Entry  Thu Mar 22 14:36:01 2018, Phan Van Chuan, Read the CalibrateWaveform 

Helo
I'm building an application for reading waveforms from the DRS4 board to PC. However, I am having problems reading calibration data from EEPROM on DRS4 board. The calibration data is read through the function reference:
void DRSBoard :: ReadCalibration (void)
...
      ReadEEPROM (1, buf, 1024 * 32);
      for (i = 0; i <8; i ++)
         for (j = 0; j <1024; j ++) {
            fCellOffset [i] [j] = buf [(i * 1024 + j) * 2];
            fCellGain [i] [j] = buf [(i * 1024 + j) * 2 + 1] / 65535.0*0.4+0.7;
         }
      
      ReadEEPROM (2, buf, 1024 * 32);
      for (i = 0; i <8; i ++)
         for (j = 0; j <1024; j ++)
            fCellOffset2 [i] [j] = buf [(i * 1024 + j) * 2];
...
The Calibrate Waveform is performed by:
int DRSBoard::CalibrateWaveform(unsigned int chipIndex, unsigned char channel, unsigned short *adcWaveform, short *waveform, bool responseCalib, int triggerCell, bool adjustToClock, float threshold, bool offsetCalib)
.....
         for (j = 0; j < n_bins; j++) {
            value = adcWaveform[j] - fCellOffset[channel+chipIndex*9][(j*skip + triggerCell) % kNumberOfBins];
            value = value / fCellGain[channel+chipIndex*9][(j*skip + triggerCell) % kNumberOfBins];
            if (offsetCalib && channel != 8)
               value = value - fCellOffset2[channel+chipIndex*9][j*skip] + 32768;
...
. Because the calibration data reads incorrectly, the Calibrate Waveform does not do it.
Can read calibration data from EEPROM by any command via Oscilloscope application or DRS Command Line Interface application?
Thank you for your help!!!!

    Reply  Fri Mar 23 09:39:55 2018, Stefan Ritt, Read the CalibrateWaveform 

You don't have to read and calibrate the waveforms in your user code, but can rely on the DRS.cpp library to do that. Just look at the drs_exam.cpp program coming with the distribution. It uses the function b->GetWave() to retrieve the calibrated waveform. If you like, you can look into that function to learn how to apply the calibration, but I can tell you that it's a bit complicated. Since each event starts at an arbitrary stop cell in the DRS4, you have to "rotate" the calibration array. Then you do actually four calibrations in a row (cell, readout, gain and range).

Stefan

Phan Van Chuan wrote:

Helo
I'm building an application for reading waveforms from the DRS4 board to PC. However, I am having problems reading calibration data from EEPROM on DRS4 board. The calibration data is read through the function reference:
void DRSBoard :: ReadCalibration (void)
...
      ReadEEPROM (1, buf, 1024 * 32);
      for (i = 0; i <8; i ++)
         for (j = 0; j <1024; j ++) {
            fCellOffset [i] [j] = buf [(i * 1024 + j) * 2];
            fCellGain [i] [j] = buf [(i * 1024 + j) * 2 + 1] / 65535.0*0.4+0.7;
         }
      
      ReadEEPROM (2, buf, 1024 * 32);
      for (i = 0; i <8; i ++)
         for (j = 0; j <1024; j ++)
            fCellOffset2 [i] [j] = buf [(i * 1024 + j) * 2];
...
The Calibrate Waveform is performed by:
int DRSBoard::CalibrateWaveform(unsigned int chipIndex, unsigned char channel, unsigned short *adcWaveform, short *waveform, bool responseCalib, int triggerCell, bool adjustToClock, float threshold, bool offsetCalib)
.....
         for (j = 0; j < n_bins; j++) {
            value = adcWaveform[j] - fCellOffset[channel+chipIndex*9][(j*skip + triggerCell) % kNumberOfBins];
            value = value / fCellGain[channel+chipIndex*9][(j*skip + triggerCell) % kNumberOfBins];
            if (offsetCalib && channel != 8)
               value = value - fCellOffset2[channel+chipIndex*9][j*skip] + 32768;
...
. Because the calibration data reads incorrectly, the Calibrate Waveform does not do it.
Can read calibration data from EEPROM by any command via Oscilloscope application or DRS Command Line Interface application?
Thank you for your help!!!!

 

Entry  Fri Mar 2 18:08:55 2018, Steven Block, ROI  

Hello,

I have a question about how ROI works. From what I have read, it will only save data that ocurs some time [ta] dictated by the user after an event is triggered as well as a small time [tb] before the event. The technical manual seems to indicated that the deadtime assciated with operating in ROI mode can be reduced by the following factor: 

\frac{t_a + t_b }{\frac{N}{Sample Speed}} .

Where N is the number of points in the time window (ex. 2048 or 1024). Is it ok to describe this as:

\frac{N'}{N} 

Where N' is the number of samples in the ROI and N is the same as before.

For example, if I were running at 5Gsps (200ps between samples), only recording 1024 samples per event and I had an signal that lasted 2ns, that means the signal would last 10 samples. If I set the ROI to only save 20 samples around this signal, would my Deadtime go to:

\frac{10}{1024} * 30ns*1024 + 2\mu s = 2.3\mu s? (The second portion of this equation comes from a response I recieved earlier, but I just want to make sure I understand this concept properly)

I recognize that the caveat is that this would work only if the signal was detected during acquistion, which leads to my next question. If no signals were detected in the 1024*200ps time frame in ROI mode, would the DRS4 go dead for 32us (using the factor = 1 from above equation), or would it dump the earliest events in the buffer for the more recent ones until it detects a signal? 

Finally, I assume this functionality can only be utilized with custom electornics with the DRS4, not the evaulation/demo board, please let me know if this is the case. 

Best,

Steven

    Reply  Fri Mar 2 20:17:17 2018, Stefan Ritt, ROI  

N'/N is correct. The 2 us "from the response you got from me" come from the fact that after readout, you have to start the DRS4 again. During this time, the power supply usually becomes slightly unstable, and it takes on the evaluation board about 2us to stabilize it again. Tha't why I add the 2 us. If you don't care about slight offset effect, or if you make a better power supply, you dead time would be 10*30ns = 300ns for 10 samples. Starting the DRS again will take one or two clock cycles from the FPGA, which might add another 30 ns or so, depending on how you program the FPGA. So the best you can achieve for 10 samples is maybe 330 ns, if you have a really good power supply (large capacitors).

You can achieve this functionality with the evaluation board, but you would have to make a special firmware for it.

Stefan

Steven Block wrote:

Hello,

I have a question about how ROI works. From what I have read, it will only save data that ocurs some time [ta] dictated by the user after an event is triggered as well as a small time [tb] before the event. The technical manual seems to indicated that the deadtime assciated with operating in ROI mode can be reduced by the following factor: 

\frac{t_a + t_b }{\frac{N}{Sample Speed}} .

Where N is the number of points in the time window (ex. 2048 or 1024). Is it ok to describe this as:

\frac{N'}{N}

Where N' is the number of samples in the ROI and N is the same as before.

For example, if I were running at 5Gsps (200ps between samples), only recording 1024 samples per event and I had an signal that lasted 2ns, that means the signal would last 10 samples. If I set the ROI to only save 20 samples around this signal, would my Deadtime go to:

\frac{10}{1024} * 30ns*1024 + 2\mu s = 2.3\mu s? (The second portion of this equation comes from a response I recieved earlier, but I just want to make sure I understand this concept properly)

I recognize that the caveat is that this would work only if the signal was detected during acquistion, which leads to my next question. If no signals were detected in the 1024*200ps time frame in ROI mode, would the DRS4 go dead for 32us (using the factor = 1 from above equation), or would it dump the earliest events in the buffer for the more recent ones until it detects a signal? 

Finally, I assume this functionality can only be utilized with custom electornics with the DRS4, not the evaulation/demo board, please let me know if this is the case. 

Best,

Steven

 

       Reply  Fri Mar 2 21:05:48 2018, Steven Block, ROI  

Great! That is very helpful. 

One more question. If no signals were detected in the 1024*200ps time frame in ROI mode, would the DRS4 go dead for 32us (or 30us depending on the supply)  for, or would it dump the earliest events in the buffer for the more recent ones until it detects a signal to readout? Or rather, does filling the buffer force a readout or can it dynamically shift out old data until it detects a signal to readout. 

Steven

Stefan Ritt wrote:

N'/N is correct. The 2 us "from the response you got from me" come from the fact that after readout, you have to start the DRS4 again. During this time, the power supply usually becomes slightly unstable, and it takes on the evaluation board about 2us to stabilize it again. Tha't why I add the 2 us. If you don't care about slight offset effect, or if you make a better power supply, you dead time would be 10*30ns = 300ns for 10 samples. Starting the DRS again will take one or two clock cycles from the FPGA, which might add another 30 ns or so, depending on how you program the FPGA. So the best you can achieve for 10 samples is maybe 330 ns, if you have a really good power supply (large capacitors).

You can achieve this functionality with the evaluation board, but you would have to make a special firmware for it.

Stefan

Steven Block wrote:

Hello,

I have a question about how ROI works. From what I have read, it will only save data that ocurs some time [ta] dictated by the user after an event is triggered as well as a small time [tb] before the event. The technical manual seems to indicated that the deadtime assciated with operating in ROI mode can be reduced by the following factor: 

\frac{t_a + t_b }{\frac{N}{Sample Speed}} .

Where N is the number of points in the time window (ex. 2048 or 1024). Is it ok to describe this as:

\frac{N'}{N}

Where N' is the number of samples in the ROI and N is the same as before.

