Entry  Wed Jun 13 13:23:17 2018, Julian Kemp, Maximum analog input voltage 

Dear all,

I have been wondering what the maximum analog input voltage for the DRS4 V5 evaluation board is. It came with a sticker indicating that it is "2.5V pk Max". On the other hand, when checking the manual (https://www.psi.ch/drs/DocumentationEN/manual_rev50.pdf), it says maximum allowed is 10V DC or even 30V for short pulses. I foresee an application where I cannot make sure that pulses stay below 2.5V, so the correct value will be quite important for me.

Best,
Julian

    Reply  Wed Jun 13 16:34:28 2018, Julian Kemp, Maximum analog input voltage 

Thank you! That solves my problem.

Stefan Ritt wrote:

In principle the numbers in the manual are correct. But they relate to pulses of a certain length, because the input protection only works for DC voltage and for pulses which are not too long. Since we could not write this all on the label of the board, we decided to put there 100% safe value as a "warning" to people, meaning that if pulses are above 2.5V, they should look into the manual and read the details. 

Stefan

Julian Kemp wrote:

Dear all,

I have been wondering what the maximum analog input voltage for the DRS4 V5 evaluation board is. It came with a sticker indicating that it is "2.5V pk Max". On the other hand, when checking the manual (https://www.psi.ch/drs/DocumentationEN/manual_rev50.pdf), it says maximum allowed is 10V DC or even 30V for short pulses. I foresee an application where I cannot make sure that pulses stay below 2.5V, so the correct value will be quite important for me.

Best,
Julian

 

 

Entry  Fri Oct 13 03:39:01 2017, Jonathan Wapman, Raspberry Pi Connection Failure 

I am currently attempting to use a raspberry pi to connect to the DRS 4 board. I whenever I try to use the DRS Command Line TOol, Revision 21435 to connect to the drs board, I get the error

"musb_open: libusb_open() error -3"

"USB successfully scanned, but no boards found"

"No DRS Boards Found".

I successfully compiled the libusb driver before compiling the drs software 5.0.6, and installed all other listed packages in the install instructions.

Entry  Thu May 8 23:23:19 2025, Jonathan Bradshaw, Clarification of full channel readout Screenshot_2025-05-09_093332.png

Hi all

We're working on a new product using the DRS4 IC, and want to do a full readout from cell 0 (not just Region of Interest).  I have a couple of questions I hope you can help me with:

  • We plan to do a full readout sequence, starting at cell 0.  Part of that sequence includes pulsing RSRLOAD and reading out the stop position as shown in v0.9 datasheet Figure 15.  What should the DRS4 address bits A3..0 be set to for reading out the stop position?  (I’m assuming it’s 1011 ‘Address Read Shift Register’)
  • What is the output delay from the falling edge of SRCLK to valid data at SROUT?
  • For channel readout, we pulse SRCLK to advance the read shift register.  The diagram shown in v0.9 datasheet Figure 12 appears to show that the analog output is updated on the rising edge of SRCLK. Is this correct or have I misread the diagram? (Other shift register transfers are clocked on the falling edge
  • The DRS4 v0.9 datasheet Figure 7 shows that the Configuration register is clocked on the falling edge of SRCLK.  Just below that is the text “The new register content becomes immediately active at the eighth rising edge of the SRCLK signal.” Should that perhaps read ‘… the eighth falling edge of the SRCLK signal’?
Entry  Thu May 8 23:41:03 2025, Jonathan Bradshaw, Handling of Write Shift Register and Write Config Register 

Hi all

We're building a product which will use two different operating modes; firstly a long capcture using channel daisy chaining (2048 samples) and secondly a segmented capture (2 separate captures of 1024 samples each). 

For the long capture, I'm looking to capture 2048 samples for 4 channels.  Therefore I configure the Write Shift Register to 0b01010101 and the Write Config Register to 0b11111111.  During capture with DWRITE=1 the Write Shift Register will update.  Am I correct that once the capture is done and DWRITE=0, I can set A3..0 to 0b1101 and simply read the value of WSROUT to tell the difference?

