DRS4 Forum
  DRS4 Discussion Forum, Page 40 of 46  Not logged in ELOG logo
Entry  Tue Sep 22 17:45:26 2020, Elmer Grundeman, External triggering 

Dear all,

I had a question about timing jitter and external triggering.

I trigger the board externally with a 3V pulse from a DG645 delay generator and as a test I use the gated charge function to integrate another pulse of the DG which goes into channel 1 (the timing jitter between different outputs of the DG is on the order of ~25 picoseconds).

The issue I’m encountering is that the signal on channel 1 is jittering in time with ~1 ns, which means the signal is jittering with respect to my integration gate (point A and B). If I look at the data it always starts at t = 0.000 but my signal (pulse) moves around in time.

If I don’t use the external trigger but trigger on channel 1 directly the signal does not move with respect to the gate, but I can see the start and end of the trace move in time. If I look at the data the first data point is not at t = 0.000 but some other time, which jitters with ~1 ns.

I did repeat the voltage and timing calibration, but that did not help either.

Do you know where this jitter comes from and if I can get rid of it?

Best regards,

 

Elmer

    Reply  Wed Oct 7 11:17:52 2020, Elmer Grundeman, External triggering 

I will try that, thanks!

Stefan Ritt wrote:

The trigger is there only to trigger the chip, but cannot be used as a precise time reference. If you want to measure precise timing, do this always BETWEEN two inputs, never between an input and the trigger. You might want to split and delay your trigger signal and feed one copy to another input of the evaluation board as your reference.

Stefan

Elmer Grundeman wrote:

Dear all,

I had a question about timing jitter and external triggering.

I trigger the board externally with a 3V pulse from a DG645 delay generator and as a test I use the gated charge function to integrate another pulse of the DG which goes into channel 1 (the timing jitter between different outputs of the DG is on the order of ~25 picoseconds).

The issue I’m encountering is that the signal on channel 1 is jittering in time with ~1 ns, which means the signal is jittering with respect to my integration gate (point A and B). If I look at the data it always starts at t = 0.000 but my signal (pulse) moves around in time.

If I don’t use the external trigger but trigger on channel 1 directly the signal does not move with respect to the gate, but I can see the start and end of the trace move in time. If I look at the data the first data point is not at t = 0.000 but some other time, which jitters with ~1 ns.

I did repeat the voltage and timing calibration, but that did not help either.

Do you know where this jitter comes from and if I can get rid of it?

Best regards,

 

Elmer

 

 

Entry  Tue May 27 13:46:18 2014, Dominik Neise, Spikes in DRS4 data on custom baord. 

We see quite some spikes in our DRS4 sampled data in FACT.  We see different types of spikes:

  • single cell spikes, usually showing a large amplitude of 200mV
  • double cell spikes, usually only in the order of 20mV.
  • Even triple and quadro cell spikes are rarely seen.

The double cell spikes often occur as symmetrical double cell spikes mirrored at cell 512. quadro cell spikes seem to be nothing else, than connected symmetrical double cell spikes. For the triple cell spikes we have no idea.
Currently we use simple filters to get rid of these spikes, this workes rather well for the large single cell spikes, but with the occurance of tripples and quadros we started to worry about higher multiples and revived our DRS4 spike investigations.

Now I was told, that you Stefan know already where these spikes come from and even a paper exisits. Unfortunately so far I was unable to find it.
I wonder if it is possible to predict the occurance of these spikes, so one does not have to search for them anymore and can get rid of the filters.

Best regards

Dominik

Entry  Fri Mar 11 19:50:18 2016, Dominik Neise,  

Hello Stefan,

I just stumbled again over a phrase in the DRS4 datasheet I never really understood, but didn't find the time to ask.

On page 8 it says: "An internal circuit ensures that the write signal is always 16 cells wide."

So when I look at a single channel, do I understand correctly, that at any given time during sampling, always 16 cells are open, i.e. 16 cells are connected to the analog inputs? So when the domino frequency is e.g. 5GHz then each cell sees the analog input not for 200ps but for 3.2ns correct?