For example, if I were running at 5Gsps (200ps between samples), only recording 1024 samples per event and I had an signal that lasted 2ns, that means the signal would last 10 samples. If I set the ROI to only save 20 samples around this signal, would my Deadtime go to:

\frac{10}{1024} * 30ns*1024 + 2\mu s = 2.3\mu s? (The second portion of this equation comes from a response I recieved earlier, but I just want to make sure I understand this concept properly)

I recognize that the caveat is that this would work only if the signal was detected during acquistion, which leads to my next question. If no signals were detected in the 1024*200ps time frame in ROI mode, would the DRS4 go dead for 32us (using the factor = 1 from above equation), or would it dump the earliest events in the buffer for the more recent ones until it detects a signal? 

Finally, I assume this functionality can only be utilized with custom electornics with the DRS4, not the evaulation/demo board, please let me know if this is the case. 

Best,

Steven

 

 

          Reply  Mon Mar 19 16:22:42 2018, Stefan Ritt, ROI  

The DRS4 has an internal storage of 1024 capacitors. They work as a ring buffer, so at 5GSPS you can store 200ns wide signals. After 200ns, the first samples are overwritten by new samples, so you always have the last 200ns of samples stored. Once you trigger the DRS4, this buffer is frozen, and the readout of this buffer causes the dead time. No trigger, no dead time. Hope this answers your question.

Stefan

Steven Block wrote:

Great! That is very helpful. 

One more question. If no signals were detected in the 1024*200ps time frame in ROI mode, would the DRS4 go dead for 32us (or 30us depending on the supply)  for, or would it dump the earliest events in the buffer for the more recent ones until it detects a signal to readout? Or rather, does filling the buffer force a readout or can it dynamically shift out old data until it detects a signal to readout. 

Steven

Stefan Ritt wrote:

N'/N is correct. The 2 us "from the response you got from me" come from the fact that after readout, you have to start the DRS4 again. During this time, the power supply usually becomes slightly unstable, and it takes on the evaluation board about 2us to stabilize it again. Tha't why I add the 2 us. If you don't care about slight offset effect, or if you make a better power supply, you dead time would be 10*30ns = 300ns for 10 samples. Starting the DRS again will take one or two clock cycles from the FPGA, which might add another 30 ns or so, depending on how you program the FPGA. So the best you can achieve for 10 samples is maybe 330 ns, if you have a really good power supply (large capacitors).

You can achieve this functionality with the evaluation board, but you would have to make a special firmware for it.

Stefan

Steven Block wrote:

Hello,

I have a question about how ROI works. From what I have read, it will only save data that ocurs some time [ta] dictated by the user after an event is triggered as well as a small time [tb] before the event. The technical manual seems to indicated that the deadtime assciated with operating in ROI mode can be reduced by the following factor: 

\frac{t_a + t_b }{\frac{N}{Sample Speed}} .

Where N is the number of points in the time window (ex. 2048 or 1024). Is it ok to describe this as:

\frac{N'}{N}

Where N' is the number of samples in the ROI and N is the same as before.

For example, if I were running at 5Gsps (200ps between samples), only recording 1024 samples per event and I had an signal that lasted 2ns, that means the signal would last 10 samples. If I set the ROI to only save 20 samples around this signal, would my Deadtime go to:

\frac{10}{1024} * 30ns*1024 + 2\mu s = 2.3\mu s? (The second portion of this equation comes from a response I recieved earlier, but I just want to make sure I understand this concept properly)

I recognize that the caveat is that this would work only if the signal was detected during acquistion, which leads to my next question. If no signals were detected in the 1024*200ps time frame in ROI mode, would the DRS4 go dead for 32us (using the factor = 1 from above equation), or would it dump the earliest events in the buffer for the more recent ones until it detects a signal? 

Finally, I assume this functionality can only be utilized with custom electornics with the DRS4, not the evaulation/demo board, please let me know if this is the case. 

Best,

Steven

 

 

 

Entry  Wed Mar 14 00:38:15 2018, Will Flanagan, sub-ms precision timestamps? 

Dear DRS4 community,

Is there a way to extract timestamps with sub-ms precision? The milliseconds of an event is clearly given when unpacking the header. I would like to determine how far apart events are when they are within the same millisecond.

Thanks,

Will

    Reply  Thu Mar 15 08:44:26 2018, Stefan Ritt, sub-ms precision timestamps? 

Putting sub-ms precision into the header does not make sense, since the USB transfer only happens in time-slots of about 2 ms. To get better timing, you would need a hardware time clock in the FPGA, which does not exist right now.

Best,
Stefan

Will Flanagan wrote:

Dear DRS4 community,

Is there a way to extract timestamps with sub-ms precision? The milliseconds of an event is clearly given when unpacking the header. I would like to determine how far apart events are when they are within the same millisecond.

Thanks,

Will

 

Entry  Tue Feb 27 16:34:26 2018, Steven Block, DRS4 Dead times 6x

Hello All,

I am currently trying to figure out how to properly characterize the dead time of the DRS4 board. My most recent experiment to try and answer this question involved using an external trigger that can range from 1Hz to 2MHz. I fed this trigger into the DRS4 and collected 1000 samples with no input to any channels. I repeated this across the range of my external trigger by a factor of ten [10Hz, 100Hz, 1kHz...etc]. After I had saved these runs in XML format, I looked at the difference between timestamps on the events. Attached are my findings. Can someone offer an explanation for the periodic peaks? I am new to the DRS4 and don't really understand how it works. My guess is that there is a buffer that has to be emptied every so often, but if so, the buffer emptying time varies with the frequency of the trigger. I would ideally like to be able to know the relation of the dead time to a particular setting I change on the DRS4 such as locking the sampling speed or changing external trigger frequency. 

Best,

Steven

    Reply  Tue Feb 27 17:04:12 2018, Stefan Ritt, DRS4 Dead times 

XML is very slow to write, and you are probably limited by that. Switch to binary mode, which is much faster. You will see in the end a maximum rate of ~500 Hz, and thus a dead time of 2ms, independent of the sampling speed. Note that you have only an evaluation board, which is optimized for ease of use. If you develop your own electronics, and do optimized readout, you can bring the deadtime down to 30ns x number of samples + 2us, or 32us if you read 1024 values from one channel.

Stefan

Steven Block wrote:

Hello All,

I am currently trying to figure out how to properly characterize the dead time of the DRS4 board. My most recent experiment to try and answer this question involved using an external trigger that can range from 1Hz to 2MHz. I fed this trigger into the DRS4 and collected 1000 samples with no input to any channels. I repeated this across the range of my external trigger by a factor of ten [10Hz, 100Hz, 1kHz...etc]. After I had saved these runs in XML format, I looked at the difference between timestamps on the events. Attached are my findings. Can someone offer an explanation for the periodic peaks? I am new to the DRS4 and don't really understand how it works. My guess is that there is a buffer that has to be emptied every so often, but if so, the buffer emptying time varies with the frequency of the trigger. I would ideally like to be able to know the relation of the dead time to a particular setting I change on the DRS4 such as locking the sampling speed or changing external trigger frequency. 

Best,

Steven

 

       Reply  Tue Feb 27 18:04:18 2018, Steven Block, DRS4 Dead times 

That is extremely helpful! Many thanks. One more question; If I were to take inputs from 2 channels at once, would that scale the dead time to 64us using your example? 

Steven

Stefan Ritt wrote:

XML is very slow to write, and you are probably limited by that. Switch to binary mode, which is much faster. You will see in the end a maximum rate of ~500 Hz, and thus a dead time of 2ms, independent of the sampling speed. Note that you have only an evaluation board, which is optimized for ease of use. If you develop your own electronics, and do optimized readout, you can bring the deadtime down to 30ns x number of samples + 2us, or 32us if you read 1024 values from one channel.

Stefan

Steven Block wrote:

Hello All,

I am currently trying to figure out how to properly characterize the dead time of the DRS4 board. My most recent experiment to try and answer this question involved using an external trigger that can range from 1Hz to 2MHz. I fed this trigger into the DRS4 and collected 1000 samples with no input to any channels. I repeated this across the range of my external trigger by a factor of ten [10Hz, 100Hz, 1kHz...etc]. After I had saved these runs in XML format, I looked at the difference between timestamps on the events. Attached are my findings. Can someone offer an explanation for the periodic peaks? I am new to the DRS4 and don't really understand how it works. My guess is that there is a buffer that has to be emptied every so often, but if so, the buffer emptying time varies with the frequency of the trigger. I would ideally like to be able to know the relation of the dead time to a particular setting I change on the DRS4 such as locking the sampling speed or changing external trigger frequency. 

Best,

Steven

 

 

          Reply  Tue Feb 27 18:12:32 2018, Stefan Ritt, DRS4 Dead times 

For applications which are critical on the dead time, one typically uses one ADC per DRS4 channel, and thus the dead time stays at 32us. If you multiplex two DRS4 channels into one ADC channel, then it goes to 32us.

Stefan

Steven Block wrote:

That is extremely helpful! Many thanks. One more question; If I were to take inputs from 2 channels at once, would that scale the dead time to 64us using your example? 

Steven

Stefan Ritt wrote:

XML is very slow to write, and you are probably limited by that. Switch to binary mode, which is much faster. You will see in the end a maximum rate of ~500 Hz, and thus a dead time of 2ms, independent of the sampling speed. Note that you have only an evaluation board, which is optimized for ease of use. If you develop your own electronics, and do optimized readout, you can bring the deadtime down to 30ns x number of samples + 2us, or 32us if you read 1024 values from one channel.