For the segmented capture, I'm looking to capture 1024 samples for 4 channels on a first tirgger pulse, followed by 1024 samples for 4 channels on a second pulse.  Therefore I configure the Write Shift Register to 0b11111111 and the Write Config Register to 0b01010101 and set DWRITE=0 to capture.  After the first trigger I set DWRITE=0 and need to update the Write Config Register.  Do I need to write in a whole 8 bits to the Write Config Register (i.e. 0b10101010), or can I just shift in a single new bit (value 0b0)?

    Reply  Tue May 13 04:10:30 2025, Jonathan Bradshaw, Handling of Write Shift Register and Write Config Register 

Hi Stefan

Just so I'm 100% clear; is there no reliable way to perform 2 segmented captures with a single DRS4 IC?

While not a showstopper, this is a bit disappointing.

 

Stefan Ritt wrote:

This is correct. Setting A0-A3 to 0b1101 multiplexes the Shift Write Register to SROUT, so you will either a "0" or a "1" depending on which of the two channels was written last.

Your segmented capture does unfortunately not work. Due to a bug in the silicon, the first (e.g. even) written channel gets half overwritten when you start sampling the second (odd) channel. I should remove that from the documentation.

Furthermore, reading the chip while writing on the "other side" introduces quite some additional noise. The recommended way to do simultaneous reading and writing is therefore to use two separate DRS4 chips and split the input signals to both chips, then read from one chip while writing to the other chip. This keeps the crosstalk at a minimum and both chips run at full performance.

Stefan

Jonathan Bradshaw wrote:

Hi all

We're building a product which will use two different operating modes; firstly a long capcture using channel daisy chaining (2048 samples) and secondly a segmented capture (2 separate captures of 1024 samples each). 

For the long capture, I'm looking to capture 2048 samples for 4 channels.  Therefore I configure the Write Shift Register to 0b01010101 and the Write Config Register to 0b11111111.  During capture with DWRITE=1 the Write Shift Register will update.  Am I correct that once the capture is done and DWRITE=0, I can set A3..0 to 0b1101 and simply read the value of WSROUT to tell the difference?

For the segmented capture, I'm looking to capture 1024 samples for 4 channels on a first tirgger pulse, followed by 1024 samples for 4 channels on a second pulse.  Therefore I configure the Write Shift Register to 0b11111111 and the Write Config Register to 0b01010101 and set DWRITE=0 to capture.  After the first trigger I set DWRITE=0 and need to update the Write Config Register.  Do I need to write in a whole 8 bits to the Write Config Register (i.e. 0b10101010), or can I just shift in a single new bit (value 0b0)?

 

 

    Reply  Thu May 15 00:01:20 2025, Jonathan Bradshaw, Handling of Write Shift Register and Write Config Register 

All right, thank you for the clarification.

Stefan Ritt wrote:

Yes this is correct. Anyhow, even if it would be working, you would not be happy with it. After having designed ~10 boards with the DRS4 chip, I learned the hard way that any digital activity on the board during the sampling phase is strictly forbidden. You see crosstalk up to 100's of mV in some cases (with a preamplifier on the board, 10-20mV without preamp). So rule #1 is to keep the board as "quite" as possible when sampling the input. If you would readout the odd channels of the DRS4 during sampling of the even channels, you would probably get so much crosstalk that the samples are almost unusable. Even if you would do this with two DRS4 chips next to each other, you have to make sure to put proper grounding between the two chips, and operate them completely independent (like each one has it's onw contol lines going to the FPGA). Designing such boards is not so easy and takes lots of experience from the layouter.

Stefan

Jonathan Bradshaw wrote:

Hi Stefan

Just so I'm 100% clear; is there no reliable way to perform 2 segmented captures with a single DRS4 IC?

While not a showstopper, this is a bit disappointing.

 

Stefan Ritt wrote:

This is correct. Setting A0-A3 to 0b1101 multiplexes the Shift Write Register to SROUT, so you will either a "0" or a "1" depending on which of the two channels was written last.

Your segmented capture does unfortunately not work. Due to a bug in the silicon, the first (e.g. even) written channel gets half overwritten when you start sampling the second (odd) channel. I should remove that from the documentation.