Entry  Wed Jun 1 22:29:01 2016, Dominik Neise, problems when stop cell >= 767 ?? stop_cell_distribution.png

Hello Stefan,

some colleages told me a story, I was neither able to confirm nor find anything in the datsheet about. According to them:

For some internal reason of the DRS4, if the “stop capacitor” of the DRS4 is >= 767, the true stop channel is one before the stop channel read from the DRS4. In other words, the stop channel which returns the DRS4 shifts after sampling to the capacitor ID 766.

Can you confirm that, or even say a few words about that matter?

I wanted to confirm this by plotting the stop cell distribution for random triggered data, taken with one of the FACT boards. I assumed (possibly misunderstanding the matter), that this would lead to missing values in the area of stop cell 767, but cannot see any significant excess or lack of entries in that area.

 

Entry  Mon Feb 29 12:58:17 2016, Dmitry Philippov, baseline shift pic1.pngpic2.pngpic3.png

Hello! My name is Dmitry. I am from SiPM Lab is NRNU MEPhI (Russia, Moscow). We bought DRS4 evaluation board V5 with firmware 21305. We use 5.0.4 build 21911 2015-11-23 software version (and before that we used 5.0.3 build 21508, 2014-10-15) with Windows 7 32bit.

We observe some strange behaviour. When we save waveforms (in xml or binary data) we see that some of them have the baseline shifted of about -5 mV.

The first picture (pic1) is 1000 waveforms which were glued in one. It is clearly see that baseline quite often has the shift.

The same effect can be seen without saving (writting): rarely when we use normal or auto trigger mode (pic3), and always in single trigger mode (pic2).

The images are attached.

Do you have any idea how it can be fixed?

 

Thanks, Dmitry.

 

Entry  Thu Feb 28 10:47:14 2013, Dmitry Hits, clock and trigger outs 
Hi,
I am considering using the DRS4 evaluation board as an ADC card for the wire chamber in the physics lab (VP) experiment at ETH. However, the wire 
chamber has 8 outputs, so I would need to have two of such boards. Is it possible to synchronise them, online or offline? From the website, it looks 
like yes, but the documentation says that these features (trigger and clock out) may not have been implemented in firmware yet. Could you tell me 
the status?

Thank you very much,

Dmitry.
Entry  Mon Apr 8 18:11:02 2013, Dmitry Hits, binary to root 

 Hi,

 

Does anyone has a program that converts a binary file from drsosc  output to a ROOT tree format?

 

Thank you,

 

Dmitry.

Entry  Tue Jul 9 11:40:00 2013, Dmitry Hits, cannot save in binary format 

Hi,

I would like to save the waveform in a binary format. When I click Save then change format from xml to dat in the menu. I still get xml format but with dat extension.

what I am missing?

Thank you,

Dmitry.

 

    Reply  Tue Jul 9 14:00:49 2013, Dmitry Hits, cannot save in binary format 

Stefan Ritt wrote:

Dmitry Hits wrote:

Hi,

I would like to save the waveform in a binary format. When I click Save then change format from xml to dat in the menu. I still get xml format but with dat extension.

what I am missing?

Thank you,

Dmitry.

Version 4.0.1 has a problem there. Please use 4.0.0. If you can compile the program yourself, just change this line:

--- DOFrame.cpp (revision 20656)
+++ DOFrame.cpp (revision 20655)
@@ -517,7 +517,7 @@
       if (!filename.empty()) {
          m_btSave->SetLabel(_T("Close"));
          m_btSave->SetToolTip(_T("Stop saving waveforms"));
-         if (filename.Find(_T(".")) != wxNOT_FOUND) {
+         if (filename.Find(_T(".xml")) != wxNOT_FOUND) {
             m_WFfd = 0;
             m_WFFile = mxml_open_file(filename.char_str());
             if (m_WFFile)

 

 Thanks that fixed it!