Stefan

Steven Block wrote:

Hello All,

I am currently trying to figure out how to properly characterize the dead time of the DRS4 board. My most recent experiment to try and answer this question involved using an external trigger that can range from 1Hz to 2MHz. I fed this trigger into the DRS4 and collected 1000 samples with no input to any channels. I repeated this across the range of my external trigger by a factor of ten [10Hz, 100Hz, 1kHz...etc]. After I had saved these runs in XML format, I looked at the difference between timestamps on the events. Attached are my findings. Can someone offer an explanation for the periodic peaks? I am new to the DRS4 and don't really understand how it works. My guess is that there is a buffer that has to be emptied every so often, but if so, the buffer emptying time varies with the frequency of the trigger. I would ideally like to be able to know the relation of the dead time to a particular setting I change on the DRS4 such as locking the sampling speed or changing external trigger frequency. 

Best,

Steven

 

 

 

Entry  Thu Jan 25 05:24:05 2018, chen wenjun, problem with the drscl(drs507) 

Hi! Stefan:

  when I change a new computer(win7,64bit),I meet a problem that the drscl app cannot found the board! It shows"USB successfully scanned,but no boards found",but the drsosc runs well . when I connect to other win7*64bits computer,only one of them runs property! Is there any driver else I need to install? 

Thank you!

Chen

    Reply  Thu Jan 25 08:00:16 2018, Stefan Ritt, problem with the drscl(drs507) 

This problem has been reported by several people, like elog:551

So far I could not solve it. On the computers at our lab it works find so I cannot reproduce and fix the problem. One suspicion I have is that the underlying libusb library needs to be updated. You can try to install the newest version from their website at http://libusb.info/, but I haven't tried it myself.

Stefan

 

chen wenjun wrote:

Hi! Stefan:

  when I change a new computer(win7,64bit),I meet a problem that the drscl app cannot found the board! It shows"USB successfully scanned,but no boards found",but the drsosc runs well . when I connect to other win7*64bits computer,only one of them runs property! Is there any driver else I need to install? 

Thank you!

Chen

 

       Reply  Thu Jan 25 08:07:32 2018, chen wenjun, problem with the drscl(drs507) 

I have tried about 4 computers,only one worked fine.I truly want to know how others get this fixed,can you get in touch with them?

Stefan Ritt wrote:

This problem has been reported by several people, like elog:551

So far I could not solve it. On the computers at our lab it works find so I cannot reproduce and fix the problem. One suspicion I have is that the underlying libusb library needs to be updated. You can try to install the newest version from their website at http://libusb.info/, but I haven't tried it myself.

Stefan

 

chen wenjun wrote:

Hi! Stefan:

  when I change a new computer(win7,64bit),I meet a problem that the drscl app cannot found the board! It shows"USB successfully scanned,but no boards found",but the drsosc runs well . when I connect to other win7*64bits computer,only one of them runs property! Is there any driver else I need to install? 

Thank you!

Chen

 

 

Entry  Tue Mar 28 21:53:12 2017, Jim Freeman, drscl doesn't find eval board but drsosc does (Windows 7) 

I cannot find the EVAL board using drscl version 5.06 while the drsosc works fine. I tried 2 different eval boards and 2 different computers and the same effect. I looked under device manager at the libusb and the drs4 was there, and checked the driver which was found to be up to date.

    Reply  Wed Apr 5 12:28:28 2017, Stefan Ritt, drscl doesn't find eval board but drsosc does (Windows 7) Screen_Shot_2017-04-05_at_12.27.46_.pngScreen_Shot_2017-04-05_at_11.45.07_.png

Two people report now this problem, while this works fine at our lab. So I'm puzzled right now.

I attach two screenshots from the device manager and the Command Line interface. Can you compare it with what you see? Which is the firmware version of your evaluaiton board?

Stefan

Jim Freeman wrote:

I cannot find the EVAL board using drscl version 5.06 while the drsosc works fine. I tried 2 different eval boards and 2 different computers and the same effect. I looked under device manager at the libusb and the drs4 was there, and checked the driver which was found to be up to date.

 

    Reply  Thu Jan 25 06:10:52 2018, chen wenjun, drscl doesn't find eval board but drsosc does (Windows 7) 

Hi! Jim:

  It seems that I meet the same question with you ,and I am confused ,have you find out the reason about this problem?Or can you tell me how you deal with it?

Thank you very much!

chen

Jim Freeman wrote:

I cannot find the EVAL board using drscl version 5.06 while the drsosc works fine. I tried 2 different eval boards and 2 different computers and the same effect. I looked under device manager at the libusb and the drs4 was there, and checked the driver which was found to be up to date.

 

Entry  Wed Jan 17 09:51:16 2018, Tran Cong Thien, The input signals recorded are different with the signal showed in oscilloscope  

Dear Stefan,

I am using an DRS4 board to record the signals from an plastic scintillator detector. It was working really good, yet a few day ago the signals became "not right". When I checked the signal using an oscilloscope it show the normal signals previously recorded. The signal amplitude are clearly reduced (from 0.3 in oscilloscope to lower than 0.1 in DRS4). Can you show me how to show this problem?

Thank you very much!

Best Regards,

Thien 

    Reply  Wed Jan 17 10:09:09 2018, Stefan Ritt, The input signals recorded are different with the signal showed in oscilloscope  

First thing is to do another voltage calibration. Disconnect input, "Config", "Execute Voltage Calibration". If this does not fix the problem, the board is probably broken. This can happen if you send very high input singals to the board (like >10V) and exceed the maximul allowed limit from the datasheet. In that case the board needs to be repaired. Please contact me directly (via email) so that we can make you a quote.

Best regards,
Stefan

Tran Cong Thien wrote:

Dear Stefan,

I am using an DRS4 board to record the signals from an plastic scintillator detector. It was working really good, yet a few day ago the signals became "not right". When I checked the signal using an oscilloscope it show the normal signals previously recorded. The signal amplitude are clearly reduced (from 0.3 in oscilloscope to lower than 0.1 in DRS4). Can you show me how to show this problem?

Thank you very much!

Best Regards,

Thien 

 

Entry  Tue Mar 26 01:17:59 2013, Jill Russek, cascading -- DRS4 Osci.cpp & DRS.cpp 

 

All I'm trying to do is cascade one input signal, though all available channels, so that I end up with 8*1024 bins per event.

Here is the read out on my board/chip:

Mezz. Board index:    0
DRS type:             DRS4
Board type:           8
Serial number:        2249
Firmware revision:    17662
Temperature:          35.2 C
Input range:          -0.5V...0.5V
Calibrated range:     -0.5V...0.5V
Calibrated frequency: 5.120 GHz
Status reg.:          0000001A
Control reg.:         00000010
  DMODE circular
Trigger bus:          00000000
Frequency:            5.120 GHz

 

What I've tried thus far:

In Osci.cpp, in the method/function  SelectSource(int board, int firstChannel, int chnSection), I added a line.. (in bold)

//----------------------------------------------------------------------------------------------------------------------------------------------

 if (b->GetBoardType() == 5 || b->GetBoardType() == 7 || b->GetBoardType() == 8) {
         
        
             if (chnSection == 2)
                b->SetChannelConfig(0, 8, 4);
             //added
             else if(chnSection == 1)
                 b->SetChannelConfig(0, 8, 2);
             //added

             else
                b->SetChannelConfig(0, 8, 8);

//----------------------------------------------------------------------------------------------------------------------------------------------

I've also tried doing settings such as SetChannelConfig(0, 8, 1);   , SetChannelConfig(0, 8, 2); , SetChannelConfig(0, 1, 2); , etc..

Which always ends up making the run fail.. and sometimes I get index errors..

As far as I understanding the program now, this is what I know:

fChannelCascading determines getChannelCascading,
this determines the  if (casc == 2) line in configDialogue.cpp, which sets:
 b->SetChannelConfig(config, 8, 4);

fChannelCascading is being set by:
 switch (nConfigChannels) {
      case 1:
         fChannelConfig = 0x01;
         fChannelCascading = 8;
         break;
      case 2:
         fChannelConfig = 0x11;
         fChannelCascading = 4;
         break;
      case 4:
         fChannelConfig = 0x55;
         fChannelCascading = 2;
         break;
      case 8:
         fChannelConfig = 0xFF;
         fChannelCascading = 1;
         break;
      default:
         printf("Invalid channel configuration\n");
         return 0;
      }

which is being set by nConfigChannels in DRS.cpp, in the method:
SetChannelConfig(int firstChannel, int lastChannel, int nConfigChannels)

SetChannelConfig is being called in the ConfigDialogue.cpp,  but the default Osci program is such that you can't do a configuration for a cascade of one signal using all the channels. At least, not that I am aware of. 

So what buttons do I need to enable, or what do I need to call, or write, so that I can cascade a signal to end up with 8*1024 bins per event?

This has had me going in circles for weeks, so thank you for your help!!!!

 

 

    Reply  Thu Apr 4 11:32:21 2013, Stefan Ritt, cascading -- DRS4 Osci.cpp & DRS.cpp 

Jill Russek wrote:

 

All I'm trying to do is cascade one input signal, though all available channels, so that I end up with 8*1024 bins per event.

Here is the read out on my board/chip:

Mezz. Board index:    0
DRS type:             DRS4
Board type:           8
Serial number:        2249
Firmware revision:    17662
Temperature:          35.2 C
Input range:          -0.5V...0.5V
Calibrated range:     -0.5V...0.5V
Calibrated frequency: 5.120 GHz
Status reg.:          0000001A
Control reg.:         00000010
  DMODE circular
Trigger bus:          00000000
Frequency:            5.120 GHz

 

What I've tried thus far:

In Osci.cpp, in the method/function  SelectSource(int board, int firstChannel, int chnSection), I added a line.. (in bold)

//----------------------------------------------------------------------------------------------------------------------------------------------

 if (b->GetBoardType() == 5 || b->GetBoardType() == 7 || b->GetBoardType() == 8) {
         
        
             if (chnSection == 2)
                b->SetChannelConfig(0, 8, 4);
             //added
             else if(chnSection == 1)
                 b->SetChannelConfig(0, 8, 2);
             //added

             else
                b->SetChannelConfig(0, 8, 8);

//----------------------------------------------------------------------------------------------------------------------------------------------

I've also tried doing settings such as SetChannelConfig(0, 8, 1);   , SetChannelConfig(0, 8, 2); , SetChannelConfig(0, 1, 2); , etc..

