DRS4 Forum
  DRS4 Discussion Forum, Page 37 of 46  Not logged in ELOG logo
    Reply  Thu Mar 18 21:38:10 2010, Hao Huan, Serial Interface Frequency of the DRS Chip 

Stefan Ritt wrote:

Hao Huan wrote:

in the DRS4 datasheet I read that the optimal frequency for SRCLK is 33MHz. However in the evaluation board firmware SRCLK is toggled at rising edges of the internal 33MHz clock, i.e. the frequency of SRCLK itself is 16.5MHz instead. Is that frequency better than 33MHz?

The reason for the 16.5 MHz is the following:

After each block of 32 bins, the DRS4 chip switches an internal segment, which causes some small spike at the analog output of the chip. This spike is a bit wider than 30ns, so if everything is digitized with 33 MHz, then you see small spiked each 32 cells. The appropriate solution would be to modify the firmware to digitize all cells with 30ns (33 MHz) and all cells with the spike with ~50 ns (20 MHz). If you do the ROI readout mode, you don't know for the first 10 cells if one of them belong to this class, since the cell address takes 10 cycles to be read out. So you would first have to read 10 cells, and then if you realize that one of them is one of the problematic ones (cell number modulo 32 is zero), you have to re-read the first 10 cells, and digitize the problematic cell with a longer settling time. Now this is a bit complicated to implement in the firmware, so I was just too lazy to do it and decided to digitize everything with 16.5 MHz. But if you are worried about the dead time, you should consider implementing the mentioned algorithm. 

 Thanks! The suggested algorithm looks promising. However, if the spikes take place only for those specific cells, is it possible to absorb them into the offset calibration?

Entry  Sun Mar 21 02:03:44 2010, Hao Huan, PLL Loop Filter Configuration 

Hi Stefan,

    in the datasheet it says at 6GSPS the typical loop filter parameters are 220Ω, 2.2nF and 27nF. If I want to run the Domino wave nominally at 1GHz, i.e. with a reference clock frequency around 0.5MHz, is there any recommended loop filter configuration? Is the setup of the evaluation board, that is, 220Ω, 3.3nF and 33nF an optimal choice?

    Thank you very much.

 

Entry  Tue Mar 30 22:57:34 2010, Hao Huan, ROFS Configuration 

Hi Stefan,

    according to the DRS4 datasheet, if we want an input range centered around U0, the ROFS should be 1.55V-U0. However when I read the codes of the evaluation board application, ROFS seems to be 1.6V-1.25*U0 where the coefficient 1.25 is said to come from sampling cell charge injection correction. Is it the right equation to use? What exactly does that charge injection correction mean?

    Thanks a lot.

 

Entry  Fri Apr 9 17:14:45 2010, Hao Huan, Baseline Variation In Data 

Hi Stefan,

    when I sample a constant input with the DRS 4 chip, there was a baseline variation showing up as a saw-tooth pattern which grows with the absolute value of the differential input. Do you think this is the kind of baseline variation mentioned in the evaluation board manual, i.e. coming from clock jitter in ADC sampling?

    Thanks a lot!

 

Entry  Thu May 13 19:14:27 2010, Hao Huan, DVDD Problem of DRS 4 

Hi Stefan,

    on our board some DRS chips draw a lot of current through DVDD after power-up and heat up significantly--it is true that our board doesn't have weak pull-down resistors at DENABLE and DWRITE output pins of FPGA, so this problem might have been caused by that, but a reinitialization of the Domino circuit doesn't help either. We tried different capacitors at DVDD and it seemed the larger the capacitance, the better the result--with a capacitor larger than 10nF some of the DRS chips could work happily in the normal way while if the capacitor is only 4.7nF all of them got very hot. Would you please provide some suggestions why there should be such a problem?

    Thanks a lot!

    Reply  Tue May 18 01:47:59 2010, Hao Huan, DVDD Problem of DRS 4 

Stefan Ritt wrote:

Hao Huan wrote:

Hi Stefan,

    on our board some DRS chips draw a lot of current through DVDD after power-up and heat up significantly--it is true that our board doesn't have weak pull-down resistors at DENABLE and DWRITE output pins of FPGA, so this problem might have been caused by that, but a reinitialization of the Domino circuit doesn't help either. We tried different capacitors at DVDD and it seemed the larger the capacitance, the better the result--with a capacitor larger than 10nF some of the DRS chips could work happily in the normal way while if the capacitor is only 4.7nF all of them got very hot. Would you please provide some suggestions why there should be such a problem?