Dmitry

Entry  Wed Nov 6 11:53:28 2013, Dmitry Hits, flickering screen for drsosc 

Hi,

 

I have install drs software on ASUS EeeBox with Ubuntu 12.04 LTS. When I try to use ./drsosc the oscilloscope window flickers. Can you suggest what might be the problem?

 

Here is some more info:

******************************************************

System:

Ubuntu 12.04 LTS

Memory: 992.9 MiB

CPU: Intel Atom CPU N270 @ 1.6GHz x 2

32 Bit

Disc: 156.5 GB

***************************************

Software:

Due to version Ubuntu I had to install the wxWidgets from source (wxWidgets with x11)

 

Thank you,

 

Dmitry.

    Reply  Mon Nov 18 11:20:15 2013, Dmitry Hits, flickering screen for drsosc 

Stefan Ritt wrote:

Dmitry Hits wrote:

Hi,

 

I have install drs software on ASUS EeeBox with Ubuntu 12.04 LTS. When I try to use ./drsosc the oscilloscope window flickers. Can you suggest what might be the problem?

 

Here is some more info:

******************************************************

System:

Ubuntu 12.04 LTS

Memory: 992.9 MiB

CPU: Intel Atom CPU N270 @ 1.6GHz x 2

32 Bit

Disc: 156.5 GB

***************************************

Software:

Due to version Ubuntu I had to install the wxWidgets from source (wxWidgets with x11)

 

Thank you,

 

Dmitry.

This problem is new. Even on a slower Raspberry Pi we did not see any flickering. Have you tried a different version of wxWidgets? 

yes the problem was in wxWidgets, I have downgraded the ubuntu version and installed packaged version of wxWidgets. Now it works without problems.

Thank you,

Dmitry. 

Entry  Mon Nov 18 15:49:01 2013, Dmitry Hits, synchronisation of readouts of two boards for offline analysis 

 Dear Stefan,

I am trying to synchronise the readout of two test boards, one is the DRS4 test board, the other is PSI46 test board (used for the readout of  CMS pixel chip) for the offline analysis. I think that the most secure way to accomplish this is to pass a trigger number from one test board to the other.

The PSI46 test board has a software which allows it to accept a 16 bit number following the trigger pulse. I was wondering whether it would be possible for DRS4 board to generate such a trigger number on the trigger out line after sending the trigger. Also would it be possible to record this trigger number for every event stored by DRS4 board?

If none of this possible or requires a lot of time, then as a minimum, would it be possible to send-through only the triggers that were recorded by the DRS4 board?

Please let me know if you have better idea how to do this.

Thank you very much,

Dmitry.

Entry  Tue Nov 26 15:36:39 2013, Dmitry Hits, reducing sampling speed 

Dear Stefan

Is there an easy way to reduce sampling speed below 0.7 GSPS? I would like to record traces up to 5 usec long.

Thank you

Dmitry 

Entry  Fri Dec 13 10:37:18 2013, Dmitry Hits, input protection in DRS4 evaluation board 

 Dear Stefan

Last month I was using a DRS4 evaluation board to digitise the signal from the charged particles in piM1 beam line at PSI. The beam in piM1 is a mixture of pions with a small percentage of protons. Pions are close to minimum ionising and were producing the signals on the order of 250 mV (Landau distributed). The protons at this momentum (250 MeV) are not minimum ionising  and produced much higher signals ( I could not exactly measure them because they were of scale for the DRS4 board). The pulses were on the order of 0.5 usec long. At low rate (~1kHz) the board was able to handle them, but when I turned the rate of particles to about ~300 kHz, the channel went flat. My question is whether the evaluation board has some type of input protection that would be possible to replace? Or does that mean that I have burned the input of the DRS4 board itself? Other channels behaving normal.

Thank you

Dmitry. 

    Reply  Mon Dec 16 11:09:25 2013, Dmitry Hits, synchronisation of readouts of two boards for offline analysis 

Stefan Ritt wrote:

Dmitry Hits wrote:

 Dear Stefan,

I am trying to synchronise the readout of two test boards, one is the DRS4 test board, the other is PSI46 test board (used for the readout of  CMS pixel chip) for the offline analysis. I think that the most secure way to accomplish this is to pass a trigger number from one test board to the other.