Which always ends up making the run fail.. and sometimes I get index errors..

As far as I understanding the program now, this is what I know:

fChannelCascading determines getChannelCascading,
this determines the  if (casc == 2) line in configDialogue.cpp, which sets:
 b->SetChannelConfig(config, 8, 4);

fChannelCascading is being set by:
 switch (nConfigChannels) {
      case 1:
         fChannelConfig = 0x01;
         fChannelCascading = 8;
         break;
      case 2:
         fChannelConfig = 0x11;
         fChannelCascading = 4;
         break;
      case 4:
         fChannelConfig = 0x55;
         fChannelCascading = 2;
         break;
      case 8:
         fChannelConfig = 0xFF;
         fChannelCascading = 1;
         break;
      default:
         printf("Invalid channel configuration\n");
         return 0;
      }

which is being set by nConfigChannels in DRS.cpp, in the method:
SetChannelConfig(int firstChannel, int lastChannel, int nConfigChannels)

SetChannelConfig is being called in the ConfigDialogue.cpp,  but the default Osci program is such that you can't do a configuration for a cascade of one signal using all the channels. At least, not that I am aware of. 

So what buttons do I need to enable, or what do I need to call, or write, so that I can cascade a signal to end up with 8*1024 bins per event?

This has had me going in circles for weeks, so thank you for your help!!!! 

Sorry for my late reply, I was away for some days.

To use channel cascading, you have to physically connect one input to all eight channels. This is not possible with the evaluation board, you have to make your own board. What you could do however is to split a signal externally and feed it to all four inputs, given that the signal delay is the same for every channel. But then you will hit the hard-wired limit in Osci.cpp. This code was never foreseen to cover 8*1024 bins (since it does not make much sense with the evaluation board). Some arrays are only 2*1024 bins wide, so you would have to rewrite code at many places.

The easiest way to get what you want is to modify drs_exam.cpp. You need SetChannelConfig(0, 8, 1) as you realised correctly, and then you have to retrieve all 8 channels via b->GetWave() and concatenate them correctly.

/Stefan 

       Reply  Fri Apr 5 02:21:33 2013, Jill Russek, cascading -- DRS4 Osci.cpp & DRS.cpp 

Stefan Ritt wrote:

Jill Russek wrote:

 

All I'm trying to do is cascade one input signal, though all available channels, so that I end up with 8*1024 bins per event.

Here is the read out on my board/chip:

Mezz. Board index:    0
DRS type:             DRS4
Board type:           8
Serial number:        2249
Firmware revision:    17662
Temperature:          35.2 C
Input range:          -0.5V...0.5V
Calibrated range:     -0.5V...0.5V
Calibrated frequency: 5.120 GHz
Status reg.:          0000001A
Control reg.:         00000010
  DMODE circular
Trigger bus:          00000000
Frequency:            5.120 GHz

 

What I've tried thus far:

In Osci.cpp, in the method/function  SelectSource(int board, int firstChannel, int chnSection), I added a line.. (in bold)

//----------------------------------------------------------------------------------------------------------------------------------------------

 if (b->GetBoardType() == 5 || b->GetBoardType() == 7 || b->GetBoardType() == 8) {
         
        
             if (chnSection == 2)
                b->SetChannelConfig(0, 8, 4);
             //added
             else if(chnSection == 1)
                 b->SetChannelConfig(0, 8, 2);
             //added

             else
                b->SetChannelConfig(0, 8, 8);

//----------------------------------------------------------------------------------------------------------------------------------------------

I've also tried doing settings such as SetChannelConfig(0, 8, 1);   , SetChannelConfig(0, 8, 2); , SetChannelConfig(0, 1, 2); , etc..

Which always ends up making the run fail.. and sometimes I get index errors..

As far as I understanding the program now, this is what I know:

fChannelCascading determines getChannelCascading,
this determines the  if (casc == 2) line in configDialogue.cpp, which sets:
 b->SetChannelConfig(config, 8, 4);

fChannelCascading is being set by:
 switch (nConfigChannels) {
      case 1:
         fChannelConfig = 0x01;
         fChannelCascading = 8;
         break;
      case 2:
         fChannelConfig = 0x11;
         fChannelCascading = 4;
         break;
      case 4:
         fChannelConfig = 0x55;
         fChannelCascading = 2;
         break;
      case 8:
         fChannelConfig = 0xFF;
         fChannelCascading = 1;
         break;
      default:
         printf("Invalid channel configuration\n");
         return 0;
      }

which is being set by nConfigChannels in DRS.cpp, in the method:
SetChannelConfig(int firstChannel, int lastChannel, int nConfigChannels)

SetChannelConfig is being called in the ConfigDialogue.cpp,  but the default Osci program is such that you can't do a configuration for a cascade of one signal using all the channels. At least, not that I am aware of. 

So what buttons do I need to enable, or what do I need to call, or write, so that I can cascade a signal to end up with 8*1024 bins per event?

This has had me going in circles for weeks, so thank you for your help!!!! 

Sorry for my late reply, I was away for some days.

To use channel cascading, you have to physically connect one input to all eight channels. This is not possible with the evaluation board, you have to make your own board. What you could do however is to split a signal externally and feed it to all four inputs, given that the signal delay is the same for every channel. But then you will hit the hard-wired limit in Osci.cpp. This code was never foreseen to cover 8*1024 bins (since it does not make much sense with the evaluation board). Some arrays are only 2*1024 bins wide, so you would have to rewrite code at many places.

The easiest way to get what you want is to modify drs_exam.cpp. You need SetChannelConfig(0, 8, 1) as you realised correctly, and then you have to retrieve all 8 channels via b->GetWave() and concatenate them correctly.

/Stefan 

 Would it be possible to just hardcode a few lines in the SetChannelConfig in DRS.cpp method as such:

     fChannelConfig = 0x01; //gives me eight
      d = fChannelConfig | (fDominoMode << 8) | (1 << 9) | (fWSRLoop << 10) | (0xF8 << 8);

      Write(T_CTRL, REG_CHANNEL_CONFIG, &d, 2);
      fChannelDepth = 8 * (fDecimation ? kNumberOfBins/2 : kNumberOfBins);// gives eight times the bins

 

then modify the GetWave method/function to include another else if statement similar to  "else if (fChannelCascading == 2) {"  but would be modifidied for fChannelCascading == 8?

 

By, "But then you will hit the hard-wired limit in Osci.cpp" do you mean  hard-coded? Would changing the hard code just amount to resizing all of the arrays, and replacing all the '2*kNumberBins"  with '8*kNumberBins' ?

 

I'm afraid of drs_exam.cpp because it doesn't come with all the perks of Osci.cpp. It seems less daunting to just modify Osci.cpp then to try understanding everything I need to include in drs_exam.cpp because I'm also using an external trigger, and saving the waveform to an external text file.

Thanks!

/Jill

 

 

          Reply  Fri Apr 5 08:54:37 2013, Stefan Ritt, cascading -- DRS4 Osci.cpp & DRS.cpp Screen_Shot_2013-04-05_at_8.51.53_.png

Jill Russek wrote:

Would it be possible to just hardcode a few lines in the SetChannelConfig in DRS.cpp method as such:

     fChannelConfig = 0x01; //gives me eight
      d = fChannelConfig | (fDominoMode << 8) | (1 << 9) | (fWSRLoop << 10) | (0xF8 << 8);

      Write(T_CTRL, REG_CHANNEL_CONFIG, &d, 2);
      fChannelDepth = 8 * (fDecimation ? kNumberOfBins/2 : kNumberOfBins);// gives eight times the bins

 

then modify the GetWave method/function to include another else if statement similar to  "else if (fChannelCascading == 2) {"  but would be modifidied for fChannelCascading == 8?

 

By, "But then you will hit the hard-wired limit in Osci.cpp" do you mean  hard-coded? Would changing the hard code just amount to resizing all of the arrays, and replacing all the '2*kNumberBins"  with '8*kNumberBins' ?

 

I'm afraid of drs_exam.cpp because it doesn't come with all the perks of Osci.cpp. It seems less daunting to just modify Osci.cpp then to try understanding everything I need to include in drs_exam.cpp because I'm also using an external trigger, and saving the waveform to an external text file.

Thanks!

/Jill

Sure it would be possible to code it, but it's not just a few lines. Besides Osci.cpp you have to massage DOScreen.cpp, Measurement.cpp and probably more since they all rely on the array size of the waveform. So if I would do it it would take me probably a couple of days including the debugging, which I don't have right now. Furthermore, as I said you have to combine all eight channels properly. For two channels, it's already pretty complicated (see lines 3537+ in DRS.cpp). I had to make myself a visual scheme in order to understand it correctly, which I attached. For eight channels, the write shift register (WSR) can have values 0-7, depending in which channel you got a trigger. Then you have to sort it out again to get one linear array with the proper order of the fragments. So you see, it's not just changing a few lines of code. In principle it's possible, but it's lots of work.