    Thanks a lot!

I found that sometimes even a reinitialization fails if the pull-down resistors are missing. So instead playing with capacitors at DVDD, I would just solder two resistors on the board which should fix the problem completely.

 Thanks! After adding pull-down resistors the voltages come back to normal.

However there is another weird problem that arises: a reset pulse seems unable to set the internal shift registers to default values. For example, after reset without addressing the Config Register the PLL will not try to lock with external reference clock. Even if I explicitly address the Config Register after reset and have the PLL locked, some channels of the chip will give null output during readout while other channels work normally. Could it be that some channels are not initiated properly with the Domino circuit?

    Reply  Wed May 19 02:24:12 2010, Hao Huan, DVDD Problem of DRS 4 

Stefan Ritt wrote:

Hao Huan wrote:

Stefan Ritt wrote:

Hao Huan wrote:

Hi Stefan,

    on our board some DRS chips draw a lot of current through DVDD after power-up and heat up significantly--it is true that our board doesn't have weak pull-down resistors at DENABLE and DWRITE output pins of FPGA, so this problem might have been caused by that, but a reinitialization of the Domino circuit doesn't help either. We tried different capacitors at DVDD and it seemed the larger the capacitance, the better the result--with a capacitor larger than 10nF some of the DRS chips could work happily in the normal way while if the capacitor is only 4.7nF all of them got very hot. Would you please provide some suggestions why there should be such a problem?

    Thanks a lot!

I found that sometimes even a reinitialization fails if the pull-down resistors are missing. So instead playing with capacitors at DVDD, I would just solder two resistors on the board which should fix the problem completely.

 Thanks! After adding pull-down resistors the voltages come back to normal.

However there is another weird problem that arises: a reset pulse seems unable to set the internal shift registers to default values. For example, after reset without addressing the Config Register the PLL will not try to lock with external reference clock. Even if I explicitly address the Config Register after reset and have the PLL locked, some channels of the chip will give null output during readout while other channels work normally. Could it be that some channels are not initiated properly with the Domino circuit?

Something is wrong. I have 800 chips, and they all start up fine. Check with your scope the RESET, DSPEED, DENABLE and DTAP signals. When RESET is applied, DSPEED should go to 2.5 V. When DENABLE goes high, the domino wave is started and you should see DTAP toggle. DSPEED is then lowered by the PLL until DTAP matches your external reference clock. I usually keep DENABLE high all the time after initialization, so the domino wave just continues running.

Another problem could however be the chip readout. If some channel gives null output, it could be that your readout has a problem. Do you use RSRLOAD to initialize the readout sequence?

 Yes; I used RSRLOAD to trigger readout of all channels in parallel so the asymmetry between channels of the same chip is really a big puzzle. Also during reset DSPEED indeed goes to 2.5V, but after reset the PLL will not try to lock with the external reference clock and lower DSPEED. Instead the Domino circuit just oscillates at the highest frequency by itself.

A more confusing discovery is that the SRIN level before starting the Domino wave could affect the behavior of the PLL. I mean the level of SRIN when the chip is at A="1111" or "1010". Is SRIN supposed to influence the chip even in these standby or transparent modes?

Entry  Wed May 26 19:18:09 2010, Hao Huan, High Frequency Input for DRS 

Hi Stefan,

    I read in the DRS datasheet that the bandwidth for the transparent mode OUT+ is only 200MHz which I think cannot be improved by any active input buffer; so if you want to operate the chip for really high frequency input, would it be better to feed on-board discriminators not from the output of DRS but from the input end?

    Thanks!

 

Entry  Sun Oct 23 23:32:28 2011, Hao Huan, Phase Shift for ADC Readout 

Dear Dr. Ritt,

    In the DRS 4 datasheet it is recommended to sample the analog output of the chip after 8~10 ns of the SRCLK edge for it to stablize and thus a phase shift between SRCLK and the ADC sampling clock is necessary. However in the latest version of the evaluation board firmware the phase-shifted clock was generated but not really used for the ADC interface. Is there any reason for that?