The PSI46 test board has a software which allows it to accept a 16 bit number following the trigger pulse. I was wondering whether it would be possible for DRS4 board to generate such a trigger number on the trigger out line after sending the trigger. Also would it be possible to record this trigger number for every event stored by DRS4 board?

If none of this possible or requires a lot of time, then as a minimum, would it be possible to send-through only the triggers that were recorded by the DRS4 board?

Please let me know if you have better idea how to do this.

Thank you very much,

Dmitry.

There are indeed several methods. You can output the trigger number at the DRS evaluation board via the trigger output, but you would have to implement this yourself in the firmware.

The send-though of recorded triggers is already implemented in the board, so you could use that. The only thing to make sure is to to readout and re-enable the PSI46 board before you readout the DRS4 board. If you would first read the DRS4 board, and re-enable the DRS4 board via StartDomino(), then there could be the next trigger going through to the PSI46 board without that board being ready. So the sequence is

- connect trigger out of DRS4 to PSI46

- arm PSI46 board

- arm DRS4 board

- wait for trigger by calling IsBusy()

- read out PSI46 board

- read out DRS4 board

- call StartDomino(), which re-enables also the trigger though

 

Best regards,

Stefan

 Dear Stefan

Thank you very much for the answer. I did not have a chance to implement this yet.

I have a  follow up question: 

Is the following sequence already implemented in the DRS oscilloscope program? Could you point me to an example of such a sequence?

- arm DRS4 board

- wait for trigger by calling IsBusy()

- read out DRS4 board

- call StartDomino(), which re-enables also the trigger through

 

Cheers

Dmitry

 

 

Entry  Fri Sep 12 11:52:21 2014, Dmitry Hits, synchronizing two DRS4 evaluation boards readout with one computer 

 Hi everyone,

Has anyone tried to synchronize 2 (two) DRS4 evaluation boards readout by the same computer? I have read about some attempts on this board in the past, but I do not know if they have succeeded. If yes, could you share your experience and/or software.

Thank you very much,

Dmitry.

 

    Reply  Fri Sep 12 13:37:42 2014, Dmitry Hits, synchronizing two DRS4 evaluation boards readout with one computer 

Stefan Ritt wrote:

Dmitry Hits wrote:

 Hi everyone,

Has anyone tried to synchronize 2 (two) DRS4 evaluation boards readout by the same computer? I have read about some attempts on this board in the past, but I do not know if they have succeeded. If yes, could you share your experience and/or software.

Thank you very much,

Dmitry.

 

Please read the manual http://www.psi.ch/drs/DocumentationEN/manual_rev50.pdf page 25 where this is described in detail.

/Stefan

 Hi Stefan,

 

Thank you for pointing me to the document. Does it apply only to version 5 of the board or can it be applied also to version 4 (which is the one I have)?

Dmitry

Entry  Fri Sep 12 14:57:22 2014, Dmitry Hits, compilation error for v5.0.2 

 Hi,

I am getting the following compilation error when trying to compile version 5.0.2 software:

src/DOFrame.cpp:617:14: error: invalid conversion from ‘char*’ to ‘wxChar {aka wchar_t}’ [-fpermissive]
 
I have wxWidgets v. 2.8.12 package on Fedora version 3.9.10-100.fc17.x86_64
 
Has anyone seen this before?
 
Thank you,
 
Dmitry 

 

---------------------------------------------------------------------------------------------

Full error report:

g++ -g -O2 -Wall -Wuninitialized -fno-strict-aliasing -Iinclude -I/usr/local/include -DOS_LINUX -DHAVE_USB -DHAVE_LIBUSB10 -DUSE_DRS_MUTEX -I/usr/lib64/wx/include/gtk