Best regards,

Stefan

             Reply  Wed Apr 10 22:41:21 2013, Jill Russek, cascading -- DRS4 Osci.cpp & DRS.cpp 

Stefan Ritt wrote:

Jill Russek wrote:

Would it be possible to just hardcode a few lines in the SetChannelConfig in DRS.cpp method as such:

     fChannelConfig = 0x01; //gives me eight
      d = fChannelConfig | (fDominoMode << 8) | (1 << 9) | (fWSRLoop << 10) | (0xF8 << 8);

      Write(T_CTRL, REG_CHANNEL_CONFIG, &d, 2);
      fChannelDepth = 8 * (fDecimation ? kNumberOfBins/2 : kNumberOfBins);// gives eight times the bins

 

then modify the GetWave method/function to include another else if statement similar to  "else if (fChannelCascading == 2) {"  but would be modifidied for fChannelCascading == 8?

 

By, "But then you will hit the hard-wired limit in Osci.cpp" do you mean  hard-coded? Would changing the hard code just amount to resizing all of the arrays, and replacing all the '2*kNumberBins"  with '8*kNumberBins' ?

 

I'm afraid of drs_exam.cpp because it doesn't come with all the perks of Osci.cpp. It seems less daunting to just modify Osci.cpp then to try understanding everything I need to include in drs_exam.cpp because I'm also using an external trigger, and saving the waveform to an external text file.

Thanks!

/Jill

Sure it would be possible to code it, but it's not just a few lines. Besides Osci.cpp you have to massage DOScreen.cpp, Measurement.cpp and probably more since they all rely on the array size of the waveform. So if I would do it it would take me probably a couple of days including the debugging, which I don't have right now. Furthermore, as I said you have to combine all eight channels properly. For two channels, it's already pretty complicated (see lines 3537+ in DRS.cpp). I had to make myself a visual scheme in order to understand it correctly, which I attached. For eight channels, the write shift register (WSR) can have values 0-7, depending in which channel you got a trigger. Then you have to sort it out again to get one linear array with the proper order of the fragments. So you see, it's not just changing a few lines of code. In principle it's possible, but it's lots of work.

Best regards,

Stefan

 Stefan, thanks for your help so far. If I go with your plan A of just modifying drs_exam.cpp, is there a quick way to get it to save the data from the wave, like how osci.cpp spits out an xml file? (Ignoring the cascading aspect for now)

Thanks again :)

/Jill

                Reply  Thu Apr 11 08:39:12 2013, Stefan Ritt, cascading -- DRS4 Osci.cpp & DRS.cpp 

Jill Russek wrote:

Stefan Ritt wrote:

Jill Russek wrote:

Would it be possible to just hardcode a few lines in the SetChannelConfig in DRS.cpp method as such:

     fChannelConfig = 0x01; //gives me eight
      d = fChannelConfig | (fDominoMode << 8) | (1 << 9) | (fWSRLoop << 10) | (0xF8 << 8);

      Write(T_CTRL, REG_CHANNEL_CONFIG, &d, 2);
      fChannelDepth = 8 * (fDecimation ? kNumberOfBins/2 : kNumberOfBins);// gives eight times the bins

 

then modify the GetWave method/function to include another else if statement similar to  "else if (fChannelCascading == 2) {"  but would be modifidied for fChannelCascading == 8?

 

By, "But then you will hit the hard-wired limit in Osci.cpp" do you mean  hard-coded? Would changing the hard code just amount to resizing all of the arrays, and replacing all the '2*kNumberBins"  with '8*kNumberBins' ?

 

I'm afraid of drs_exam.cpp because it doesn't come with all the perks of Osci.cpp. It seems less daunting to just modify Osci.cpp then to try understanding everything I need to include in drs_exam.cpp because I'm also using an external trigger, and saving the waveform to an external text file.

Thanks!

/Jill

Sure it would be possible to code it, but it's not just a few lines. Besides Osci.cpp you have to massage DOScreen.cpp, Measurement.cpp and probably more since they all rely on the array size of the waveform. So if I would do it it would take me probably a couple of days including the debugging, which I don't have right now. Furthermore, as I said you have to combine all eight channels properly. For two channels, it's already pretty complicated (see lines 3537+ in DRS.cpp). I had to make myself a visual scheme in order to understand it correctly, which I attached. For eight channels, the write shift register (WSR) can have values 0-7, depending in which channel you got a trigger. Then you have to sort it out again to get one linear array with the proper order of the fragments. So you see, it's not just changing a few lines of code. In principle it's possible, but it's lots of work.

Best regards,

Stefan

 Stefan, thanks for your help so far. If I go with your plan A of just modifying drs_exam.cpp, is there a quick way to get it to save the data from the wave, like how osci.cpp spits out an xml file? (Ignoring the cascading aspect for now)

Thanks again :)

/Jill

Well, you have to learn C programming, I won't do it for you. drs_exam.cpp contains already code to write to the ASCII file data.txt, so you just can use that or modify it to your needs.

/Stefan

                   Reply  Thu Apr 11 23:32:57 2013, Jill Russek, cascading -- DRS4 Osci.cpp & DRS.cpp 

Stefan Ritt wrote:

Jill Russek wrote:

Stefan Ritt wrote:

Jill Russek wrote:

Would it be possible to just hardcode a few lines in the SetChannelConfig in DRS.cpp method as such:

     fChannelConfig = 0x01; //gives me eight
      d = fChannelConfig | (fDominoMode << 8) | (1 << 9) | (fWSRLoop << 10) | (0xF8 << 8);

      Write(T_CTRL, REG_CHANNEL_CONFIG, &d, 2);
      fChannelDepth = 8 * (fDecimation ? kNumberOfBins/2 : kNumberOfBins);// gives eight times the bins

 

then modify the GetWave method/function to include another else if statement similar to  "else if (fChannelCascading == 2) {"  but would be modifidied for fChannelCascading == 8?

 

By, "But then you will hit the hard-wired limit in Osci.cpp" do you mean  hard-coded? Would changing the hard code just amount to resizing all of the arrays, and replacing all the '2*kNumberBins"  with '8*kNumberBins' ?

 

I'm afraid of drs_exam.cpp because it doesn't come with all the perks of Osci.cpp. It seems less daunting to just modify Osci.cpp then to try understanding everything I need to include in drs_exam.cpp because I'm also using an external trigger, and saving the waveform to an external text file.

Thanks!

/Jill

Sure it would be possible to code it, but it's not just a few lines. Besides Osci.cpp you have to massage DOScreen.cpp, Measurement.cpp and probably more since they all rely on the array size of the waveform. So if I would do it it would take me probably a couple of days including the debugging, which I don't have right now. Furthermore, as I said you have to combine all eight channels properly. For two channels, it's already pretty complicated (see lines 3537+ in DRS.cpp). I had to make myself a visual scheme in order to understand it correctly, which I attached. For eight channels, the write shift register (WSR) can have values 0-7, depending in which channel you got a trigger. Then you have to sort it out again to get one linear array with the proper order of the fragments. So you see, it's not just changing a few lines of code. In principle it's possible, but it's lots of work.

Best regards,

Stefan

 Stefan, thanks for your help so far. If I go with your plan A of just modifying drs_exam.cpp, is there a quick way to get it to save the data from the wave, like how osci.cpp spits out an xml file? (Ignoring the cascading aspect for now)

Thanks again :)

/Jill

Well, you have to learn C programming, I won't do it for you. drs_exam.cpp contains already code to write to the ASCII file data.txt, so you just can use that or modify it to your needs.

/Stefan

 Ha! So then the answer is no, there isn't a ready made function/method to pull out the timing and voltage,  like how it was done in osci.cpp. That's all I wanted to know. (Not whether you would write it for me! Only trying to save time!) Thanks!

/Jill

                      Reply  Fri Apr 12 08:25:05 2013, Stefan Ritt, cascading -- DRS4 Osci.cpp & DRS.cpp 

Jill Russek wrote:

Stefan Ritt wrote:

Jill Russek wrote:

 Stefan, thanks for your help so far. If I go with your plan A of just modifying drs_exam.cpp, is there a quick way to get it to save the data from the wave, like how osci.cpp spits out an xml file? (Ignoring the cascading aspect for now)

Thanks again :)

/Jill

Well, you have to learn C programming, I won't do it for you. drs_exam.cpp contains already code to write to the ASCII file data.txt, so you just can use that or modify it to your needs.

/Stefan

 Ha! So then the answer is no, there isn't a ready made function/method to pull out the timing and voltage,  like how it was done in osci.cpp. That's all I wanted to know. (Not whether you would write it for me! Only trying to save time!) Thanks!

/Jill

You misunderstood. The answer is yes. drs_exam.cpp contains already code to write to an ASCII file. If you actually look into the file, you see:

   f = fopen("data.txt", "w");
   ...
   b->GetTime(0, b->GetTriggerCell(0), time_array);
   ...
   b->GetWave(0, 0, wave_array[0]);
   ...
   fprintf(f, "%5.2f %6.2f\n", time_array[i], wave_array[0][i]);

which actually pulls out the timing and voltage and writes it to the file.

                         Reply  Wed Dec 20 15:30:38 2017, Yoni Sher, cascading -- DRS4 Osci.cpp & DRS.cpp 

Hi, 

I'm trying to do the same thing (get 1 channel with 8192 bins), but I'm having some trouble with it. When I call SetChannelConfig(0, 8, 1) as suggeted, I get output that looks like noise on all readouts. Could you please explain what is supposed to happen in this case? 

I will happily write the code to combine the channels correctly (and debug it) if I can understand what needs to be done. 

(I should mention that my primary concern is a MATLAB interface which I have already written and don't mind sharing when it's complete). 

 

Yoni Sher

Stefan Ritt wrote:

 

Jill Russek wrote:

 

Stefan Ritt wrote:

 

Jill Russek wrote:

 Stefan, thanks for your help so far. If I go with your plan A of just modifying drs_exam.cpp, is there a quick way to get it to save the data from the wave, like how osci.cpp spits out an xml file? (Ignoring the cascading aspect for now)

Thanks again :)

/Jill

Well, you have to learn C programming, I won't do it for you. drs_exam.cpp contains already code to write to the ASCII file data.txt, so you just can use that or modify it to your needs.