Furthermore, reading the chip while writing on the "other side" introduces quite some additional noise. The recommended way to do simultaneous reading and writing is therefore to use two separate DRS4 chips and split the input signals to both chips, then read from one chip while writing to the other chip. This keeps the crosstalk at a minimum and both chips run at full performance.

Stefan

Jonathan Bradshaw wrote:

Hi all

We're building a product which will use two different operating modes; firstly a long capcture using channel daisy chaining (2048 samples) and secondly a segmented capture (2 separate captures of 1024 samples each). 

For the long capture, I'm looking to capture 2048 samples for 4 channels.  Therefore I configure the Write Shift Register to 0b01010101 and the Write Config Register to 0b11111111.  During capture with DWRITE=1 the Write Shift Register will update.  Am I correct that once the capture is done and DWRITE=0, I can set A3..0 to 0b1101 and simply read the value of WSROUT to tell the difference?

For the segmented capture, I'm looking to capture 1024 samples for 4 channels on a first tirgger pulse, followed by 1024 samples for 4 channels on a second pulse.  Therefore I configure the Write Shift Register to 0b11111111 and the Write Config Register to 0b01010101 and set DWRITE=0 to capture.  After the first trigger I set DWRITE=0 and need to update the Write Config Register.  Do I need to write in a whole 8 bits to the Write Config Register (i.e. 0b10101010), or can I just shift in a single new bit (value 0b0)?

 

 

 

 

Entry  Wed Oct 25 19:44:25 2023, John Westmoreland, WaveDREAM Design 

Hello All,

Are there any design resources available for the WaveDREAM PCBA's?

Thanks In Advance,
John W.

    Reply  Wed Oct 25 19:52:33 2023, John Westmoreland, WaveDREAM Design 

Stefan,

Oh, didn't realize that.

Thanks!
John

Stefan Ritt wrote:

No. This is a proprietary design.

Best,
Stefan

 

Entry  Wed Oct 23 17:56:26 2019, John Jendzurski, Computing corrected time from binary data...what is t_0,0? Screenshot.png

In the equations for computing the corrected time for channels other than channel 1, does anyone know what the term t0,0 refers to?  This is the last term in the last equation on page 24 of DRS4 Evaluation Board User’s Manual, Board Revision 5 as of January 2014, Last revised: April 27, 2016.

Screenshot from User's Manual is attached below.

Thank you!

Entry  Fri Oct 16 09:51:03 2009, Jinhong Wang, DSR4 Full Readout Mode 

Hello Mr. Stefan Ritt

          In DSR4 DATASHEET Rev.0.8 Page13, I noticed you metioned the samping should occur after 38 ns after the rising edge of SRCLK when the multiplexer is used. So what is suggested value(delay time between sampling and the rising edge of SRCLK) for the parallel mode,in which the multiplexer is not used?

          Best wishes!

                                                       Jinhong Wang

Entry  Mon Oct 19 09:06:43 2009, Jinhong Wang, BIAS Pin of DRS4 

Dear Mr. Stefan Ritt.

         Thank u for your timely response on "DSR4 Full Readout Mode", I received it from Professor Qi An, who is my PhD supervisor.

        I am currently going through the DRS4 datasheet. Well, can you give some specification on the usage of "BIAS" pin of DRS4? It is just metioned in the datasheet as bias of internal buffer. What is the internal buffer exactly reffered to here? The MUXOUT buffer of channel 8 or else? Does it have some relationship to O_OFS? I mean, if the reference voltage to BIAS is changed, how will the output be influenced?

       Looking forward to hearing from you soon.

                                                                       Jinhong Wang

                                                                    Fast Electronics LAB. of University of Science and Technology of China.

