016 benjamin legeytincrement write config register on the fly?

If I may trouble you for a little more information, the critical point then is that there should not be any zeroes in the write config register while the sampling is active?  In case it was unclear I would only be reading out once sampling was stopped (dwrite = 0).  

As for the readout, I know that I would have to read out all 1024 samples each time, and keep track of where each channel stopped in the FPGA.  I would never know the exact cell where sampling stopped but I hoped that if I discard some number of cells on each side of the expected stopping point that I would be OK.  

Thanks again

Stefan Ritt wrote:

The issue with "stopping at cell 767" would also affect this mode of operation. Furthermore, the DRS4 chip has only 10 bit register which records in which cell the event has occured, and where the readout must be started. If you record 8 separate events, you don't know where to start the readout.

The DRS5 chip will have all this possibilitied, but unfortunately it won't be ready before 2-3 years from now.

Stefan

 

benjamin legeyt wrote:

Hello,

I have a question about using the write config register to enable/disable sampling on the fly.  I am looking to instrument an experiment at EPFL where multiple short events need to be captured during a 20us period followed by an 80us quiet period during which we could read out the chip.  Would it be possible to start an acquisition with all channels seeing the same signal and the write config register set to 111111111 and then shift a zero into the write config reg after each event is detected to freeze the channels in time one-by-one?  In this way we could measure up to 8 different events during the active period and then read them all out together during the quiet period.  I have read the posts about the simultaneous read-write mode and the issue with waveforms stopping at cell 767.  not knowing the exact details of what causes this issue I wonder if it would effect this sort of operation?  Also, I would like to know if dwrite must be de-asserted while the write config register is being updated or if it could be done while the sampling is active?  The latter would obviously be preferable as we would not incur any dead-time during the active period.

Thanks in advance for the information,

Benjamin LeGeyt

 

 

  534   Mon Aug 29 10:57:33 2016 Stefan Rittincrement write config register on the fly?

The issue with "stopping at cell 767" would also affect this mode of operation. Furthermore, the DRS4 chip has only 10 bit register which records in which cell the event has occured, and where the readout must be started. If you record 8 separate events, you don't know where to start the readout.

The DRS5 chip will have all this possibilitied, but unfortunately it won't be ready before 2-3 years from now.

Stefan

 

benjamin legeyt wrote:

Hello,

I have a question about using the write config register to enable/disable sampling on the fly.  I am looking to instrument an experiment at EPFL where multiple short events need to be captured during a 20us period followed by an 80us quiet period during which we could read out the chip.  Would it be possible to start an acquisition with all channels seeing the same signal and the write config register set to 111111111 and then shift a zero into the write config reg after each event is detected to freeze the channels in time one-by-one?  In this way we could measure up to 8 different events during the active period and then read them all out together during the quiet period.  I have read the posts about the simultaneous read-write mode and the issue with waveforms stopping at cell 767.  not knowing the exact details of what causes this issue I wonder if it would effect this sort of operation?  Also, I would like to know if dwrite must be de-asserted while the write config register is being updated or if it could be done while the sampling is active?  The latter would obviously be preferable as we would not incur any dead-time during the active period.

Thanks in advance for the information,

Benjamin LeGeyt

 

  533   Mon Aug 29 09:36:34 2016 benjamin legeytincrement write config register on the fly?

Hello,

I have a question about using the write config register to enable/disable sampling on the fly.  I am looking to instrument an experiment at EPFL where multiple short events need to be captured during a 20us period followed by an 80us quiet period during which we could read out the chip.  Would it be possible to start an acquisition with all channels seeing the same signal and the write config register set to 111111111 and then shift a zero into the write config reg after each event is detected to freeze the channels in time one-by-one?  In this way we could measure up to 8 different events during the active period and then read them all out together during the quiet period.  I have read the posts about the simultaneous read-write mode and the issue with waveforms stopping at cell 767.  not knowing the exact details of what causes this issue I wonder if it would effect this sort of operation?  Also, I would like to know if dwrite must be de-asserted while the write config register is being updated or if it could be done while the sampling is active?  The latter would obviously be preferable as we would not incur any dead-time during the active period.