/Stefan

 Ha! So then the answer is no, there isn't a ready made function/method to pull out the timing and voltage,  like how it was done in osci.cpp. That's all I wanted to know. (Not whether you would write it for me! Only trying to save time!) Thanks!

/Jill

You misunderstood. The answer is yes. drs_exam.cpp contains already code to write to an ASCII file. If you actually look into the file, you see:

   f = fopen("data.txt", "w");
   ...
   b->GetTime(0, b->GetTriggerCell(0), time_array);
   ...
   b->GetWave(0, 0, wave_array[0]);
   ...
   fprintf(f, "%5.2f %6.2f\n", time_array[i], wave_array[0][i]);

which actually pulls out the timing and voltage and writes it to the file.

 

                            Reply  Wed Dec 20 16:21:42 2017, Stefan Ritt, cascading -- DRS4 Osci.cpp & DRS.cpp 

First you need a board which is modified in hardware to support channel cascading. Basically there are internal resistors which connect each input connector to two channels. You have to specify this when you order the board. Then you can use the new drs_exam_2048.cpp file contains in the git repository which correctly configures and reads out the board in two-channel cascading mode. Putting all 8 channels together is not supported by the evaluation boards.

Stefan

Yoni Sher wrote:

Hi, 

I'm trying to do the same thing (get 1 channel with 8192 bins), but I'm having some trouble with it. When I call SetChannelConfig(0, 8, 1) as suggeted, I get output that looks like noise on all readouts. Could you please explain what is supposed to happen in this case? 

I will happily write the code to combine the channels correctly (and debug it) if I can understand what needs to be done. 

(I should mention that my primary concern is a MATLAB interface which I have already written and don't mind sharing when it's complete). 

 

Yoni Sher

Stefan Ritt wrote:

 

Jill Russek wrote:

 

Stefan Ritt wrote:

 

Jill Russek wrote:

 Stefan, thanks for your help so far. If I go with your plan A of just modifying drs_exam.cpp, is there a quick way to get it to save the data from the wave, like how osci.cpp spits out an xml file? (Ignoring the cascading aspect for now)

Thanks again :)

/Jill

Well, you have to learn C programming, I won't do it for you. drs_exam.cpp contains already code to write to the ASCII file data.txt, so you just can use that or modify it to your needs.

/Stefan

 Ha! So then the answer is no, there isn't a ready made function/method to pull out the timing and voltage,  like how it was done in osci.cpp. That's all I wanted to know. (Not whether you would write it for me! Only trying to save time!) Thanks!

/Jill

You misunderstood. The answer is yes. drs_exam.cpp contains already code to write to an ASCII file. If you actually look into the file, you see:

   f = fopen("data.txt", "w");
   ...
   b->GetTime(0, b->GetTriggerCell(0), time_array);
   ...
   b->GetWave(0, 0, wave_array[0]);
   ...
   fprintf(f, "%5.2f %6.2f\n", time_array[i], wave_array[0][i]);

which actually pulls out the timing and voltage and writes it to the file.

 

 

                               Reply  Wed Dec 20 16:30:45 2017, Yoni Sher, cascading -- DRS4 Osci.cpp & DRS.cpp 

Hi, 

The board is modified (and checks out with the DRSScope program). Could you please point me to the drs_exam_2048.cpp file? I can't seem to fine the most up-to-date git repository....

 

Thanks, 

Yoni

Stefan Ritt wrote:

First you need a board which is modified in hardware to support channel cascading. Basically there are internal resistors which connect each input connector to two channels. You have to specify this when you order the board. Then you can use the new drs_exam_2048.cpp file contains in the git repository which correctly configures and reads out the board in two-channel cascading mode. Putting all 8 channels together is not supported by the evaluation boards.

Stefan

Yoni Sher wrote:

Hi, 

I'm trying to do the same thing (get 1 channel with 8192 bins), but I'm having some trouble with it. When I call SetChannelConfig(0, 8, 1) as suggeted, I get output that looks like noise on all readouts. Could you please explain what is supposed to happen in this case? 

I will happily write the code to combine the channels correctly (and debug it) if I can understand what needs to be done. 

(I should mention that my primary concern is a MATLAB interface which I have already written and don't mind sharing when it's complete). 

 

Yoni Sher

Stefan Ritt wrote:

 

Jill Russek wrote:

 

Stefan Ritt wrote:

 

Jill Russek wrote:

 Stefan, thanks for your help so far. If I go with your plan A of just modifying drs_exam.cpp, is there a quick way to get it to save the data from the wave, like how osci.cpp spits out an xml file? (Ignoring the cascading aspect for now)

Thanks again :)

/Jill

Well, you have to learn C programming, I won't do it for you. drs_exam.cpp contains already code to write to the ASCII file data.txt, so you just can use that or modify it to your needs.

/Stefan

 Ha! So then the answer is no, there isn't a ready made function/method to pull out the timing and voltage,  like how it was done in osci.cpp. That's all I wanted to know. (Not whether you would write it for me! Only trying to save time!) Thanks!

/Jill

You misunderstood. The answer is yes. drs_exam.cpp contains already code to write to an ASCII file. If you actually look into the file, you see:

   f = fopen("data.txt", "w");
   ...
   b->GetTime(0, b->GetTriggerCell(0), time_array);
   ...
   b->GetWave(0, 0, wave_array[0]);
   ...
   fprintf(f, "%5.2f %6.2f\n", time_array[i], wave_array[0][i]);

which actually pulls out the timing and voltage and writes it to the file.

 

 

 

                                  Reply  Wed Dec 20 22:14:35 2017, Stefan Ritt, cascading -- DRS4 Osci.cpp & DRS.cpp 

https://bitbucket.org/ritt/drs4eb

 

Entry  Tue Dec 12 00:25:50 2017, Diego Yankelevich, External trigger using Raspberry Pi 

Dear Steffan:

We have been able to use the DRS4 using a Raspberry Pi but we have not been able to use the external trigger. What we are doing is basically comment out the code shown below (downloaded from PSI) to use the hardware trigger and uncomment the code to use the external trigger. We have not been able to get external trigger to work. Could you see what could be wrong?

Thanks

Diego

/* use following line to turn on the internal 100 MHz clock connected to all channels  */
   //b->EnableTcal(1);

   /* use following lines to enable hardware trigger on CH1 at 50 mV positive edge */

   /*
   if (b->GetBoardType() >= 8) {        // Evaluaiton Board V4&5
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<0);        // set CH1 as source
   } else if (b->GetBoardType() == 7) { // Evaluation Board V3
      b->EnableTrigger(0, 1);           // lemo off, analog trigger on
      b->SetTriggerSource(0);           // use CH1 as source
   }
   b->SetTriggerLevel(0.05);            // 0.05 V
   b->SetTriggerPolarity(false);        // positive edge
   */

   /* use following lines to set individual trigger elvels */
   //b->SetIndividualTriggerLevel(1, 0.1);
   //b->SetIndividualTriggerLevel(2, 0.2);
   //b->SetIndividualTriggerLevel(3, 0.3);
   //b->SetIndividualTriggerLevel(4, 0.4);
   //b->SetTriggerSource(15);

   b->SetTriggerDelayNs(0);             // zero ns trigger delay

   /* use following lines to enable the external trigger */
   if (b->GetBoardType() == 8) {     // Evaluaiton Board V4
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<4);        // set external trigger as source
   } else {                          // Evaluation Board V3
      b->EnableTrigger(1, 0);           // lemo on, analog trigger off
    }
    Reply  Tue Dec 12 13:58:06 2017, Stefan Ritt, External trigger using Raspberry Pi 

Indeed the code does not work for the current evaluation board, it has been written for a previous version and never been updated. Please use following code to enable the external trigger

   /* use following lines to enable the external trigger */
   if (b->GetBoardType() >= 8) {        // Evaluaiton Board V4&5
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerConfig(1<<4);        // set external trigger as source
   } else {                             // Evaluation Board V3
      b->EnableTrigger(1, 0);           // lemo on, analog trigger offf
   }

Please also make sure that the signal on the external trigger input is strong enough. You need at least 2.5V at 50 Ohms, and not every driver is capable of driving 50 Ohms.

Stefan

Diego Yankelevich wrote:

Dear Steffan:

We have been able to use the DRS4 using a Raspberry Pi but we have not been able to use the external trigger. What we are doing is basically comment out the code shown below (downloaded from PSI) to use the hardware trigger and uncomment the code to use the external trigger. We have not been able to get external trigger to work. Could you see what could be wrong?

Thanks

Diego

/* use following line to turn on the internal 100 MHz clock connected to all channels  */
   //b->EnableTcal(1);

   /* use following lines to enable hardware trigger on CH1 at 50 mV positive edge */

   /*
   if (b->GetBoardType() >= 8) {        // Evaluaiton Board V4&5
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<0);        // set CH1 as source
   } else if (b->GetBoardType() == 7) { // Evaluation Board V3
      b->EnableTrigger(0, 1);           // lemo off, analog trigger on
      b->SetTriggerSource(0);           // use CH1 as source
   }
   b->SetTriggerLevel(0.05);            // 0.05 V
   b->SetTriggerPolarity(false);        // positive edge
   */

   /* use following lines to set individual trigger elvels */
   //b->SetIndividualTriggerLevel(1, 0.1);
   //b->SetIndividualTriggerLevel(2, 0.2);
   //b->SetIndividualTriggerLevel(3, 0.3);
   //b->SetIndividualTriggerLevel(4, 0.4);
   //b->SetTriggerSource(15);

   b->SetTriggerDelayNs(0);             // zero ns trigger delay

   /* use following lines to enable the external trigger */
   if (b->GetBoardType() == 8) {     // Evaluaiton Board V4
      b->EnableTrigger(1, 0);           // enable hardware trigger
      b->SetTriggerSource(1<<4);        // set external trigger as source
   } else {                          // Evaluation Board V3
      b->EnableTrigger(1, 0);           // lemo on, analog trigger off
    }

 

Entry  Thu Nov 16 02:55:44 2017, Diego Yankelevich, Averaging capabilities  

The Display window in the Oscilloscope software shows averaging capabilites but I have not been able to activate these. Is it possible to activate averaging with the existing oscilloscope software? Thanks

    Reply  Wed Nov 22 14:52:31 2017, Stefan Ritt, Averaging capabilities  

This feature is not yet implemented. The (disabled) software swtich is more like a kind of a reminder to myself to work on that one day...