2-unicode-release-2.8 -I/usr/include/wx-2.8 -D_FILE_OFFSET_BITS=64 -D_LARGE_FILES -D__WXGTK__ -pthread -c src/DOFrame.cpp
src/DOFrame.cpp: In member function ‘void DOFrame::LoadConfig(char*, int)’:
src/DOFrame.cpp:617:14: error: invalid conversion from ‘char*’ to ‘wxChar {aka wchar_t}’ [-fpermissive]
In file included from /usr/include/wx-2.8/wx/memory.h:16:0,
                 from /usr/include/wx-2.8/wx/object.h:20,
                 from /usr/include/wx-2.8/wx/wx.h:16,
                 from include/DRSOscInc.h:9,
                 from src/DOFrame.cpp:9:
/usr/include/wx-2.8/wx/string.h:1413:13: error:   initializing argument 1 of ‘wxString& wxString::operator+=(wxChar)’ [-fpermissive]
    Reply  Fri Sep 12 16:38:24 2014, Dmitry Hits, compilation error for v5.0.2 

Stefan Ritt wrote:

Dmitry Hits wrote:

 Hi,

I am getting the following compilation error when trying to compile version 5.0.2 software:

src/DOFrame.cpp:617:14: error: invalid conversion from ‘char*’ to ‘wxChar {aka wchar_t}’ [-fpermissive]
 
I have wxWidgets v. 2.8.12 package on Fedora version 3.9.10-100.fc17.x86_64
 
Has anyone seen this before?
 
Thank you,
 
Dmitry 

 

---------------------------------------------------------------------------------------------

Full error report:

g++ -g -O2 -Wall -Wuninitialized -fno-strict-aliasing -Iinclude -I/usr/local/include -DOS_LINUX -DHAVE_USB -DHAVE_LIBUSB10 -DUSE_DRS_MUTEX -I/usr/lib64/wx/include/gtk

2-unicode-release-2.8 -I/usr/include/wx-2.8 -D_FILE_OFFSET_BITS=64 -D_LARGE_FILES -D__WXGTK__ -pthread -c src/DOFrame.cpp
src/DOFrame.cpp: In member function ‘void DOFrame::LoadConfig(char*, int)’:
src/DOFrame.cpp:617:14: error: invalid conversion from ‘char*’ to ‘wxChar {aka wchar_t}’ [-fpermissive]
In file included from /usr/include/wx-2.8/wx/memory.h:16:0,
                 from /usr/include/wx-2.8/wx/object.h:20,
                 from /usr/include/wx-2.8/wx/wx.h:16,
                 from include/DRSOscInc.h:9,
                 from src/DOFrame.cpp:9:
/usr/include/wx-2.8/wx/string.h:1413:13: error:   initializing argument 1 of ‘wxString& wxString::operator+=(wxChar)’ [-fpermissive]

I don't get this error under gcc 4.4.7, so I guess you have a newer version. Each one becomes more picky. Just try to replace

str += filename; 

with

str += (wxString) filename;

in line 617 of DOFrame.cpp

/Stefan

 

Hi Stefan,

 

Unfortunately that did not work and from suggestions in the error I do see a good solution:

----------------------------

src/DOFrame.cpp: In member function ‘void DOFrame::LoadConfig(char*, int)’:

src/DOFrame.cpp:617:25: error: call of overloaded ‘wxString(char [1024])’ is ambiguous

src/DOFrame.cpp:617:25: note: candidates are:

/usr/include/wx-2.8/wx/string.h:722:3: note: wxString::wxString(const wxWCharBuffer&) <near match>

/usr/include/wx-2.8/wx/string.h:722:3: note:   no known conversion for argument 1 from ‘char [1024]’ to ‘const wxWCharBuffer&’

/usr/include/wx-2.8/wx/string.h:692:3: note: wxString::wxString(wxChar, size_t) <near match>

/usr/include/wx-2.8/wx/string.h:692:3: note:   no known conversion for argument 1 from ‘char [1024]’ to ‘wxChar {aka wchar_t}’

/usr/include/wx-2.8/wx/string.h:690:3: note: wxString::wxString(const wxString&) <near match>

/usr/include/wx-2.8/wx/string.h:690:3: note:   no known conversion for argument 1 from ‘char [1024]’ to ‘const wxString&’

/usr/include/wx-2.8/wx/string.h:682:3: note: wxString::wxString(int) <near match>

/usr/include/wx-2.8/wx/string.h:682:3: note:   no known conversion for argument 1 from ‘char [1024]’ to ‘int’

---------------------------------

 

let me know if you see one.