Thanks in advance for the information,

Benjamin LeGeyt

  531   Wed Jun 29 09:10:01 2016 Stefan RittNegative input signals

Hello everybody,

I get often asked if the DRS4 evaluation board can accomodate negative input pulses going to -1V. This is unfortunately not possible, since the board is mainly for evaluation of the DRS4 chip and should not be seen as a complete oscilloscope with flexible input stage. So the maximum it can do is -0.5V to +0.5V or 0V to 1V. For -1V signals, one can use however a passive inverter like this one:

http://www.phillipsscientific.com/pdf/460ds.pdf

And for signals going furhter (-2V, -10V) one can use a passive attenuator like this one:

http://www.pomonaelectronics.com/pdf/d4108_K5513_101.pdf

 

Best regards,

Stefan

 

  530   Wed Jun 15 14:49:00 2016 Stefan Rittproblems of DRS4

1. Simultaneous writing and reading is not possible with the DRS4 chip. The manual says differently on p. 14, but due to a bug in the chip waveforms get clipped at the end if one does that. We hopt to fix this problem in a future version of the chip.

2. You can cascade 2,4 or 8 channels. If you cascade 8 channels and run at 1 GSPS, you digitize a window of 8 us. If you have 16 signals, you then need 16 chips.

/Stefan

Michael wrote:

Hi

I want to use DRS4 to digitize 16 channels of signals. The width of signal is about 20 ns, with frequency of 50Hz. The time differences between these 16 signals are not constant, arranging from 3us to 0. I am confused about this in some aspects.

  1. Can I use SIMULTANEOUS WRITINT AND READING to realize this? I saw the VHDL program, and if I understand it correctly, it did not work at this state.
  2. Or sampling at 1GSPS, using CASCADING OF CHANNELS, I can sample signal at most 4us or 8us, then digitizing all signals of one chip. Have you tested 4 or more channels cascading before?

Besides, any advice will be helpful!

Thank you.

 

  Draft   Sun Jun 12 08:49:54 2016 Michaelproblems of DRS4

Hi

I want to use DRS4 to digitize 16 channels of signals. The width of signal is about 20 ns, with frequency of 50Hz. The time differences between these 16 signals are not constant, arranging from 3us to 0. I am confused about this in some aspects.

  1. Can I use SIMULTANEOUS WRITINT AND READING to realize this? I saw the VHDL program, and if I understand it correctly, it did not work at this state.
  2. Or sampling at 1GSPS, using CASCADING OF CHANNELS, I can sample signal at most 4us or 8us, then digitizing all signals of one chip. Have you tested 4 or more channels cascading before?

Besides, any advice will be helpful!

Thank you.

  528   Sun Jun 12 08:45:52 2016 Michaelproblems of DRS4

Hi

I want to use DRS4 to digitize 16 channels of signals. The width of signal is about 20 ns, with frequency of 50Hz. The time differences between these 16 signals are not constant, arranging from 3us to 0. I am confused about this in some aspects.

  1. Can I use SIMULTANEOUS WRITINT AND READING to realize this? I saw the VHDL program, and if I understand it correctly, it did not work at this state.
  2. Or sampling at 1GSPS, using CASCADING OF CHANNELS, I can sample signal at most 4us or 8us, then digitizing all signals of one chip. Have you tested 4 or more channels cascading before?

Besides, any advice will be helpful!

Thank you.

  527   Wed Jun 1 23:16:01 2016 Stefan Rittproblems when stop cell >= 767 ??

I cannot confirm the story with the "stop capacitor > 767". It can be seen from your plots that the distribution of stop cells are even, no holes or bins with double height.

There is an issue with cell 767, but this is when one tries to do simultaneous reading/writing to the chip. This does not really work as writen in the data sheet. Waveforms sometimgs get cut off at cell 767. But the stop cell is always correct, otherwise one could not calibrate the data. If you use the evaluation board for example, which is perfectly calibrated, and introduce an "artifical" shift like

if stop cell > 767 then
  stop cell = stop cell + 1

then you would see that the voltage calibration would become wrong and very noisy.

Stefan

Dominik Neise wrote:

Hello Stefan,

some colleages told me a story, I was neither able to confirm nor find anything in the datsheet about. According to them:

For some internal reason of the DRS4, if the “stop capacitor” of the DRS4 is >= 767, the true stop channel is one before the stop channel read from the DRS4. In other words, the stop channel which returns the DRS4 shifts after sampling to the capacitor ID 766.

Can you confirm that, or even say a few words about that matter?

I wanted to confirm this by plotting the stop cell distribution for random triggered data, taken with one of the FACT boards. I assumed (possibly misunderstanding the matter), that this would lead to missing values in the area of stop cell 767, but cannot see any significant excess or lack of entries in that area.

 

 

  526   Wed Jun 1 22:29:01 2016 Dominik Neiseproblems when stop cell >= 767 ??

Hello Stefan,

some colleages told me a story, I was neither able to confirm nor find anything in the datsheet about. According to them:

For some internal reason of the DRS4, if the “stop capacitor” of the DRS4 is >= 767, the true stop channel is one before the stop channel read from the DRS4. In other words, the stop channel which returns the DRS4 shifts after sampling to the capacitor ID 766.