Diego Yankelevich wrote:

The Display window in the Oscilloscope software shows averaging capabilites but I have not been able to activate these. Is it possible to activate averaging with the existing oscilloscope software? Thanks

 

Entry  Wed Nov 22 08:31:03 2017, chen wenjun , using of the DRS Command Line Interface Î¢ÐÅͼƬ_20171122153834.png

Hello! I'm using DRS4 evaluation board V5 with the drs command line interface,but the mannal only explained the meaning of the command--"info".And I can't get the hang of the use of other commands through "help",so is there anywhere can I learn more about other commands?Or I can only learn it through the datasheet of DRS4 chip.

Thanks!

 

 

    Reply  Wed Nov 22 08:48:36 2017, Stefan Ritt, using of the DRS Command Line Interface 

The command line interface is more a debugging tool for experts, and you are not supposed to use it except to test the connection to the evaluation board. The programs for the user are the DRS Oscilloscope and the drs_exam.cpp example program to read out the board with your own program.

Stefan

chen wenjun wrote:

Hello! I'm using DRS4 evaluation board V5 with the drs command line interface,but the mannal only explained the meaning of the command--"info".And I can't get the hang of the use of other commands through "help",so is there anywhere can I learn more about other commands?Or I can only learn it through the datasheet of DRS4 chip.

Thanks!

 

 

 

       Reply  Wed Nov 22 08:58:33 2017, chen wenjun , using of the DRS Command Line Interface Î¢ÐÅͼƬ_20171122160245.png

OK!Thank you! One more question,when I use the Oscillocope ,I found that the actual speed is a constant value of 1.007G,how can change this speed.

Stefan Ritt wrote:

The command line interface is more a debugging tool for experts, and you are not supposed to use it except to test the connection to the evaluation board. The programs for the user are the DRS Oscilloscope and the drs_exam.cpp example program to read out the board with your own program.

Stefan

chen wenjun wrote:

Hello! I'm using DRS4 evaluation board V5 with the drs command line interface,but the mannal only explained the meaning of the command--"info".And I can't get the hang of the use of other commands through "help",so is there anywhere can I learn more about other commands?Or I can only learn it through the datasheet of DRS4 chip.

Thanks!

 

 

 

 

          Reply  Wed Nov 22 09:14:18 2017, Stefan Ritt, using of the DRS Command Line Interface 

Remove the check mark from the "Lock" box and enter a different value in the sampling speed box and hit return.

chen wenjun wrote:

OK!Thank you! One more question,when I use the Oscillocope ,I found that the actual speed is a constant value of 1.007G,how can change this speed.

Stefan Ritt wrote:

The command line interface is more a debugging tool for experts, and you are not supposed to use it except to test the connection to the evaluation board. The programs for the user are the DRS Oscilloscope and the drs_exam.cpp example program to read out the board with your own program.

Stefan

chen wenjun wrote:

Hello! I'm using DRS4 evaluation board V5 with the drs command line interface,but the mannal only explained the meaning of the command--"info".And I can't get the hang of the use of other commands through "help",so is there anywhere can I learn more about other commands?Or I can only learn it through the datasheet of DRS4 chip.

Thanks!

 

 

 

 

 

             Reply  Wed Nov 22 09:19:11 2017, chen wenjun , using of the DRS Command Line Interface 

Thank you very much !! All my fault for I thought it too comlicated. Thank you sincerely!

Stefan Ritt wrote:

Remove the check mark from the "Lock" box and enter a different value in the sampling speed box and hit return.

chen wenjun wrote:

OK!Thank you! One more question,when I use the Oscillocope ,I found that the actual speed is a constant value of 1.007G,how can change this speed.

Stefan Ritt wrote:

The command line interface is more a debugging tool for experts, and you are not supposed to use it except to test the connection to the evaluation board. The programs for the user are the DRS Oscilloscope and the drs_exam.cpp example program to read out the board with your own program.

Stefan

chen wenjun wrote:

Hello! I'm using DRS4 evaluation board V5 with the drs command line interface,but the mannal only explained the meaning of the command--"info".And I can't get the hang of the use of other commands through "help",so is there anywhere can I learn more about other commands?Or I can only learn it through the datasheet of DRS4 chip.

Thanks!

 

 

 

 

 

 

Entry  Fri Nov 3 12:11:14 2017, Håkan Wennlöf, Triggering using AND 

Hi!

I'm using the DRSOsc program, and I have a question that I need a bit clarified;

When triggering using AND between two channels, am I then triggering on rising/falling edge of both channels, or on the actual values?

That is, is it the change in value that it triggers on, or when the actual value goes above a certain threshold?

Using AND, it seems like I get a trigger when one (CH2) is above its trigger value (sometimes quite far above), and the other (CH1) changes to go above its trigger value. This works for my purposes, which is nice, but I just want to be sure I understand why it works. Ideally, I'd like to trigger when one of my channels is above a certain value, and the other has a rising edge above a certain value.

I'm sorry if it's a silly question! I've just noticed that the Keysight oscilloscopes I've used can only do one channel with a rising edge at a time when doing a logic trigger, and I thought the same thing might be going on in the background here (which would be ideal for my purposes).

 

Kind regards,

Håkan Wennlöf

    Reply  Fri Nov 3 13:28:04 2017, Stefan Ritt, Triggering using AND 

Think about: How would you make a coincidence (AND) between two edges? Since an edge is infinitesimally small, there is no way to make a meaningful coincidence between edges. Therefore, the DRS4 EB firmware makes a simple AND of levels. If you trigger on rising signals and do an AND, then you get a trigger if both values are above their threshold. For falling edge trigger (arrow goes down in the trigger configuration) the board triggers when both signals are BELOW the trigger threshold. So actually the singnal which crosses the threshold last determines the timing of the trigger. I see no other way how to implement an AND.

Stefan

Håkan Wennlöf wrote:

Hi!

I'm using the DRSOsc program, and I have a question that I need a bit clarified;

When triggering using AND between two channels, am I then triggering on rising/falling edge of both channels, or on the actual values?

That is, is it the change in value that it triggers on, or when the actual value goes above a certain threshold?

Using AND, it seems like I get a trigger when one (CH2) is above its trigger value (sometimes quite far above), and the other (CH1) changes to go above its trigger value. This works for my purposes, which is nice, but I just want to be sure I understand why it works. Ideally, I'd like to trigger when one of my channels is above a certain value, and the other has a rising edge above a certain value.

I'm sorry if it's a silly question! I've just noticed that the Keysight oscilloscopes I've used can only do one channel with a rising edge at a time when doing a logic trigger, and I thought the same thing might be going on in the background here (which would be ideal for my purposes).

 

Kind regards,

Håkan Wennlöf

 

Entry  Tue Oct 17 14:58:58 2017, Vadym Denysenko, Time offset  

Hello.

 

I have a simple question, can I set SetTriggerDelayNs() more than 1631 ns?

I need to set this delay to about 5 us... Can I do this? 

 

Thank you in advance! 

 

With best regards, 

Vadym

    Reply  Wed Oct 18 09:12:26 2017, Stefan Ritt, Time offset  

No this is not possible. But you can delay your signal externally (like with a delay cable or electronically) and then send the dealyed signal to the evaluation board for triggering.

Stefan

Vadym Denysenko wrote:

Hello.

 

I have a simple question, can I set SetTriggerDelayNs() more than 1631 ns?

I need to set this delay to about 5 us... Can I do this? 

 

Thank you in advance! 

 

With best regards, 

Vadym

 

       Reply  Wed Oct 18 11:48:14 2017, Vadym Denysenko, Time offset  

Thank you for your reply!

Stefan Ritt wrote:

No this is not possible. But you can delay your signal externally (like with a delay cable or electronically) and then send the dealyed signal to the evaluation board for triggering.

Stefan

Vadym Denysenko wrote:

Hello.

 

I have a simple question, can I set SetTriggerDelayNs() more than 1631 ns?

I need to set this delay to about 5 us... Can I do this? 

 

Thank you in advance! 

 

With best regards, 

Vadym

 

 

Entry  Fri Oct 13 03:39:01 2017, Jonathan Wapman, Raspberry Pi Connection Failure 

I am currently attempting to use a raspberry pi to connect to the DRS 4 board. I whenever I try to use the DRS Command Line TOol, Revision 21435 to connect to the drs board, I get the error

"musb_open: libusb_open() error -3"

"USB successfully scanned, but no boards found"

"No DRS Boards Found".

I successfully compiled the libusb driver before compiling the drs software 5.0.6, and installed all other listed packages in the install instructions.

    Reply  Mon Oct 16 15:35:22 2017, Stefan Ritt, Raspberry Pi Connection Failure 

Have you tried as root? Maybe you miss some permissions.

Stefan

Jonathan Wapman wrote:

I am currently attempting to use a raspberry pi to connect to the DRS 4 board. I whenever I try to use the DRS Command Line TOol, Revision 21435 to connect to the drs board, I get the error

"musb_open: libusb_open() error -3"

"USB successfully scanned, but no boards found"

"No DRS Boards Found".