Entry  Wed Dec 14 00:44:37 2011, Hao Huan, Synchronization Delay in the Firmware for 8051 Controller 

Hi Stefan,

    I have a question regarding the DRS 4 evaluation board firmware for the 8051 controller embedded in the CY7C68013 USB chip: on the board the controller is running at 12 MHz and the FIFO interface of the USB chip is running at 30 MHz, so the number of delay cycles for synchronization as defined in fx2sdly.h should be (3*12000+5*30000-1)/(2*30000)=3, but the actual number used in drs_eval.c is (3*12000+5*48000-1)/(2*48000)=2, so there is a mismatch between the IFCLK frequency used in this calculation and the actual IFCLK frequency configured. Am I misunderstanding something or is there an explanation for that?

 

    Thanks,

Hao Huan

Entry  Sat Feb 2 00:13:12 2019, Hans Steiger, Saving Rate (only 15Acq/s) 

Dear All,

 

when I use my Evaluation Board with some PMTs I can digitize 450 Acq/s or so. But when I want to save the waveforms the rate goes down. The Acqu. rate with saving is in the range of 14Hz up to 24 Hz.

I normally use the .txt file. I try to use the 5GS/s but also with much lower sampling rate the saving rate is not getting much better. 

Is this a problem of my McBook connected to the Evaluation Board?

 

All the best,

 

Hans 

Entry  Mon Feb 4 16:42:08 2019, Hans Steiger, Different Distances between the sampling points 

Dear All,

with the older software for my V5 Board i did not have the problem, that the distance between the sampling points (in time) is not the same (e.g. a sampling point all 200ps for 5GS/s). 

How can i fix this?

Can someone provide me the software for the board which is old enough to not have this problem. All my Root interpreters produce problems with this new data format.  Which version would be old enough?

 

All the best and thanks a lot,

 

Hans

    Reply  Mon Feb 4 17:36:49 2019, Hans Steiger, Different Distances between the sampling points 

Sorry.... but is there a solution or a Root Macro, that reads the waveforms into a Root-Tree? I simply can not work anymore with the data. 

Can you tell me, which software was in use in early 2017?

All the best,

 

Hans

 

Stefan Ritt wrote:

The sampling points are NOT equidestant, they have varying bin widths of 150ps to 250ps at 5GS/s. That's due the way the DRS4 chip works. You might have neglected that fact in the past, but that would have led to poor timing resolutions (typically 1-2ns resolution only). To get bins with the same width, you have to treat your waveform as a real X/Y points (or better U/T), and the re-sample that cure, maybe spline-interpolated, at 200ps bins.

Stefan

Hans Steiger wrote:

Dear All,

with the older software for my V5 Board i did not have the problem, that the distance between the sampling points (in time) is not the same (e.g. a sampling point all 200ps for 5GS/s). 

How can i fix this?

Can someone provide me the software for the board which is old enough to not have this problem. All my Root interpreters produce problems with this new data format.  Which version would be old enough?

 

All the best and thanks a lot,

 

Hans

 

 

Entry  Sat Aug 29 22:00:30 2020, Hans Steiger, Dynamic Range Evaluation Board and Software 

Dear Evaluation Board Team,

 

currently I am facing the problem of digitizing pulses with an amplitude of -0.6V to -0.8V. As the dynamic range of the board is 1Vpp, this should be feasible. However, I do not know how to set in the software a correct range. I see only -0.5V/0.5V, and the two positive options. Normally I would use -0.5V/0.5V and give the thing an offset of 0.4V or so? Is this possible? Where can I set such a offset?

 

All the best,

Hans

Entry  Mon Aug 31 16:44:12 2020, Hans Steiger, Channel Cascading 

Dear All,

I have a board with Channel Cascading Option. I have the problem, that it seems to be impossible to run all 4 Channels simultaneously for digitizing pulses. I can just run even or odd channels but not even and odd ones? If I run in combined option, My question: If a board comes with this combined option, is it still usable as a 4Ch Digitizer but with 1024bin traces?