Entry  Mon Oct 19 11:26:29 2009, Jinhong Wang, output common mode voltage of DRS4 
Hello Mr. Stifan.Ritt
       In the DSR4 datasheet, it is mentioned that there is an additional buffer at each analog output, this buffer shifts the the differential range of -0.5V~0.5V to 0.8V~1.8V. Does it mean that this buffer shifts a voltage of about 1.3V for the primary differential range? 
       Again for the differential range of -0.5V~0.5V, can the common mode voltage of the analog output at OUT+/OUT-  be chaned? In the example presented in the datasheet, OUT+ is 0.8V~1.8V and OUT- is 1.8V~0.8V. So for an output swing of 2V p-p, can the common mode voltage be modified to the desired value? Supposed that the input ranges from -0.5V~0.5V.
                                                      
     Thank you!
                                             Jinhong Wang
Entry  Fri Oct 30 03:31:54 2009, Jinhong Wang, outline dimension of DRS4 QFN_package.jpg

QFN_package.jpg

                                                                                                            Fig.1        typical dimension of QFN package

Above is the typical dimension specification for QFN package. I cann't find the corresponding "T1" as in Fig.1 in the DRS4 documents, nor any of the tolerance of the dimensions, which are usually expressed in the form of  a range between a min. value and a max. value.

So will you specify the dimension of "T1" and "W1", and the dimension tolerance of them?

Thanks and best wishes!

                                                                               Jinhong Wang       University of Science and Technology of China

Entry  Mon Dec 14 10:14:16 2009, Jinhong Wang, Trigger of DRS4 

Dear Mr. S. Ritt

     The following is my confusion about the trigger of DRS4. It mainly concertrates on the generation of trigger signal to stop DRS4 sampling process for readout of sampled waveform.

     As metioned in the datasheet of DRS4, the chip samples the analog input every domino sampling period.  After finished sampling a waveform, the sampling process can be stoped by lowering the DWRITE while keeping DENABLE high. But the analog input is asychronous to the Domino CLK.  Then, how can we know when to  stop the domino sampling process to read out the sampled waveform? Of course, a trigger can be used. But from my present knowledge of DRS4, trigger can only be generated from analog input. Analog input is splited into two channels, one to DRS4 analog input, the other to FPGA as the trigger. However, splitting analog inputs increases the system design complexity, and may lower the total performace. So what is your suggestion?

    In our system, there are 8 analog inputs to a signal DRS4 chip, the outputs of DRS4 chip are connected to an 8-channel 14 bit ADC ( AD9252). It wold be kind of you to inform me about the most applicable approach for readout of DRS4 sampled wavefrom.

   Best regards.

   Sincerely,

   Jinhong Wang (wangjinh@mail.ustc.edu.cn)

    Reply  Mon Dec 21 10:17:05 2009, Jinhong Wang, Trigger of DRS4 

Stefan Ritt wrote:

Jinhong Wang wrote:

Dear Mr. S. Ritt

     The following is my confusion about the trigger of DRS4. It mainly concertrates on the generation of trigger signal to stop DRS4 sampling process for readout of sampled waveform.

     As metioned in the datasheet of DRS4, the chip samples the analog input every domino sampling period.  After finished sampling a waveform, the sampling process can be stoped by lowering the DWRITE while keeping DENABLE high. But the analog input is asychronous to the Domino CLK.  Then, how can we know when to  stop the domino sampling process to read out the sampled waveform? Of course, a trigger can be used. But from my present knowledge of DRS4, trigger can only be generated from analog input. Analog input is splited into two channels, one to DRS4 analog input, the other to FPGA as the trigger. However, splitting analog inputs increases the system design complexity, and may lower the total performace. So what is your suggestion?

    In our system, there are 8 analog inputs to a signal DRS4 chip, the outputs of DRS4 chip are connected to an 8-channel 14 bit ADC ( AD9252). It wold be kind of you to inform me about the most applicable approach for readout of DRS4 sampled wavefrom.

   Best regards.

   Sincerely,

   Jinhong Wang (wangjinh@mail.ustc.edu.cn)

Indeed you have to make an external trigger. The evaluation board uses the "transparent mode" of the DRS4 to "mirror" the input signal at the output, then puts a comparator there. The schematics of the evaluation board is in the manual. This does then not degrate the analog performance. You can of course also split the signal at the input, this will only add a minor additional load to the input signal, since the load of the DRS4 chips itself is much bigger than that of any comparator.

An alternative is to turn on the transparent mode and continuously digitize all 8 outputs with your AD9252. Then you make the trigger purely digital in your FPGA. You can put there a comparator, or even more complex logic like multiplicity etc. Note however that this causes some latency, since the ADC has a pipeline which is quite long, so you have to buffer the latency of your trigger in the analog window of the DRS4 sampling cells. Like if you run the DRS4 at 1 GSPS, you can accomodate 1024 ns of sampling depth, which is good for maybe 500 ns of trigger latency plus 500 ns of the waveform of interest.