I successfully compiled the libusb driver before compiling the drs software 5.0.6, and installed all other listed packages in the install instructions.

 

Entry  Wed Sep 27 16:11:03 2017, Yoni Sher, Event acquisition pace for irregular timing 

Hi, 

I'm running a LIDAR application that requires that every outgoing pulse be captured. My current setup firess sets of 20-50 pulses at 1 ms intervals, about 10 times a second, but only 10-20 pulses a second are captured. 

When I fire at full speed (1KHz - one pulse every ms), about 500-600 pulses a second are captured. 

At the moment, I'm triggering on channel 1 and captureing the data on channel 2. Would it help if I used the external trigger? Is there anything else I can do?

 

Yoni

    Reply  Mon Oct 2 16:08:05 2017, Stefan Ritt, Event acquisition pace for irregular timing 

As written in the documentation, the DRS evaluaiton board has a maximum trigger capability of ~500 Hz. This is limited by the USB bus which has a finite data transfer rate. If you build your own electronics around the chip (like many other groups are doing), you can squeeze this to a few kHz, but it is some development effort.

Stefan

Yoni Sher wrote:

Hi, 

I'm running a LIDAR application that requires that every outgoing pulse be captured. My current setup firess sets of 20-50 pulses at 1 ms intervals, about 10 times a second, but only 10-20 pulses a second are captured. 

When I fire at full speed (1KHz - one pulse every ms), about 500-600 pulses a second are captured. 

At the moment, I'm triggering on channel 1 and captureing the data on channel 2. Would it help if I used the external trigger? Is there anything else I can do?

 

Yoni

 

Entry  Sun Aug 27 12:44:16 2017, Yuvaraj Elangovan, DRS4 version Support 

Hi i am using DRS4 Eval Board V2, How to acquire data to a bin file using it.    

Entry  Fri Jul 21 09:16:02 2017, Volodymyr Rodin, Time output 

Hello Stefan

I tried to convert binary to a simple txt file and found next problem - strange time output.

Here is output from little modification for read_binary.cpp (Its last output line also is strange: dT = -1.#IOns +- -1.$ps)

Found data for board #0
Found timing calibration for channel #1
Found boards#  1
     event    channel   waveform       time      point
         1          0  -0.000092   0.000000          0
         1          0   0.030548   0.000000          1
         1          0   0.059418   0.000000          2
         1          0   0.080200   0.000000          3
         1          0   0.094223   0.000000          4
         1          0   0.097702   0.000000          5
         1          0   0.094055   0.000000          6
         1          0   0.079117   0.000000          7
         1          0   0.060364   0.000000          8
         1          0   0.030960   0.000000          9
         1          0   0.000504   0.000000         10
         1          0  -0.031555   0.000000         11
         1          0  -0.057465   0.000000         12
         1          0  -0.080536   0.000000         13
         1          0  -0.095413   0.000000         14
         1          0  -0.099365   0.000000         15

I used output string in the following places, but it didn't help:

// reach channel data

:for (i=0 ; i<1024 ; i++) {
               // convert data to volts
               waveform[0][chn_index][i] = (voltage[i] / 65536. + eh.range/1000.0 - 0.5);
               // calculate time for this cell
               for (j=0,time[b][chn_index][i]=0 ; j<i ; j++)
                  time[b][chn_index][i] += bin_width[b][chn_index][(j+tch.trigger_cell) % 1024];

printf("%10d %10d %10f %10f %10d\n", eh.event_serial_number , chn_index , waveform[0][chn_index][i]  ,  time[0][chn_index][i]  , i);

And after alignment procedure:

t1 = time[b][0][(1024-tch.trigger_cell) % 1024];
         for (chn=1 ; chn<4 ; chn++) {
            t2 = time[b][chn][(1024-tch.trigger_cell) % 1024];
            dt = t1 - t2;
            for (i=0 ; i<1024 ; i++)
               time[b][chn][i] += dt;

printf("%10d %10d %10f %10f %10d\n", eh.event_serial_number , chn_index , waveform[0][chn_index][i]  ,  time[0][chn][i]  , i);
         }

Does it caused by some software or drivers changes?

Best regards,

Volodymyr

 

 

 

 

    Reply  Tue Jul 25 14:47:05 2017, Volodymyr Rodin, Time output 

Hi again.

Okay, it works with 5.05 version very good and it is enough for me.

Besides,

What do I need to fix in this code for 2048 board?

Best wishes,

Volodymyr

Volodymyr Rodin wrote:

Hello Stefan

I tried to convert binary to a simple txt file and found next problem - strange time output.

Here is output from little modification for read_binary.cpp (Its last output line also is strange: dT = -1.#IOns +- -1.$ps)

Found data for board #0
Found timing calibration for channel #1
Found boards#  1
     event    channel   waveform       time      point
         1          0  -0.000092   0.000000          0
         1          0   0.030548   0.000000          1
         1          0   0.059418   0.000000          2
         1          0   0.080200   0.000000          3
         1          0   0.094223   0.000000          4
         1          0   0.097702   0.000000          5
         1          0   0.094055   0.000000          6
         1          0   0.079117   0.000000          7
         1          0   0.060364   0.000000          8
         1          0   0.030960   0.000000          9
         1          0   0.000504   0.000000         10
         1          0  -0.031555   0.000000         11
         1          0  -0.057465   0.000000         12
         1          0  -0.080536   0.000000         13
         1          0  -0.095413   0.000000         14
         1          0  -0.099365   0.000000         15

I used output string in the following places, but it didn't help:

// reach channel data

:for (i=0 ; i<1024 ; i++) {
               // convert data to volts
               waveform[0][chn_index][i] = (voltage[i] / 65536. + eh.range/1000.0 - 0.5);
               // calculate time for this cell
               for (j=0,time[b][chn_index][i]=0 ; j<i ; j++)
                  time[b][chn_index][i] += bin_width[b][chn_index][(j+tch.trigger_cell) % 1024];

printf("%10d %10d %10f %10f %10d\n", eh.event_serial_number , chn_index , waveform[0][chn_index][i]  ,  time[0][chn_index][i]  , i);

And after alignment procedure:

t1 = time[b][0][(1024-tch.trigger_cell) % 1024];
         for (chn=1 ; chn<4 ; chn++) {
            t2 = time[b][chn][(1024-tch.trigger_cell) % 1024];
            dt = t1 - t2;
            for (i=0 ; i<1024 ; i++)
               time[b][chn][i] += dt;

printf("%10d %10d %10f %10f %10d\n", eh.event_serial_number , chn_index , waveform[0][chn_index][i]  ,  time[0][chn][i]  , i);
         }

Does it caused by some software or drivers changes?

Best regards,

Volodymyr

 

 

 

 

 

Entry  Fri Jun 16 17:34:20 2017, Laura Gonella, Driver installation on Windows 10 

Hello,

I am trying to get a DRS4 board to run on Windows 10. I am having problems with the driver installation. I am getting the follwoing message

"There is no driver selected for the device information set or element"

I had specified the path to look for the driver as C:\ProgramFilesx86\DRS\driver\. I also tried the option to look online for the driver. None works. Can anyone help?

Thanks,

Laura

    Reply  Thu Jul 20 13:00:44 2017, Volodymyr Rodin, Driver installation on Windows 10 

Dear Laura

You need to disable driver signature enforcement.  Then try again with path option.

 It helped me.

http://www.drivethelife.com/windows-drivers/how-to-disable-driver-signature-enforcement-on-windows-10-8-7-xp-vista.html

Best regards,

Volodymyr

Laura Gonella wrote:

Hello,

I am trying to get a DRS4 board to run on Windows 10. I am having problems with the driver installation. I am getting the follwoing message

"There is no driver selected for the device information set or element"

I had specified the path to look for the driver as C:\ProgramFilesx86\DRS\driver\. I also tried the option to look online for the driver. None works. Can anyone help?

Thanks,

Laura

 

Entry  Wed Jul 12 04:24:39 2017, Toshihiro Nonaka, Time resolution between boards 

Hello,

I 'm using four evaluation boards v.3 to construct the multi-board DAQ system. One channel for each board is used as reference clock, then calibrate timing offline, which allow below 10ps resolution between boards.

Is it possible to keep the time resolution between boards below 10ps in daisy-chain mode with v.5 boards?

Thank you in adcance.

Best regards,

Toshihiro Nonaka

    Reply  Wed Jul 12 20:16:05 2017, Stefan Ritt, Time resolution between boards 

Yes this should  be possible.

Stefan

Toshihiro Nonaka wrote:

Hello,

I 'm using four evaluation boards v.3 to construct the multi-board DAQ system. One channel for each board is used as reference clock, then calibrate timing offline, which allow below 10ps resolution between boards.

Is it possible to keep the time resolution between boards below 10ps in daisy-chain mode with v.5 boards?

Thank you in adcance.

Best regards,

Toshihiro Nonaka

 

Entry  Thu Jul 6 15:10:48 2017, Esperienza Giove, Trigger setting (AND AND) OR (AND AND) 

Hello there,

is it possible to setup trigger in double AND configuration (a pair in and or other pair in and). 

eg (CH 1 AND CH 2 ) OR ( CH 3 AND CH4)

Thank you

    Reply  Fri Jul 7 10:31:47 2017, Stefan Ritt, Trigger setting (AND AND) OR (AND AND) 

Unfortunately not with the current firmware.

Stefan

Esperienza Giove wrote:

Hello there,

is it possible to setup trigger in double AND configuration (a pair in and or other pair in and). 

eg (CH 1 AND CH 2 ) OR ( CH 3 AND CH4)

Thank you

 

ELOG V3.1.5-fe60aaf