Can you confirm that, or even say a few words about that matter?

I wanted to confirm this by plotting the stop cell distribution for random triggered data, taken with one of the FACT boards. I assumed (possibly misunderstanding the matter), that this would lead to missing values in the area of stop cell 767, but cannot see any significant excess or lack of entries in that area.

 

  525   Thu May 12 12:38:17 2016 Stefan RittDRS4 Macro to save events

Dear Maksat,

If your car does not run, and you call the car dealer and tell him "my car does not run", what will the car dealer ask you? Eh... ? Right ! He will ask "what are the symptoms, what did you try, what did and what did not work". Here it's the same. "was not able to get it work" is not a valid statement, since I have absolutely no idea what did not work and what you did try.

The official way is to follow the instruction in the evlauation board manual on section 2.4 - Installation under Linux. If that does not work, please be a bit more precise what errors you get.

Cheers,
Stefan

Maksat wrote:

Dear Stefan,

I am trying to setup DRS inside radiation enclosure and would like to write a simple script that will automatically save certain number of events.

Could you please point to me an example that can I use for Mac OS? I saw there is drs_exam.cpp in the directory but was not able to get work in Mac OS. Any help would be greatly appreciated.

Thanks

 

 

 

  524   Thu May 12 08:16:41 2016 Stefan RittProblem For Software Download

Can you tell me (screendump) what is the problem on the web site https://www.psi.ch/drs/software-download ? It should redirect you to

https://www.dropbox.com/sh/qul1cgtm4x7zx13/AADKQ-qGQGdAHPu6OR3vTNY0a?dl=0

for the Windows download.

I cannot send executables via email, that won't go though any spam filter.

Stefan

Yu wrote:

Hi

 I can't download the software for windows on this website 'www.psi.ch/drs/software-download', there is some mistake when i click on download. 

 If convenient, can you send the software Version 5.0.5 for windows to me? My E-mail address is 'yuhaiyang421@163.com'. Thank you!

 

Best Regards

Yu

 

  523   Thu May 12 05:18:47 2016 YuProblem For Software Download

Hi

 I can't download the software for windows on this website 'www.psi.ch/drs/software-download', there is some mistake when i click on download. 

 If convenient, can you send the software Version 5.0.5 for windows to me? My E-mail address is 'yuhaiyang421@163.com'. Thank you!

 

Best Regards

Yu

  522   Wed May 11 15:48:57 2016 SANDJONG Saturnin OrlyProbléme de Calibration de la DRS4

Bonjour, Je suis en stage dans un laboratoire ou on utilise pour echantillonnage des données, une cartes DRS4 5GSPS avec 1024 cell, mon probléme réside dans la partie Calibration en tension selon l'article "Novel Calibration Method for Switched Capacitor Arrays Enables Time Measurements with Sub-Picosecond Resolution". 

En fait je ne comprends pas précisément ces 3 parties de la calibration en tension. Quelqu'un pourras t-il s'il vous plait m'expliquer assez clairement avec des exemples comment il faut s'y prendre? 

Merci et bien Cordialement. 

  521   Wed May 11 04:01:14 2016 MaksatDRS4 Macro to save events

Dear Stefan,

I am trying to setup DRS inside radiation enclosure and would like to write a simple script that will automatically save certain number of events.

Could you please point to me an example that can I use for Mac OS? I saw there is drs_exam.cpp in the directory but was not able to get work in Mac OS. Any help would be greatly appreciated.

Thanks

 

 

  520   Mon May 2 14:31:28 2016 Dmitry Hitstwo DRS4 boards configuration with 2048 samples each

Hi Stefan

Any chance you have time to fix the software for multiboard configuration with 2048 samples each. I tried 5.0.5, but drsosc still shows only half of the waveform.

Dmitry

Stefan Ritt wrote:

The multi-board mode has never been tested with 2048 samples, so is very likely not to work. I don't know yet how much work this will be to fix, but I'm on a business trip the next three weeks and probably will only have time to look at it when I return.

Stefan

Dmitry Hits wrote:

Dear Stefan,

I daisy-chained two boards (master sn#: 2514 - slave sn#: 2513) each with 2048 samples. However, when I use drsosc and put check mark in "configure multi-board daisy-chain" I see only 1024 samples. Namely, the first 1024 samples, the last part is missing. When I remove this check mark, I see all 2048 samples. Is there a simple software fix for this or is it a more involved firmware limitation? 

Other parameters: software version: 5.0.4, firmware version 21305, configured for 0.7 GSPS, display at 500 ns/div

Thank you,

Dmitry Hits.

 

 