Thank you. The transparent mode can be really helpful. Can you provide me in more details of the chip's transparent mode? I am still confused about the following aspects.

I notice that DRS4 samples the analog wave in the way "clear before write", and in the transparent mode, there will be certain delay before the trigger logic stops the sampling process. So,does it mean that the waveform recording process per Domino sampling cycle will not degrade the amplitude of the analog signal? Hence, for two idential analog inputs, one with a trigger latency of 500 ns and the other of 510 ns, the sampled waveform is identical, what differs is the starting number of the first active sampling cell, where the reading process considered to be started.   Is that right? Looking forward to your insight.

Best regrads.

Sincerely,

         Jinhong Wang (wangjinh@mail.ustc.edu.cn)

    Reply  Tue Dec 22 01:30:55 2009, Jinhong Wang, Trigger of DRS4 

Stefan Ritt wrote:

Jinhong Wang wrote:

Stefan Ritt wrote:

Jinhong Wang wrote:

Dear Mr. S. Ritt

     The following is my confusion about the trigger of DRS4. It mainly concertrates on the generation of trigger signal to stop DRS4 sampling process for readout of sampled waveform.

     As metioned in the datasheet of DRS4, the chip samples the analog input every domino sampling period.  After finished sampling a waveform, the sampling process can be stoped by lowering the DWRITE while keeping DENABLE high. But the analog input is asychronous to the Domino CLK.  Then, how can we know when to  stop the domino sampling process to read out the sampled waveform? Of course, a trigger can be used. But from my present knowledge of DRS4, trigger can only be generated from analog input. Analog input is splited into two channels, one to DRS4 analog input, the other to FPGA as the trigger. However, splitting analog inputs increases the system design complexity, and may lower the total performace. So what is your suggestion?

    In our system, there are 8 analog inputs to a signal DRS4 chip, the outputs of DRS4 chip are connected to an 8-channel 14 bit ADC ( AD9252). It wold be kind of you to inform me about the most applicable approach for readout of DRS4 sampled wavefrom.

   Best regards.

   Sincerely,

   Jinhong Wang (wangjinh@mail.ustc.edu.cn)

Indeed you have to make an external trigger. The evaluation board uses the "transparent mode" of the DRS4 to "mirror" the input signal at the output, then puts a comparator there. The schematics of the evaluation board is in the manual. This does then not degrate the analog performance. You can of course also split the signal at the input, this will only add a minor additional load to the input signal, since the load of the DRS4 chips itself is much bigger than that of any comparator.

An alternative is to turn on the transparent mode and continuously digitize all 8 outputs with your AD9252. Then you make the trigger purely digital in your FPGA. You can put there a comparator, or even more complex logic like multiplicity etc. Note however that this causes some latency, since the ADC has a pipeline which is quite long, so you have to buffer the latency of your trigger in the analog window of the DRS4 sampling cells. Like if you run the DRS4 at 1 GSPS, you can accomodate 1024 ns of sampling depth, which is good for maybe 500 ns of trigger latency plus 500 ns of the waveform of interest.

Thank you. The transparent mode can be really helpful. Can you provide me in more details of the chip's transparent mode? I am still confused about the following aspects.

I notice that DRS4 samples the analog wave in the way "clear before write", and in the transparent mode, there will be certain delay before the trigger logic stops the sampling process. So,does it mean that the waveform recording process per Domino sampling cycle will not degrade the amplitude of the analog signal? Hence, for two idential analog inputs, one with a trigger latency of 500 ns and the other of 510 ns, the sampled waveform is identical, what differs is the starting number of the first active sampling cell, where the reading process considered to be started.   Is that right? Looking forward to your insight.

Best regrads.