 

Thank you,

 

Dmitry.

 

 

 

_____________________________________________________________________________________________________________________________________________

Full error:

++ -g -O2 -Wall -Wuninitialized -fno-strict-aliasing -Iinclude -I/usr/local/include -DOS_LINUX -DHAVE_USB -DHAVE_LIBUSB10 -DUSE_DRS_MUTEX -I/usr/lib64/wx/include/gtk

2-unicode-release-2.8 -I/usr/include/wx-2.8 -D_FILE_OFFSET_BITS=64 -D_LARGE_FILES -D__WXGTK__ -pthread -c src/DOFrame.cpp

src/DOFrame.cpp: In member function ‘void DOFrame::LoadConfig(char*, int)’:

src/DOFrame.cpp:617:14: error: invalid conversion from ‘char*’ to ‘wxChar {aka wchar_t}’ [-fpermissive]

In file included from /usr/include/wx-2.8/wx/memory.h:16:0,

                 from /usr/include/wx-2.8/wx/object.h:20,

                 from /usr/include/wx-2.8/wx/wx.h:16,

                 from include/DRSOscInc.h:9,

                 from src/DOFrame.cpp:9:

/usr/include/wx-2.8/wx/string.h:1413:13: error:   initializing argument 1 of ‘wxString& wxString::operator+=(wxChar)’ [-fpermissive]

[dmitry@kitkat ~]$ more error-drs4v5_2

g++ -g -O2 -Wall -Wuninitialized -fno-strict-aliasing -Iinclude -I/usr/local/include -DOS_LINUX -DHAVE_USB -DHAVE_LIBUSB10 -DUSE_DRS_MUTEX -I/usr/lib64/wx/include/gtk

2-unicode-release-2.8 -I/usr/include/wx-2.8 -D_FILE_OFFSET_BITS=64 -D_LARGE_FILES -D__WXGTK__ -pthread -c src/DOFrame.cpp

src/DOFrame.cpp: In member function ‘void DOFrame::LoadConfig(char*, int)’:

src/DOFrame.cpp:617:25: error: call of overloaded ‘wxString(char [1024])’ is ambiguous

src/DOFrame.cpp:617:25: note: candidates are:

In file included from /usr/include/wx-2.8/wx/memory.h:16:0,

                 from /usr/include/wx-2.8/wx/object.h:20,

                 from /usr/include/wx-2.8/wx/wx.h:16,

                 from include/DRSOscInc.h:9,

                 from src/DOFrame.cpp:9:

/usr/include/wx-2.8/wx/string.h:722:3: note: wxString::wxString(const wxWCharBuffer&) <near match>

/usr/include/wx-2.8/wx/string.h:722:3: note:   no known conversion for argument 1 from ‘char [1024]’ to ‘const wxWCharBuffer&’

/usr/include/wx-2.8/wx/string.h:692:3: note: wxString::wxString(wxChar, size_t) <near match>

/usr/include/wx-2.8/wx/string.h:692:3: note:   no known conversion for argument 1 from ‘char [1024]’ to ‘wxChar {aka wchar_t}’

/usr/include/wx-2.8/wx/string.h:690:3: note: wxString::wxString(const wxString&) <near match>

/usr/include/wx-2.8/wx/string.h:690:3: note:   no known conversion for argument 1 from ‘char [1024]’ to ‘const wxString&’

/usr/include/wx-2.8/wx/string.h:682:3: note: wxString::wxString(int) <near match>

/usr/include/wx-2.8/wx/string.h:682:3: note:   no known conversion for argument 1 from ‘char [1024]’ to ‘int’

make: *** [DOFrame.o] Error 1 

ELOG V3.1.5-3fb85fa6