 

All the best,

 

Hans

Entry  Mon Sep 15 16:24:41 2014, Hannes Wachter, Timing Calibration Fail 

Hi,

has anyone experienced a shutdown of the DRSosc.exe or DRScl.exe when executing a Timing Calibration? Also, when we add the command b->CalibrateTiming(NULL); to the drs_exam.cpp and run the exe, our program shuts down immediately and windows shows an error message (identical to DRSosc and DRScl).

Any help is appreciated.

 

    Reply  Wed Sep 7 17:28:25 2011, Hannes Friederich, DRS4 and AD9222 

Guillaume Blanchard wrote:

Hello,

I am designing a DAQ board with both DRS4 + AD9222 and a  FPGA to monitor.

Do I have to change the default value of O-OFS ?

Does a simple low-pass filter (series resistor + capacitor) on each AD9222 input is enough to limit the noise ?

I am planning to use the (DRS4,AD9222,FPGA) group as both a trigger and digitizing system (as shown in the DRS4 datasheet). The DRS4 will be working at 5Ghz with 8 active channels.
So each channel will have a time depth of 1/5Ghz x 1024 = 204.8ns. So, in order to miss nothing, the ADC latency + the trigger decision must be inferior to 204.8ns, am I correct ?
This leads me to implement on my board the 65Mhz version of the AD9222 as this converter has a 8 clock period latency, i.e. 123ns and it left me 81ns to perform a trigger decision ?

Cordially,

G.Blanchard

 Like Stefan pointed out, your time constraints are quite tight. In those 81 ns, you also need to deserialize the AD9222 output. Unless you implement some really fancy input comparison logic, this will consume another 1-2 ADC clock cycles. Perhaps you should first verify that your FPGA design actually can do its job within those 81 ns. In our system, we sample at only 1-2 GHz and have enough margin to implement really complex triggers in FPGA. But the total latency (ADC + FPGA deserialization) takes 250 ns.

Depending on the application, you do need a low-pass filter. Not only because of the noise, but also in order to be able to trigger reliably. Using fast PMTs for example, you will not be able to see all pulses in full size if the bandwidth is 50 MHz and you're only sampling at 65 MSPS.

Hannes

Entry  Wed Nov 21 08:34:52 2012, Gyuhee Kim, Question for using Multi board 

 Hi.

 

I have 2 DRS4 evaluation V4 boards, and I want to use these 2 board to multi board DAQ system for 4 ch vs 4 ch DAQ.

But there is no option for multi board use. I just only find the multi board trigger mode check button on DRS4 Oscilloscope program, but I couldn`t check. 

Is there any method to use multi board?

 

Best regards.

Gyuhee.

    Reply  Wed Nov 21 08:48:00 2012, Gyuhee Kim, Question for using Multi board 

Stefan Ritt wrote:

Gyuhee Kim wrote:

 Hi.

 

I have 2 DRS4 evaluation V4 boards, and I want to use these 2 board to multi board DAQ system for 4 ch vs 4 ch DAQ.

But there is no option for multi board use. I just only find the multi board trigger mode check button on DRS4 Oscilloscope program, but I couldn`t check. 

Is there any method to use multi board?

 

Best regards.

Gyuhee.

This mode is not yet implemented in firmware. Maybe I find some time towards the end of this year to add this. At the moment, you have to build and external trigger to synchronize the two boards. There are also 16-channel boards on the market where you would not need a multi-board mode. Just Google for "DT5742".

/Stefan

 Thanks Stefan.

I will build external trigger system. 

Entry  Mon Dec 3 08:32:28 2012, Gyuhee Kim, Another question about using multi boards. 

 Hi.

 

I asked about using multi boards some days ago, and I got answer to use external trigger. (Thanks Stefan!)

And here is another question. I made two external triggers and try to acquire coincidence data using two boards. but DRS Oscilloscope program can connect only one board and don`t acquire both of them simultaneously.

So I tried to use two computer for each board separately, but, well, you already know, I failed to acquire because two computers don`t promise to synchronize the two boards acquisition.

Is there any method to solve this problem?

 

1. I want to acquire coincidence data from the two DRS 4 Evaluation board V4 simultaneosly.

2. I have external trigger to provide the two boards at the same time.

3. How can I get data from the two boards?

 

Best regards.

Gyuhee.

ELOG V3.1.5-3fb85fa6