Sincerely,

         Jinhong Wang (wangjinh@mail.ustc.edu.cn)

The amplitude of the analog signal is not degraded by the transparent mode, since the signal is buffered on the chip, and the output of this buffer is send off the chip. The waveform digitizing of course requires quite some current to charge up all capacitors, so there is  maximum current of ~1mA for 5 GSPS. If you only have a weak signal source, your bandwidth might be limited by that. On the evaluation board for example we use passive transformers to produce the differential input signal from a single-ended signal. Although the transformers are rated 1 GHz Bandwidth, we only achieve 200 MHz with the passive transformers. By using active high speed differential drivers, you can get about 700 MHz right now.

If you have two channels with 500 ns and 510 ns trigger latency, there is no difference in the "domino stop position" since there is only one domino circuit per chip which can be stopped. So the stop position is the same for all eight channels on a chip. 

 So you mean there is an analog buffer per channel? The analog signal is buffered there, before entering the sampling cells? Then, when will the buffer content be released and cleared? How shall I handle "Dwite" and "Denable" during a complete operation when an analog signal arrives in the transparent mode? I cannot find more information beyond the datasheet,  a detailed description of the transparent mode (and the analog buffer, if possible) will be really helpful for me.

Best,

Sincerely,

Jinhong Wang (wangjinh@mail.ustc.edu.cn)

    Reply  Wed May 12 11:47:39 2010, Jinhong Wang, DRS4 chip model 

Stefan Ritt wrote:

Ignacio Diéguez Estremera wrote:

So i guess i won't be able to include drs4 in my simulations :-(. Any other suggestions? Maybe the S-params model you where working on? Anything is better than nothing :-)

Please find attached the S-parameters. 

 Hi, we plan to do a time interpolating among the eight channels on a single chip to obtain a maximum 40 GSPS (or, maybe 30 GSPS ) sampling rate.  Hence RF behavior of the anlog input is very important for us.

Will you give us some advice on the modeling of  the anlog input circuit of the chip?  Perhaps just the Spice model of the analog input?

The attached S parameters I found  here is for fs =1 GSPS, what about fs=5GSPS?

thanks in advance,

                                                                               Jinhong Wang (wangjinh@mail.ustc.edu.cn  ;  alleyor.wang@gmail.com)

    Reply  Fri Jun 18 11:31:20 2010, Jinhong Wang, DVDD Problem of DRS 4 

Stefan Ritt wrote:

Hao Huan wrote:

Stefan Ritt wrote:

Hao Huan wrote:

Stefan Ritt wrote:

Hao Huan wrote:

Hi Stefan,

    on our board some DRS chips draw a lot of current through DVDD after power-up and heat up significantly--it is true that our board doesn't have weak pull-down resistors at DENABLE and DWRITE output pins of FPGA, so this problem might have been caused by that, but a reinitialization of the Domino circuit doesn't help either. We tried different capacitors at DVDD and it seemed the larger the capacitance, the better the result--with a capacitor larger than 10nF some of the DRS chips could work happily in the normal way while if the capacitor is only 4.7nF all of them got very hot. Would you please provide some suggestions why there should be such a problem?

    Thanks a lot!

I found that sometimes even a reinitialization fails if the pull-down resistors are missing. So instead playing with capacitors at DVDD, I would just solder two resistors on the board which should fix the problem completely.

 Thanks! After adding pull-down resistors the voltages come back to normal.

However there is another weird problem that arises: a reset pulse seems unable to set the internal shift registers to default values. For example, after reset without addressing the Config Register the PLL will not try to lock with external reference clock. Even if I explicitly address the Config Register after reset and have the PLL locked, some channels of the chip will give null output during readout while other channels work normally. Could it be that some channels are not initiated properly with the Domino circuit?

Something is wrong. I have 800 chips, and they all start up fine. Check with your scope the RESET, DSPEED, DENABLE and DTAP signals. When RESET is applied, DSPEED should go to 2.5 V. When DENABLE goes high, the domino wave is started and you should see DTAP toggle. DSPEED is then lowered by the PLL until DTAP matches your external reference clock. I usually keep DENABLE high all the time after initialization, so the domino wave just continues running.

Another problem could however be the chip readout. If some channel gives null output, it could be that your readout has a problem. Do you use RSRLOAD to initialize the readout sequence?

 Yes; I used RSRLOAD to trigger readout of all channels in parallel so the asymmetry between channels of the same chip is really a big puzzle. Also during reset DSPEED indeed goes to 2.5V, but after reset the PLL will not try to lock with the external reference clock and lower DSPEED. Instead the Domino circuit just oscillates at the highest frequency by itself.

A more confusing discovery is that the SRIN level before starting the Domino wave could affect the behavior of the PLL. I mean the level of SRIN when the chip is at A="1111" or "1010". Is SRIN supposed to influence the chip even in these standby or transparent modes?

Just some ideas:

  • Is DENABLE really kept high all the time?
  • Is DRESET only applied once during initialization, after that it should stay high
  • Does  REFCLK+/REFCLK- really toggle at the required sampling speed / 2048?
  • Is the REFCLK really a good differential signal? Note that it must be biased properly since the DRS4 inputs are high impedance
  • Is the Bit1 in the Config Register really at "1" to enable the PLL?

The only way the SRIN level could affect the PLL is if you address the Config Register (A="1100") and you clock in a few bits with SRCLK.

Have you thought about 'crazy' things such as:

  • Defining the DRS4 chip wronly in your CAD software so that the pins are different from what you think?
  • Some soldering problem of the DRS4 chips (we had this in the past) so that some pins are not connected at all and others have shorts

I guess you checked most of the things, so I'm just wildly guessing in order to stimulate some thoughts.

The ultimate check would be to get one of the evaluation boards (I sent a few to Jean-Francois Genat some time ago...) and compare the DRS4 signals pin by pin.

 Hi Stefan

       I designed the evaluation board for our experiment. On our boards, I  encountered the similar problem when working on the PLL of DRS4. I compared the following two configuration process, which on with PLL locked, the other not,

   Process1: 

       step 1: Set DEnable and DWrite low, 

      Step2 : Reset DRS4 with a negative pulse of about 900 ns

      Step3: Set DEnable high,  thus do nothing but wait

       I found DRS4 PLL working and locked. 

  

  Process 2:

   Step 1: Set DEnable and DWrite low, 

  Step2 : Reset DRS4 with a negative pulse of about 900 ns 

  Step3: Set Config. Register( "11111111" .of course, this step was not necessary, just to see whether SPI was working properly from DTAP when set to "11111110")

  Step4: Set The read shift Register ( full read out mode)

  Step5: Set DEnable high, 

  Step6: Set DWrite high  , thus low it , and prepare to read the waveform.

  Well, I found in this case, the PLL was not locked, I am sure there was no problem with my SPI configuration process of DRS4. 

  toggle from DTAP could be viewed, but not stable. 

 Any Suggestions ?

  thanks.

    Reply  Sat Jun 19 10:09:18 2010, Jinhong Wang, DVDD Problem of DRS 4 

Stefan Ritt wrote:

Jinhong Wang wrote:

 

 Hi Stefan

       I designed the evaluation board for our experiment. On our boards, I  encountered the similar problem when working on the PLL of DRS4. I compared the following two configuration process, which on with PLL locked, the other not,

   Process1: 

       step 1: Set DEnable and DWrite low, 

      Step2 : Reset DRS4 with a negative pulse of about 900 ns

      Step3: Set DEnable high,  thus do nothing but wait

       I found DRS4 PLL working and locked. 

  

  Process 2:

   Step 1: Set DEnable and DWrite low, 

  Step2 : Reset DRS4 with a negative pulse of about 900 ns 

  Step3: Set Config. Register( "11111111" .of course, this step was not necessary, just to see whether SPI was working properly from DTAP when set to "11111110")

  Step4: Set The read shift Register ( full read out mode)

  Step5: Set DEnable high, 

  Step6: Set DWrite high  , thus low it , and prepare to read the waveform.

  Well, I found in this case, the PLL was not locked, I am sure there was no problem with my SPI configuration process of DRS4. 

  toggle from DTAP could be viewed, but not stable. 

 Any Suggestions ?

  thanks.

So the main difference, if I understand correctly, is the setting of the Config Register. Actually I never tried that, I always went with the default settings (all "1"). What happens if you write "00000000"? You know Bit1 controls the PLL, maybe there is a bug and the signal needs to be inverted. 

 Hi, Stefan,

       The problem was fixed by setting Reg_addr "1001" instead of "1111" when in idle state, I was confused. 